LIFEPAK 20 and LIFEPAK 20e External Defibrillator/Monitors by
Physio-Control Inc.: Class I Recall Due To Power Supply Failure

Audience: Emergency Medicine, Cardiology, Risk Manager

ISSUE: A failure on the power supply assembly can result in either "No
DC power" or "No DC or AC power". A failure of DC (battery) power can
result in the inability to deliver defibrillation therapy if the
device will not turn on using DC (battery) power and no AC (line)
power is available.

BACKGROUND: The LIFEPAK 20 and LIFEPAK 20e defibrillator/monitor is
designed for use by trained medical personnel in hospitals and clinic
settings to monitor patient heart rhythms and to treat patients
experiencing cardiac arrest. Approximately 42,943 devices were
distributed worldwide between September 16, 2002 and September 27,
2007. These devices were manufactured from July 31, 2002 to September
19, 2007.

RECOMMENDATION: All affected power supplies will be updated. Customers
are advised to keep the defibrillators in service and follow
recommended daily Operator Checklist steps while service updates are
scheduled. See Recall Notice for contact information.

Read the MedWatch safety alert, including a link to the FDA Class I
Recall Notice, at:

<http://www.fda.gov/Safety/MedWatch/SafetyInformation/SafetyAlertsforHumanMedicalProducts/ucm217980.htm>

-- 