Fermi is in the process of putting the Canberra iSolo alpha/beta counter into 
service.  We plan to use it only for determining compensated alpha activity on 
air sample filters.  It looks like we will only use it if the air sample has 
been taken on a PTFE filter: our testing has shown that iSolo radon rejection 
is greatly superior using a PTFE filter compared to that of glass fiber 
filters.  Another advantage is that the iSolo alpha background appears to be 
lower than that of other alpha counters, corresponding to a lower MDA and 
smaller required sample volumes.  The iSolo spectrum display also provides a 
qualitative indication of alpha energy which should be helpful in 
distinguishing natural from plant related alpha activity. 
 
We also have a surveillance every refueling outage in which we take high 
activity smears for sendout to GEL for analysis for specific alpha isotopes. 
 
 
Tom VanderMey, CHP
Principal Radiological Engineer
DTE Energy, Fermi 2
734-586-1539



-----owner-power...@hps1.org wrote: -----


To: "'powernet@hps1.org'" <powernet@hps1.org>
From: "Holmes, Stephen J" <stephen.hol...@cengllc.com>
Sent by: owner-power...@hps1.org
Date: 05/04/2012 12:41PM
Subject: Powernet: Benchmark: Alpha Spec


At Ginna, we are evaluating purchasing Alpha Spec equipment (such as the 
Canberra Alpha Analyst with Alpha Apex software). Part of our justification for 
this expense would be to reduce our reliance on independent laboratories for 
performing alpha characterization (i.e. per the EPRI alpha monitoring 
guidelines). We currently have Ludlum 3030Ps and MGPI ABPMs which use alpha 
spec in a limited manner to perform radon rejection, but no true alpha 
spectroscopy equipment. We would like to know what the rest of the industry is 
doing; any experience, comments, or feedback are appreciated.  
 
Do you have Alpha Spec capabilities at your facility?
 
 
If so, what equipment do you use?
 
 
What do you use the equipment for (e.g. radon rejection or a quantitative alpha 
spectrum of individual nuclides)?
 
 
 
If using for alpha spectrum analysis, what energy resolution can you achieve?
 
 
 
What type of sample preparation is required?
 
 
Any other equipment strengths/limitations or lessons learned you would like to 
share?
 
 
Thanks,
-Steve
 
Stephen J Holmes, PE
Sr. Plant Health Physicist, CENG-RE Ginna NPP
stephen.hol...@cengllc.com 
585-771-3577; 585-545-1661 (cell)
 
>>> This e-mail and any attachments are confidential, may contain legal,
professional or other privileged information, and are intended solely for the
addressee.  If you are not the intended recipient, do not use the information
in this e-mail in any way, delete this e-mail and notify the sender. -IP2

Reply via email to