Hi Brian,

We used hioki in the past. The problem is how to calibrate internal circuit up 
to 1MHz. You can calibrate voltage up to 1MHz, however instrument shows current 
and this is not an easy task. Many labs have the same issue.

Calibration up to 1MHz is required by IEC 60990 standard and by CB scheme. See 
OD-5013.

Instrument shall allow rms and peak measurements.

One customer was using in production Chroma. They were not very happy with it 
since they could not integrate it into their system.

We prefer now to use normal network circuit. Our metrology department built 
different networks for different applications and  made ISO 17025 calibration 
up to 1MHz. You need to use osc with the network. This is good since you can 
see the signal on osc. It is not so good for medical measurement since you need 
to do a lot of measurements manually. Therefore we built a second switch box, 
which makes measurement easier.

Of course, such method is not useful if you are doing routine tests in 
production.

Best regards,
Boštjan


Boštjan Glavič
Vodja laboratorija za elektroniko / Head of Laboratory
Varnost in elektromagnetika / Safety and Electromagnetics

SIQ Ljubljana, Mašera-Spasićeva ulica 10,
SI-1000 Ljubljana, Slovenija
T: +386 1 4778 265
M: +386 41 391 283
[email protected]<mailto:[email protected]>; 
www.siq.si<https://www.siq.si/>

[SIQ]<https://www.siq.si/>
To elektronsko sporočilo je namenjeno izključno naslovniku. Pravno 
obvestilo.<https://www.siq.si/pravno-obvestilo-za-prejemnike-elektronske-poste/>
The information transmitted by this email is intended only for the person or 
entity to which it is addressed. 
Disclaimer.<https://www.siq.si/en/legal-notice-for-e-mail-recipients/>

[SIQ Facebook]<https://www.facebook.com/SIQLjubljana/>  [SIQ LinkedIn] 
<https://www.linkedin.com/company/siq>   [SIQ YouTube] 
<https://www.youtube.com/channel/UCURzJFvP5CDVtL_vdvWjbQA/featured>








From: Brian Kunde <[email protected]>
Sent: Wednesday, July 31, 2024 2:51 PM
To: [email protected]
Subject: [PSES] Touch/Leakage Current Test Equipment

CAUTION: This email originated from outside of our organisation. Do not click 
links or open attachments unless you recognise the sender and know the content 
is safe.

Greetings to all.

I would like to hear your recommendations for test equipment for performing 
Touch/Leakage Current.  If it is ok to say, we currently have an ED&D LT-952 
with the 30A option and it has been great. However, we would like to get a 
second tester that has a LAN interface so we can automate the testing and data 
collection.

We like the Meter Reference feature so the measurement is not only referenced 
to ground, but also to Line 1 or Line 2.  We also like the 30 amp ability since 
we do test products that are rated between 20A and 30A.  Even higher current 
rating would be nice.

I would love to hear your recommendations for a good Leakage Current Tester.  
If it is not ok to post here, please send it directly to me.

Validation and Calibration:  On another related topic,  how do you all Validate 
your Leakage Current Tester?  Does anyone self calibrate their own Tester?  How 
hard can this be?  The company we use for calibration says they cannot 
calibrate it so they send it to another company, resulting in my tester being 
gone for a long period of time. So I am trying to find an acceptable solution.

I also have the Human Body Model circuit as a separate box that I connect to a 
calibrated DMM to make measurements. This box is easy to validate just with a 
meter.  Our NRTL Inspector says we can just measure the resistance and 
capacitance of the Box ourselves and do not have to send it out for calibration 
(that is what they do). So can I use this Box to compare the results I get with 
my ED&D Leakage Current Tester and call it good?

Thank you very much.
The Other Brian
________________________________

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
[email protected]<mailto:[email protected]>
All emc-pstc postings are archived and searchable on the web at:
https://www.mail-archive.com/[email protected]/<https://www.mail-archive.com/[email protected]/%20>

Website: https://ewh.ieee.org/soc/pses/
Instructions: https://ewh.ieee.org/soc/pses/list.html (including how to 
unsubscribe)<https://ewh.ieee.org/soc/pses/list.html>
List rules: https://ewh.ieee.org/soc/pses/listrules.html

For help, send mail to the list administrators:
Mike Sherman at: [email protected]<mailto:[email protected]>
Rick Linford at: [email protected]<mailto:[email protected]>

For policy questions, send mail to:
Jim Bacher at: [email protected]<mailto:[email protected]>

________________________________

To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
[email protected]

All emc-pstc postings are archived and searchable on the web at:
https://www.mail-archive.com/[email protected]/

Website:  https://ewh.ieee.org/soc/pses/
Instructions:  https://ewh.ieee.org/soc/pses/list.html (including how to 
unsubscribe)
List rules: https://ewh.ieee.org/soc/pses/listrules.html

For help, send mail to the list administrators:
Mike Sherman at: [email protected]
Rick Linford at: [email protected]

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
_________________________________________________
To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1

Reply via email to