We evaluate the safety of Laboratory Equipment by applying the IEC/EN/UL
61010-1 safety standard. For equipment that has accessable moving parts,
such as sample loaders, we apply section 7.3 "Moving Parts".  Under 7.3.4,
it states,

   - " The physical levels specified below are *not considered to be
   hazardous.*"
   - " The maximum tolerable continuous contact pressure is 50 N/cm2 , with
   a maximum force of 150 N.


So for continuous contact pressures greater than 50 N/cm2, we have to apply
the Risk Assessment in section 7.3.3.

The First step in the Risk Assessment is to define the Severity of the
hazard or injury.  You have two choices; Moderate and Serious.  The
standard does attempt to define the potential injury, but this is very
subjective. How do you test for a dismembered body part?  My Engineers want
numbers, not guesswork.

I was hoping there are numbers, such as a Contact Pressure that would draw
a line between Moderate and Serious.  For pinch, crush, and maybe shear
hazards, there should be a way to come up with a N/cm2 number or test that
can make this determination.  Are there such numbers, rules of thumb,
industry standards, that would better define the difference?  Having
something like this would really help.

Thanks to all,
The Other Brian

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
[email protected]

All emc-pstc postings are archived and searchable on the web at:
https://www.mail-archive.com/[email protected]/

Website:  https://ewh.ieee.org/soc/pses/
Instructions:  https://ewh.ieee.org/soc/pses/list.html (including how to 
unsubscribe)
List rules: https://ewh.ieee.org/soc/pses/listrules.html

For help, send mail to the list administrators:
Mike Sherman at: [email protected]
Rick Linford at: [email protected]

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
_________________________________________________
To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1

Reply via email to