Could you please share your experience and/or knowledge for the ground
plane grounding in the immunity lab?

We plan to layout a L shape ground plane in the 24' x 18' area for ESD,
etc. tests.  The corner of the L shape is right next to a 10 wide building
collar.  Is a ground strap tied to the collar good enough?

Do we need (copper) ground rods?  If yes, what is the minimum length
(deep)?  What is the recommended space between rods?

How to choose metal sheet material (galvanized steel, stainless steel,
etc.)?

When weldering the metal sheet, is there anything that need to be cautious?

Thank you very much and look forward to your help.

Best regards,
Grace Lin

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Ground Plane Grounding in the Immunity Lab

2013-08-23 Thread ce-test, qualified testing bv - Gert Gremmen
You do not need any of them for EMC.

Do as you deem necessary for electrical safety.

 

Regards,

Ing.  Gert Gremmen, BSc

 

 

 

Van: emc-p...@ieee.org [mailto:emc-p...@ieee.org] Namens Grace Lin
Verzonden: Friday, August 23, 2013 3:04 PM
Aan: emc-p...@ieee.org
Onderwerp: Ground Plane Grounding in the Immunity Lab

 

Dear Members,

 

Could you please share your experience and/or knowledge for the ground
plane grounding in the immunity lab?  

 

We plan to layout a L shape ground plane in the 24' x 18' area for ESD,
etc. tests.  The corner of the L shape is right next to a 10 wide
building collar.  Is a ground strap tied to the collar good enough?

 

Do we need (copper) ground rods?  If yes, what is the minimum length
(deep)?  What is the recommended space between rods?

 

How to choose metal sheet material (galvanized steel, stainless steel,
etc.)?  

 

When weldering the metal sheet, is there anything that need to be
cautious?

 

Thank you very much and look forward to your help.

 

Best regards,

Grace Lin

-


This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list. To post a message to the list, send your
e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site
at http://product-compliance.oc.ieee.org/ can be used for graphics (in
well-used formats), large files, etc.

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to:
Jim Bacher j.bac...@ieee.org
David Heald dhe...@gmail.com 


-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Ground Plane Grounding in the Immunity Lab

2013-08-23 Thread Ed Price
Grace:

 

Welding galvanized steel is not a good idea; toxic zinc fumes.

What is a building collar?

Your immunity lab must be within a shielded enclosure, so it should already
be safety grounded. Since you will be doing ESD testing INSIDE the
enclosure, you will not have to be draining any charges of the exterior of
that chamber.

Metal sheet material for where; the chamber walls, the ground plane table?

 

Ed Price
WB6WSN
Chula Vista, CA USA



 

From: Grace Lin [mailto:graceli...@gmail.com] 
Sent: Friday, August 23, 2013 6:04 AM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: [PSES] Ground Plane Grounding in the Immunity Lab

 

Dear Members,

 

Could you please share your experience and/or knowledge for the ground plane
grounding in the immunity lab?  

 

We plan to layout a L shape ground plane in the 24' x 18' area for ESD, etc.
tests.  The corner of the L shape is right next to a 10 wide building
collar.  Is a ground strap tied to the collar good enough?

 

Do we need (copper) ground rods?  If yes, what is the minimum length (deep)?
What is the recommended space between rods?

 

How to choose metal sheet material (galvanized steel, stainless steel,
etc.)?  

 

When weldering the metal sheet, is there anything that need to be cautious?

 

Thank you very much and look forward to your help.

 

Best regards,

Grace Lin

-


This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at
http://product-compliance.oc.ieee.org/ can be used for graphics (in
well-used formats), large files, etc.

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to:
Jim Bacher j.bac...@ieee.org
David Heald dhe...@gmail.com 


-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Ground Plane Grounding in the Immunity Lab

2013-08-23 Thread T.Sato
Grace,

On Fri, 23 Aug 2013 09:03:35 -0400,
  Grace Lin graceli...@gmail.com wrote:

 Could you please share your experience and/or knowledge for the ground
 plane grounding in the immunity lab?
 
 We plan to layout a L shape ground plane in the 24' x 18' area for ESD,
 etc. tests.  The corner of the L shape is right next to a 10 wide building
 collar.  Is a ground strap tied to the collar good enough?
 
 Do we need (copper) ground rods?  If yes, what is the minimum length
 (deep)?  What is the recommended space between rods?

I think you can simply connect the ground plane to the grounding system
already available in the building.

 How to choose metal sheet material (galvanized steel, stainless steel,
 etc.)?

If mechanical strength is also required, thick sheet of galvanized steel
or stainless steel may be used.
Galvanized steel is cheap, but, personally, I don't like this kind of
plated material as we may need to take care so that those thin plating
will not be damaged.
Recently, we replaced galvanized steel sheets in our old shielded rooms
with austenite stainless steel sheets.

If mechanical strength is not important, thin sheet of aluminum or copper
may work well.
In the case of IEC 61000-4-2, minimum thickness of the ground plane is
0.25 mm for copper and aluminum.
Copper is electrically good but its surface may not looks so good, and
can be heavy and expensive than aluminum.

 When weldering the metal sheet, is there anything that need to be cautious?

It can become troublesome to weld thick metal sheets placed on the floor,
and some material may not suitable for welding.

I rather like not to weld and bond them with other means.

Regards,
Tom

-- 
Tomonori Sato  vef00...@nifty.ne.jp
URL: http://homepage3.nifty.com/tsato/

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Ground Plane Grounding in the Immunity Lab

2013-08-23 Thread Flavin, John

Grace,

Our lab is used for testing of ITE products to the IEC 55024 (CISPR24) immunity 
standard.

We built ours from sheets of 1/4 (6mm) 5052 aluminum, screwed down onto 1 
(25mm) plywood, with the plywood then fastened to the building floor. (Our 
products are large, heavy, rack-based systems, and this aluminum is hard enough 
to prevent the rack casters from doing any damage to the plate.)  A 4 (10cm) 
wide backing plate is placed under each butt joint, and fasted to the plates 
with countersunk screws about every 4 along the joint. This insures good 
electrical continuity across the entire plane. The backing plate does require 
the plywood to be routed/notched at the butt joints, to accommodate the extra 
thickness of the backing plate.

We have two such floors, one which is about 15 years old, and have had no 
issues with them (electrical, mechanical, auditors).

These are grounded to the building steel. One uses a flat braid, the other uses 
6 gauge wire.

John D. Flavin
Teradata TCP Engineering
17095 Via del Campo
San Diego, CA 92127
john.fla...@teradata.commailto:john.fla...@teradata.com
V: (858) 485-3874
F: (213) 337-5432

From: Grace Lin [mailto:graceli...@gmail.com]
Sent: Friday, August 23, 2013 6:04 AM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: [PSES] Ground Plane Grounding in the Immunity Lab

Dear Members,

Could you please share your experience and/or knowledge for the ground plane 
grounding in the immunity lab?

We plan to layout a L shape ground plane in the 24' x 18' area for ESD, etc. 
tests.  The corner of the L shape is right next to a 10 wide building collar.  
Is a ground strap tied to the collar good enough?

Do we need (copper) ground rods?  If yes, what is the minimum length (deep)?  
What is the recommended space between rods?

How to choose metal sheet material (galvanized steel, stainless steel, etc.)?

When weldering the metal sheet, is there anything that need to be cautious?

Thank you very much and look forward to your help.

Best regards,
Grace Lin
-


This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.orgmailto:emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.netmailto:emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.orgmailto:mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher j.bac...@ieee.orgmailto:j.bac...@ieee.org
David Heald dhe...@gmail.commailto:dhe...@gmail.com

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Ground Plane Grounding in the Immunity Lab

2013-08-23 Thread Pettit, Ghery
Keep in mind that a primary purpose of grounding the plane is electrical safety.

Ghery S. Pettit

From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Flavin, John
Sent: Friday, August 23, 2013 11:55 AM
To: emc-p...@ieee.org
Subject: RE: [PSES] Ground Plane Grounding in the Immunity Lab


Grace,

Our lab is used for testing of ITE products to the IEC 55024 (CISPR24) immunity 
standard.

We built ours from sheets of 1/4 (6mm) 5052 aluminum, screwed down onto 1 
(25mm) plywood, with the plywood then fastened to the building floor. (Our 
products are large, heavy, rack-based systems, and this aluminum is hard enough 
to prevent the rack casters from doing any damage to the plate.)  A 4 (10cm) 
wide backing plate is placed under each butt joint, and fasted to the plates 
with countersunk screws about every 4 along the joint. This insures good 
electrical continuity across the entire plane. The backing plate does require 
the plywood to be routed/notched at the butt joints, to accommodate the extra 
thickness of the backing plate.

We have two such floors, one which is about 15 years old, and have had no 
issues with them (electrical, mechanical, auditors).

These are grounded to the building steel. One uses a flat braid, the other uses 
6 gauge wire.

John D. Flavin
Teradata TCP Engineering
17095 Via del Campo
San Diego, CA 92127
john.fla...@teradata.commailto:john.fla...@teradata.com
V: (858) 485-3874
F: (213) 337-5432

From: Grace Lin [mailto:graceli...@gmail.com]
Sent: Friday, August 23, 2013 6:04 AM
To: EMC-PSTC@LISTSERV.IEEE.ORGmailto:EMC-PSTC@LISTSERV.IEEE.ORG
Subject: [PSES] Ground Plane Grounding in the Immunity Lab

Dear Members,

Could you please share your experience and/or knowledge for the ground plane 
grounding in the immunity lab?

We plan to layout a L shape ground plane in the 24' x 18' area for ESD, etc. 
tests.  The corner of the L shape is right next to a 10 wide building collar.  
Is a ground strap tied to the collar good enough?

Do we need (copper) ground rods?  If yes, what is the minimum length (deep)?  
What is the recommended space between rods?

How to choose metal sheet material (galvanized steel, stainless steel, etc.)?

When weldering the metal sheet, is there anything that need to be cautious?

Thank you very much and look forward to your help.

Best regards,
Grace Lin
-


This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.orgmailto:emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.netmailto:emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.orgmailto:mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher j.bac...@ieee.orgmailto:j.bac...@ieee.org
David Heald dhe...@gmail.commailto:dhe...@gmail.com
-


This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.orgmailto:emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.netmailto:emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.orgmailto:mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher j.bac...@ieee.orgmailto:j.bac...@ieee.org
David Heald dhe...@gmail.commailto:dhe...@gmail.com

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html

Re: [PSES] Ground Plane Grounding in the Immunity Lab

2013-08-23 Thread John Woodgate
In message 
beec1fc38f37bc4da6df70db9d7d0e3608d9c...@sushdc8002.td.teradata.com, 
dated Fri, 23 Aug 2013, Flavin, John john.fla...@teradata.com 
writes:


Our lab is used for testing of ITE products to the IEC 55024 (CISPR24) 
immunity standard.


Not IEC 55024, which does not exist, but EN 55024.
--
OOO - Own Opinions Only. With best wishes. See www.jmwa.demon.co.uk
Why is the stapler always empty just when you want it?

John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com