1 and 2 performed similarly when we compared them.  3 is not a good idea, as 
GRE images have signal dropout that is asymmetric and will mess up the 
registration that underlies the reversed phase encoding direction method.

Matt.

From: 
<hcp-users-boun...@humanconnectome.org<mailto:hcp-users-boun...@humanconnectome.org>>
 on behalf of Chris Filo Gorgolewski 
<krzysztof.gorgolew...@gmail.com<mailto:krzysztof.gorgolew...@gmail.com>>
Date: Monday, October 5, 2015 at 6:25 PM
To: "Harms, Michael" <mha...@wustl.edu<mailto:mha...@wustl.edu>>
Cc: Hcp Users 
<hcp-users@humanconnectome.org<mailto:hcp-users@humanconnectome.org>>
Subject: Re: [HCP-Users] Use of spin-echo EPI scans for distortion correction 
of fMRI data

Thank you for the clarification!

A quick follow up question. Do you know of any systematic comparison of:

1) standard fieldmap (two pairs of phase and magnitude images)
2) fieldmap derived from a pair of spin echo images with opposite phase 
encoding directions
3) fieldmap derived from a pair of gradient echo images with opposite phase 
encoding directions (for example the reference images - SBRef)

?

Best regards,
Chris

On Mon, Oct 5, 2015 at 4:04 PM, Harms, Michael 
<mha...@wustl.edu<mailto:mha...@wustl.edu>> wrote:

Yes, that is correct.  (The SBRef for the BOLD scans are gradient-echo images).

--
Michael Harms, Ph.D.
-----------------------------------------------------------
Conte Center for the Neuroscience of Mental Disorders
Washington University School of Medicine
Department of Psychiatry, Box 8134
660 South Euclid Ave. Tel: 314-747-6173<tel:314-747-6173>
St. Louis, MO  63110 Email: mha...@wustl.edu<mailto:mha...@wustl.edu>

From: Chris Filo Gorgolewski 
<krzysztof.gorgolew...@gmail.com<mailto:krzysztof.gorgolew...@gmail.com>>
Date: Monday, October 5, 2015 5:13 PM
To: Hcp Users 
<hcp-users@humanconnectome.org<mailto:hcp-users@humanconnectome.org>>
Subject: [HCP-Users] Use of spin-echo EPI scans for distortion correction of 
fMRI data

Hi,
After reading the preprocessing scripts (see 
https://github.com/Washington-University/Pipelines/blob/master/fMRIVolume/scripts/DistortionCorrectionAndEPIToT1wReg_FLIRTBBRAndFreeSurferBBRbased.sh#L271)
 and the paper (see 
http://www.sciencedirect.com/science/article/pii/S1053811913005053) I figured 
that HCP is using a pair of spin-echo EPI scans to estimate a fieldmap later 
used to correct fMRI data for distortions (in contrast to using a pair of SBRef 
images). We had a bit of a discussion in the lab about this and I just wanted 
someone from the HCP team to confirm this. Thanks in advance!

Best,
Chris

_______________________________________________
HCP-Users mailing list
HCP-Users@humanconnectome.org<mailto:HCP-Users@humanconnectome.org>
http://lists.humanconnectome.org/mailman/listinfo/hcp-users

________________________________
The materials in this message are private and may contain Protected Healthcare 
Information or other information of a sensitive nature. If you are not the 
intended recipient, be advised that any unauthorized use, disclosure, copying 
or the taking of any action in reliance on the contents of this information is 
strictly prohibited. If you have received this email in error, please 
immediately notify the sender via telephone or return mail.


_______________________________________________
HCP-Users mailing list