HI Haifeng,

Suppose you measure Fe foil three times, and all three scans overlap. They are aligned. But suppose the edge positions are at 7116 eV instead of 7112 eV. The calibration is off. You can adjust to it, but, as Carlo mentioned, it is really an angle correction, not simply a shift in energy. It is better to have an accurate calibration to begin with, then aligned and
calibrated would be the same.

Note: Choice of target calibration can differ - some use the X-ray data booklet values
http://xdb.lbl.gov/
and some use the Kraft  et al. values
https://aip.scitation.org/doi/abs/10.1063/1.1146657

Just state which calibration you are using when presenting/publishing.

-R.


On 2018-07-27 10:32 AM, Haifeng Li wrote:
Hi, Robert,

Thanks for your help.

One more question. What is the real difference between calibration and alignment? Do they have specific meaning?

Thanks,

Haifeng

On Sun, Jul 22, 2018 at 5:45 PM, Robert Gordon <ragor...@alumni.sfu.ca <mailto:ragor...@alumni.sfu.ca>> wrote:

    Hi Haifeng,

    Here's my take on what you described:

    Two sample to be studied at the same edge using the same
    reference. Three scans on A and three on
    B are done with simultaneous reference.

    First: compare the three reference scans for A. If they agree,
    then compare the data scans. If they also look
    similar (i.e. no evidence of changing in the beam), you can merge
    the three data scans
    on A right away. If the references don't agree, then you determine
    by how much they differ (how much a correction
    would be needed to bring them into alignment) and apply that same
    correction to the data before merging.
    It is not meaningful to merge data that is not aligned. (If the
    sample scans show changes from scans 1  to 3,
    then you need to rethink how to do the measurements)

    Repeat for B.

    Now compare the merged (corrected beforehand if necessary)
    references for A and B. If they agree,
    you can compare data for A and B (merged) directly. If they do
    not, determine how much one reference
    differs from the other and apply that same correction to, say, B,
    that brings its reference into alignment with
    A's reference, and then compare A and B

    You align the references between samples to the same value in
    order to do a meaningful comparison
    between them. If you report energy positions of features in your
    near-edge spectra, the reader
    would need to know to what energy those positions are referenced.

    When possible, I recommend references that have tabulated edge
    values (i.e. metal foils). A reference
    need not be the same edge as the one being studied. For arsenic,
    as an example, the gold L3-edge is
    quite close to the As K-edge and serves as a good reference. If
    not possible to use a metal foil, use a
    reference that another interested researcher could readily obtain
    or has used. This allows for
    comparison of reproducibility. If your reference looks nothing
    like literature, you may have a problem
    with the beamline or in how you processed the data. This should be
    one of the first things you
    check when you start taking data at the beamlne.

    You should also note how the beamline was calibrated when you did
    the measurements.

    regards,
    Robert


    On 2018-07-22 1:21 PM, Haifeng Li wrote:
    Dear ALL,

    I am a beginner in Athena. Recently I got the spectra and I am
    confusing about the data calibration and alignment. The manual
    shows that calibrate the reference data of one scan and align
    other reference data to that calibrated one.

    Here I want to show examples. I have two samples A and B, Each
    sample has three scans with the corresponding reference data. For
    sample A, 1st scan is calibrated and the other two scans are
    aligned to 1st scan. Then merge them into merged A. The same
    procedures for sample B and get merged B. If I want to compare
    XANES of sample A and B, do I need to align the merged reference
    data between A and B? If so, why? My understanding is that all
    scans (original data and merged data) in sample A and B are
    calibrated to standard edge energy. Why do they need to align?

    I appreciate your help.

    Thanks,

    Haifeng


    _______________________________________________
    Ifeffit mailing list
    Ifeffit@millenia.cars.aps.anl.gov
    <mailto:Ifeffit@millenia.cars.aps.anl.gov>
    http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
    <http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit>
    Unsubscribe:http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
    <http://millenia.cars.aps.anl.gov/mailman/options/ifeffit>


    _______________________________________________
    Ifeffit mailing list
    Ifeffit@millenia.cars.aps.anl.gov
    <mailto:Ifeffit@millenia.cars.aps.anl.gov>
    http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
    <http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit>
    Unsubscribe:
    http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
    <http://millenia.cars.aps.anl.gov/mailman/options/ifeffit>




_______________________________________________
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit

_______________________________________________
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit

Reply via email to