I would like to present a qualified disagreement with John and Jim for this 
specific case of WBC QA/QC.

I agree with Jim's comments that from a diagnostic standpoint more peaks are 
better and low energy peaks could be useful.   But modern Ge systems are quite 
reliable and a competent spectrum reviewer can identify most all common failure 
modes by simply looking at the WBC spectrum and result.  My disagreement is 
using Eu-152 as the primary QC source in a program designed to prove to 
outsiders that the plant working environment has been safe.

A very important purpose of the WBC QA program is to prove to an outside 
observer [regulator, litigant] that the system was operating properly when it 
is in use in the normal mode - i.e. to count people.  That requires the same 
nuclide library and analysis process be in place during the QC count as is used 
for people counts.  With NaI systems [e.g. FastScan] putting Eu152 in the 
People library is asking for trouble.  Yes with HPGe systems you could get away 
with it - although it can cause problems if there is also Co-57 in your 
library.  So unless you have Eu-152 in your plant inventory in sufficient 
concentrations as to have a meaningful contribution to internal body activity, 
it should not be in the people library.  NOTE:  John at Oyster Creek perhaps 
could soon be in that category as neutrons + concrete make Eu-152.  Even of one 
chooses to use Eu-152 as a check source you should track each of several lines 
separately, not simply nuclide activity which is normally the weighted average 
of all lines.

Eu-152 is a very practical and useful source for calibrations [energy, FWHM, 
efficiency], although there are some quirks for NaI systems that must be 
avoided.  But we use it routinely for that purpose.  However we only supply and 
recommend Co-60 and Cs-137 as the standard check source for our WBC products.  
Those two nuclides in a QC program demonstrate to an outside observer 
[generally non-technical] the consistent stability of the system when measuring 
two of the major potential dose contributors in NPPs.  That is a far easier 
task than convincing a skeptical non-technical observer [think lawyer] why a 
non-NPP nuclide is relevant and especially why a different library is relevant.

While I am on a roll, a proper QC program should have all 3 elements included:  
1] check source counts to show that positive activity will be measured 
reliably; 2] blank counts to show that the measurement system does not falsely 
report activity in the sample when a sample is not present; and 3] Duplicates 
to demonstrate the reproducibility of real samples [people].  Those are the 
elements that Quality Control experts will look at as evidence of the validity 
of the results.

Fortunately today modern NPP internal doses are so low and WBCs are so reliable 
that this interesting technical discussion does not have significant dose 
consequences - but it could have expensive litigation consequences if not done 
properly.

Have fun

frazier

Frazier Bronson   CHP
Scientific Director
Meriden Technologies - Canberra
800 Research Parkway
Meriden CT  06450   USA
[email protected]



From: Powernet [mailto:[email protected]] On Behalf Of McCarthy, 
John:(GenCo-Nuc) via Powernet
Sent: Wednesday, January 9, 2019 12:08 PM
To: [email protected]; [email protected]
Cc: McCarthy, John:(GenCo-Nuc)
Subject: Re: [Powernet] Benchmark - WBC Check Source

I used 152Eu in the mid 80's with Pilgrim's HPGe WBC.

Jack McCarthy, DCHP
Decommissioning Health Physics
Oyster Creek
609.971.4764

From: Powernet <[email protected]> On Behalf Of Jim Key via 
Powernet
Sent: Wednesday, January 09, 2019 11:31 AM
To: [email protected]
Cc: Jim Key <[email protected]>
Subject: Re: [Powernet] Benchmark - WBC Check Source

Eu-152 would be the preferred source. HPIG energy response is slightly 
non-linear so the lower energies from Eu-152 are advantageous. Also, the lower 
energy peaks provide resolution info that can be useful in recognizing signal 
noise.

Jim

Jim Key
Key Solutions, Inc.
4350 Big Springs Road,
Lebanon, TN 37090
615-453-3712 (Office)
[email protected]<mailto:[email protected]>
413-403-9805 (e-Fax)

From: Powernet 
<[email protected]<mailto:[email protected]>> On 
Behalf Of Holmes, Stephen J:(GenCo-Nuc) via Powernet
Sent: Wednesday, January 9, 2019 5:35 AM
To: [email protected]<mailto:[email protected]>
Cc: Holmes, Stephen J:(GenCo-Nuc) 
<[email protected]<mailto:[email protected]>>
Subject: [Powernet] Benchmark - WBC Check Source

This benchmark is primarily for those of you using a Germanium based whole body 
counter...I welcome feedback from everyone, however

Questions (with Ginna info)


  1.  What WBC system are you using?
Mirion/Canberra Accuscan-II with dual HPGe detectors

  1.  What nuclides/peaks are you using to perform your routine QC checks?
Cs-137/Co-60 source. The 662 and 1332 peaks are used for QCC. We are in need of 
a new source and are considering Eu-152 (122 keV and 1408 keV peaks)

  1.  Have you changed this source in recent history? If so, why and did you 
encounter any issues when you changed sources?

No, but Mirion has recommended we consider changing to Eu-152

  1.  Have you been challenged by a regulator/auditor on the source you are 
choosing to use?

We have not. Benefit of the Co-60/Cs-137 mixture is that we are directly 
validating two key nuclides each time we do a QC check. Benefit of Eu-152 is 
ability to validate the lower end of the energy spectrum during QC checks. 
Eu-152 will also have a longer working life.

Thanks

Stephen J Holmes, CHP
Rad Prot Tech Support Manager
[email protected]<mailto:[email protected]>
315-791-3577 | 845-476-0796 (mobile)

[cid:[email protected]]
R.E. Ginna Nuclear Power Plant
1503 Lake Rd
Ontario, NY 14519



This Email message and any attachment may contain information that is 
proprietary, legally privileged, confidential and/or subject to copyright 
belonging to Exelon Corporation or its affiliates ("Exelon"). This Email is 
intended solely for the use of the person(s) to which it is addressed. If you 
are not an intended recipient, or the employee or agent responsible for 
delivery of this Email to the intended recipient(s), you are hereby notified 
that any dissemination, distribution or copying of this Email is strictly 
prohibited. If you have received this message in error, please immediately 
notify the sender and permanently delete this Email and any copies. Exelon 
policies expressly prohibit employees from making defamatory or offensive 
statements and infringing any copyright or any other legal right by Email 
communication. Exelon will not accept any liability in respect of such 
communications. -EXCIP
_______________________________________________
Powernet mailing list
[email protected]
http://hpspowernet.org/mailman/listinfo/powernet_hpspowernet.org

Reply via email to