Thanks Martin,  Notice 42 makes it clear.  I can only think the guy I talked
to at CDRH either didn't know about it or he is interpreting it to only
include the examples noted in the notice (CD players, etc. where the laser
is totally buried in the product).  My application is similar to yours where
I provide an optical communications interface which uses a Class 1 laser
module.

Regards, 

Scott Lemon 
email: sle...@caspiannetworks.com 

 -Original Message-
From: Martin Lindquist [mailto:mlind...@cisco.com]
Sent: Thursday, March 22, 2001 10:07 AM
To: emc-p...@majordomo.ieee.org
Subject: RE: FDA laser classes 



Scott,
 
  My understanding is that (for our Class I components (OC3 transceivers) at
least),  according to Notice 42 (Dec. 18, 1989), initial and annual reports
are not required if you comply with the four points listed (see actual
notice for more information):
1. No modification of performance or intended use and incorporation of
the certified product results in only the concealment of the original
manufacturer's certification and identification labels required by 1010
2. The labeling requirements of 1010 and 1040.10 would be met if the
certified product were removed from the product in which it was incorporated
3. The labeling requirements of 1040.10(g) would be met in any service
configuration of the certified product
4. The laser safety information provided by the certifying manufacturer
is distributed with the product
 
  Jumping into Table 1 of Part 1002 there is no requirements for
Distribution records.  I even went so far as to verify this (via email) with
the CDRH last year.  Note that this is for a Class I OC-3 laser module
component used by us in its intended use and manufactured by someone else
(HP, etc), where the manufacturer has done all the filing, attestations,
etc., and where the module does not contain a higher Class laser.  In other
words, according to the CDRH and Notice 42 we are a 'distributor' of the
module and thus not subject to initial and annual reports, nor is any
additional certification and identification labels required on the outside
of the final product.  
 
  Your situation may well require the 5 points you list, but take a look at
Notice 42 (and Notice 41 as well (Aug. 9, 1988)), along with the other
notices in the following PDF (link provided to me by John Juhasz).  Your
compliance contact may not have been aware of these, or he may be holding
you to a higher standard than is required by the CDRH, I can't say.  
 

Notices: 
http://www.fda.gov/cdrh/radhlth/pdf/laspol01.pdf
http://www.fda.gov/cdrh/radhlth/pdf/laspol01.pdf 
(Notice 42 is page 87 of 105)
 
Reporting form:

http://www.fda.gov/cdrh/radhlth/pdf/lasrpt0p.pdf
http://www.fda.gov/cdrh/radhlth/pdf/lasrpt0p.pdf 
 
  If nothing else, you may have learned something new today and be able to
go home early...  :)
 
Martin.
 
 
 -Original Message-
From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On Behalf Of
Scott Lemon
Sent: Thursday, March 22, 2001 8:00 AM
To: 'Bandele Adepoju'; 'John Juhasz'; 'Martin Lindquist'; '|Moshe Valdman'
Cc: emc-p...@majordomo.ieee.org
Subject: RE: FDA laser classes 



Bandele,
 
This is not my understanding after talking to a compliance specialist in
CDRH yesterday.  He informed me that manufacturers that incorporate class 1
lasers (which have themselves followed the required reporting requirements,
etc. per the CFR and CDRH guidelines) must still:
 
1.  file an initial product report per 1002.10
2.  file an annual report per 1002.13
3.  file a report of accidental radiation occurrences (as applicable) per
1002.20
4.  end product must have a certification label per 1010.2
5.  end product must have an ID label per 1010.3
 
He indicated that all other reporting requirements (including supplemental
reports) are not required of the end product manufacturer.
 
This seems to correlate somewhat with the Aug 23, 1985 notice (at least the
second to the last paragraph).
 
I contacted Mr. Manuel Karos at CDRH (non-medical lasers) at (301) 594-4646.
This was my understanding - it might not be a bad idea if someone else wants
to contact the office as well to substantiate and report back to the group.

Regards, 

Scott Lemon 
sle...@caspiannetworks.com 

 -Original Message-
From: Bandele Adepoju [mailto:badep...@jetstream.com]
Sent: Wednesday, March 21, 2001 8:09 PM
To: 'John Juhasz'; 'Martin Lindquist'; '|Moshe Valdman'
Cc: emc-p...@majordomo.ieee.org
Subject: RE: FDA laser classes 



Hello all,
 
Manufacturers who only incorporate certified Class 1 lasers into their
products
only need to comply with the record keeping requirements of 21CFR 1002.  
They do not need to submit initial and annual reports.
 
Regards,
 
Bandele 
Jetstream Communications, Inc. 
badep...@jetstream.com mailto:badep...@jetstream.com  

 

 

-Original Message-
From: John Juhasz [mailto:jjuh...@fiberoptions.com]
Sent: Wednesday, March 21, 2001 8:19 AM
To: 

RE: EMC documentation for Australia

2001-03-22 Thread WOODS

The following requirements are for those products that must be C-tick marked

*   Your Australian representative must apply for approval to use the
C-tick marking
*   Anyone may apply the marking once approval is granted
*   Depending on the method of identification of the manufacture as
noted in the application, you may have to also apply an N number adjacent
to the C-tick marking. 
*   The DoC must be issued by the Australian representative

--
From:  plaw...@west.net
Sent:  Thursday, March 22, 2001 3:51 PM
To:  EMC-PSTC
Subject:  EMC documentation for Australia


Our company makes open frame linear and switch-mode component power
supplies in the range of 5-500W.  Applications include industrial, and
medical equipment. 

The sales distributor we use in Australia has contacted us asking for
EMC documentation to support C-Tick marking of our products, as
required by the ACA (Australian Communications Authority).
They are asking for an EMC test report and declaration of conformity
showing that the converter complies with the relevant EMC standard.
I want to verify what, if any, documentation is required for a product
like ours.

I have copies of several EMC-related documents from the ACA web site.
Unfortunately, I can't tell if component power supplies require the
same level of documentation as completed systems.

1) It appears that the regulations do covers power supplies, as
discussed in the ACA EMC FAQ at
http://www.aca.gov.au/consumer/faq/emc.htm.
Is this accurate?

1) What is the definition of a 'Converter'?
The EMC FAQ located at http://www.aca.gov.au/consumer/faq/emc.htm says
that Converters that use linear technology fall within the scope of
AS 1044 if they are designed for connection to a low voltage
electrical supply.
Does this mean that our linear-mode power supply (i.e., isolation
transformer with bulk storage capacitor, series pass regulator, LM7805
regulator, etc.) must comply with that CISPR14-like standard, while
the ITE-related end use may need to conform with a CISPR22-like
standard?

2) Some of our power supplies were designed to meet only the conducted
emission portion of the EMC specification, not the radiated emission
portion.  Therefore, we can't claim full compliance with the emission
standard.
Does this imply that the power supply must be redesigned, or no units
can shipped to Australia?

3) Do we put the C-Tick mark on our products, or is this an issue for
our Australian partner?

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




RE: Translation Requirements

2001-03-22 Thread David_Sterner

Safety related instructions for the user should be in the vernacular under
73/23/EEC.

There is more latitude on language for safety-related instructions for an
'installer'.  However if installer's instructions have life-threatening
consequences you reduce legal exposure by using a widely-understood language
of your target country.

David

-Original Message-
From: Finn, Paul [mailto:fi...@pan0.panametrics.com]
Sent: Thursday, March 22, 2001 1:35 PM
To: 'emc-pstc'
Subject: Translation Requirements




Can some one point out the specific requirements, if any, that are imposed
on translating manuals into the language of the country they are to be used.
Specifically if you CE label in accordance with 89/336/EEC (EN 61326) and
73/23/EEC (1010-1).  In reviewing the archives I could only find mention
that the manufacturer is to supply a version of the manual which is agreed
upon in that country and English is most common. 

Your comments would be greatly appreciated.

Paul Finn
Panametrics Inc. 

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




RE: Translation Requirements

2001-03-22 Thread Ned Devine

Paul,

This is not from the EMC or LVD , but you mentioned 1010-1 which includes
laboratory equipment, so for the medical directives, including the one for
in vitro diagnostic medical devices, it is very clearly specified. 

For the MDD, See article 4, paragraph 4

Member States may require the information, which must be made available to
the user and the patient in accordance with Annex I, point 13, to be in
their national language(s) or in another Community language, when a device
reaches the final user, regardless of whether it is for professional or
other use.


For the IVDD, See article 4, paragraph 4

Member States may require the information to be supplied pursuant to Annex
I, part B, section 8 to be in their official language(s) when a device
reaches the final user.

Ned


Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI  49548

616 248 9671 Phone
616 574 9752 Fax
ndev...@entela.com  e-mail





-Original Message-
From: Finn, Paul [mailto:fi...@pan0.panametrics.com]
Sent: Thursday, March 22, 2001 1:35 PM
To: 'emc-pstc'
Subject: Translation Requirements




Can some one point out the specific requirements, if any, that are imposed
on translating manuals into the language of the country they are to be used.
Specifically if you CE label in accordance with 89/336/EEC (EN 61326) and
73/23/EEC (1010-1).  In reviewing the archives I could only find mention
that the manufacturer is to supply a version of the manual which is agreed
upon in that country and English is most common. 

Your comments would be greatly appreciated.

Paul Finn
Panametrics Inc. 

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Michael Garretson:pstc_ad...@garretson.org
 Dave Healddavehe...@mediaone.net

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.rcic.com/  click on Virtual Conference Hall,




Re: FDA laser classes

2001-03-22 Thread Andrew Carson
All

I contacted the CDRH in Feb 2000 on this very issue of incorporated
certified Class I lasers into a product. After much passing up the
management chain until I found some one who could answer the question ,
I got the following reply. As it comes form the chief of the Electronics
Products Branch, I will put faith in it being the correct CDRH view
point on the issue.

Dear Andrew Carson:

You my meet the condition specified in our laser notice 42 (attached).
If so
you would not need to report or certify you product.  Please review the
attachment, to determine if your products qualify.  Sorry for the delay
in
responding.


 LASER NOTICE 42.DOC

Collin L. Figueroa
Chief, Electronic Products Branch (HFZ-342)
Center for Devices and Radiological Health
Office of Compliance
2094 Gaither Road
Rockville, MD 20850
Phone: 301-594-4654
Fax: 301-594-4672


Martin Lindquist wrote:

  Scott,  My understanding is that (for our Class I components (OC3
 transceivers) at least),  according to Notice 42 (Dec. 18, 1989),
 initial and annual reports are not required if you comply with the
 four points listed (see actual notice for more information):1. No
 modification of performance or intended use and incorporation of the
 certified product results in only the concealment of the original
 manufacturer's certification and identification labels required by
 10102. The labeling requirements of 1010 and 1040.10 would be met
 if the certified product were removed from the product in which it was
 incorporated3. The labeling requirements of 1040.10(g) would be
 met in any service configuration of the certified product4. The
 laser safety information provided by the certifying manufacturer is
 distributed with the product  Jumping into Table 1 of Part 1002 there
 is no requirements for Distribution records.  I even went so far as to
 verify this (via email) with the CDRHlast year.  Note that this is for
 a Class I OC-3 laser module component used by us in its intended use
 and manufactured by someone else (HP, etc), where the manufacturer has
 done all the filing, attestations, etc., and where the module does not
 contain a higher Class laser.  In other words, according to the CDRH
 and Notice 42 we are a 'distributor' of the module and thus not
 subject to initial and annual reports, nor is any additional
 certification and identification labels required on the outside of the
 final product.   Your situation may well require the 5 points you
 list, but take a look at Notice 42 (and Notice 41 as well (Aug. 9,
 1988)), along with the other notices in the following PDF (link
 provided to me by John Juhasz).  Your compliance contact may not have
 been aware of these, or he may be holding you to a higher standard
 than is required by the CDRH, I can't
 say. Notices:http://www.fda.gov/cdrh/radhlth/pdf/laspol01.pdf(Notice
 42 is page 87 of 105)Reporting
 form:http://www.fda.gov/cdrh/radhlth/pdf/lasrpt0p.pdf  If nothing
 else, you may have learned something new today and be able to go home
 early...  :)Martin.-Original Message-
 From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On
 Behalf Of Scott Lemon
 Sent: Thursday, March 22, 2001 8:00 AM
 To: 'Bandele Adepoju'; 'John Juhasz'; 'Martin Lindquist'; '|Moshe
 Valdman'
 Cc: emc-p...@majordomo.ieee.org
 Subject: RE: FDA laser classes


  Bandele,This is not my understanding after talking to a
  compliance specialist in CDRH yesterday.  He informed me
  that manufacturers that incorporate class 1 lasers (which
  have themselves followed the required reporting
  requirements, etc. per the CFR and CDRH guidelines) must
  still:1.  file an initial product report per 1002.102.  file
  an annual report per 1002.133.  file a report of accidental
  radiation occurrences (as applicable) per 1002.204.  end
  product must have a certification label per 1010.25.  end
  product must have an ID label per 1010.3He indicated that
  all other reporting requirements (including supplemental
  reports) are not required of the end product
  manufacturer.This seems to correlate somewhat with the Aug
  23, 1985 notice (at least the second to the last
  paragraph).I contacted Mr. Manuel Karos at CDRH (non-medical
  lasers) at (301) 594-4646.  This was my understanding - it
  might not be a bad idea if someone else wants to contact the
  office as well to substantiate and report back to the
  group.Regards,

  Scott Lemon
  sle...@caspiannetworks.com

  -Original Message-
  From: Bandele Adepoju [mailto:badep...@jetstream.com]
  Sent: Wednesday, March 21, 2001 8:09 PM
  To: 'John Juhasz'; 'Martin Lindquist'; '|Moshe Valdman'
  Cc: emc-p...@majordomo.ieee.org
  Subject: RE: FDA laser classes


   Hello all,Manufacturers who only incorporate
   certified Class 1 lasers into their productsonly
   need to comply with the record keeping