Great, thank you for the clarification.

Best,
Ben

On Wed, May 11, 2016 at 9:08 AM, Harms, Michael <[email protected]> wrote:

>
> Hi Ben,
> The following replies to your questions are from Steen Moeller (cc’ed) at
> CMRR.
>
> 1/ the reconstruction for all HCP data is performed using the slice-grappa
> algorithm. There was a change in Q3 that improved on the implementation of
> it relative to some low-level sequence detail concerning the use of the
> blipping scheme in blipped-CAIPI reference data. The choosen algorithm
> however has stayed the same. The SpSg algorithm has not been used, but is
> available in new versions of the sequence.
>
> 2/ the individual coil data are not available. If this is only needed for
> a single session, that can then be made available for a representative
> dataset. If so then email me directly to make this available.
>
> cheers,
> --
> Michael Harms, Ph.D.
> -----------------------------------------------------------
> Conte Center for the Neuroscience of Mental Disorders
> Washington University School of Medicine
> Department of Psychiatry, Box 8134
> 660 South Euclid Ave. Tel: 314-747-6173
> St. Louis, MO  63110 Email: [email protected]
>
> From: <[email protected]> on behalf of Benjamin Risk <
> [email protected]>
> Date: Monday, May 9, 2016 at 10:41 PM
> To: "Glasser, Matthew" <[email protected]>
> Cc: "[email protected]" <[email protected]>
> Subject: Re: [HCP-Users] Clarification in change in fMRI reconstruction
> algorithm
>
> They didn't use in-plane GRAPPA but they used a multi-band variant on
> GRAPPA. Setsompop 2012 is called Slice-GRAPPA. I am asking about the
> details of the multiband reconstruction.
>
> Thanks,
> Ben
>
> On Mon, May 9, 2016 at 10:51 PM, Glasser, Matthew <[email protected]>
> wrote:
>
>> fMRI didn’t use GRAPPA, rather it was an update to the multi-band
>> reconstruction.
>>
>> Peace,
>>
>> Matt.
>>
>> From: <[email protected]> on behalf of Benjamin Risk
>> <[email protected]>
>> Date: Monday, May 9, 2016 at 12:55 PM
>> To: "[email protected]" <[email protected]>
>> Subject: [HCP-Users] Clarification in change in fMRI reconstruction
>> algorithm
>>
>> Dear HCP experts,
>>
>> I have two questions:
>>
>> 1) I understand that the reconstruction algorithm for fMRI changed with
>> the Q3 data release. I believe there are three main GRAPPA reconstruction
>> approaches for multiband imaging: SENSE/GRAPPA from Blaimer et al (2006);
>> Slice-GRAPPA from Setsompop et al (2012); and Split-Slice-GRAPPA from
>> Cauley et al (2014). Do the HCP reconstruction algorithms correspond to
>> these methods? From my reading of Ugurbil et al 2013, p. 89, I thought the
>> change was from SENSE/GRAPPA to Slice-GRAPPA, but I now see that the
>> overview document
>> http://www.cmrr.umn.edu/multiband/Multi-Band_C2P_Instructions_R013a.pdf
>> includes a citation to Cauley.
>>
>> 2) I ultimately want to simulate data generated by coil sensitivities
>> similar to the HCP scanners and head coils. Are single-band calibration
>> data for each coil available?
>>
>>
>> Thank you,
>> Ben Risk
>> Postdoc at SAMSI / Department of Biostatistics UNC Chapel Hill
>>
>> _______________________________________________
>> HCP-Users mailing list
>> [email protected]
>> http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>
>>
>> ------------------------------
>>
>> The materials in this message are private and may contain Protected
>> Healthcare Information or other information of a sensitive nature. If you
>> are not the intended recipient, be advised that any unauthorized use,
>> disclosure, copying or the taking of any action in reliance on the contents
>> of this information is strictly prohibited. If you have received this email
>> in error, please immediately notify the sender via telephone or return mail.
>>
>
> _______________________________________________
> HCP-Users mailing list
> [email protected]
> http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>
>
> ------------------------------
>
> The materials in this message are private and may contain Protected
> Healthcare Information or other information of a sensitive nature. If you
> are not the intended recipient, be advised that any unauthorized use,
> disclosure, copying or the taking of any action in reliance on the contents
> of this information is strictly prohibited. If you have received this email
> in error, please immediately notify the sender via telephone or return mail.
>

_______________________________________________
HCP-Users mailing list
[email protected]
http://lists.humanconnectome.org/mailman/listinfo/hcp-users

Reply via email to