Hi James,

Please find below some additional thoughts on the subject that have 
accidentally been discussed off-list.

Kind regards,

Bernd

Von: Dürrer Bernd
Gesendet: Donnerstag, 6. April 2023 11:36
An: John Woodgate <j...@woodjohn.uk>
Betreff: AW: AW: [PSES] AW: [PSES] Safety: 60335-1 conflicting clauses question 
- earthing (or not) of handles

Agreed. The standard would be clearer if clauses 22.35 and 22.36 would start 
with "For class II constructions, [...]". However, I am neither a member of the 
national nor international committee for IEC 60335-1.

Kind regards,

Bernd

Von: John Woodgate <j...@woodjohn.uk<mailto:j...@woodjohn.uk>>
Gesendet: Donnerstag, 6. April 2023 11:14
An: Dürrer Bernd <bernd.duer...@wilo.com<mailto:bernd.duer...@wilo.com>>
Betreff: Re: AW: [PSES] AW: [PSES] Safety: 60335-1 conflicting clauses question 
- earthing (or not) of handles

[EXTERNAL E-MAIL] This email originated from outside of the organization. Do 
not click links or open attachments unless you recognize the sender.

Noted, but the standard doesn't explicitly address that situation, which is 
surely very rare, and I understand that the handles could be reliably earthed. 
I still recommend the simple solution of covering them with reinforced 
insulation, but the standard needs to be clarified.
======================================================================================
Best wishes John Woodgate OOO-Own Opinions Only
www.woodjohn.uk<http://www.woodjohn.uk/>
Rayleigh, Essex UK

I hear, and I forget. I see, and I remember. I do, and I understand. Xunzi (340 
- 245 BC)
On 2023-04-06 10:00, Dürrer Bernd wrote:
As I understand from James' explanations, the metal handles of the appliance in 
question are not reliably earthed. As such, these parts of the appliance do not 
comply with the requirements for a class I appliance (IEC 60335-1, 3.3.9) that 
all conductive accessible parts shall be connected to the protective earthing 
conductor in the fixed wiring of the installation. Therefore, these handles 
shall be assessed as class II construction (IEC 60335-1, 3.3.11 and 5.14) for 
which the requirements of clauses 22.35 and 22.36 are applicable.

Kind regards,

Bernd

Von: John Woodgate <j...@woodjohn.uk><mailto:j...@woodjohn.uk>
Gesendet: Donnerstag, 6. April 2023 10:46
An: Dürrer Bernd <bernd.duer...@wilo.com><mailto:bernd.duer...@wilo.com>
Betreff: Re: [PSES] AW: [PSES] Safety: 60335-1 conflicting clauses question - 
earthing (or not) of handles

[EXTERNAL E-MAIL] This email originated from outside of the organization. Do 
not click links or open attachments unless you recognize the sender.

It still doesn't make sense to require double or reinforced insulation  for a 
Class I product. Doing so, causes the safety of Class 1 construction to be 
questioned.
======================================================================================
Best wishes John Woodgate OOO-Own Opinions Only
www.woodjohn.uk<http://www.woodjohn.uk/>
Rayleigh, Essex UK

I hear, and I forget. I see, and I remember. I do, and I understand. Xunzi (340 
- 245 BC)

On 2023-04-06 09:40, Dürrer Bernd wrote:
Hi James,

actually, these two clauses address two different, but related risks: Clause 
22.35 addresses the risk that the handle itself may become live. In any case, a 
conductive surface (like the handle) has to be separated from live parts by 
basic insulation. Due to the higher risk that handles are actually touched in 
the event of failure of basic insulation, the standard requires additional 
precautions. Either the handles itself shall be covered by insulating material 
(i.e. supplementary insulation) or they shall be separated by supplementary 
insulation from their fixing. In both cases, the handle will be protected by 
double insulation from live parts.

Clause 22.36 addresses the risk that other parts than the handle in the 
vicinity of the handle may become live. Again, there is a higher risk for the 
user in the event of failure of basic insulation that these surfaces may be 
actually touched, so also for these neighbouring conductive surfaces either 
double or reinforced insulation is required.

Kind regards,

Bernd

Von: James Pawson (U3C) 
<ja...@unit3compliance.co.uk><mailto:ja...@unit3compliance.co.uk>
Gesendet: Mittwoch, 5. April 2023 12:23
An: EMC-PSTC@LISTSERV.IEEE.ORG<mailto:EMC-PSTC@LISTSERV.IEEE.ORG>
Betreff: [PSES] Safety: 60335-1 conflicting clauses question - earthing (or 
not) of handles

[EXTERNAL E-MAIL] This email originated from outside of the organization. Do 
not click links or open attachments unless you recognize the sender.
Hello experts,

In running through some safety checks on a customer's product I've come across 
what appears to be a conflict between two clauses. This resulted in some head 
scratching! I'm sure it is my misunderstanding or misreading, but I could do 
with some pointers in how to unpick this.

Context

  1.  Standard is IEC EN 60335-1:2012 (+amendments)
  2.  Context is an exercise machine (the appliance)
  3.  Supply is AC mains Class I
  4.  Appliance has metal handles that are held during normal use. These are 
connected through bearings to the metal frame but are not "reliably earthed" 
(no explicit wired connection)

Earthing of metal parts is acceptable
22.35 For constructions other than those of class III, handles, levers and 
knobs that are held or actuated in normal use shall not become live in the 
event of a failure of basic insulation. If these handles, levers and knobs are 
of metal and if their shafts or fixings are likely to become live in the event 
of a failure of basic insulation, they shall be adequately covered by 
insulating material or their accessible parts shall be separated from their 
shafts or fixings by supplementary insulation.

For stationary appliances and cordless appliances, this requirement does not 
apply to handles, levers and knobs, other than those of electrical components, 
provided that they are reliably connected to an earthing terminal or earthing 
contact or separated from live parts by earthed metal.


  1.  Class I appliance, with metal handles held during use so this requirement 
is applicable
  2.  Handles are not reliably earthed so we are advising manufacturer to make 
an explicit bond to PE
  3.  With Basic insulation and Earthing we have our two safeguards between 
live and user = reduced risk of shock
  4.  No complaints with this clause

Earthing of metal parts is not accpetable
Then the next clause seems to contradict this:

22.36 For appliances other than those of class III, handles which are 
continuously held in the hand in normal use shall be constructed so that when 
gripped in normal use, the operator's hand is not likely to touch metal parts 
unless they are separated from live parts by double insulation or reinforced 
insulation


  1.  This clause appears to not allow for earthing of the metal handles
  2.  Instead, only rouble or reinforced insulation are acceptable
  3.  Thoughts / speculation:

     *   Would this clause make more sense if it read "For appliances other 
than those of class I..."? i.e. unearthed
     *   Talks about "appliances" (whole product) rather than "constructions" 
(parts of appliance)

Conclusion
I'm happy that the approach of reliably earthing the metal handles will reduce 
the electric shock risk in the context of the equipment, this approach being 
consistent with other safety standards I know like 62368-1 and 61010-1. I just 
can't rationalise this with clause 22.36. As I say, I'm sure I've misread or 
misunderstood something so any help would be greatly appreciated.

Otherwise, if anyone has a link to an explanatory document or contact for 
someone who could help at standards or committee level again that would help.

All the best
James

James Pawson
Managing Director & EMC Problem Solver

Office hours:
My mornings are reserved for full attention on consultancy, testing, and 
troubleshooting activities for our customers' projects. I am otherwise 
contactable between 1300h to 1730h from Monday to Friday.
For inquiries, bookings, and testing updates please send us an email on 
he...@unit3compliance.co.uk<mailto:he...@unit3compliance.co.uk> or call 01274 
911747. Our lead times for testing and consultancy are typically 4-5 weeks.

Unit 3 Compliance Ltd
EMC : Environmental & Vibration : Electrical Safety : CE & UKCA : Consultancy

www.unit3compliance.co.uk<http://www.unit3compliance.co.uk/> | 
ja...@unit3compliance.co.uk<mailto:ja...@unit3compliance.co.uk>
+44(0)1274 911747  |  +44(0)7811 139957
2 Wellington Business Park, New Lane, Bradford, BD4 8AL
Registered in England and Wales # 10574298

-
----------------------------------------------------------------

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org<mailto:emc-p...@ieee.org>>

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html

Website: http://www.ieee-pses.org/
Instructions: http://www.ieee-pses.org/list.html (including how to 
unsubscribe)<http://www.ieee-pses.org/list.html>
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Mike Cantwell <mcantw...@ieee.org<mailto:mcantw...@ieee.org>>

For policy questions, send mail to:
Jim Bacher <j.bac...@ieee.org<mailto:j.bac...@ieee.org>>
David Heald <dhe...@gmail.com<mailto:dhe...@gmail.com>>

________________________________

To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1

________________________________

WILO SE
Wilopark 1, 44263 Dortmund
Amtsgericht Dortmund, HRB 21356
www.wilo.com<http://www.wilo.com/>

Vorstand/Executive Board:
Oliver Hermes (President & CEO), Dr. Patrick Niehr, Georg Weber, Mathias Weyers
Aufsichtsratsvorsitzender/Chairman of the Supervisory Board:
Lars Roßner

Dieses Dokument ist vertraulich zu behandeln. Die Weitergabe sowie 
Vervielfaeltigung, Verwertung und Mitteilung seines Inhalts ist nur mit unserer 
ausdruecklichen Genehmigung gestattet. Alle Rechte vorbehalten, insbesondere 
fuer den Fall der Schutzrechtsanmeldung.
This document has to be treated confidentially. Its contents are not to be 
passed on, duplicated, exploited or disclosed without our expressed permission. 
All rights reserved, especially the right to apply for protective rights.

Datenschutz ist uns wichtig! Daher gehen wir sorgsam mit Ihren Daten um. 
Näheres finden Sie in unseren 
Datenschutzhinweisen<https://wilo.com/de/de/Wilo-Deutschland/Datenschutzhinweise.html>.
 Sollten Sie keinen Kontakt mehr mit uns wünschen, wenden Sie sich bitte an 
datensch...@wilo.com<mailto:datensch...@wilo.com>.
Data Protection is important to us! Therefore we handle your data carefully. 
Please find the details in our Note on Data 
Protection<https://wilo.com/de/de/Note-on-data-protection.html>. If you do not 
wish to stay in touch with us anymore, please write to 
data-priv...@wilo.com<mailto:data-priv...@wilo.com>. -
----------------------------------------------------------------

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org<mailto:emc-p...@ieee.org>>

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html

Website: http://www.ieee-pses.org/
Instructions: http://www.ieee-pses.org/list.html (including how to 
unsubscribe)<http://www.ieee-pses.org/list.html>
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Mike Cantwell <mcantw...@ieee.org<mailto:mcantw...@ieee.org>>

For policy questions, send mail to:
Jim Bacher <j.bac...@ieee.org<mailto:j.bac...@ieee.org>>
David Heald <dhe...@gmail.com<mailto:dhe...@gmail.com>>

________________________________

To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1

________________________________

WILO SE
Wilopark 1, 44263 Dortmund
Amtsgericht Dortmund, HRB 21356
www.wilo.com<http://www.wilo.com/>

Vorstand/Executive Board:
Oliver Hermes (President & CEO), Dr. Patrick Niehr, Georg Weber, Mathias Weyers
Aufsichtsratsvorsitzender/Chairman of the Supervisory Board:
Lars Roßner

Dieses Dokument ist vertraulich zu behandeln. Die Weitergabe sowie 
Vervielfaeltigung, Verwertung und Mitteilung seines Inhalts ist nur mit unserer 
ausdruecklichen Genehmigung gestattet. Alle Rechte vorbehalten, insbesondere 
fuer den Fall der Schutzrechtsanmeldung.
This document has to be treated confidentially. Its contents are not to be 
passed on, duplicated, exploited or disclosed without our expressed permission. 
All rights reserved, especially the right to apply for protective rights.

Datenschutz ist uns wichtig! Daher gehen wir sorgsam mit Ihren Daten um. 
Näheres finden Sie in unseren 
Datenschutzhinweisen<https://wilo.com/de/de/Wilo-Deutschland/Datenschutzhinweise.html>.
 Sollten Sie keinen Kontakt mehr mit uns wünschen, wenden Sie sich bitte an 
datensch...@wilo.com<mailto:datensch...@wilo.com>.
Data Protection is important to us! Therefore we handle your data carefully. 
Please find the details in our Note on Data 
Protection<https://wilo.com/de/de/Note-on-data-protection.html>. If you do not 
wish to stay in touch with us anymore, please write to 
data-priv...@wilo.com<mailto:data-priv...@wilo.com>.
________________________________

WILO SE
Wilopark 1, 44263 Dortmund
Amtsgericht Dortmund, HRB 21356
www.wilo.com

Vorstand/Executive Board:
Oliver Hermes (President & CEO), Dr. Patrick Niehr, Georg Weber, Mathias Weyers
Aufsichtsratsvorsitzender/Chairman of the Supervisory Board:
Lars Roßner

Dieses Dokument ist vertraulich zu behandeln. Die Weitergabe sowie 
Vervielfaeltigung, Verwertung und Mitteilung seines Inhalts ist nur mit unserer 
ausdruecklichen Genehmigung gestattet. Alle Rechte vorbehalten, insbesondere 
fuer den Fall der Schutzrechtsanmeldung.
This document has to be treated confidentially. Its contents are not to be 
passed on, duplicated, exploited or disclosed without our expressed permission. 
All rights reserved, especially the right to apply for protective rights.

Datenschutz ist uns wichtig! Daher gehen wir sorgsam mit Ihren Daten um. 
Näheres finden Sie in unseren 
Datenschutzhinweisen<https://wilo.com/de/de/Wilo-Deutschland/Datenschutzhinweise.html>.
 Sollten Sie keinen Kontakt mehr mit uns wünschen, wenden Sie sich bitte an 
datensch...@wilo.com<mailto:datensch...@wilo.com>.
Data Protection is important to us! Therefore we handle your data carefully. 
Please find the details in our Note on Data 
Protection<https://wilo.com/de/de/Note-on-data-protection.html>. If you do not 
wish to stay in touch with us anymore, please write to 
data-priv...@wilo.com<mailto:data-priv...@wilo.com>.

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>
_________________________________________________
To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1

Reply via email to