Duke Energy Answers in Red:

In addition to calibration labeling, does your site have a label on in use 
portable equipment (friskers, survey meters) that is used to indicate 
instrument has been source checked?
At Duke Energy we have a Central Calibration Facility that attaches an Accepted 
(Calibration) Sticker at the time of calibration. If a meter has some 
restriction (i.e. a meter must be used with a specified detector serial number) 
we will also attach a restricted sticker indicating the use restrictions. We 
support 6 fleet stations. Of the fleet stations 4 do not currently use similar 
stickers as described above and two do. Of the two sites that use the stickers, 
one site uses them only for portable instruments, typically extended use 
instruments that remain in the field (i.e. friskers, area monitors, etc.). The 
second site uses the sticker on all extended use equipment both portable 
instruments and fixed contamination monitors. We are currently trying to 
standardize on one method and as you can imagine those who use the stickers do 
not want to eliminate them and those who do not use the stickers, do not want 
to start using them. For those that use the stickers the sticker is not the QA 
Record. All sites use a separate log to record the checks for QA Record 
purposes. The remaining 4 sites do not use the stickers.

All HP / RP equipment (PCMs SAMs, lab counters)?
One site some portables, primarily those instruments that remain in the field 
all of the time (i.e. Friskers). The other site adds the fixed contamination 
monitors, as well. The remaining 4 sites do not use the stickers.

Some equipment?
At two sites "YES" some equipment but not all equipment. The remaining four 
sites do not use these stickers.

Value added?
There are two schools of thought:

1)    The user or auditor can look at the sticker at any time and tell if the 
instrument has been source checked. Some view this as a plus so they use it.

2)   Instruments are checked at the frequency required and the log sheet is the 
QA Record. This is all the user needs to know and if anyone wants to verify 
then they go to the QA Record. These see the sticker as additional work and an 
unnecessary Human Error Trap (i.e. Technician forgets to initial the sticker).
As you can see either case can be argued as correct. The question then becomes 
is the value added real or perceived. Some would even argue that there is no 
value added. For this reason your questions are good and it will be good to see 
how others respond. Duke Energy does want to standardize this one way or the 
other across the fleet.

Separate record (QA record) for documenting source checks?
All sites have a log that is the QA Record. The Sticker is disposed of when all 
blanks are full.

Thanks,
Bill Faulkenberry, NRRPT
Senior Staff Scientist, EnRad Labs
Duke Energy

[EnRad Logo.jpg]

13339 Hagers Ferry Rd
Mail Stop MG03A2
Huntersville, NC  28078
Office: 980-875-6893
Mobile: 803-627-4832



From: [email protected] [mailto:[email protected]] On Behalf Of 
Benjy P Bertossi
Sent: Thursday, October 16, 2014 9:06 AM
To: [email protected]
Subject: Powernet: Labeling of HP / RP portable equipment



*** This is an EXTERNAL email. Exercise caution. DO NOT open attachments or 
click links from unknown senders or unexpected email. ***

In addition to calibration labeling, does your site have a label on in use 
portable equipment (friskers, survey meters) that is used to indicate 
instrument has been source checked?

All HP / RP equipment (PCMs SAMs, lab counters)?

Some equipment?

Value added?

Seperate record (QA record) for documenting source checks?


Thanks, Benjy


Benjy Bertossi
Fermi2 - RP Supervisor
734-586-4935
[email protected]<mailto:[email protected]>

Reply via email to