2008-12-08 Thread Pius Padayatti
nanodrop system is wonderful. it helps very much in solutions with
high interference from detergents etc etc.

i highly receommend nanodrop specs

PSP

On Thu, Dec 4, 2008 at 10:16 AM, Tim Gruene [EMAIL PROTECTED] wrote:
 Dear all,

 we would like to purchase a UV spectrometer for measuring protein
 concentrations (280nm), and I would like to here your comments and
 especially recommendations.

 We don't need anything fancy, a small, fast device would be sufficient.

 Tim


 --
 Tim Gruene
 Institut fuer anorganische Chemie
 Tammannstr. 4
 D-37077 Goettingen

 GPG Key ID = A46BEE1A




-- 
Pius S Padayatti
Scientist,
Polgenix, Inc.
11000 Cedar Ave, Suite 260
Cleveland, OH 44106
Phone: 216-658-4528
Fax: 216-658-4529


Re: [ccp4bb] AW: [ccp4bb] suggestions for UV spectrometer

2008-12-08 Thread Jose Antonio Cuesta-Seijo

I also have to come in defense of the nanodrop here.
I have measured up to A280 = 98 and the curve is always reasonably  
smooth, spikes normally mean that bubbles have formed. And proper  
cleaning seems to be rubbing with a kimwipe three or four times  
after each drop. If the last user does not clean after the last  
measurement, then things will dry up in the pedestals and that can of  
course be a problem. Also bubbles do not form if you are careful. I  
always use 2uL, 1uL can be too little, and even with 20% detergent I  
get a column and no bubbles nearly every time. You have to pipet  
carefully and lower the lever slowly (think of a vinyl record here).  
For tricky samples, use 2.2uL instead. A bubble might still form  
occasionally, but looking at the spectrum will quickly tell you that  
something went wrong.


Cheers,

Jose.

**
Jose Antonio Cuesta-Seijo
Cancer Genomics and Proteomics
Ontario Cancer Institute, UHN
MaRS TMDT Room 4-902M
101 College Street
M5G 1L7 Toronto, ON, Canada
Phone:  (416)581-7544
Fax: (416)581-7562
email: [EMAIL PROTECTED]
**


On Dec 5, 2008, at 7:00 PM, wangsa tirta ismaya wrote:


Dear all,

Thanks a lot for raising the issue with the not reproducibility of  
protein measurement with Nanodrop. We use the instrument as a  
workhorse in the lab. Indeed, recently I observed that the protein  
concentration suggested by Nanodrop is sometimes differ to the  
usual colorimetric measurement (Bradford method, measured with our  
Pharmacia's Ultrospec 2000 spectrometer). Since the cell in  
Nanodrop is very small, could it be due to the homogeneity of the  
sample in the cell? Also what I have observed, we have to be sure  
that the cell is cleaned properly before use. Another thing is, at  
high protein concentration I obtained noisy absorption curve at the  
peak (like a seismograf ) thus the protein concentration  
measured is doubtful, I have to dilute the sample to have good  
curve (thank God it requires only 2 mikroliter for a measurement).  
Well, I think nanodrop is a good, fast, and powerful instrument,  
however it would be better if we established a reference of our  
daily practice to a normal spectrometer measurement.


cheers,


Wangsa

2008/12/5 Martin Hallberg [EMAIL PROTECTED]
Which brings us back to the Hellma TrayCell solution where you  
can, from the same spectrometer, have both the cuvette option and  
the quickness of the NanoDrop/NanoVue system.


Anyone that can comment on the performance of the TrayCell from  
Hellma?


Cheers,

Martin


On Dec 5, 2008, at 9:06 AM, Gregor Witte wrote:

Agree!
I think for crystallographic use the nanodrop is perfectly okay to  
see if the protein is 5mg/ml or 30mg/ml. But in fact I also do not  
trust our instrument if it comes to more important issues like  
preparing solutions for titrations or assays. And due to the small  
pathlength I do not trust absorptions of small concentrated samples  
at all. I always prefer a real 2-beam spectrophotometer  
(monochromators) with a quarz-cuvette and a nice pathlength. Of  
course, you cannot reliably measure solutions exceeding Abs 1 or  
maybe 1.5 OD in a spectrophotometer with 1cm pathlength.


There's also one quite strange thing about the nanodrop – they sell  
the calibration check solution (which is some kind of yellow  
chromate-solution with known absorption), then you check your  
nanodrop with it and maybe find out that it's off to some certain  
extend: But then you're stuck(!), because you cannot calibrate it  
on your own. I guess it would be quite easy to integrate a  
calibration-option into the software, but at the moment the  
instrument tells you calibration failure and you have to call the  
service guys who then carry it home and calibrate it by turning of  
the two small screws at the top of it and then glue them with  
locktite.
Anyway, at least for our mid to high concentrated samples the  
nanodrop is not showing large fluctuations so we are happy with it.  
But everyone using a nanodrop should check it from time to time –  
as I found out that ours was off more than 20% at one day - which  
raised some trouble of course…


Cheers,

Gregor


Von: CCP4 bulletin board [mailto:[EMAIL PROTECTED] Im Auftrag  
von Filip Van Petegem

Gesendet: Donnerstag, 4. Dezember 2008 22:20
An: CCP4BB@JISCMAIL.AC.UK
Betreff: Re: [ccp4bb] suggestions for UV spectrometer

I want to add I absotely hate the nanodrop.  We've had a demo for  
it, and found the readouts to be very unreliable.  Fluctuations of  
20% and more. Just leaving the same drop in and measuring the  
sample multiple times gives different values (going in both  
directions, so not only due to evaportations). Sure, it's easy and  
fast, and maybe good to have a rough idea about your protein  
concentration, but I would never want to use it for exact  
measurements such as needed for e.g. a CD or an ITC instrument.  
I've heard

Re: [ccp4bb] AW: [ccp4bb] suggestions for UV spectrometer

2008-12-08 Thread Edward Snell
Getting even further off topic, we had a Nanodrop demo for trial shortly
after hearing that a significant number of crystallization groups were
using them. I won't go into the trial details, but I'd be interested in
hearing from people (via email rather then the ccp4 postings) who did
systematic trials in comparison to more conventional spectrometers;
especially concerning the accuracy and reproducibility. The ease of use
was very nice but we did not buy one.

Thanks,

Eddie.

Edward Snell Ph.D.
Assistant Prof. Department of Structural Biology, SUNY Buffalo,
Hauptman-Woodward Medical Research Institute
700 Ellicott Street, Buffalo, NY 14203-1102
Phone: (716) 898 8631 Fax: (716) 898 8660
Email: [EMAIL PROTECTED]  Telepathy: 42.2 GHz
 
Heisenberg was probably here!
 

-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Jose Antonio Cuesta-Seijo
Sent: Monday, December 08, 2008 12:54 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] AW: [ccp4bb] suggestions for UV spectrometer

I also have to come in defense of the nanodrop here.
I have measured up to A280 = 98 and the curve is always reasonably  
smooth, spikes normally mean that bubbles have formed. And proper  
cleaning seems to be rubbing with a kimwipe three or four times  
after each drop. If the last user does not clean after the last  
measurement, then things will dry up in the pedestals and that can of  
course be a problem. Also bubbles do not form if you are careful. I  
always use 2uL, 1uL can be too little, and even with 20% detergent I  
get a column and no bubbles nearly every time. You have to pipet  
carefully and lower the lever slowly (think of a vinyl record here).  
For tricky samples, use 2.2uL instead. A bubble might still form  
occasionally, but looking at the spectrum will quickly tell you that  
something went wrong.

Cheers,

Jose.

**
Jose Antonio Cuesta-Seijo
Cancer Genomics and Proteomics
Ontario Cancer Institute, UHN
MaRS TMDT Room 4-902M
101 College Street
M5G 1L7 Toronto, ON, Canada
Phone:  (416)581-7544
Fax: (416)581-7562
email: [EMAIL PROTECTED]
**


On Dec 5, 2008, at 7:00 PM, wangsa tirta ismaya wrote:

 Dear all,

 Thanks a lot for raising the issue with the not reproducibility of  
 protein measurement with Nanodrop. We use the instrument as a  
 workhorse in the lab. Indeed, recently I observed that the protein  
 concentration suggested by Nanodrop is sometimes differ to the  
 usual colorimetric measurement (Bradford method, measured with our  
 Pharmacia's Ultrospec 2000 spectrometer). Since the cell in  
 Nanodrop is very small, could it be due to the homogeneity of the  
 sample in the cell? Also what I have observed, we have to be sure  
 that the cell is cleaned properly before use. Another thing is, at  
 high protein concentration I obtained noisy absorption curve at the  
 peak (like a seismograf ) thus the protein concentration  
 measured is doubtful, I have to dilute the sample to have good  
 curve (thank God it requires only 2 mikroliter for a measurement).  
 Well, I think nanodrop is a good, fast, and powerful instrument,  
 however it would be better if we established a reference of our  
 daily practice to a normal spectrometer measurement.

 cheers,


 Wangsa

 2008/12/5 Martin Hallberg [EMAIL PROTECTED]
 Which brings us back to the Hellma TrayCell solution where you  
 can, from the same spectrometer, have both the cuvette option and  
 the quickness of the NanoDrop/NanoVue system.

 Anyone that can comment on the performance of the TrayCell from  
 Hellma?

 Cheers,

 Martin


 On Dec 5, 2008, at 9:06 AM, Gregor Witte wrote:

 Agree!
 I think for crystallographic use the nanodrop is perfectly okay to  
 see if the protein is 5mg/ml or 30mg/ml. But in fact I also do not  
 trust our instrument if it comes to more important issues like  
 preparing solutions for titrations or assays. And due to the small  
 pathlength I do not trust absorptions of small concentrated samples  
 at all. I always prefer a real 2-beam spectrophotometer  
 (monochromators) with a quarz-cuvette and a nice pathlength. Of  
 course, you cannot reliably measure solutions exceeding Abs 1 or  
 maybe 1.5 OD in a spectrophotometer with 1cm pathlength.

 There's also one quite strange thing about the nanodrop - they sell  
 the calibration check solution (which is some kind of yellow  
 chromate-solution with known absorption), then you check your  
 nanodrop with it and maybe find out that it's off to some certain  
 extend: But then you're stuck(!), because you cannot calibrate it  
 on your own. I guess it would be quite easy to integrate a  
 calibration-option into the software, but at the moment the  
 instrument tells you calibration failure and you have to call the  
 service guys who then carry it home and calibrate it by turning of  
 the two small screws at the top of it and then glue them

Re: [ccp4bb] AW: [ccp4bb] suggestions for UV spectrometer

2008-12-06 Thread Chun Luo
We have used Nanodrop for several years and found the readings are always
accurate. The highest concentration we have measured is around 30 mg/ml. The
differences between diluted and concentrated samples are within dilution
error. Nanodrop spectra at low concentration are noisier. We actually prefer
to have protein concentrations over 0.5 mg/ml.

 

The key to get good Nanodrop reading is to have a nice liquid column. Take a
look whiling waiting for the results. High concentration samples with
detergent or glycerol may produce bubbles (giving an error message
sometimes) or improperly formed liquid column (giving incorrect numbers).
Use more samples (up to 5 ul) and gently push down the head will help the
liquid column formation. 1 ul is not enough for most protein samples. If the
reading is not good, wipe clean the surface and load a new sample. Repeated
reading of same sample gave huge variation. The liquid column probably has
difficulty to form properly after the first reading.

 

We provide protein purification contract services and send proteins to many
users around the globe. The proteins have been used for crystallography,
ITC, DSC, Biacore, and activity assays. So far we haven’t heard any complain
about our concentration measurement using Nanodrop.

 

We also use Bradford to measure concentration if required by our customers
or when concentration can’t be measured by A280. The error range for
Bradford is much wider and easily off by a factor of 2 (by repeating or
comparing to Nanodrop). There are different brands of Bradford reagent. Each
gives a different number. We once had a big discrepancy with a customer. It
turns out the home-made BSA standard lot was off. After adjusting the BSA
concentration with Nanodrop and using the same Bradford reagent and same
curve fitting method, the concentrations agreed. The lesson leaned is to
have good fresh standard solution.

 

The color of Bradford changes over time, it’s better to use a plate reader
than single sample spectrometer. Estimation by eye is just as good as a
spectrometer:-)

 

--Chun

 

  _  

From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of
Juergen Bosch
Sent: Saturday, December 06, 2008 12:29 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] AW: [ccp4bb] suggestions for UV spectrometer

 

Hi,

 

really strange, I always dilute my protein when taking an absorption
spectra. I try to adjust my expected concentration to a readout of ~ 0.2-0.3
OD280. And the Bradford is just as well as 'picking house numbers',
depending on your protein, you can underestimate your protein concentration
by a factor of 2.

For my current approach I use a 100 µl volume cuvette and 2 µl of my
concentrated protein (most of the times). The Structural Genomics lab
(MSGPP) has a Nanodrop and I was also always happy with those results (but
again I dilute my protein sample before taking a measurement) 

 

Jürgen

 

On 5 Dec 2008, at 16:00, wangsa tirta ismaya wrote:





Dear all,

Thanks a lot for raising the issue with the not reproducibility of protein
measurement with Nanodrop. We use the instrument as a workhorse in the lab.
Indeed, recently I observed that the protein concentration suggested by
Nanodrop is sometimes differ to the usual colorimetric measurement (Bradford
method, measured with our Pharmacia's Ultrospec 2000 spectrometer). Since
the cell in Nanodrop is very small, could it be due to the homogeneity of
the sample in the cell? Also what I have observed, we have to be sure that
the cell is cleaned properly before use. Another thing is, at high protein
concentration I obtained noisy absorption curve at the peak (like a
seismograf ) thus the protein concentration measured is doubtful, I have
to dilute the sample to have good curve (thank God it requires only 2
mikroliter for a measurement). Well, I think nanodrop is a good, fast, and
powerful instrument, however it would be better if we established a
reference of our daily practice to a normal spectrometer measurement.

cheers,


Wangsa

2008/12/5 Martin Hallberg [EMAIL PROTECTED]

Which brings us back to the Hellma TrayCell solution where you can, from
the same spectrometer, have both the cuvette option and the quickness of the
NanoDrop/NanoVue system.

Anyone that can comment on the performance of the TrayCell from Hellma?

Cheers,

Martin



On Dec 5, 2008, at 9:06 AM, Gregor Witte wrote:

Agree!
I think for crystallographic use the nanodrop is perfectly okay to see if
the protein is 5mg/ml or 30mg/ml. But in fact I also do not trust our
instrument if it comes to more important issues like preparing solutions for
titrations or assays. And due to the small pathlength I do not trust
absorptions of small concentrated samples at all. I always prefer a real
2-beam spectrophotometer (monochromators) with a quarz-cuvette and a nice
pathlength. Of course, you cannot reliably measure solutions exceeding Abs 1
or maybe 1.5 OD in a spectrophotometer with 1cm pathlength.

There's also

[ccp4bb] AW: [ccp4bb] suggestions for UV spectrometer

2008-12-05 Thread Gregor Witte
Agree! 

I think for crystallographic use the nanodrop is perfectly okay to see if
the protein is 5mg/ml or 30mg/ml. But in fact I also do not trust our
instrument if it comes to more important issues like preparing solutions for
titrations or assays. And due to the small pathlength I do not trust
absorptions of small concentrated samples at all. I always prefer a real
2-beam spectrophotometer (monochromators) with a quarz-cuvette and a nice
pathlength. Of course, you cannot reliably measure solutions exceeding Abs 1
or maybe 1.5 OD in a spectrophotometer with 1cm pathlength.



There's also one quite strange thing about the nanodrop - they sell the
calibration check solution (which is some kind of yellow chromate-solution
with known absorption), then you check your nanodrop with it and maybe find
out that it's off to some certain extend: But then you're stuck(!), because
you cannot calibrate it on your own. I guess it would be quite easy to
integrate a calibration-option into the software, but at the moment the
instrument tells you calibration failure and you have to call the service
guys who then carry it home and calibrate it by turning of the two small
screws at the top of it and then glue them with locktite.

Anyway, at least for our mid to high concentrated samples the nanodrop is
not showing large fluctuations so we are happy with it. But everyone using a
nanodrop should check it from time to time - as I found out that ours was
off more than 20% at one day - which raised some trouble of course.

 

Cheers,

 

Gregor

 

 

Von: CCP4 bulletin board [mailto:[EMAIL PROTECTED] Im Auftrag von Filip
Van Petegem
Gesendet: Donnerstag, 4. Dezember 2008 22:20
An: CCP4BB@JISCMAIL.AC.UK
Betreff: Re: [ccp4bb] suggestions for UV spectrometer

 

I want to add I absotely hate the nanodrop.  We've had a demo for it, and
found the readouts to be very unreliable.  Fluctuations of 20% and more.
Just leaving the same drop in and measuring the sample multiple times gives
different values (going in both directions, so not only due to
evaportations). Sure, it's easy and fast, and maybe good to have a rough
idea about your protein concentration, but I would never want to use it for
exact measurements such as needed for e.g. a CD or an ITC instrument. I've
heard other labs in our department have similar issues.  We've also had a
demo for the Nanovue from GE Healthcare:  same issues - very large
fluctuations from one sample to another.  I suppose this is simply an
inherent problem with small volumes...

Cheers

Filip Van Petegm




On Thu, Dec 4, 2008 at 12:48 PM, Patrick Loll [EMAIL PROTECTED] wrote:

At the risk of dragging this discussion even further afield from
crystallography:

 

How can you get realistic numbers for concentrated solutions using the
Nanodrop?  I understand that the instrument reduces absorbance by using a
very short path length. However, I thought that in order for the
Beer-Lambert formalism to be applicable, the solution needs to be
sufficiently dilute so that the chance of molecules shadowing one another
is negligible. Isn't this condition violated for concentrated solutions
(even with short path lengths)?

 

Pat

 

On 4 Dec 2008, at 1:27 PM, Michael Giffin wrote:





We also like the Nanodrop...


---

Patrick J. Loll, Ph. D.  

Professor of Biochemistry  Molecular Biology

Director, Biochemistry Graduate Program

Drexel University College of Medicine

Room 10-102 New College Building

245 N. 15th St., Mailstop 497

Philadelphia, PA  19102-1192  USA

 

(215) 762-7706

[EMAIL PROTECTED]

 




-- 
Filip Van Petegem, PhD
Assistant Professor
The University of British Columbia
Dept. of Biochemistry and Molecular Biology
2350 Health Sciences Mall - Rm 2.356
Vancouver, V6T 1Z3

phone: +1 604 827 4267
email: [EMAIL PROTECTED]
http://crg.ubc.ca/VanPetegem/



Re: [ccp4bb] AW: [ccp4bb] suggestions for UV spectrometer

2008-12-05 Thread Martin Hallberg
Which brings us back to the Hellma TrayCell solution where you can,  
from the same spectrometer, have both the cuvette option and the  
quickness of the NanoDrop/NanoVue system.


Anyone that can comment on the performance of the TrayCell from Hellma?

Cheers,

Martin

On Dec 5, 2008, at 9:06 AM, Gregor Witte wrote:


Agree!
I think for crystallographic use the nanodrop is perfectly okay to  
see if the protein is 5mg/ml or 30mg/ml. But in fact I also do not  
trust our instrument if it comes to more important issues like  
preparing solutions for titrations or assays. And due to the small  
pathlength I do not trust absorptions of small concentrated samples  
at all. I always prefer a “real” 2-beam spectrophotometer  
(monochromators) with a quarz-cuvette and a nice pathlength. Of  
course, you cannot reliably measure solutions exceeding Abs 1 or  
maybe 1.5 OD in a spectrophotometer with 1cm pathlength.


There’s also one quite strange thing about the nanodrop – they sell  
the “calibration check solution” (which is some kind of yellow  
chromate-solution with known absorption), then you check your  
nanodrop with it and maybe find out that it’s off to some certain  
extend: But then you’re stuck(!), because you cannot calibrate it on  
your own. I guess it would be quite easy to integrate a calibration- 
option into the software, but at the moment the instrument tells you  
“calibration failure” and you have to call the service guys who then  
carry it home and calibrate it by turning of the two small screws at  
the top of it and then glue them with locktite.
Anyway, at least for our mid to high concentrated samples the  
nanodrop is not showing large fluctuations so we are happy with it.  
But everyone using a nanodrop should check it from time to time – as  
I found out that ours was off more than 20% at one day - which  
raised some trouble of course…


Cheers,

Gregor


Von: CCP4 bulletin board [mailto:[EMAIL PROTECTED] Im Auftrag  
von Filip Van Petegem

Gesendet: Donnerstag, 4. Dezember 2008 22:20
An: CCP4BB@JISCMAIL.AC.UK
Betreff: Re: [ccp4bb] suggestions for UV spectrometer

I want to add I absotely hate the nanodrop.  We've had a demo for  
it, and found the readouts to be very unreliable.  Fluctuations of  
20% and more. Just leaving the same drop in and measuring the sample  
multiple times gives different values (going in both directions, so  
not only due to evaportations). Sure, it's easy and fast, and maybe  
good to have a rough idea about your protein concentration, but I  
would never want to use it for exact measurements such as needed for  
e.g. a CD or an ITC instrument. I've heard other labs in our  
department have similar issues.  We've also had a demo for the  
Nanovue from GE Healthcare:  same issues - very large fluctuations  
from one sample to another.  I suppose this is simply an inherent  
problem with small volumes...


Cheers

Filip Van Petegm


On Thu, Dec 4, 2008 at 12:48 PM, Patrick Loll [EMAIL PROTECTED]  
wrote:
At the risk of dragging this discussion even further afield from  
crystallography:


How can you get realistic numbers for concentrated solutions using  
the Nanodrop?  I understand that the instrument reduces absorbance  
by using a very short path length. However, I thought that in order  
for the Beer-Lambert formalism to be applicable, the solution needs  
to be sufficiently dilute so that the chance of molecules  
shadowing one another is negligible. Isn't this condition violated  
for concentrated solutions (even with short path lengths)?


Pat

On 4 Dec 2008, at 1:27 PM, Michael Giffin wrote:


We also like the Nanodrop...
---
Patrick J. Loll, Ph. D.
Professor of Biochemistry  Molecular Biology
Director, Biochemistry Graduate Program
Drexel University College of Medicine
Room 10-102 New College Building
245 N. 15th St., Mailstop 497
Philadelphia, PA  19102-1192  USA

(215) 762-7706
[EMAIL PROTECTED]




--
Filip Van Petegem, PhD
Assistant Professor
The University of British Columbia
Dept. of Biochemistry and Molecular Biology
2350 Health Sciences Mall - Rm 2.356
Vancouver, V6T 1Z3

phone: +1 604 827 4267
email: [EMAIL PROTECTED]
http://crg.ubc.ca/VanPetegem/


.
B. Martin Hallberg, PhD
Assistant professor
Department of Cell and Molecular Biology
Medical Nobel Institute
Karolinska Institutet
Von Eulersv. 1
SE-171 77 Stockholm
Sweden
Fax: +46-8-339380


[ccp4bb] suggestions for UV spectrometer

2008-12-04 Thread Tim Gruene

Dear all,

we would like to purchase a UV spectrometer for measuring protein 
concentrations (280nm), and I would like to here your comments and 
especially recommendations.


We don't need anything fancy, a small, fast device would be sufficient.

Tim


--
Tim Gruene
Institut fuer anorganische Chemie
Tammannstr. 4
D-37077 Goettingen

GPG Key ID = A46BEE1A


Re: [ccp4bb] suggestions for UV spectrometer

2008-12-04 Thread Chavas Leo

Dear Tim --

On 4 Dec 2008, at 15:16, Tim Gruene wrote:
we would like to purchase a UV spectrometer for measuring protein  
concentrations (280nm), and I would like to here your comments and  
especially recommendations.


I love the Nanodrop system... I know you said you don't want anything  
fancy, but I like to spend as little protein as possible when  
measuring its concentration, and I like it fast and reproducible.  
Plus, in my hands, the Nanodrop system give very reproducible  
results. Small problem... expensive...

You can find infos there:
http://www.nanodrop.com/nd-1000-overview.html

HTH
Kind regards.

-- Leo --

Chavas Leonard, Ph.D. @ home
Research Associate
Marie Curie Actions Fellow

Faculty of Life Sciences
The University of Manchester
The Michael Smith Building
Oxford Road
Manchester Lancashire
M13 9PT

Tel: +44(0)161-275-1586
e-mail: [EMAIL PROTECTED]
http://personalpages.manchester.ac.uk/staff/leonard.chavas/




Re: [ccp4bb] suggestions for UV spectrometer

2008-12-04 Thread James M. Vergis
I would also recommend the nanodrop.  It takes a whole spectra every
measurement and there is no need to dilute your sample.  You can demo it for
a week and try it out.  



James M. Vergis, Ph.D.
University of Virginia Molecular Physiology and Biological Physics
MKWEINR 360A Snyder Building
480 Ray C. Hunt Drive
PO Box  800886
Charlottesville, VA 22908-0886
phone: 434-243-2730   FAX: 434-243-8271
[EMAIL PROTECTED]



-Original Message-
From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of Tim
Gruene
Sent: Thursday, December 04, 2008 10:16 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: [ccp4bb] suggestions for UV spectrometer

Dear all,

we would like to purchase a UV spectrometer for measuring protein 
concentrations (280nm), and I would like to here your comments and 
especially recommendations.

We don't need anything fancy, a small, fast device would be sufficient.

Tim


--
Tim Gruene
Institut fuer anorganische Chemie
Tammannstr. 4
D-37077 Goettingen

GPG Key ID = A46BEE1A


Re: [ccp4bb] suggestions for UV spectrometer

2008-12-04 Thread Jim Fairman
We use a Beckman Coulter DU730.  It has a small footprint if lab space is an
issue.  It will do single-wavelength, multi-wavelength, or take an entire
spectrum in 0.5 nm steps if you desire.  It comes with the standard 1 ml
cuvette holder, but we also purchased the microcuvette accessory for volumes
in the 100 uL range.  We've been using the instrument for over a year now
with no problems.  Results for determining protein concentrations using the
method of Pace et al * *Protein
Sci.javascript:AL_get(this,%20'jour',%20'Protein%20Sci.');1995
Nov;4(11):2411-23 are very reproducible using this spec.

On Thu, Dec 4, 2008 at 10:37 AM, James M. Vergis [EMAIL PROTECTED]wrote:

 I would also recommend the nanodrop.  It takes a whole spectra every
 measurement and there is no need to dilute your sample.  You can demo it
 for
 a week and try it out.


 
 James M. Vergis, Ph.D.
 University of Virginia Molecular Physiology and Biological Physics
 MKWEINR 360A Snyder Building
 480 Ray C. Hunt Drive
 PO Box  800886
 Charlottesville, VA 22908-0886
 phone: 434-243-2730   FAX: 434-243-8271
 [EMAIL PROTECTED]
 


 -Original Message-
 From: CCP4 bulletin board [mailto:[EMAIL PROTECTED] On Behalf Of Tim
 Gruene
 Sent: Thursday, December 04, 2008 10:16 AM
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: [ccp4bb] suggestions for UV spectrometer

 Dear all,

 we would like to purchase a UV spectrometer for measuring protein
 concentrations (280nm), and I would like to here your comments and
 especially recommendations.

 We don't need anything fancy, a small, fast device would be sufficient.

 Tim


 --
 Tim Gruene
 Institut fuer anorganische Chemie
 Tammannstr. 4
 D-37077 Goettingen

 GPG Key ID = A46BEE1A




-- 
Jim Fairman
Graduate Research Assistant
Department of Biochemistry, Cellular, and Molecular Biology (BCMB)
University of Tennessee -- Knoxville
216-368-3337 [EMAIL PROTECTED] [EMAIL PROTECTED]


Re: [ccp4bb] suggestions for UV spectrometer

2008-12-04 Thread Mischa Machius
Tim - I would recommend a spectrometer that records entire spectra,  
instead of one that takes readings at just 280 nm. Contributions from  
light scattering can be very strong and can give results that deviate  
from the true value by a factor of two or more. One cannot detect  
scattering without recording spectra. The most severe case we have had  
was someone who thought the protein concentration was 10 mg/mL (based  
on 280) when in reality (after subtraction of the scattering  
contribution) it was only 4 mg/mL. Lots of other, less severe cases as  
well. Hope that helps. Best - MM



Mischa Machius, PhD
Associate Professor
Department of Biochemistry
UT Southwestern Medical Center at Dallas
5323 Harry Hines Blvd.; ND10.214A
Dallas, TX 75390-8816; U.S.A.
Tel: +1 214 645 6381
Fax: +1 214 645 6353



On Dec 4, 2008, at 9:16 AM, Tim Gruene wrote:


Dear all,

we would like to purchase a UV spectrometer for measuring protein  
concentrations (280nm), and I would like to here your comments and  
especially recommendations.


We don't need anything fancy, a small, fast device would be  
sufficient.


Tim


--
Tim Gruene
Institut fuer anorganische Chemie
Tammannstr. 4
D-37077 Goettingen

GPG Key ID = A46BEE1A


Re: [ccp4bb] suggestions for UV spectrometer

2008-12-04 Thread Patrick Loll
At the risk of dragging this discussion even further afield from  
crystallography:


How can you get realistic numbers for concentrated solutions using  
the Nanodrop?  I understand that the instrument reduces absorbance by  
using a very short path length. However, I thought that in order for  
the Beer-Lambert formalism to be applicable, the solution needs to be  
sufficiently dilute so that the chance of molecules shadowing one  
another is negligible. Isn't this condition violated for concentrated  
solutions (even with short path lengths)?


Pat

On 4 Dec 2008, at 1:27 PM, Michael Giffin wrote:


We also like the Nanodrop...


 
---

Patrick J. Loll, Ph. D. 
Professor of Biochemistry  Molecular Biology
Director, Biochemistry Graduate Program
Drexel University College of Medicine
Room 10-102 New College Building
245 N. 15th St., Mailstop 497
Philadelphia, PA  19102-1192  USA

(215) 762-7706
[EMAIL PROTECTED]



Re: [ccp4bb] suggestions for UV spectrometer

2008-12-04 Thread hari jayaram
We too have a nano-drop. We really like it , but have not  yet fully
switched over.

I agree with all the good things said about it , but here are the few times
the nano drop falls short:

1) We still use the old spec ( 1 cm path length ) for things at a very low
concentration , i.e for the monitoring free thiols in protein with ellman
reagent , the absorbances are very low and give poor reproducibility on the
nanodrop because of small path-length . Of course this can be overcome by
using a lot of protein at a higher concentration and modifying the assay.
2) For very concentrated membrane protein samples which tend to have a large
concentration of detergent . The reproducibility is not very good because of
the high concentration of deteregent preventing a proper meniscus from
forming. The solution to this is to dilute your sample so the dteergent
concentration is manageable ( a few times the CMC instead of tens of times
the CMC

Other than for these issues we almost entirely use the nanodrop and would
gladly recommend it
Hari

On Thu, Dec 4, 2008 at 1:27 PM, Michael Giffin [EMAIL PROTECTED] wrote:

 We also like the Nanodrop.  Very fast, no cuvettes (breaking, washing,
 cleaning, uh nitric acid bath anyone?), and the .ndv data file is a
 delimited text file.  Open in a text editor, copy and paste into a
 spreadsheet, and you have a convenient record of all of your stocks,
 including date, sample name, concentration, and full spectra.

 It is expensive, but so are good cuvettes.


 Mike


 Michael Giffin
 The Scripps Research Institute
 Department of Molecular and Experimental Medicine
 10550 North Torrey Pines Road, MEM-131
 La Jolla, CA 92037
 email:  [EMAIL PROTECTED]
 lab:  858-784-7758

 On Thu, Dec 4, 2008 at 7:16 AM, Tim Gruene [EMAIL PROTECTED]
 wrote:
  Dear all,
 
  we would like to purchase a UV spectrometer for measuring protein
  concentrations (280nm), and I would like to here your comments and
  especially recommendations.
 
  We don't need anything fancy, a small, fast device would be sufficient.
 
  Tim
 
 
  --
  Tim Gruene
  Institut fuer anorganische Chemie
  Tammannstr. 4
  D-37077 Goettingen
 
  GPG Key ID = A46BEE1A
 



[ccp4bb] suggestions for UV spectrometer

2008-12-04 Thread Clemens Steegborn
 

Hi Tim,

 

The Shimadzu UV2401PC comes at a reasonable price and has all features you
might possible need in a

Biochemistry lab. And if you like the option to use small sample volumes, I
would suggest to by a TrayCell

from Helma - you put it in a regular photometer, it guides the light through
a microliter chamber and back

into the photometer (it is essentially a cuvette with optics and a 1 or 5
microliter chamber). Costs a few

hundred Euros .

If you're sure you will never need regular cuvettes - go for Nanodrop. If
you want more flexibility, Shimadzu

plus TrayCell . 

 

Best

Clemens

 


---

  

Jun.-Prof. Dr. Clemens Steegborn 
Ruhr-University Bochum 
Physiological Chemistry, MA 2/141
Universitaetsstr. 150
44801 Bochum, Germany
phone: ++49 234 32 27041
fax: ++49 234 32 14193 
email:  [EMAIL PROTECTED]   

 



Re: [ccp4bb] suggestions for UV spectrometer

2008-12-04 Thread Filip Van Petegem
I want to add I absotely hate the nanodrop.  We've had a demo for it, and
found the readouts to be very unreliable.  Fluctuations of 20% and more.
Just leaving the same drop in and measuring the sample multiple times gives
different values (going in both directions, so not only due to
evaportations). Sure, it's easy and fast, and maybe good to have a rough
idea about your protein concentration, but I would never want to use it for
exact measurements such as needed for e.g. a CD or an ITC instrument. I've
heard other labs in our department have similar issues.  We've also had a
demo for the Nanovue from GE Healthcare:  same issues - very large
fluctuations from one sample to another.  I suppose this is simply an
inherent problem with small volumes...

Cheers

Filip Van Petegm



On Thu, Dec 4, 2008 at 12:48 PM, Patrick Loll [EMAIL PROTECTED] wrote:

 At the risk of dragging this discussion even further afield from
 crystallography:
 How can you get realistic numbers for concentrated solutions using the
 Nanodrop?  I understand that the instrument reduces absorbance by using a
 very short path length. However, I thought that in order for the
 Beer-Lambert formalism to be applicable, the solution needs to be
 sufficiently dilute so that the chance of molecules shadowing one another
 is negligible. Isn't this condition violated for concentrated solutions
 (even with short path lengths)?

 Pat

 On 4 Dec 2008, at 1:27 PM, Michael Giffin wrote:

 We also like the Nanodrop...


 ---

 Patrick J. Loll, Ph. D.

 Professor of Biochemistry  Molecular Biology

 Director, Biochemistry Graduate Program

 Drexel University College of Medicine

 Room 10-102 New College Building

 245 N. 15th St., Mailstop 497

 Philadelphia, PA  19102-1192  USA


 (215) 762-7706

 [EMAIL PROTECTED]




-- 
Filip Van Petegem, PhD
Assistant Professor
The University of British Columbia
Dept. of Biochemistry and Molecular Biology
2350 Health Sciences Mall - Rm 2.356
Vancouver, V6T 1Z3

phone: +1 604 827 4267
email: [EMAIL PROTECTED]
http://crg.ubc.ca/VanPetegem/