[Over a decade ago, a hospital in the Texas Medical Center killed 
several patients due to a firmware bug in a therapeutic radiation 
device. The bug was triggered only when a combination of certain valid, 
but rarely occurring, settings were selected in the device control 
software. When the bug was triggered the device delivered more than 100 
times the intended dose of radiation, killing the patient. It took a 
while to figure out that it was a software bug, rather than operator 
error, that was producing the unintended & lethal result. Before  
investigators got a handle on this several patients had been killed and 
several others were seriously injured.]

‘Known Software Bug’ Disrupts Brain-Tumor Zapping

By Kevin Poulsen
Wired.com

October 16, 2009  |  4:19 pm 

http://www.wired.com/threatlevel/2009/10/gamma/


The maker of a life-saving radiation therapy device has patched a 
software bug that could cause the system’s emergency stop button to fail 
to stop, following an incident at a Cleveland hospital in which medical 
staff had to physically pull a patient from the maw of the machine.

The bug affected the Gamma Knife, a device resembling a CT scan machine 
that focuses radiation on a patient’s brain tumor while leaving 
surrounding tissue untouched. A patient lies down on a motorized couch 
that glides into a chamber, where 201 emitters focus radiation on the 
treatment area from different angles. The patient wears a specialized 
helmet screwed onto his skull to ensure that his head doesn’t move and 
expose the wrong part of the brain to the machine’s pinpoint 
tumor-zapping beams.

Positioning is vital in the procedure, so when the couch moved out of 
position during a treatment at an university hospital in Cleveland last 
December, staffers hit the “emergency stop” button, expecting the couch 
to pull the patient out of the Gamma Knife, and the radiation shields at 
the mouth of the machine to automatically close. Instead, according to a 
report eventually filed with the Nuclear Regulatory Agency, nothing 
happened.

“Staff had to manually pull out the couch from the Gamma Knife and 
manually close the doors to the Gamma Knife to shield the source,” reads 
the report, which states that neither the patient nor the workers were 
harmed. “Radiation exposure to all individuals involved with the 
incident was minimal.”

When the hospital called the company that makes the Gamma Knife, it 
learned that there was a “known software bug problem” affecting the 
unit’s couch sensors. Known, anyway, to the company, Stockholm-based 
Elekta AB.

“Elekta was aware of the software ‘bug’ at the time of the December 2008 
event and had implemented actions to correct the ‘bug’ in a future 
software release,” says Thomas Valentine, director of quality assurance 
and regulatory affairs for the Elekta’s U.S. arm, in an e-mail.

Since then, he adds, “The ‘bug’ has been corrected in software upgrades 
that have been implemented to all of the affected sites in the U.S. The 
U.S. NRC was notified of the completed status of software upgrades to 
correct the identified ‘bug’.”

We don’t know why “bug” is in quotes; surely this wasn’t a feature. In 
any case, Valentine says the Ohio incident was the only one of its kind 
“in the U.S.,” and that the bug had been triggered by an unusual 
combination of events.

It’s worth noting that Gamma Knife has been used to treat about 
half-a-million people without trouble. But the bug is another reminder 
that increasingly smart medical devices are susceptible to the same kind 
of programming errors that have long afflicted less critical 
applications. This week, the Los Angeles Times reported that 
Cedars-Sinai Medical Center made an error while tinkering with the 
settings on a hospital CT scan machine in February 2008, resulting in 
about 80 patients temporarily losing patches of hair due to radiation 
overdoses.

The most notorious medical bug was a “race condition” in the software 
powering the Therac-25 medical accelerator in the 1980s, which resulted 
in three patients dying from radiation overdoses from 1985 to 1987.

The far less serious Gamma Knife bug came to light in the medical 
community four months after the incident, after an inspector with the 
Ohio Department of Health spotted a discussion of the Cleveland incident 
in the minutes of the hospital’s radiation safety committee meeting. The 
hospital is not named in public filings, but had apparently failed to 
report the incident to the state, as required by law.

The Department of Health went on to report the matter to the NRC, which 
in April alerted hospitals around the country in an e-mail to its 
medical mailing list.

-- 
================================
George Antunes, Political Science Dept
University of Houston; Houston, TX 77204 
Voice: 713-743-3923  Fax: 713-743-3927
Mail: antunes at uh dot edu

***********************************
* POST TO [email protected] *
***********************************

Medianews mailing list
[email protected]
http://lists.etskywarn.net/mailman/listinfo/medianews

Reply via email to