Dear Ron,

Your original dataset provided clearance estimates for 100 patients, which were 
all within the range 5 to 30 L/h.  When you simulated 4800 clearance values (= 
2 cross-over observations/subject x 12 subject/trial x 200 trials), I would 
expect to see a wider range of clearance values in the simulation dataset than 
in the original analysis, simply because there are many more observations in 
the simulation dataset.

However, arbitrarily truncating clearance to the range of 5 to 30 L/h may not 
be wise.  Consider the subject whose simulated clearance values are 30 L/h and 
32 L/h, a 6.7% difference.  If both of these values are truncated to 30 L/h, 
then there is zero within-patient difference in clearance values (and zero 
within-patient difference in AUC), and the truncation procedure both (1) 
reduces the mean within-subject treatment difference and (2) reduces the 
within-subject treatment variability.  Either one of these effects would 
arbitrarily increase the statistical power of attaining bioequivalence.

Statistical power for a bioequivalence test depends on the projected mean 
within-subject treatment difference (say <=5%), the projected mean 
within-subject variability (estimated in your original population PK analysis), 
and the sample size.  Instead of truncating clearance values, I think you 
should examine the effects of larger sample sizes on the statistical power; 
maybe run simulations with 16, 24, 36, and 48 subjects per study.  In reality, 
very few bioequivalence trials will be successful with only 12 subjects.

Good luck,
Steve

Steven Troy
Sr. Director,
Clinical Pharmacology and Pharmacokinetics
Shire Pharmaceuticals


-----Original Message-----
From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On Behalf Of Ron Mathôt
Sent: Tuesday, April 22, 2008 3:18 PM
To: [email protected]
Subject: [NMusers] truncation & simulation

Dear NONMEM users,

Currently I am working on the simulation of a bio-equavalence trial. For 
the reference compound a population PK model has been derived on basis 
of data from 100 patients. Values for between-and within-patient 
variability are available for all PK parameters. The simulation 
comprises a randomized cross-over study with 12 patients taking  the 
test and reference compound.  Two-hunderd trials are simulated and 
summarized. During the simulations I noticed that truncation of the 
simulated of PK parameters significantly influences the power of the 
study to confirm bio-equivalence. For instance truncation of simulated 
oral clearances of both compounds from a range of 1-300 L/hr to 5 - 30 
L/hr doubled the number of positive trials (due to decreased within- 
patient variability). Post-hoc estimates form the popPK study indicated 
that clearance values of the reference compound are all within the 
latter range of 5 to 30 L/hr. I expect that oral clearance of the test 
compound will not differ more than 5% from the reference compound. In my 
opinion simulation of trials with the smallest range will produce more 
reliable estimates of the power to detect bio-equivalence.

I would greatly appreciate your comments on this subject.
Best regards,

Ron Mathôt

Department of Hospital Pharmacy and Clincal Pharmacology
Erasmus University Medical Center
Rotterdam
The Netherlands




______________________________________________________________________
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email 
______________________________________________________________________

Please consider the environment before printing this e-mail



This email and any files transmitted with it are confidential and
may be legally privileged and are intended solely for the use of
the individual or entity to whom they are addressed. If you are 
not the intended recipient please note that any disclosure, 
distribution, or copying of this email is strictly prohibited and may 
be unlawful.  If received in error, please delete this email and any 
attachments and confirm this to the sender.

www.shire.com

Reply via email to