Hi,

Ok, I made the assumption that deltadof2 was not the spin-lock offset
(in ppm) but rather the spin-lock field strength (in Hz).  The ppm
values as printed out from the r1rho_1_ini.py script are far too low
for 15N.  They should be in the range of 100-120 ppm!  I.e. around the
15N chemical shifts.  They can be much higher and much lower for
off-resonance data, but having many at an offset of 0.0 ppm seems a
little strange.

Regards,

Edward

On 22 October 2013 13:18, Troels Emtekær Linnet <[email protected]> wrote:
> Hi Edward.
>
> Thanks for looking at this.
>
> The offset for N15 spinlock is given in Hz as parameter deltadof2
> for our pulse sequence.
>
> So I guess that this conversions should be appropriate:
> omega_rf_ppm = float(deltadof2) / (set_sfrq * 1E6) * 1E6
>
> where set_sfrq=799.7773991 is the spectrometer frequency.
>
> Best
> Troels
>
>
> 2013/10/22 Edward d'Auvergne <[email protected]>
>>
>> Hi,
>>
>> In the relax prompt, have a look at:
>>
>> relax> help(relax_disp.spin_lock_offset)
>>
>> The unit you need is ppm.  You can find this number directly from the
>> Bruker acqus or Varian procpar files.  It is the ppm offset of the
>> spin-lock pulse which is manually set by the person recording for each
>> spectrum.  You may need to go to the original pulse sequence to know
>> which pulse this is.  It will be a fixed value for each spectrum
>> collected and you can use that value directly.  The calculations in
>> your scripts seem far to complicated and it looks like the internal
>> conversions performed within relax.  You should never need
>> gyromagnetic ratios, factors of pi, etc.  If you do, you should
>> probably go back to the original experiments and pull out the correct
>> offset number in ppm for each spectrum.  I hope this helps.
>>
>> Regards,
>>
>> Edward
>>
>>
>>
>>
>> On 22 October 2013 10:46, Troels Emtekær Linnet <[email protected]>
>> wrote:
>> > Hi Edward.
>> >
>> > I wonder if would have time to look at the settings script in:
>> > test_suite/shared_data/dispersion/Kjaergaard_et_al_2013
>> >
>> > It is the scripts:
>> > r1rho_1_ini.py
>> > r1rho_3_spectra_settings.py
>> >
>> > In r1rho, I am unsure if I do the correct conversion from Hz to ppm for
>> > omega_rf.
>> >
>> > Again, if there is hidden radian units?
>> >
>> > It is when I set:
>> > relax_disp.spin_lock_offset
>> >
>> > And I wonder, how to start modifying, so R1 rates can be read per
>> > spectra?
>> >
>> > Best
>> > Troels
>> >
>> >
>> >
>> >
>> >
>> > _______________________________________________
>> > relax (http://www.nmr-relax.com)
>> >
>> > This is the relax-devel mailing list
>> > [email protected]
>> >
>> > To unsubscribe from this list, get a password
>> > reminder, or change your subscription options,
>> > visit the list information page at
>> > https://mail.gna.org/listinfo/relax-devel
>> >
>
>

_______________________________________________