I note that the SCC SG04 has a 2.5v excitation voltage and the force
unit indicated is recommended at 10v excitation.

There is a 10v excitation unit which is the SCC SG24.
Anyway with a lower excitation voltage you get less mV output for a
force input which could mean that you loose ability to measure the
desired force as it would be 4 times smaller than recommended.

The smallest force unit is 50lb force full scale. This is quite a lot
for pressing a syringe isn't it? An alternative way of looking at this
is that you don't get enough resolution in the measurement if you
sensor range is too big.

I could have the wrong end of the stick for your application?

Best advice is to perform a simulation with some wieghts.