Dear all,

Recently I have measure a set of kinetic data of a receptor ligand
interaction using BIAcore 3000. To my surprise, the dissociation rate is
very low (~ 10e-6). During the measurement, I use a long dissociation time
(2 hours) .I repeat several time which give very similar results. So I am
wondering if the BIAcore can measure such a low off-rate kinetic. What is
the limitation of BIAcore? Any review about that?

Thanks in advance.

Xianchi Dong
Research Fellow
Children's Hospital Boston
Harvard Medical School


Re: [ccp4bb] off topic question BIAcore problem

2013-02-20 Thread Bosch, Juergen
I assume you use CM5 chips ?
I further assume you run at pH 7.5 perhaps ? What's the pI of your analyte ?  
7.5 ? Do you get significant binding to your reference cell under the 
conditions you are running ?
You might get rebinding to your negatively charged surface and the dissociation 
you are measuring might not really be correct (masked through rebinding)
Check that first I would say.
You can measure low picomolar dissociations. I recently had one with ~80 pM.

Jürgen

..
Jürgen Bosch
Johns Hopkins University
Bloomberg School of Public Health
Department of Biochemistry  Molecular Biology
Johns Hopkins Malaria Research Institute
615 North Wolfe Street, W8708
Baltimore, MD 21205
Office: +1-410-614-4742
Lab:  +1-410-614-4894
Fax:  +1-410-955-2926
http://lupo.jhsph.edu

On Feb 20, 2013, at 12:03 PM, xianchi dong wrote:

Dear all,

Recently I have measure a set of kinetic data of a receptor ligand interaction 
using BIAcore 3000. To my surprise, the dissociation rate is very low (~ 
10e-6). During the measurement, I use a long dissociation time (2 hours) .I 
repeat several time which give very similar results. So I am wondering if the 
BIAcore can measure such a low off-rate kinetic. What is the limitation of 
BIAcore? Any review about that?

Thanks in advance.

Xianchi Dong
Research Fellow
Children's Hospital Boston
Harvard Medical School






Re: [ccp4bb] off topic question BIAcore problem

2013-02-20 Thread Tomas Malinauskas
Dear Xianchi,
unfortunately, dissociation rate constant kd 10^-6 s^-1 was just
beyond the limit of Biacore in 1999 (e.g. see Fig. 1 in
http://www.ncbi.nlm.nih.gov/pubmed/10556876). I am not sure about
these days.
As Juergen noted, you may have a problem with rebinding (you could try
to reduce amount of the ligand on the chip, see other tricks here
http://users.path.ox.ac.uk/~vdmerwe/internal/spr.pdf).
Regarding reviews, David Myszka writes enjoyable reviews on SPR.
Hope that helps,
Tomas

On Wed, Feb 20, 2013 at 5:03 PM, xianchi dong dongxian...@gmail.com wrote:
 Dear all,

 Recently I have measure a set of kinetic data of a receptor ligand
 interaction using BIAcore 3000. To my surprise, the dissociation rate is
 very low (~ 10e-6). During the measurement, I use a long dissociation time
 (2 hours) .I repeat several time which give very similar results. So I am
 wondering if the BIAcore can measure such a low off-rate kinetic. What is
 the limitation of BIAcore? Any review about that?

 Thanks in advance.

 Xianchi Dong
 Research Fellow
 Children's Hospital Boston
 Harvard Medical School

Dea


Re: [ccp4bb] off topic question BIAcore problem

2013-02-20 Thread Zhijie Li
Hi Xianchi,

First of all: a very slow dissociation rate can also be an artifact: the 
analyte can be simply precipitating on the surface. You do need to rule this 
out by proper controls. 

But there is no rule saying that 10e-6 s-1 off rate is not realistic. Even with 
protein-small molecule binding, one can get extremely slow dissociation if the 
interaction is very strong. For example, the dissociation of biotin from 
streptavidin has a rate constant of the 10e-6 s(-1) order. A very slow 
dissociation rate is often (if not always) correlated with very tight binding 
(for example KD in picomolar range or even smaller). What is your calculated 
KD? The binding phase of the BIAcore curves should also reflect the fact that 
the off rate is low (the Kon-obs has an off rate term).

We have measured some pico molar bindings with BIAcore in the past. It is 
doable, but difficult. You may also want to consider in-solution methods (so 
that you do not need to worry about artifacts caused by the surface), such as 
ITC (by competition method), or fluorescence-based methods. 

For BIAcore, when working with very strong bindings (KD 100pM), there are a 
few things to consider:
1) You may find great difficulties regenerating the chip - probably the biggest 
concern with BIAcore (considering the ridiculous price of chips).
2) How much RU should you conjugate to the chip? With strong interactions, as 
low as possible amount of your ligand should be labeled on the chip, for 4 
purposes: a) to make regeneration easier; b) to reduce rebinding effect; c) to 
reduce mass transfer effect; d) to make sure you do not take away significant 
amount of analyte from the solution (discussed in 3)).
3) If you plan to span the KD range with the analyte, then if the KD is in 
picomolar molar range, you are supplying the surface with extremely dilute 
analyte solutions. In such case, you need to calculate if your flow rate is 
high enough, to compensate the loss of solute due to binding to surface, 
otherwise the real concentration of the analyte in the mobile phase will be 
much lower than the assumed analyte concentration.
4) The association phase of the BIAcore experiment is also affected by the 
dissociation rate.  The observed binding rate Kon-obs contains a Koff term. 
When you are loading the analyte at near KD concentrations, the binding will 
take similar amount of time as the dissociation phase to reach plateau. The 
Kon-obs also contains a concentration terml, so the time required for reaching 
plateau will be shorter and shorter when you load with higher concentrations of 
the analytes.
5) The only proper way of labeling chip for slow dissociations is by covalent 
means.

HTH,

Zhijie




From: xianchi dong 
Sent: Wednesday, February 20, 2013 12:03 PM
To: CCP4BB@JISCMAIL.AC.UK 
Subject: [ccp4bb] off topic question BIAcore problem


Dear all, 


Recently I have measure a set of kinetic data of a receptor ligand interaction 
using BIAcore 3000. To my surprise, the dissociation rate is very low (~ 
10e-6). During the measurement, I use a long dissociation time (2 hours) .I 
repeat several time which give very similar results. So I am wondering if the 
BIAcore can measure such a low off-rate kinetic. What is the limitation of 
BIAcore? Any review about that?


Thanks in advance.


Xianchi Dong
Research Fellow 
Children's Hospital Boston
Harvard Medical School

Re: [ccp4bb] off topic question BIAcore problem

2013-02-20 Thread xianchi dong
Thanks a lot for the kind reply.

I used CM5 chip at pH 7.4. I am not quite sure about the PI of my protein
but I have different truncations of my protein which behaved similarly.
I can fit my data quite good with the Chi2 around 1 and Rmax 150. The kon
is about 10e4 and the KD is around 0.2 nM. The concentration of analytes I
used was from 100 nM to 5 nM. I  used a control buffer which can disrupt
the binding. In that buffer, no binding was observed.

I also used fluorescence anisotropy to measure the Ki of the interaction
which showed a 1 nM binding compared with the 0.2 nM binding in the SPR
assay.


On Wed, Feb 20, 2013 at 12:03 PM, xianchi dong dongxian...@gmail.comwrote:

 Dear all,

 Recently I have measure a set of kinetic data of a receptor ligand
 interaction using BIAcore 3000. To my surprise, the dissociation rate is
 very low (~ 10e-6). During the measurement, I use a long dissociation time
 (2 hours) .I repeat several time which give very similar results. So I am
 wondering if the BIAcore can measure such a low off-rate kinetic. What is
 the limitation of BIAcore? Any review about that?

 Thanks in advance.

 Xianchi Dong
 Research Fellow
 Children's Hospital Boston
 Harvard Medical School



Re: [ccp4bb] off topic question BIAcore problem

2013-02-20 Thread Bosch, Juergen
Well that looks pretty real then. You might have wrong concentrations in one or 
the other experiment perhaps hence the difference.

Jürgen

On Feb 20, 2013, at 5:45 PM, xianchi dong wrote:

Thanks a lot for the kind reply.

I used CM5 chip at pH 7.4. I am not quite sure about the PI of my protein but I 
have different truncations of my protein which behaved similarly.
I can fit my data quite good with the Chi2 around 1 and Rmax 150. The kon is 
about 10e4 and the KD is around 0.2 nM. The concentration of analytes I used 
was from 100 nM to 5 nM. I  used a control buffer which can disrupt the 
binding. In that buffer, no binding was observed.

I also used fluorescence anisotropy to measure the Ki of the interaction which 
showed a 1 nM binding compared with the 0.2 nM binding in the SPR assay.


On Wed, Feb 20, 2013 at 12:03 PM, xianchi dong 
dongxian...@gmail.commailto:dongxian...@gmail.com wrote:
Dear all,

Recently I have measure a set of kinetic data of a receptor ligand interaction 
using BIAcore 3000. To my surprise, the dissociation rate is very low (~ 
10e-6). During the measurement, I use a long dissociation time (2 hours) .I 
repeat several time which give very similar results. So I am wondering if the 
BIAcore can measure such a low off-rate kinetic. What is the limitation of 
BIAcore? Any review about that?

Thanks in advance.

Xianchi Dong
Research Fellow
Children's Hospital Boston
Harvard Medical School


..
Jürgen Bosch
Johns Hopkins University
Bloomberg School of Public Health
Department of Biochemistry  Molecular Biology
Johns Hopkins Malaria Research Institute
615 North Wolfe Street, W8708
Baltimore, MD 21205
Office: +1-410-614-4742
Lab:  +1-410-614-4894
Fax:  +1-410-955-2926
http://lupo.jhsph.edu






Re: [ccp4bb] off topic question BIAcore problem

2013-02-20 Thread Zhijie Li
Hi Xianchi,

How did you treat the control flow cell? And what is the composition of the 
control buffer that you use to disrupt binding? Is it denaturing or a mild 
condition? 

If the Rmax=150 and Koff=1e-6, then at 7200 s (2 hr), the signal will only drop 
1RU from 150RU (assuming you can reach the Rmax), is that what you see on the 
dissociation curve? In reality, when loading analyte at 100nM, assuming you 
have Kon=1e4, Koff=1e-6, binding for 30min would only let you reach ~125 RU, 
and at the lower concentrations will be much worse. Consequently the total 
dissociation after 7200s would be less than 1 RU on most curves - I am not sure 
if the Biacore 3000 will have a stable enough baseline for you to confidently 
measure that. If you try to fit the dissociation phase alone to get the Koff, 
the reported Koff won't be too accurate due to the low S/N ratio. On the other 
hand the 5-100nM spanning would generate significant changes on the binding 
phase, thus the global fitting should be able to get a relatively accurate KD. 
0.2nM from SPR and 1nM from in-solution experiment also sounds reasonable.

Zhijie



From: xianchi dong 
Sent: Wednesday, February 20, 2013 5:45 PM
To: CCP4BB@JISCMAIL.AC.UK 
Subject: Re: [ccp4bb] off topic question BIAcore problem


Thanks a lot for the kind reply. 


I used CM5 chip at pH 7.4. I am not quite sure about the PI of my protein but I 
have different truncations of my protein which behaved similarly. 
I can fit my data quite good with the Chi2 around 1 and Rmax 150. The kon is 
about 10e4 and the KD is around 0.2 nM. The concentration of analytes I used 
was from 100 nM to 5 nM. I  used a control buffer which can disrupt the 
binding. In that buffer, no binding was observed. 


I also used fluorescence anisotropy to measure the Ki of the interaction which 
showed a 1 nM binding compared with the 0.2 nM binding in the SPR assay. 



On Wed, Feb 20, 2013 at 12:03 PM, xianchi dong dongxian...@gmail.com wrote:

  Dear all, 


  Recently I have measure a set of kinetic data of a receptor ligand 
interaction using BIAcore 3000. To my surprise, the dissociation rate is very 
low (~ 10e-6). During the measurement, I use a long dissociation time (2 hours) 
.I repeat several time which give very similar results. So I am wondering if 
the BIAcore can measure such a low off-rate kinetic. What is the limitation of 
BIAcore? Any review about that?


  Thanks in advance.


  Xianchi Dong
  Research Fellow 
  Children's Hospital Boston
  Harvard Medical School