Re: [ccp4bb] data processing problem with ice rings

2011-10-26 Thread ChenTiantian
Hi there,
Thank you for all your suggestions and generous help, I tried some methods
you guys mentioned and learned something new . I really appreciate it.
With Kay's help,(after exclusion of the ice rings he found that the data are
P1, not P2(1). ) I got my structure solved, there are four copies in the AU,
I cut the high resolution to 2.4, and now the R/Rfree is  0.2254/0.2648,
this is not the final result. I'm still working on it.
Thank you so much.

Best Regards,

Tiantian


Re: [ccp4bb] data processing problem with ice rings

2011-10-17 Thread Sanishvili, Ruslan
> In fact, just estimating ORGX and ORGY from the first frame, using adxv or > 
> XDS-viewer, seems to do a good job.

The silver lining of having ice (or any other kind) rings is that one can 
determine the origin rather accurately by just finding the center of these 
rings.
Cheers,
N.


Ruslan Sanishvili (Nukri), Ph.D.

GM/CA-CAT
Biosciences Division, ANL
9700 S. Cass Ave.
Argonne, IL 60439

Tel: (630)252-0665
Fax: (630)252-0667
rsanishv...@anl.gov

-Original Message-
From: CCP4 bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK] On Behalf Of Kay 
Diederichs
Sent: Saturday, October 15, 2011 6:10 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] data processing problem with ice rings

Hi ChenTiantian,

the R-factors and I/sigma are bad even at low resolution where the first 
icering does not influence the results.
Thus, the problem with your data processing has little to do with the 
icerings. I guess that the indexing is not correct.
My suggestion:
1) using adxv or a similar display program, note what the inner and 
outer limits of the ice rings are. These values should be used as 
parameters for the EXCLUDE_RESOLUTION_RANGE= keywords in XDS.INP, not 
the provided ones (which are meant for hexagonal ice).
2) start XDS from the INIT step
3) use at least half of your DATA_RANGE as SPOT_RANGE
4) make sure that ORGX and ORGY are correct - mis-indexing is in 90% of 
the cases due to a wrong origin. In fact, just estimating ORGX and ORGY 
from the first frame, using adxv or XDS-viewer, seems to do a good job.

HTH,

Kay


Am 20:59, schrieb ChenTiantian:
> Hi there,
> I am processing a dataset which has bad ice rings (as you can see in the
> attach png file).
> I tried both XDS and imosflm, and got similar results, it seems that
> adding " EXCLUDE_RESOLUTION_RANGE" cannot get rid of the effects of the
> ice rings.
> the following is part of the CORRECT.LP which is the second attached
> file, you can find more details there.
>
>SUBSET OF INTENSITY DATA WITH SIGNAL/NOISE >= -3.0 AS FUNCTION OF
> RESOLUTION
>   RESOLUTION NUMBER OF REFLECTIONSCOMPLETENESS R-FACTOR
> R-FACTOR COMPARED I/SIGMA   R-meas  Rmrgd-F  Anomal  SigAno   Nano
> LIMIT OBSERVED  UNIQUE  POSSIBLE OF DATA   observed
> expected  Corr
>
>   4.24   371525537  5545   99.9%  46.9%
> 52.7%371502.4850.8%19.4%   -28%   0.5135136
>   3.01   553449002  9840   91.5%  62.7%
> 65.1%551161.7668.3%48.1%   -28%   0.5207760
>   2.46   84636   12699 12703  100.0%  67.4%
> 84.7%846341.5573.0%54.2%   -19%   0.513   12104
>   2.13   97910   14743 14987   98.4% 254.5%
> 199.3%979080.16   276.2%  4899.9%   -23%   0.473   14037
>   1.90  110260   16846 16940   99.4% 299.2%
> 303.3%   1102450.06   325.0%   -99.9%   -17%   0.422   15995
>   1.74  118354   18629 18744   99.4%1062.0%
> 1043.6%   118317   -0.20  1156.4%   -99.9%   -13%   0.380   17414
>   1.61  122958   20193 20331   99.3% 967.5%
> 1571.1%   1228680.10  1059.7%   987.3%-2%   0.402   18348
>   1.51  125075   21554 21794   98.9% 838.9%
> 1355.1%   1249330.08   922.6%  1116.9%-1%   0.402   18977
>   1.42   72057   17042 23233   73.4% 640.8%
> 775.3%703910.08   732.5%   826.7%-8%   0.425   10003
>  total  823746  136245144117   94.5% 166.4%
> 166.7%   8215620.40   181.1%   296.7%   -15%   0.435  119774
>
> Note that I/SIGMA of each resolution shell is <2.5, so how should I do
> to process the dataset properly? Any suggestion about this super ice rings?
> Thanks!
>
> Tiantian
>
> --
> Shanghai Institute of Materia Medica, Chinese Academy of Sciences
> Address: Room 101, 646 Songtao Road, Zhangjiang Hi-Tech Park,
> Shanghai, 201203


-- 
Kay Diederichshttp://strucbio.biologie.uni-konstanz.de
email: kay.diederi...@uni-konstanz.deTel +49 7531 88 4049 Fax 3183
Fachbereich Biologie, Universität Konstanz, Box 647, D-78457 Konstanz

This e-mail is digitally signed. If your e-mail client does not have the
necessary capabilities, just ignore the attached signature "smime.p7s".


Re: [ccp4bb] data processing problem with ice rings

2011-10-15 Thread Nian Huang
Hi,
I agree with other people. You must have a wrong index here. Can you tell us
what is the unit cell for this crystal from your determination? I can see
very close spots in the high resolution shell from your image, which are
overlapped into one spot in the low resolution shell. Try to use other
frames to do the indexing. If it is still not working, it might be easier to
collect another dataset with better crystal alignment.

Best,

Nian


On Fri, Oct 14, 2011 at 12:12 AM, ChenTiantian
wrote:

> Hi there,
> I am processing a dataset which has bad ice rings (as you can see in the
> attach png file).
> I tried both XDS and imosflm, and got similar results, it seems that adding
> " EXCLUDE_RESOLUTION_RANGE" cannot get rid of the effects of the ice rings.
> the following is part of the CORRECT.LP which is the second attached file,
> you can find more details there.
>
>   SUBSET OF INTENSITY DATA WITH SIGNAL/NOISE >= -3.0 AS FUNCTION OF
> RESOLUTION
>  RESOLUTION NUMBER OF REFLECTIONSCOMPLETENESS R-FACTOR  R-FACTOR
> COMPARED I/SIGMA   R-meas  Rmrgd-F  Anomal  SigAno   Nano
>LIMIT OBSERVED  UNIQUE  POSSIBLE OF DATA   observed
> expected  Corr
>
>  4.24   371525537  5545   99.9%  46.9% 52.7%
> 371502.4850.8%19.4%   -28%   0.5135136
>  3.01   553449002  9840   91.5%  62.7% 65.1%
> 551161.7668.3%48.1%   -28%   0.5207760
>  2.46   84636   12699 12703  100.0%  67.4% 84.7%
> 846341.5573.0%54.2%   -19%   0.513   12104
>  2.13   97910   14743 14987   98.4% 254.5%199.3%
> 979080.16   276.2%  4899.9%   -23%   0.473   14037
>  1.90  110260   16846 16940   99.4% 299.2%303.3%
> 1102450.06   325.0%   -99.9%   -17%   0.422   15995
>  1.74  118354   18629 18744   99.4%1062.0%   1043.6%
> 118317   -0.20  1156.4%   -99.9%   -13%   0.380   17414
>  1.61  122958   20193 20331   99.3% 967.5%   1571.1%
> 1228680.10  1059.7%   987.3%-2%   0.402   18348
>  1.51  125075   21554 21794   98.9% 838.9%   1355.1%
> 1249330.08   922.6%  1116.9%-1%   0.402   18977
>  1.42   72057   17042 23233   73.4% 640.8%
> 775.3%703910.08   732.5%   826.7%-8%   0.425   10003
> total  823746  136245144117   94.5% 166.4%166.7%
> 8215620.40   181.1%   296.7%   -15%   0.435  119774
>
> Note that I/SIGMA of each resolution shell is <2.5, so how should I do to
> process the dataset properly? Any suggestion about this super ice rings?
> Thanks!
>
> Tiantian
>
> --
> Shanghai Institute of Materia Medica, Chinese Academy of Sciences
> Address: Room 101, 646 Songtao Road, Zhangjiang Hi-Tech Park,
> Shanghai, 201203
>


Re: [ccp4bb] data processing problem with ice rings

2011-10-15 Thread Harry

Hi

I'd agree with Kay here - I would think that the original indexing is  
incorrect.


One thing I notice on the original image as posted - there's a red  
cross on it - if that's supposed to mark the beam position, I think  
it's about 4mm or so away from the true position.


So -

(1) check the beam position carefully (it may be wrong in the image  
header)


(2) after indexing, make sure that the predictions match the spot  
positions


(3) if the predictions don't match the spot positions, don't try to  
integrate - find out what's wrong (wrong wavelength, beam position,  
distance???). If you can't work it out, ask one of the experts to look  
at a sample of your original images (iMosflm ask Andrew or me, XDS ask  
Kay, HKL Wladek or ZO...).


(4) If the predictions do match the spot positions, integrate the  
dataset in P1 (i.e. triclinic) and see what Pointless suggests as the  
symmetry. You may just be trying to impose too much symmetry. If you  
can't work out what the issue is, ask an expert to help directly -  
we're all happy to help out!


(5) Worry about the ice rings after you've sorted out the above  
problems, not before.


HTH

On 15 Oct 2011, at 12:09, Kay Diederichs wrote:


Hi ChenTiantian,

the R-factors and I/sigma are bad even at low resolution where the  
first icering does not influence the results.
Thus, the problem with your data processing has little to do with  
the icerings. I guess that the indexing is not correct.

My suggestion:
1) using adxv or a similar display program, note what the inner and  
outer limits of the ice rings are. These values should be used as  
parameters for the EXCLUDE_RESOLUTION_RANGE= keywords in XDS.INP,  
not the provided ones (which are meant for hexagonal ice).

2) start XDS from the INIT step
3) use at least half of your DATA_RANGE as SPOT_RANGE
4) make sure that ORGX and ORGY are correct - mis-indexing is in 90%  
of the cases due to a wrong origin. In fact, just estimating ORGX  
and ORGY from the first frame, using adxv or XDS-viewer, seems to do  
a good job.


HTH,

Kay


Am 20:59, schrieb ChenTiantian:

Hi there,
I am processing a dataset which has bad ice rings (as you can see  
in the

attach png file).
I tried both XDS and imosflm, and got similar results, it seems that
adding " EXCLUDE_RESOLUTION_RANGE" cannot get rid of the effects of  
the

ice rings.
the following is part of the CORRECT.LP which is the second attached
file, you can find more details there.

  SUBSET OF INTENSITY DATA WITH SIGNAL/NOISE >= -3.0 AS FUNCTION OF
RESOLUTION
 RESOLUTION NUMBER OF REFLECTIONSCOMPLETENESS R-FACTOR
R-FACTOR COMPARED I/SIGMA   R-meas  Rmrgd-F  Anomal  SigAno   Nano
   LIMIT OBSERVED  UNIQUE  POSSIBLE OF DATA   observed
expected  Corr

 4.24   371525537  5545   99.9%  46.9%
52.7%371502.4850.8%19.4%   -28%   0.5135136
 3.01   553449002  9840   91.5%  62.7%
65.1%551161.7668.3%48.1%   -28%   0.5207760
 2.46   84636   12699 12703  100.0%  67.4%
84.7%846341.5573.0%54.2%   -19%   0.513   12104
 2.13   97910   14743 14987   98.4% 254.5%
199.3%979080.16   276.2%  4899.9%   -23%   0.473   14037
 1.90  110260   16846 16940   99.4% 299.2%
303.3%   1102450.06   325.0%   -99.9%   -17%   0.422   15995
 1.74  118354   18629 18744   99.4%1062.0%
1043.6%   118317   -0.20  1156.4%   -99.9%   -13%   0.380   17414
 1.61  122958   20193 20331   99.3% 967.5%
1571.1%   1228680.10  1059.7%   987.3%-2%   0.402   18348
 1.51  125075   21554 21794   98.9% 838.9%
1355.1%   1249330.08   922.6%  1116.9%-1%   0.402   18977
 1.42   72057   17042 23233   73.4% 640.8%
775.3%703910.08   732.5%   826.7%-8%   0.425   10003
total  823746  136245144117   94.5% 166.4%
166.7%   8215620.40   181.1%   296.7%   -15%   0.435  119774

Note that I/SIGMA of each resolution shell is <2.5, so how should I  
do
to process the dataset properly? Any suggestion about this super  
ice rings?

Thanks!

Tiantian

--
Shanghai Institute of Materia Medica, Chinese Academy of Sciences
Address: Room 101, 646 Songtao Road, Zhangjiang Hi-Tech Park,
Shanghai, 201203



--
Kay Diederichshttp://strucbio.biologie.uni-konstanz.de
email: kay.diederi...@uni-konstanz.deTel +49 7531 88 4049 Fax 3183
Fachbereich Biologie, Universität Konstanz, Box 647, D-78457 Konstanz

This e-mail is digitally signed. If your e-mail client does not have  
the
necessary capabilities, just ignore the attached signature  
"smime.p7s".




Harry
--
Dr Harry Powell, MRC Laboratory of Molecular Biology, MRC Centre,  
Hills Road, Cambridge, CB2 0QH


Re: [ccp4bb] data processing problem with ice rings

2011-10-15 Thread Kay Diederichs

Hi ChenTiantian,

the R-factors and I/sigma are bad even at low resolution where the first 
icering does not influence the results.
Thus, the problem with your data processing has little to do with the 
icerings. I guess that the indexing is not correct.

My suggestion:
1) using adxv or a similar display program, note what the inner and 
outer limits of the ice rings are. These values should be used as 
parameters for the EXCLUDE_RESOLUTION_RANGE= keywords in XDS.INP, not 
the provided ones (which are meant for hexagonal ice).

2) start XDS from the INIT step
3) use at least half of your DATA_RANGE as SPOT_RANGE
4) make sure that ORGX and ORGY are correct - mis-indexing is in 90% of 
the cases due to a wrong origin. In fact, just estimating ORGX and ORGY 
from the first frame, using adxv or XDS-viewer, seems to do a good job.


HTH,

Kay


Am 20:59, schrieb ChenTiantian:

Hi there,
I am processing a dataset which has bad ice rings (as you can see in the
attach png file).
I tried both XDS and imosflm, and got similar results, it seems that
adding " EXCLUDE_RESOLUTION_RANGE" cannot get rid of the effects of the
ice rings.
the following is part of the CORRECT.LP which is the second attached
file, you can find more details there.

   SUBSET OF INTENSITY DATA WITH SIGNAL/NOISE >= -3.0 AS FUNCTION OF
RESOLUTION
  RESOLUTION NUMBER OF REFLECTIONSCOMPLETENESS R-FACTOR
R-FACTOR COMPARED I/SIGMA   R-meas  Rmrgd-F  Anomal  SigAno   Nano
LIMIT OBSERVED  UNIQUE  POSSIBLE OF DATA   observed
expected  Corr

  4.24   371525537  5545   99.9%  46.9%
52.7%371502.4850.8%19.4%   -28%   0.5135136
  3.01   553449002  9840   91.5%  62.7%
65.1%551161.7668.3%48.1%   -28%   0.5207760
  2.46   84636   12699 12703  100.0%  67.4%
84.7%846341.5573.0%54.2%   -19%   0.513   12104
  2.13   97910   14743 14987   98.4% 254.5%
199.3%979080.16   276.2%  4899.9%   -23%   0.473   14037
  1.90  110260   16846 16940   99.4% 299.2%
303.3%   1102450.06   325.0%   -99.9%   -17%   0.422   15995
  1.74  118354   18629 18744   99.4%1062.0%
1043.6%   118317   -0.20  1156.4%   -99.9%   -13%   0.380   17414
  1.61  122958   20193 20331   99.3% 967.5%
1571.1%   1228680.10  1059.7%   987.3%-2%   0.402   18348
  1.51  125075   21554 21794   98.9% 838.9%
1355.1%   1249330.08   922.6%  1116.9%-1%   0.402   18977
  1.42   72057   17042 23233   73.4% 640.8%
775.3%703910.08   732.5%   826.7%-8%   0.425   10003
 total  823746  136245144117   94.5% 166.4%
166.7%   8215620.40   181.1%   296.7%   -15%   0.435  119774

Note that I/SIGMA of each resolution shell is <2.5, so how should I do
to process the dataset properly? Any suggestion about this super ice rings?
Thanks!

Tiantian

--
Shanghai Institute of Materia Medica, Chinese Academy of Sciences
Address: Room 101, 646 Songtao Road, Zhangjiang Hi-Tech Park,
Shanghai, 201203



--
Kay Diederichshttp://strucbio.biologie.uni-konstanz.de
email: kay.diederi...@uni-konstanz.deTel +49 7531 88 4049 Fax 3183
Fachbereich Biologie, Universität Konstanz, Box 647, D-78457 Konstanz

This e-mail is digitally signed. If your e-mail client does not have the
necessary capabilities, just ignore the attached signature "smime.p7s".



smime.p7s
Description: S/MIME Kryptografische Unterschrift


Re: [ccp4bb] data processing problem with ice rings

2011-10-14 Thread James Holton
These rings are nanocrystalline cubic ice (ice Ic, as opposed to the 
"usual" ice Ih).  It is an interesting substance in that noone has ever 
prepared a large single crystal of it.  In fact, for very small crystals 
it can be hard to distinguish it from amorphous ice (or "glassy 
water").  The three main rings that you see from ice Ic coincide almost 
exactly with the centroids of the three main diffuse rings of glassy 
water, and as the ice Ic crystals get smaller, the rings get fatter 
(Scherrer broadening).  You can even measure the size of the 
crystallites by measuring the width of the rings.  At the limit of 1-2 
unit cells wide, the diffraction pattern of ice Ic powder looks almost 
exactly like that of glassy water, so I suppose one could say that there 
is a continuum of phases between the two.


And yes, there are crystals that "like" a certain mixture of cubic ice 
and amorphous water in their solvent channels.  Other's don't like it at 
all.  But I agree with JS below that the problem here is not the ice 
rings.  Probably overlaps?  Best to look only at spots inside the 3.8A 
circle until you figure out what is going on.


-James Holton
MAD Scientist

On 10/13/2011 11:20 PM, James Stroud wrote:

First of all, are you sure those are ice rings? They do not look typical. I 
think you might have salt crystals from dehydration *before* freezing. 
Otherwise, I think your freezing went well. Maybe try a humidity controlled 
environment when you freeze.

Second, I'm not so sure the bad stats come from the contaminating rings. The 
lattice seems to have some sort of problem, like a split lattice. You might be 
able to tackle this problem by increasing your spot size or skewing it's shape 
to compensate for the split. You need to investigate several images throughout 
the run to see whether and how to manipulate your spot size. Sometimes, the 
split lengthens the spots in the direction of the phi axis and you get lucky. 
But I think the phi axis might be horizontal in this picture, which makes 
things a little trickier. From one image, it is difficult to tell the pathology 
of this crystal.

In principle, if you can accurately measure the most high-resolution spots 
visible (which appear to be about 1.9 Å, guessing from your log file) then you 
will have a pretty good data set, even with the contaminating rings.

Personally, I'd use Denzo for this data, but I don't know what is vogue with 
the community right now. I still use O, so my tastes might be somewhat 
antiquated.

James



On Oct 13, 2011, at 11:12 PM, ChenTiantian wrote:


Hi there,
I am processing a dataset which has bad ice rings (as you can see in the attach 
png file).
I tried both XDS and imosflm, and got similar results, it seems that adding " 
EXCLUDE_RESOLUTION_RANGE" cannot get rid of the effects of the ice rings.
the following is part of the CORRECT.LP which is the second attached file, you 
can find more details there.

   SUBSET OF INTENSITY DATA WITH SIGNAL/NOISE>= -3.0 AS FUNCTION OF RESOLUTION
  RESOLUTION NUMBER OF REFLECTIONSCOMPLETENESS R-FACTOR  R-FACTOR 
COMPARED I/SIGMA   R-meas  Rmrgd-F  Anomal  SigAno   Nano
LIMIT OBSERVED  UNIQUE  POSSIBLE OF DATA   observed  expected   
   Corr

  4.24   371525537  5545   99.9%  46.9% 52.7%
371502.4850.8%19.4%   -28%   0.5135136
  3.01   553449002  9840   91.5%  62.7% 65.1%
551161.7668.3%48.1%   -28%   0.5207760
  2.46   84636   12699 12703  100.0%  67.4% 84.7%
846341.5573.0%54.2%   -19%   0.513   12104
  2.13   97910   14743 14987   98.4% 254.5%199.3%
979080.16   276.2%  4899.9%   -23%   0.473   14037
  1.90  110260   16846 16940   99.4% 299.2%303.3%   
1102450.06   325.0%   -99.9%   -17%   0.422   15995
  1.74  118354   18629 18744   99.4%1062.0%   1043.6%   
118317   -0.20  1156.4%   -99.9%   -13%   0.380   17414
  1.61  122958   20193 20331   99.3% 967.5%   1571.1%   
1228680.10  1059.7%   987.3%-2%   0.402   18348
  1.51  125075   21554 21794   98.9% 838.9%   1355.1%   
1249330.08   922.6%  1116.9%-1%   0.402   18977
  1.42   72057   17042 23233   73.4% 640.8%775.3%
703910.08   732.5%   826.7%-8%   0.425   10003
 total  823746  136245144117   94.5% 166.4%166.7%   
8215620.40   181.1%   296.7%   -15%   0.435  119774

Note that I/SIGMA of each resolution shell is<2.5, so how should I do to 
process the dataset properly? Any suggestion about this super ice rings?
Thanks!

Tiantian

--
Shanghai Institute of Materia Medica, Chinese Academy of Sciences
Address: Room 101, 646 Songtao Road, Zhangjiang Hi-Tech Park,
Shanghai, 201203



Re: [ccp4bb] data processing problem with ice rings

2011-10-14 Thread vandana kukshal
Hello ,
Can any one send me pdf of this paper as its a old paper and not
accessible here .
   M.F. Perutz, Preparation of haemoglobin crystals. *J. Cryst. Growth*
, * 2 * (1968), pp. 54–56.


On Fri, Oct 14, 2011 at 10:42 AM, ChenTiantian
wrote:

> Hi there,
> I am processing a dataset which has bad ice rings (as you can see in the
> attach png file).
> I tried both XDS and imosflm, and got similar results, it seems that adding
> " EXCLUDE_RESOLUTION_RANGE" cannot get rid of the effects of the ice rings.
> the following is part of the CORRECT.LP which is the second attached file,
> you can find more details there.
>
>   SUBSET OF INTENSITY DATA WITH SIGNAL/NOISE >= -3.0 AS FUNCTION OF
> RESOLUTION
>  RESOLUTION NUMBER OF REFLECTIONSCOMPLETENESS R-FACTOR  R-FACTOR
> COMPARED I/SIGMA   R-meas  Rmrgd-F  Anomal  SigAno   Nano
>LIMIT OBSERVED  UNIQUE  POSSIBLE OF DATA   observed
> expected  Corr
>
>  4.24   371525537  5545   99.9%  46.9% 52.7%
> 371502.4850.8%19.4%   -28%   0.5135136
>  3.01   553449002  9840   91.5%  62.7% 65.1%
> 551161.7668.3%48.1%   -28%   0.5207760
>  2.46   84636   12699 12703  100.0%  67.4% 84.7%
> 846341.5573.0%54.2%   -19%   0.513   12104
>  2.13   97910   14743 14987   98.4% 254.5%199.3%
> 979080.16   276.2%  4899.9%   -23%   0.473   14037
>  1.90  110260   16846 16940   99.4% 299.2%303.3%
> 1102450.06   325.0%   -99.9%   -17%   0.422   15995
>  1.74  118354   18629 18744   99.4%1062.0%   1043.6%
> 118317   -0.20  1156.4%   -99.9%   -13%   0.380   17414
>  1.61  122958   20193 20331   99.3% 967.5%   1571.1%
> 1228680.10  1059.7%   987.3%-2%   0.402   18348
>  1.51  125075   21554 21794   98.9% 838.9%   1355.1%
> 1249330.08   922.6%  1116.9%-1%   0.402   18977
>  1.42   72057   17042 23233   73.4% 640.8%
> 775.3%703910.08   732.5%   826.7%-8%   0.425   10003
> total  823746  136245144117   94.5% 166.4%166.7%
> 8215620.40   181.1%   296.7%   -15%   0.435  119774
>
> Note that I/SIGMA of each resolution shell is <2.5, so how should I do to
> process the dataset properly? Any suggestion about this super ice rings?
> Thanks!
>
> Tiantian
>
> --
> Shanghai Institute of Materia Medica, Chinese Academy of Sciences
> Address: Room 101, 646 Songtao Road, Zhangjiang Hi-Tech Park,
> Shanghai, 201203
>



-- 
Vandana kukshal


Re: [ccp4bb] data processing problem with ice rings

2011-10-14 Thread Petri Kursula
Your main problem is not the ice rings but a wrong lattice/indexing solution. R 
factors are very high for even low res shells and I/sigma very low. To me this 
tells you are not finding your diffraction spots at all.

First thing to try: Take more images for the indexing step and use only the 
strongest spots. And do not refine distance during indexing, as you probably 
have a pretty high mosaicity. 

Petri

On Oct 14, 2011, at 7:12 AM, ChenTiantian wrote:

> Hi there,
> I am processing a dataset which has bad ice rings (as you can see in the 
> attach png file).
> I tried both XDS and imosflm, and got similar results, it seems that adding " 
> EXCLUDE_RESOLUTION_RANGE" cannot get rid of the effects of the ice rings.
> the following is part of the CORRECT.LP which is the second attached file, 
> you can find more details there. 
> 
>   SUBSET OF INTENSITY DATA WITH SIGNAL/NOISE >= -3.0 AS FUNCTION OF RESOLUTION
>  RESOLUTION NUMBER OF REFLECTIONSCOMPLETENESS R-FACTOR  R-FACTOR 
> COMPARED I/SIGMA   R-meas  Rmrgd-F  Anomal  SigAno   Nano
>LIMIT OBSERVED  UNIQUE  POSSIBLE OF DATA   observed  expected  
> Corr
> 
>  4.24   371525537  5545   99.9%  46.9% 52.7%
> 371502.4850.8%19.4%   -28%   0.5135136
>  3.01   553449002  9840   91.5%  62.7% 65.1%
> 551161.7668.3%48.1%   -28%   0.5207760
>  2.46   84636   12699 12703  100.0%  67.4% 84.7%
> 846341.5573.0%54.2%   -19%   0.513   12104
>  2.13   97910   14743 14987   98.4% 254.5%199.3%
> 979080.16   276.2%  4899.9%   -23%   0.473   14037
>  1.90  110260   16846 16940   99.4% 299.2%303.3%   
> 1102450.06   325.0%   -99.9%   -17%   0.422   15995
>  1.74  118354   18629 18744   99.4%1062.0%   1043.6%   
> 118317   -0.20  1156.4%   -99.9%   -13%   0.380   17414
>  1.61  122958   20193 20331   99.3% 967.5%   1571.1%   
> 1228680.10  1059.7%   987.3%-2%   0.402   18348
>  1.51  125075   21554 21794   98.9% 838.9%   1355.1%   
> 1249330.08   922.6%  1116.9%-1%   0.402   18977
>  1.42   72057   17042 23233   73.4% 640.8%775.3%
> 703910.08   732.5%   826.7%-8%   0.425   10003
> total  823746  136245144117   94.5% 166.4%166.7%   
> 8215620.40   181.1%   296.7%   -15%   0.435  119774
> 
> Note that I/SIGMA of each resolution shell is <2.5, so how should I do to 
> process the dataset properly? Any suggestion about this super ice rings?
> Thanks!
> 
> Tiantian
> 
> -- 
> Shanghai Institute of Materia Medica, Chinese Academy of Sciences
> Address: Room 101, 646 Songtao Road, Zhangjiang Hi-Tech Park,
> Shanghai, 201203 
> 


---
Petri Kursula, PhD
Group Leader, Docent of Neurobiochemistry
Department of Biochemistry, University of Oulu, Finland
Department of Chemistry, University of Hamburg, Germany
Visiting Scientist (CSSB-HZI, DESY, Hamburg, Germany)
www.biochem.oulu.fi/kursula
www.desy.de/~petri
petri.kurs...@oulu.fi
petri.kurs...@desy.de
---



Re: [ccp4bb] data processing problem with ice rings

2011-10-13 Thread James Stroud
First of all, are you sure those are ice rings? They do not look typical. I 
think you might have salt crystals from dehydration *before* freezing. 
Otherwise, I think your freezing went well. Maybe try a humidity controlled 
environment when you freeze.

Second, I'm not so sure the bad stats come from the contaminating rings. The 
lattice seems to have some sort of problem, like a split lattice. You might be 
able to tackle this problem by increasing your spot size or skewing it's shape 
to compensate for the split. You need to investigate several images throughout 
the run to see whether and how to manipulate your spot size. Sometimes, the 
split lengthens the spots in the direction of the phi axis and you get lucky. 
But I think the phi axis might be horizontal in this picture, which makes 
things a little trickier. From one image, it is difficult to tell the pathology 
of this crystal.

In principle, if you can accurately measure the most high-resolution spots 
visible (which appear to be about 1.9 Å, guessing from your log file) then you 
will have a pretty good data set, even with the contaminating rings.

Personally, I'd use Denzo for this data, but I don't know what is vogue with 
the community right now. I still use O, so my tastes might be somewhat 
antiquated.

James



On Oct 13, 2011, at 11:12 PM, ChenTiantian wrote:

> Hi there,
> I am processing a dataset which has bad ice rings (as you can see in the 
> attach png file).
> I tried both XDS and imosflm, and got similar results, it seems that adding " 
> EXCLUDE_RESOLUTION_RANGE" cannot get rid of the effects of the ice rings.
> the following is part of the CORRECT.LP which is the second attached file, 
> you can find more details there. 
> 
>   SUBSET OF INTENSITY DATA WITH SIGNAL/NOISE >= -3.0 AS FUNCTION OF RESOLUTION
>  RESOLUTION NUMBER OF REFLECTIONSCOMPLETENESS R-FACTOR  R-FACTOR 
> COMPARED I/SIGMA   R-meas  Rmrgd-F  Anomal  SigAno   Nano
>LIMIT OBSERVED  UNIQUE  POSSIBLE OF DATA   observed  expected  
> Corr
> 
>  4.24   371525537  5545   99.9%  46.9% 52.7%
> 371502.4850.8%19.4%   -28%   0.5135136
>  3.01   553449002  9840   91.5%  62.7% 65.1%
> 551161.7668.3%48.1%   -28%   0.5207760
>  2.46   84636   12699 12703  100.0%  67.4% 84.7%
> 846341.5573.0%54.2%   -19%   0.513   12104
>  2.13   97910   14743 14987   98.4% 254.5%199.3%
> 979080.16   276.2%  4899.9%   -23%   0.473   14037
>  1.90  110260   16846 16940   99.4% 299.2%303.3%   
> 1102450.06   325.0%   -99.9%   -17%   0.422   15995
>  1.74  118354   18629 18744   99.4%1062.0%   1043.6%   
> 118317   -0.20  1156.4%   -99.9%   -13%   0.380   17414
>  1.61  122958   20193 20331   99.3% 967.5%   1571.1%   
> 1228680.10  1059.7%   987.3%-2%   0.402   18348
>  1.51  125075   21554 21794   98.9% 838.9%   1355.1%   
> 1249330.08   922.6%  1116.9%-1%   0.402   18977
>  1.42   72057   17042 23233   73.4% 640.8%775.3%
> 703910.08   732.5%   826.7%-8%   0.425   10003
> total  823746  136245144117   94.5% 166.4%166.7%   
> 8215620.40   181.1%   296.7%   -15%   0.435  119774
> 
> Note that I/SIGMA of each resolution shell is <2.5, so how should I do to 
> process the dataset properly? Any suggestion about this super ice rings?
> Thanks!
> 
> Tiantian
> 
> -- 
> Shanghai Institute of Materia Medica, Chinese Academy of Sciences
> Address: Room 101, 646 Songtao Road, Zhangjiang Hi-Tech Park,
> Shanghai, 201203 
> 


Re: [ccp4bb] data processing problem with ice rings

2011-10-13 Thread Stefan Gerhardt
try a frozen xtal ...

On Fri, 14 Oct 2011 13:12:12 +0800
 ChenTiantian  wrote:
> Hi there,
> I am processing a dataset which has bad ice rings (as you
> can see in the
> attach png file).
> I tried both XDS and imosflm, and got similar results, it
> seems that adding
> " EXCLUDE_RESOLUTION_RANGE" cannot get rid of the effects
> of the ice rings.
> the following is part of the CORRECT.LP which is the
> second attached file,
> you can find more details there.
> 
>   SUBSET OF INTENSITY DATA WITH SIGNAL/NOISE >= -3.0 AS
> FUNCTION OF
> RESOLUTION
>  RESOLUTION NUMBER OF REFLECTIONSCOMPLETENESS
> R-FACTOR  R-FACTOR
> COMPARED I/SIGMA   R-meas  Rmrgd-F  Anomal  SigAno   Nano
>LIMIT OBSERVED  UNIQUE  POSSIBLE OF DATA
>   observed
> expected  Corr
> 
>  4.24   371525537  5545   99.9%
>  46.9% 52.7%
> 371502.4850.8%19.4%   -28%   0.5135136
>  3.01   553449002  9840   91.5%
>  62.7% 65.1%
> 551161.7668.3%48.1%   -28%   0.5207760
>  2.46   84636   12699 12703  100.0%
>  67.4% 84.7%
> 846341.5573.0%54.2%   -19%   0.513   12104
>  2.13   97910   14743 14987   98.4%
> 254.5%199.3%
> 979080.16   276.2%  4899.9%   -23%   0.473   14037
>  1.90  110260   16846 16940   99.4%
> 299.2%303.3%
> 1102450.06   325.0%   -99.9%   -17%   0.422   15995
>  1.74  118354   18629 18744   99.4%
>1062.0%   1043.6%
> 118317   -0.20  1156.4%   -99.9%   -13%   0.380   17414
>  1.61  122958   20193 20331   99.3%
> 967.5%   1571.1%
> 1228680.10  1059.7%   987.3%-2%   0.402   18348
>  1.51  125075   21554 21794   98.9%
> 838.9%   1355.1%
> 1249330.08   922.6%  1116.9%-1%   0.402   18977
>  1.42   72057   17042 23233   73.4%
> 640.8%775.3%
> 703910.08   732.5%   826.7%-8%   0.425   10003
> total  823746  136245144117   94.5%
> 166.4%166.7%
> 8215620.40   181.1%   296.7%   -15%   0.435  119774
> 
> Note that I/SIGMA of each resolution shell is <2.5, so
> how should I do to
> process the dataset properly? Any suggestion about this
> super ice rings?
> Thanks!
> 
> Tiantian
> 
> -- 
> Shanghai Institute of Materia Medica, Chinese Academy of
> Sciences
> Address: Room 101, 646 Songtao Road, Zhangjiang Hi-Tech
> Park,
> Shanghai, 201203

Dr Stefan Gerhardt
Albert-Ludwigs-Universität Freiburg
Inst.f.Org.Chem.u.Biochem
Albertstrasse 21
79104 Freiburg
Tel. +49 761 2035970
Fax. +49 761 2036161