Jools,

You are right. “Lab grade” bottled N2 may have enough residual oxygen to cause 
problems, plus it is hard to purge the chamber properly to get the pO2 low 
enough to avoid issues.  But an H2/N2 mix might not be suitable, especially if 
it destabilizes his sample. Here are some other ideas:


  1.  I used to flow high purity N2 through a gettering furnace in series with 
the HTXRD chamber, and that, coupled with multiple purges of the chamber prior 
to starting the experiment, worked pretty well.
  2.  I have in the past used a Bühler HT furnace that had a secondary 
“surround heater” option. I had some success putting a gettering wire across 
these leads and running it at a higher temperature than the heater strip in 
order to remove traces of oxygen from the chamber. This has the advantage that 
as you heat up the walls of the chamber and release adsorbed oxygen it goes to 
the getter rather than the strip. It can be a bit tricky ensuring that the 
getter is hotter than the strip across the range of temperatures that you run.

However, I fully agree with you that a platinum foil is probably a better 
option for Blaise’s experiment. If he has one handy.

There are some caveats regarding platinum though. It is generally more 
expensive. It is readily attacked by silicon, iron, and other elements. You can 
get recrystallization and/or rapid grain growth at high temperatures, which 
dramatically changes the diffraction pattern from the strip during a HTXRD 
experiment. The relatively high CTE means that the strip height will change as 
the temperature increases, which will cause large peak shifts unless you are 
using parallel beam optics and even then will cause intensity variations at the 
lower 2thetas. If you work at a US government lab, platinum is a precious metal 
and there is some paperwork required every time you change heater strips. Also, 
if your sample melts or sinters, you may have some “fun” getting the platinum 
clean.

Andrew

*****************************************************

Dr. E. Andrew Payzant
Materials Engineering Group Leader
Neutron Scattering Division
Neutron Sciences Directorate

Mailing address:
Oak Ridge National Laboratory
Dr. E. Andrew Payzant
P.O. Box 2008
Bldg. 8600, MS 6475
Oak Ridge, TN 37831

cell: (865) 235-4981
email: payza...@ornl.gov
https://www.ornl.gov/staff-profile/e-andrew-payzant

******************************************************



From: Julian Richard Tolchard <julianrichard.tolch...@sintef.no>
Date: Thursday, July 25, 2019 at 10:47 AM
To: "Mibeck, Blaise" <bmib...@undeerc.org>, Andrew Payzant <payza...@ornl.gov>
Cc: Rietveld list <rietveld_l@ill.fr>
Subject: RE: [EXTERNAL] Mounting Sample on High Temperature XRD Ribbon Heater

Actually, i wouldn't recommend that either 😊

My experience with the Ta heaters is that they react with even the smallest 
amount of oxygen, and bottled N2 has more than enough to kill a Ta strip. The 
best result we had was with a H2/N2 mix, but that still failed (probably 
reaction with the sample).  Personally I think the only reliable way to use 
them is with a high vacuum system (turbo pump), and with relatively inert 
samples.

Ultimately you are better off getting a Pt strip instead.


Jools


From: rietveld_l-requ...@ill.fr <rietveld_l-requ...@ill.fr> On Behalf Of 
Mibeck, Blaise
Sent: torsdag 25. juli 2019 16.27
To: Payzant, Andrew <payza...@ornl.gov>
Cc: rietveld_l@ill.fr
Subject: Re: [EXTERNAL] Mounting Sample on High Temperature XRD Ribbon Heater

Thank you -- I will try nitrogen then!


________________________________
From: Payzant, Andrew <payza...@ornl.gov<mailto:payza...@ornl.gov>>
Sent: Thursday, July 25, 2019 9:08 AM
To: Mibeck, Blaise <bmib...@undeerc.org<mailto:bmib...@undeerc.org>>
Cc: rietveld_l@ill.fr<mailto:rietveld_l@ill.fr> 
<rietveld_l@ill.fr<mailto:rietveld_l@ill.fr>>
Subject: Re: [EXTERNAL] Mounting Sample on High Temperature XRD Ribbon Heater


STOP!!!



The tantalum foil should not be used for high temperature in air. Above a few 
hundred degrees Celsius it will oxidize rapidly and disintegrate.

I accidently learned this the hard way many years ago!  😉



Either use an inert atmosphere or vacuum. Or use a different heater ribbon, 
suitable for air atmosphere.



Andrew



*****************************************************



Dr. E. Andrew Payzant

Materials Engineering Group Leader

Neutron Scattering Division

Neutron Sciences Directorate



Mailing address:

Oak Ridge National Laboratory

Dr. E. Andrew Payzant

P.O. Box 2008

Bldg. 8600, MS 6475

Oak Ridge, TN 37831



cell: (865) 235-4981

email: payza...@ornl.gov<mailto:payza...@ornl.gov>

https://www.ornl.gov/staff-profile/e-andrew-payzant



******************************************************







From: <rietveld_l-requ...@ill.fr<mailto:rietveld_l-requ...@ill.fr>> on behalf 
of "Mibeck, Blaise" <bmib...@undeerc.org<mailto:bmib...@undeerc.org>>
Reply-To: "Mibeck, Blaise" <bmib...@undeerc.org<mailto:bmib...@undeerc.org>>
Date: Thursday, July 25, 2019 at 10:03 AM
To: Rietveld list <rietveld_l@ill.fr<mailto:rietveld_l@ill.fr>>
Subject: [EXTERNAL] Mounting Sample on High Temperature XRD Ribbon Heater





I am looking for advice on how to prepar a HTXRD sample (mounting techniques, 
stage preparation, etc...). The ribbon heater is tantalum (brand new). The 
sample contained ammonium di-hydrogen phosphate. The atmosphere is air.



I hope to reach at least 1100C, but have had the ribbon break before 750C. 
Corrosion? or can the sample lower electrical resistance?