Two possible answers for you, based on traditional NI-DAQ.  Either
should work, though I'd recommend #2 unless there's a specific reason
you need to do #1.

1. Add a call to 'Counter Set Attribute.vi' to the configuration
chain.  The attribute and the attribute value type should be 'up down'
and 'hardware' to specify hardware control of count direction.
   For this to work, the call to 'Counter Group Config' should specify
   'event counting', not 'position measurement.'  Note that this prevents
   you from using the encoder's z-index pulse (if it has one) to hardware
   reset the count value.

2. Bypass the LS7084 and feed the quadrature A&B signals directly to
your counter (assuming they are already TTL compatible).  Your 6601 is
designed to interface directly to quadrature signals.  They just need
to be wired to the counter's default Source and Up/Down pins.
   With this method, your call to 'Counter Group Config' should
   specify 'position measurement'.  There's a handy config vi under the
   examples directory (look for something like
   "examples\daq\counter\ni_tio.llb") with a name like 'Counter Encoder
   (NI-TIO).vi' that'll help you configure for direct encoder position
   measurement.
   This method can also allow you to hardware reset your count value
   on a z-index pulse (wired to the counter's default Gate input).