STP response:

We perform a daily check with at least 10 TLDs irradiated to Cs-137 with 
approximately 300-500 mrem of dose prior to reading the TLDs to check the TLD 
reader response for the photon counter. In addition, the tray has run controls 
in the 1st, 26th, and 49th position in the tray and a back ground controls 
(less then 10 mrem) in the 2nd, 27th, and 50th position with the remaining 
slots for personnel TLDs to ensure that the TLD reader is still responding 
correctly while reading personnel TLDs. We participate in an inter-comparison 
test (Battelle or NVLAP) on an annual basis and have QC tests performed each 
quarter which we irradiate the TLDs ourselves to various ranges of gamma, beta, 
and neutron mixtures. Since we are on a semi annual TLD change out period we 
also run a set of blind controls at least once during each processing cycle. 
All of these tests are in addition to the TLD reader calibration which is 
verified or calibrated annually.

Laura Stoicescu
South Texas Project
[email protected]<mailto:[email protected]>
361-972-4095


________________________________
From: [email protected] [mailto:[email protected]] On Behalf Of 
Courtenay, Christopher C
Sent: Wednesday, October 27, 2010 11:46 AM
To: [email protected]
Subject: Powernet: Dosimetry Processing Practices

For those who process TLDs:

Do you perform a daily/weekly/monthly QC check on the TLD readers by 
irradiating TLDs to various levels (i.e. 3 sets of 5 TLDs to 10, 100, and 1000 
mrem or some other combination) and ensure TLD group statistics are within 
acceptable limits?  This would be in addition to any TLD reader calibrations 
and calibration checks/verifications.

Chris Courtenay
Engineer, Duke Energy
980-875-5005