RE: Medical Grade Transformers

2007-07-16 Thread emc-p...@ieee.org
 From: jeff collins
 Sent: Friday, July 13, 2007 5:00 PM
  
 Can anyone recommend a Transformer manufacturer that makes 
 custom medical grade transformers?

Speaking of on-line databases, try searching on UL's category named
Transformers, Construction Only - Component (XORU2).  Add a keyword for
a standard number (e.g. 60601, 2601, 544).

Construction only transformers meet the insulation requirements (within
the limitations identified in the reports) of the identified standard(s)
and the product category is designed to allow for custom builds, but
will still need thorough testing in the end-product.


Regards,

Peter L. Tarver, PE
ptar...@ieee.org 

CONFIDENTIALITY
This e-mail message and any attachments thereto, is intended only for use by
the addressee(s) named herein and may contain legally privileged and/or
confidential information. If you are not the intended recipient of this e-mail
message, you are hereby notified that any dissemination, distribution or
copying of this e-mail message, and any attachments thereto, is strictly
prohibited.  If you have received this e-mail message in error, please
immediately notify the sender and permanently delete the original and any
copies of this email and any prints thereof.
ABSENT AN EXPRESS STATEMENT TO THE CONTRARY HEREINABOVE, THIS E-MAIL IS NOT
INTENDED AS A SUBSTITUTE FOR A WRITING.  Notwithstanding the Uniform
Electronic Transactions Act or the applicability of any other law of similar
substance and effect, absent an express statement to the contrary hereinabove,
this e-mail message its contents, and any attachments hereto are not intended
to represent an offer or acceptance to enter into a contract and are not
otherwise intended to bind the sender, Sanmina-SCI Corporation (or any of its
subsidiaries), or any other person or entity.

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:emc-p...@daveheald.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc

__
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email 
__



RE: Medical Grade Transformers

2007-07-13 Thread emc-p...@ieee.org
List Members,
 
Can anyone recommend a Transformer manufacturer that makes custom medical
grade transformers?
 
We will obviously need to get this component Approved/Listed/Certified and
want to work with a vendor that is established and reputable in handling
agency listings and manufacturing requirements.
 
Thanks in advance,
 
Jeff Collins
 
 
-  This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 


Instructions: http://listserv.ieee.org/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org 


For policy questions, send mail to: 


Jim Bacher: j.bac...@ieee.org David Heald: emc-p...@daveheald.com 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc 
__
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email 
__





Re: medical-grade transformers

2002-08-02 Thread Rich Nute






Hi Ged:


   In transformers with REINFORCED INSULATION or DOUBLE INSULATION the
   insulation between 1ary and 2ary windings shalll consist of -
   -  1 insulation layer having thickness at least 1mm, or
   -  at least 2 insulation layers with total thickness not less than 0.3mm, or
   -  3 layers provided each combination of 2 layers can withstand the
   dielectric strenght test for REINFORCED INSULATION
   
   Can anyone explain ther rationale behind these requirements ?

Thin insulation is presumed to be subject to 
pin-holes.

If 1 layer of insulation, then the insulation
must be thick rather than thin, hence 1 mm 
thickness.  We presume no pin-holes in thick
insulation, and the insulation comprises
reinforced insulation.

If 2 layers of thin insulation, then presume
a pin-hole in one layer, and that the pin-hole
will not overlay a pin-hole in the other layer.  
If that pin-hole is 0.15 mm, then we know that 
the air in that pin-hole will at least withstand 
the mains voltage.  Therefore, the construction
still has two insulations in series, one being
the 0.15 mm air of the pin-hole, the other being 
the 0.15 mm thick solid insulation.  The 2 
insulations in series comprise double insulation.

If 3 layers of thin insulation, then presume 
a pin-hole in one layer, and that the pin-hole
will not ovelay a pin-hole in the adjacent
layer.  The remaining 2 layers comprise double 
insulation.  

(In IEC 60960, the requirement for 1 layer in
a 2-layer system is 0.4 mm thickness.)


Best regards,
Rich


ps:  I'm not certain of my explanation for 2
 layers.  This rationale came from 60950,
 and I have applied it to 60601.





---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


Re: medical-grade transformers

2002-07-29 Thread Dean Gerard (gdean)

Regarding my query on medical equipment transformer construction
(EN-60601-1, clause 57.9.4(e)), can I be a bit more specific.

- How were insulation thickness requirements decided, and why is the
thickness requirement lower if insulation is 2-layer rather than
single-layer ?.
- Why are single-layer and 2-layer insulation requirements defined in terms
of thickness, yet 3-layer insulaton requirements defined in terms of
dielectric strength ?.

Thanks

Ged Dean


**
This email and any files transmitted with it are confidential and
intended solely for the use of the individual or entity to whom they
are addressed. If you have received this email in error please notify
the system manager.

This footnote also confirms that this email message has been swept by
MIMEsweeper for the presence of computer viruses.

NCHT
**


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


RE: medical-grade transformers

2002-07-29 Thread Ned Devine

Hi,

From Appendix A, General Guidance and Rationale, for Clause 57.9

...For reasons of PATIENT safety additional requirements must be applied to
the construction
of such transformers, e.g. restriction of LEAKAGE CURRENTS flowing to
PATIENT CIRCUITS.


57.9.4 e) In transformers with REINFORCED INSULATION or DOUBLE INSULATION
the insulation
between the primary and secondary winding shall consist of:
- one insulation layer having a thickness of at least 1 mm, or
- at least two insulation layers with a total thickness of not less than 0.3
mm, or
- three layers provided that each combination of two layers can withstand
the dielectric
strength test for REINFORCED INSULATION.

Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI  49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail

Entela, Inc. A Certified Woman Owned Business
www.entela.com 






-Original Message-
From: Dean Gerard (gdean) [mailto:gd...@ncht.trent.nhs.uk]
Sent: Monday, July 29, 2002 9:29 AM
To: 'emc-p...@majordomo.ieee.org'
Subject: medical-grade transformers



With regard to transformer construction, CLAUSE-57.9.4(e) of EN-60601-1
General Requirements for Safety of Medical Electrical Equipment requires:-

In transformers with REINFORCED INSULATION or DOUBLE INSULATION the
insulation between 1ary and 2ary windings shalll consist of -
-  1 insulation layer having thickness at least 1mm, or
-  at least 2 insulation layers with total thickness not less than 0.3mm, or
-  3 layers provided each combination of 2 layers can withstand the
dielectric strenght test for REINFORCED INSULATION

Can anyone explain ther rationale behind these requirements ?

Thanks.

Ged Dean
Nottingham City Hospital.



**
This email and any files transmitted with it are confidential and
intended solely for the use of the individual or entity to whom they
are addressed. If you have received this email in error please notify
the system manager.

This footnote also confirms that this email message has been swept by
MIMEsweeper for the presence of computer viruses.

NCHT
**


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


medical-grade transformers

2002-07-29 Thread Dean Gerard (gdean)

With regard to transformer construction, CLAUSE-57.9.4(e) of EN-60601-1
General Requirements for Safety of Medical Electrical Equipment requires:-

In transformers with REINFORCED INSULATION or DOUBLE INSULATION the
insulation between 1ary and 2ary windings shalll consist of -
-  1 insulation layer having thickness at least 1mm, or
-  at least 2 insulation layers with total thickness not less than 0.3mm, or
-  3 layers provided each combination of 2 layers can withstand the
dielectric strenght test for REINFORCED INSULATION

Can anyone explain ther rationale behind these requirements ?

Thanks.

Ged Dean
Nottingham City Hospital.



**
This email and any files transmitted with it are confidential and
intended solely for the use of the individual or entity to whom they
are addressed. If you have received this email in error please notify
the system manager.

This footnote also confirms that this email message has been swept by
MIMEsweeper for the presence of computer viruses.

NCHT
**


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list