If you are intent on using thermocouples, then I would highly recomend
signal conditioning, as at room temperatures, the millivolt output of
almost all thermocouples is very low (depends on the type of
thermocuple). Thermocouple-specific signal conditioning will also
provide for the required signal characterization (thermocouples are
not linear devices).
The other problem you mention, calibration, is also critical with
thermocouples especially if you are operating in the room temperature
and lower ranges. You have to have a method of compensating for the
reference junction temperature (assuming that it remains constant,
which in an outdoor environment, is questionable).
Having said all of the above, I would highly recomend that you elect
to use RTD's to measure your temperatures. The first gain is that you
can calibrate them with a fixed value resistor substitution. There is
no reference junction ambiguity/possible-variable to deal with.
You will require signal conditioning with and RTD, but the cost is
similar to that for a thermocouple, and you circumvent all of the
problems of thermocouples. The second big advantage with RTD's is that
you can wire to them with ordinary copper wire. With thermocouples,
you must use specific thermocouple wire (read expensive) from the
thermocouple terminals to the point at which you measure your
reference junction. If there is some distance to your thermocouple,
the cost of wire can outstrip the cost of the thermocouple itself!
Which ever path you decide to take, NI has good signal conditioners
listed on their site/in their catalog.
Hope this is of some helpt to you.
Dave