Brett,

If your apparatus is already preloaded and you want to use the
preloaded value as the initial condition reference point, you will
want to use either hardware or software offset nulling.

Offset nulling is a method used to offset a series of measurements
with respect to an initial condition (I.C.). If your load cell is
preloaded, you can reference all future load conditions with respect
to the preloaded I.C.

On a hardware level, offset nulling is possible if you're using a
SCXI-1520 or PXI-4220. In this case, you will turn on offset nulling
in LabVIEW or Measurement Explorer.

On a software level, offset nulling is accomplished by taking a series
of initial condition readings from the load cell. Then, you average
these readings and subtract this value from all subsequent readings in
order to offset the initial preloaded value.

If the load cell is initially unloaded, you will not want to
incorporate any form of offset nulling (assuming I.C. voltage reading
is 0V).

Now, I understand you're looking to have the arrow blink up and down
with loading and unloading, respectively. In order to accomplish this
effect, you will want to incorporate a series of Boolean tests. First,
measure the initial load reading (with or without the offset nulling
dependent on whether/not preloading is existent). Now, use the
imbedded LabVIEW Boolean tests to measure whether each subsequent
reading is greater than the previous measurement. Make a case
structure with the switch condition 'true' and 'false' dependent on
the Boolean test condition. In the 'true' condition structure, place
the property node coding (answered in your last post) that causes the
up arrow to appear. In the 'false' condition structure, place the
property node coding for the down arrow to appear. If you need
instruction on case structure programming, please refer to this <a
href=
"http://sine.ni.com/apps/we/niepd_web_display.display_epd4?p_guid=BD18D9452FC21649E034080020E74861&p_node=DZ52000_US&p_source=External";>
link </a>.

In effect, your program should now display the up and down arrow with
respect to increasing and decreasing loading of the load cell.

If you store all of the signal values to an array as they are
acquired, you can use the Array Max & Min function to determine the
signal value respective to the maximum displacement. To access this
function, [right-click] on your LabVIEW block diagram, click [All
Functions]>>[Array]>>[Array Max & Min].