>From Eberline's Procedure 10429-A164 Sheet 5 of 8

37. Set the ALARM DELAY control (R33 at the top center of the Logic Board) 
fully counterclockwise.

38. Check the time to alarm. Must be two seconds or less.

39. Set the ALARM DELAY fully clockwise.

40. Check the time to alarm. Must be from thirty to fifty seconds.

Vic Johnson
Central Calibration Facility
Entergy South
601.437.2161

Accomplishing the impossible means only that the boss will add it to your 
regular duties.

________________________________
From: [email protected] [mailto:[email protected]] On Behalf Of 
Roger Schanzenbach
Sent: Thursday, January 17, 2013 1:38 PM
To: [email protected]
Subject: Re: Powernet: AMS-3 Alarm delay times

Rob:

I don't recall the time range, but it should be testable by simply inserting a 
source several times at different delay settings.

In practice, the delay setting might prevent an alarm completely by smoothing 
out a short term event.  Such an outcome may or may not be desirable.

Roger Schanzenbach, CHP
(949) 361-3996


----- Original Message -----
From: Robert J Gunter<mailto:[email protected]>
To: [email protected]<mailto:[email protected]>
Sent: Thursday, January 17, 2013 10:30 AM
Subject: Powernet: AMS-3 Alarm delay times

Greetings All,

I am trying to find out how much delay is possible between reaching an alarm 
setpoint and the alarm actually going off.  Reading the manual indicates there 
is a delay circuit, and you can adjust it.  What it does not say is how long.  
Anyone out there know the possible range of delay time for alarm on an AMS-3?

Thanks,

Rob

Robert J. Gunter, MSc, CHP
CHP Consultants/CHP Dosimetry
www.chpconsultants.com
www.chpdosimetry.com
Toll Free: (888) 766-4833
Fax:  (866) 491-9913
Cel:  (865) 387-0028
[email protected]