I seem to remember i asked the same question about Goebel mirrors last year. 
The outcome of that, from the list and a private comm or two, was that they are 
theoretically challenging to calculate because they effectively act as a series 
of monochromators with varying acceptance angles.

The good news is that the acceptance angle is so low that you'll be fine just 
assuming that there is no polarization. I did try refining the polarization on 
a standard from our focusing Mo instrument, and it refused to budge from zero.


jools


________________________________
From: [email protected] [[email protected]] on behalf of 
Leopoldo Suescun [[email protected]]
Sent: 03 June 2015 16:54
To: Matthey, Björn
Cc: Davide Levy; [email protected]
Subject: Re: polarization

Dear Björn,

Since the Goebel mirror uses diffraction it should require the same polization 
factor than a crystal monochromator.

http://www.ccp14.ac.uk/ccp/web-mirrors/dirdif/xtal/documents/equipment/d8/accessories/goebel.html

you can get the equations from:

http://reference.iucr.org/dictionary/Lorentz%E2%80%93polarization_correction

Hope this helps,
Leo

2015-06-03 10:39 GMT-03:00 Matthey, Björn 
<[email protected]<mailto:[email protected]>>:
Dear all,

what is he right factor in case of a laboratory diffractometer with „ sealed 
tube or rotating anode source” and especially for a goebel mirror?

Thank you in advance

Björn Matthey

--

Mit freundlichen Grüßen | Best Regards
Björn Matthey

Fraunhofer-Institut für Keramische Technologien und Systeme, IKTS Dresden
Dipl.-Ing. Björn Matthey
Keramografie, Phasenanalyse / Ceramography, Phase Analysis
Winterbergstraße 28, 01277 Dresden, Germany
Telefon +49 351 2553-7973<tel:%2B49%20351%202553-7973>
Fax     +49 351 2554-280<tel:%2B49%20351%202554-280>
mailto: 
[email protected]<mailto:[email protected]>
http://www.ikts.fraunhofer.de

Von: [email protected]<mailto:[email protected]> 
[mailto:[email protected]<mailto:[email protected]>] Im Auftrag 
von Leopoldo Suescun
Gesendet: Mittwoch, 3. Juni 2015 15:31
An: Davide Levy
Cc: [email protected]<mailto:[email protected]>
Betreff: Re: polarization

Dear Davide,

The polarization factor depends on the incident X-rays not on the type of 
detector.
If your radiation comes from a sealed tube or rotating anode source without 
monochromator there is no polarization of light and the factor is 0.5, the use 
of a filter does not change the polarization state.
If  you have an incident beam monochromator it depends on the material of the 
monochromator (or better on the diffraction angle).
If you are using synchrotron radiation you will need something in the range 
0.95 to 1 depending on the optics of the beamline.
I hope this helps,
Leo


2015-06-03 5:56 GMT-03:00 Davide Levy 
<[email protected]<mailto:[email protected]>>:
Hi to everybody,
Which polarization correction is better to use in GSAS for data collected with 
a linear detector without monochromator?
thanks
Davide

++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
Please do NOT attach files to the whole list 
<[email protected]<mailto:[email protected]>>
Send commands to <[email protected]<mailto:[email protected]>> eg: HELP as the 
subject with no body text
The Rietveld_L list archive is on http://www.mail-archive.com/[email protected]/
++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++




--
Dr. Leopoldo Suescun
Prof. Agr (Assoc. Prof.) de Física          Tel: (+598) 
29290705<tel:%28%2B598%29%2029290705>/29249859
Cryssmat-Lab./Cátedra de Fisica/DETEMA          Fax: (+598) 
29241906<tel:%28%2B598%29%2029241906>*
Facultad de Quimica, Universidad de la Republica. Montevideo, Uruguay

Ahora la cristalografía importa más (www.iucr.org<http://www.iucr.org/>) 
Crystallography Matters more.



--
Dr. Leopoldo Suescun
Prof. Agr (Assoc. Prof.) de Física          Tel: (+598) 29290705/29249859
Cryssmat-Lab./Cátedra de Fisica/DETEMA          Fax: (+598) 29241906*
Facultad de Quimica, Universidad de la Republica. Montevideo, Uruguay

Ahora la cristalografía importa más (www.iucr.org<http://www.iucr.org/>) 
Crystallography Matters more.
++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
Please do NOT attach files to the whole list <[email protected]>
Send commands to <[email protected]> eg: HELP as the subject with no body text
The Rietveld_L list archive is on http://www.mail-archive.com/[email protected]/
++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++