Dear Sir,

Thank you very much for your reply.

I minutely checked the things and following is the
accurate data I have  

No Load Ni Ti Sample:

calculated d spacing from ring is 
3.07,
2.417(this is a faint line in ring pattern I couldn't
index this. )
2.1040,
1.4883,
1.21017 
out of these five lines I have indexed four accuraely
as in front std 'd' spacing given
(1 0 0) 3.00926
(1 1 0) 2.12787
(2 0 0) 1.50463
(211) 1.22853

2.417 d spacing ring disappears as I stress the SMA
sample (Aus to Martensite transformation) and when I
agian unload the sample it reappears at 2% strain
continues to appear upto full unloading (0% strain)
when you again load the sample you will see the same
ring upto 3% strain.After that it disappears.

I have seen the pearsons data to locate that ring 
Ni4Ti3 phase    Pd4Pu3  hR14    hexagonal

9.734125539             1       0       0
5.62                    1       1       0
4.867062769             2       0       0
4.867062769             2       0       0
4.501508735             1       0       1
3.679153629             2       1       0
3.513407213             2       0       1
2.979147198             2       1       1
2.5385                  0       0       2
2.456348032             1       0       2
2.313447019             1       1       2

NiTi phase      NiTi str        mp4

0       1       1       3.070299698
1       0       0       2.85522823
-1      1       0       2.344914921
1       1       0       2.344914921
0       0       2       2.309230182


Above both std are having Hexagonal and monoclinic
str.
In Austenite of Ni-Ti B2 Str is there,In B2 str std D
spacing is 

(1 0 0) 3.00926
(1 1 0) 2.12787
(1 1 1) 1.7374
(2 0 0) 1.50463
(2 1 0) 1.34578 means its not contaning 2.41 d spacing
reflection....

About My previous questions, I found the answeres now.
Only this 2.41 d spacing faint line I am following
now.

Any help regarding this will be appreciated.

Thank you once again.

Regadrs

Chandra








 
--- Robin Shirley <[EMAIL PROTECTED]> wrote:
> Dear Chandra
> 
> Firstly, you don't have enough lines in either of
> your two observed (?)
> patterns to index them ab initio from internal
> evidence alone.
> 
> Five or six lines are not enough, even to index a
> supposedly cubic
> pattern without additional information either from
> other measurements or
> from the literature.
> 
> Also, the status of your data is unclear:
> 
> a)  How accurate are they?
> b)  What is the wavelength and how well do you know
> it?
> c)  Are they observed data, or calculated, or a
> mixture of the two?
> 
> Referring to your paragraph numbers:
> 
> A  The 5-line NiTi SMA ring-diffraction pattern
> 
> Your second line "d=2.544763147" appears not to be
> an observed line but
> presumably a calculated one, since the ratio of its
> d with that of the
> first line is exactly 0.8 to 8 significant figures.
> 
> The remaining 3 lines have values of cubic N (=Q/Q1)
> of 2.08, 3.02 and
> 4.05, so they could be 110, 111 and 200 if the
> precision of your
> measurements was only about 3%.
> 
> In order for your line 2 to fit a cubic sequence,
> the values of N would
> need to be 16, 25, 32, 48 and 64 (i.e. 400, 500,
> 440, 444 and 800).
> 
> But, as I've said, line 2 doesn't seem to be an
> observed line anyway.
> 
> For your four lines (omitting line 2) to match the
> calculated
> austenite-b2 NiTi pattern that you quote, either
> your wavelength would
> need to be some 5% larger than you think it is, or
> considerable lattice
> expansion would need to be present (from thermal
> expansion or some other
> cause), or some combination of these.
> 
> B  The stressed NiTi pattern
> 
> This appears to be unrelated to the monoclinic
> martensite-b19' NiTi
> pattern that you quote, and contains far too few
> lines to index ab
> initio, so your only hope is to find some
> appropriate reference
> pattern(s).
> 
> I suspect that you will need a better knowledge of
> the wavelength(s) used
> and of the purity and nature of your sample(s) to
> get further with these
> investigations.
> 
> Perhaps someone in the list can point you towards
> reference patterns
> that may help.
> 
> With best wishes
> 
> Robin Shirley
> 
> ---------------------------------
> 
> Date:          Thu, 19 Jul 2001 11:08:20 -0700 (PDT)
> From:          CHANDRASEN RATHOR
> <[EMAIL PROTECTED]>
> Subject:       NiTi SMA diffraction pattern indexing
> To:            [EMAIL PROTECTED]
> Reply-to:      [EMAIL PROTECTED]
> 
> Hi,
> 
> I am having problem with indexing of ring
> diffraction
> pattern obtained by Synchrotron radiation NiTi SMA
> alloy in no stress and stressed condition.
> 
> 1.I have calculated 'd' spacing for each ring.
> 
> A. For no load condition (austenite-b2 structure)
> calculated 'd(in angstrom)' data is
> 
> 3.180953934,2.544763147,2.20295878,1.830587683
> 1.580730311 
> 
> by STD. hkl reflection d spacing for austenite Ni-Ti
> is 
> (1 0 0)       3.00926
> (1 1 0)       2.12787
> (1 1 1)       1.7374
> (2 0 0)       1.50463
> (2 1 0)       1.34578
> 
> I can understand ordered b2 structure of austenite
> acts as simple cubic in diffraction behavior.I can
> index the data easily, but 2.544763147 this
> calculated
> 'd' i couldn't fit with STD data.
> 
> Also in stressed condition i am getting first
> reflection 'd'spacing claculated is 4.266783322 and
> by
> STD what i have is (0 1 1) 3.07927.which is not at
> all
> matching.
> 
> B.For stressed condition (martenisite- b19'
> monoclinic)of NiTi sample calculated 'd(in
> angstrom)'from diffarction ring data is
> 
>
4.266783322,3.224330579,2.556226044,2.25191342,1.818853147,1.585145759
> 
> by STD. hkl reflection d spacing for martensite
> Ni-Ti
> is 
> 
> (0 1 1)       3.07927,(1 0 0) 2.86543,(-1 1 0)2.52841,(1 0
> 1)2.35966,(1 1 0)     2.34098,(0 2 0) 2.29035,(-1 1
> 1)2.16035
> (0 0 2)       2.07962,(1 1 1) 2.04004,(1 2 1) 2.00628,(0 1
> 2)1.89361,(-1 2 0)1.86867,(1 0 2)     1.68307,(1 2 0)
> 1.71884
> (-1 2 1)1.70454,(-1 1 2)1.60613,(1 2 1)       1.58854
> 
> 
> WHEN I SEARCHED FOR CRYSTALLOGRAPIC DATA FOR NI-TI
> intermettalic PHASE, I CAME TO KNOW VARIOUS
> INTERMEDIATE PHASES AS Ni3Ti(hexagonal DO24 str) ,
> NiTi (cubic) ,
> piTi-Ni (hexagonal?)   ,NiTi2(cubic)
> 
> have anyone gone through indexing diffraction
> pattern
> for SMA NiTi alloy ?where I will find standard d
> spacing for various h,k,l reflection in in above all
> phase structures.?
> 
> Any help would be greatly appreciated.
> 
> Chandra
> University of central florida.
> 


__________________________________________________