Hi,

Have any of  you used the Thermo Stainmate Mix? It has 10 dishes holding 200ml 
of fluid? Am looking for opinions. We are considering getting this for our rush 
FNAs.

Thanks.

__
Lynn O'Donnell, CT (ASCP), MHA
Technical Specialist, Cytology
Danbury Hospital
203-739-7846
lynn.o'donn...@wchn.org



___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: formalin substitutes. HELP

2012-11-09 Thread Connolly, Brett M
Hello Nieves,

Here in the US Anatech LTD sells a gyloxal fixative called Prefer. I have 
attached a link to the website. Go to the MSDS menu and click on 'Prefer' to 
pull up the MSDS. 

http://www.anatechltdusa.com/

Below is an old  post from the HistoNet that might be helpful. I suggest you 
contact Ada Feldman at Anatech, as she should be able to address your issues

Regards,
Brett


Brett M. Connolly, Ph.D.
Principal Scientist, Imaging Dept.
Merck  Co., Inc.
PO Box 4, WP-44K
West Point, PA 19486
brett_conno...@merck.com
T- 215-652-2501
F- 215-993-6803



As the developers of the Prefer fixative we would like to address  
some of the issues.

Interchangeability of glyoxal products: The manufacturer of each  
fixative should be able to provide the information necessary to work  
with their fixative in conjunction with any other fixative that a lab  
may be using. Just as an example, Hollandes is not compatible with  
NBF and requires special attention. As for Prefer we can say it is  
compatible with the majority of other fixatives, glyoxal or not.

Limited time in the fixative: There is a slight reduction in staining  
intensity after several weeks, but increasing staining time corrects  
this. So tissues are not rendered unstainable.

Transition period: This is true any time you are switching fixatives  
or processing methods.

Eosinophila: Your statements here are true.

Lysis of erythrocytes: True again. It is often seen as an advantage  
because diagnostics cells are easier to see.

Breast cancer: It would seem that the improved nuclear morphology  
would make marked nuclear variation easier to determine.

Prostate biopsies: Hope you can get a response from any of the  
glyoxal users with prostate biopsies. We would be interested  in this  
answer also.


Ada T. Feldman, MS, HT/HTL(ASCP)
ANATECH LTD.
1020 Harts Lake Road
Battle Creek, MI 49015

Phone: 800.262.8324
Fax: 269.964.8084
email: adafeld...@anatechltdusa.com
website: www.anatechltdusa.com


On Jul 14, 2006, at 4:13 PM, rsrichm...@aol.com wrote:

 The people at Anatech, makers of Prefer fixative, have published a  
 review of
 glyoxal fixation that every pathologist and histotechnologist ought  
 to read.
 This working surgical pathologist would like to add - and solicit -  
 some
 comments on Histonet.

 Glyoxal Fixation and Its Relationship to Immunohistochemistry.  
 Richard W.
 Dapson, Ada T. Feldman, and Dee Wolfe. Anatech Limited, Battle  
 Creek MI. The
 Journal of Histotechnology June 2006;29:65-76.

 I don't want to change, but I think we all need to be prepared for  
 the day
 when a manager walks into our laboratory, or a letter from a  
 regulatory agency
 arrives in the mail, telling us that we have to get rid of  
 formaldehyde right
 now. Probably glyoxal is the only acceptable substitute, and we all  
 need to
 have a look at it. I have a number of questions.

 Interchangeability of glyoxal products: Prefer is described as a  
 buffered
 solution of glyoxal with a pH of about 4. The formula is a trade  
 secret.
 Competing glyoxal products probably also have trade-secret  
 formulas. So can a lab
 change brands of buffered glyoxal without problems, or does it have  
 to stay with a
 particular brand with its trade-secret buffering? - We've seen a  
 similar
 problem with distilling aliphatic xylene substitutes: every one of  
 them requires a
 separate distillation routine, at least on a spinning-band still. -  
 I'll
 leave it to John Kiernan to comment on the appropriateness of trade- 
 secret
 reagents in histopathology.

 Limited time in the fixative: Tissue can be left in neutral  
 buffered formalin
 for quite a long time and still be stainable, but tissue stored in  
 glyoxal
 becomes unstainable after about two weeks. Can glyoxal fixed tissue be
 transferred to 70% ethanol for more prolonged storage? - A very  
 occasional surgical
 specimen requires additional blocks after a week - a bigger problem  
 will be the
 pathologist who doesn't trim his autopsies promptly.

 Transition period: A laboratory changing to glyoxal would have to  
 keep IHC
 procedures for both fixatives working for some time. There would  
 have to be some
 way to identify whether a block was fixed in formaldehyde or glyoxal.

 Eosinophilia: One ought to be able to distinguish eosinophils from
 neutrophils in tissue sections by nuclear morphology, without  
 having to see granules.
 But quantitation of eosinophils - needed in an increasing number of  
 GI biopsy
 situations - could be a problem. We might need an IHC for  
 eosinophils in some of
 these settings.

 Lysis of erythrocytes: Not much of a problem, since we're used to  
 it with
 acid fixatives anyway.

 Breast cancer: Elimination of nuclear bubble artifact in breast biopsy
 specimens may raise the apparent nuclear grade of tumors, and thus  
 increase
 Nottingham (Elston-Ellis) scores.

 Prostate biopsies: I'd want to see some prostate biopsies - is  
 somebody from
 OURLab in 

Re: [Histonet] formalin substitutes. HELP

2012-11-09 Thread Rene J Buesa
Nieves:
NO formalin substitute will work in the same way as formalin and the solution 
is not to start testing other substitutes that will make your life miserable 
and your sections and staining procedures of sub-standard quality.
The solution is:
1- to use LESS amounts of formalin (a 5:1 volume is more than enough);
2- have a well ventilated area;
3- do NOT prepare your own buffered formalin; buy pre-filled sample bottles; and
4- never handle formalin more than absolutely necessary.
Under separate cover I am sending you 2 articles I wrote on this subject
René J.



From: Nieves Gomez ngo...@neiker.net
To: histonet@lists.utsouthwestern.edu 
Sent: Friday, November 9, 2012 8:55 AM
Subject: [Histonet] formalin substitutes. HELP

Dear Histonetters,

I'm new in the net. I work as Pathologist in a Vet Lab in Spain.

Because formalin is toxic our Lab is for the practice of using
alternative fixatives. I think the main viable alternative is glyoxal
based formulas, but I have so many questions that Commercials don't know
or don't want to answer me. For example, have a MSDS or is it
accessible? Really is less hazardous than formalin or just is not
checked? (the advantages and desadvantages of formalin are known for at
least 100 years). Related to this, I think the glyoxal is suggested as a
formalin substitute in an article in 1940's and now it is sold as a new
product and most of the products are sold as green, no-toxic or
non harmful. In my opinion, a fixative can not be non-toxic if you
want it fixed tissues. 

Another question is the time needed to fix tissues or the ratio volume
specimen/fixative. To the first point, I have read an article that
mentions there is mould growth in specimens over time. Are we changing a
chemical risk to a biological risk? In my lab we have a specifically
workstation for the gross examination and sectioning of specimens, and
we wear all the Personal Protective Equipment needed (formalin chemical
filters, gloves, googles...) that minimizes the risk (the chemical risk
not the biological risk).  It is believed that formalin given time will
kill any microorganisms (or spores) present in tissues, mycobacteria
also.. what about these new products? Are they germicidal?

I do not get to appreciate the morfological changes (nuclear changes,
lysis of erithrocites, eosinophilia...) because they are well
documented. My aim is to know if there is any Lab that works with any
formalin substitute routinely to aks these questions. Please, help me.

Thanks and have a nice weekend


Nieves Gomez
Veterinary Pathologist
Animal Health Department
NEIKER-Tecnalia 
Berreaga, 1.
48160 Derio. Bizkaia. Spain.
ngo...@neiker.net

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Stain for osteoporotic bone in Spurr plastic

2012-11-09 Thread Patsy Ruegg
Allison,

I had a great von kossa protocol for this but it was on glycol methacrylate
embedded bone, here it is, it uses the fluorescence property of eosin which
will stain the osteoid with the calcified bone covered by the silver von
kossa.

GMA is not removed
Rinse 5 micron sections in dih20, use clean glassware
Make 2% silver nitrate, put on the slides and expose them to uv (a lamp or
even in the window will do, for about 20 min.
The calcified bone should turn black
Rinse well in dih20, you can fix the silver with sodium thiosulfate but do
it quickly as it does remove some silver, rinse well in dih20
Hematoxylin nuclear stain, we used Gills#3 for 5 min., wash in tap water,
blue in ammonia water, wash well in tap water then to dih20
Make a 0.2% solution of aqueous eosin y, stain slides in this for 20 min.,
rinse in dih20 and airdry, coverslip with permount.
You can measure the calcified bone and nuclei in the light microscope, then
use a fluorescent scope with a fitc filter to see the osteoid, it will be
green just on the surface of the calcified bone which is black from the
silver.  We got a total sample area measurement, calcified area as a
percentage of total area, and osteoid area as a percentage of total area.
We did other measurements as well such as number of osteoclasts (we used a
Acid phosphotase (TRAP) enzyme histochemical stain for this) and number of
osteoblast (can use alkaline phosphotase enzyme histochemical stain for
these or we just counted them from the VK slide with the heme nuclear stain
by morphology, they will be lining the surface of the bone, very flat and
run together.)

Hope this helps.  Goldner TC is also used to measure osteoid but this worked
best for us.

Patsy Ruegg, HT(ASCP)QIHC
Ruegg IHC Consulting, LLC
40864 Arkansas Ave
Bennett, CO 80102
Phone: 303-644-4538
Fax: 720-859-4110
pru...@ihctech.net
-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Jack Ratliff
Sent: Thursday, November 08, 2012 4:52 PM
To: Malandra, Allison E; Histonet
Subject: RE: [Histonet] Stain for osteoporotic bone in Spurr plastic

Allison,
I am not exactly certain if this protocol will work for Spurr's embedded
specimens, but here is the protocol that I use for deplasticized sections of
methyl methacrylate embedded undemineralized bone (non-decalcified):
Xylenes - warmed @ 50 C for 30 min w/ dip  dunk @ 15 min interval
Xylenes - warmed @ 50 C for 30 min w/ dip  dunk @ 15 min interval
Xylenes - warmed @ 50 C for 30 min w/ dip  dunk @ 15 min interval
100% EtOH - room temp for 5 min
95% EtOH - room temp for 5 min
70% EtOH - room temp for 5 min
DI H2O - room temp for 5 min
Silver Nitrate (20 g) + DI H2O (400 ml) - room temp (in dark protected from
light) for 5 min
DI H2O rinse - room temp (protected from light) for 1 min
DI H2O rinse - room temp (protected from light) for 1 min
DI H2O rinse - room temp (protected from light) for 1 min
Sodium Carbonate (22.5 G) + DI H2O (337.5 ml) + 37% Formaldehyde (112.5 ml)
- room temp (protected from light) for 2 min
DI H2O rinse - room temp for 1 min
DI H2O rinse - room temp for 1 min
Sodium Thiosulfate (40 g) + Potassium Ferricyanide (2 g) + DI H2O (420 ml)
Solution - room temp for 30 sec(solution stable for 30-60 min after addition
of potassium ferricyanide)
Running Tap Water Rinse - 10 min
Counterstain w/ MacNeal's Tetrachrome (12 g) + DI H2O (600 ml) - room temp
for 6-8 min(stir continuously w/ heat @ 60C for 48 hours covered, then
filter with paper towel)
DI H2O rinse - room temp for 1 min
DI H2O rinse - room temp for 1 min
DI H2O rinse - room temp for 1 min
70% EtOH - room temp for 1 min
100% EtOH - room temp for 1 min
Xylenes - room temp for 3 min
Xylenes - room temp for 3 min
Coverslip w/ Eukitt


Hope this helps! Please feel free to contact me if you have any questions or
concerns.
Best Regards,
Jack

Jack L RatliffOwner, Ratliff Histology Consultants, LLCChairman, Hard Tissue
Committee - National Society for Histotechnology
(317) 281-1975jratl...@ratliffhistology.com LinkedIn Profile:
http://www.linkedin.com/profile/edit?trk=hb_tab_pro_top




 From: allison-malan...@uiowa.edu To: histonet@lists.utsouthwestern.edu
 Date: Thu, 8 Nov 2012 22:52:42 +
 Subject: [Histonet] Stain for osteoporotic bone in Spurr plastic
 
 Hi there -
 
   We are trying to stain osteoporotic bone that was embedded using Spurr
plastic.  We need to be able to differentiate osteoid from mineralized bone.
We have tried Gio's trichrome and Goldner's trichrome with no success, both
surface staining and on deplasticized slides.  We are thinking of trying Von
Kossa, does anyone have a recipe/protocol for this?  Or other types of
stains that you have had success with?  Thank you!
 
 Allison Malandra, DVM
 University of Iowa, Bone Healing Research Lab
 
 
 
 Notice: This UI Health Care e-mail (including attachments) is covered by
the Electronic Communications Privacy 

[Histonet] decal affecting IHC

2012-11-09 Thread Diana McCaig
Does decalcifying tissue  (Calex II) affect the antibody reaction for IHC in 
bony tissue?

Diana

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] decal affecting IHC

2012-11-09 Thread Jack Ratliff
I do not know what is in Calex II but high concentration HCl can have a 
negative affect. We use 5% formic acid without issue.

Jack


Jack L Ratliff
Owner/Histologist, Ratliff Histology Consultants, LLC
Chairman, Hard Tissue Committee - National Society for Histotechnology



On Nov 9, 2012, at 11:34 AM, Diana McCaig dmcc...@ckha.on.ca wrote:

 Does decalcifying tissue  (Calex II) affect the antibody reaction for IHC in 
 bony tissue?
 
 Diana
 
 ___
 Histonet mailing list
 Histonet@lists.utsouthwestern.edu
 http://lists.utsouthwestern.edu/mailman/listinfo/histonet
 

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] decal affecting IHC

2012-11-09 Thread Rene J Buesa
IHC reactions can be affected by decalcification especially if:
1- the tissue is not perfectly fixed before going into decalcification
2- decalcification is done at temperature above room temp.
3- decal solution is too acid, specially made with inorganic acids.
If at all possible decalcification should be done with a chelating agent, like 
EDTA
René J.



From: Diana McCaig dmcc...@ckha.on.ca
To: 'histonet@lists.utsouthwestern.edu' histonet@lists.utsouthwestern.edu 
Sent: Friday, November 9, 2012 12:34 PM
Subject: [Histonet] decal affecting IHC

Does decalcifying tissue  (Calex II) affect the antibody reaction for IHC in 
bony tissue?

Diana

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] decal affecting IHC

2012-11-09 Thread Elizabeth Chlipala
Diana

I agree with Jack, formic acid is the way to go in my opinion, we use 10% 
formic acid routinely but have pushed it up to 20% on occasion when time was an 
issue.  HCL can work if its controlled really well but you don't have much 
wiggle room, you can over decal quite quickly, formic acid is a bit more 
forgiving.  As for Rene's comment we have seen the exact opposite with respects 
to the EDTA decal part.  Granted we do not run a lot of IHC on EDTA decaled 
samples but on several occasions and with several antibodies we were able to 
obtain good IHC staining in formic acid decaled samples but those antibodies 
did not work well in EDTA decaled samples.  These were not direct comparisons 
of the decalcification method on the same study, but on samples from different 
studies so other factors could have affected the results.

Liz

Elizabeth A. Chlipala, BS, HTL(ASCP)QIHC
Manager
Premier Laboratory, LLC
PO Box 18592
Boulder, CO 80308-1592
(303) 682-3949 office
(303) 682-9060 fax
(303) 881-0763 cell
www.premierlab.com

Ship to address:

1567 Skyway Drive, Unit E
Longmont, CO 80504

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Rene J Buesa
Sent: Friday, November 09, 2012 10:47 AM
To: Diana McCaig; 'histonet@lists.utsouthwestern.edu'
Subject: Re: [Histonet] decal affecting IHC

IHC reactions can be affected by decalcification especially if:
1- the tissue is not perfectly fixed before going into decalcification
2- decalcification is done at temperature above room temp.
3- decal solution is too acid, specially made with inorganic acids.
If at all possible decalcification should be done with a chelating agent, like 
EDTA
René J.



From: Diana McCaig dmcc...@ckha.on.ca
To: 'histonet@lists.utsouthwestern.edu' histonet@lists.utsouthwestern.edu
Sent: Friday, November 9, 2012 12:34 PM
Subject: [Histonet] decal affecting IHC

Does decalcifying tissue  (Calex II) affect the antibody reaction for IHC in 
bony tissue?

Diana

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Becky Orr

2012-11-09 Thread Andrea T. Hooper
I'm looking to contact Rebecca (Becky) Orr. Does anyone know her current email 
address? The one I have bounced back.
 
Thanks, Andrea
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Microtome knives

2012-11-09 Thread Jon Krupp
Greetings

I need some advice regarding microtome knives. I am not  histo tech, I did all 
my sectioning in a plant research lab, but now I find myself needing to learn 
more about histo type methods.

We have microtomes, AO 820's, and we have a bunch of donated knives. I need 
advice about whether it would be better to find a knife sharpener and use the 
microtome knives we have, or check into getting a disposable knife holder. 

When I was sectioning, we just used a simple razor blade holder. Now I see 
references to high profile and low profile blades and holders, and I don't know 
the difference. 

Anyone willing to help me out?

Thanks

Jon

Jonathan Krupp
Delta College
5151 Pacific Ave.
Box 212
Stockton, CA  95207
209-954-5284
jkr...@deltacollege.edu

Find us on Facebook @
Electron Microscopy at SJ Delta College







___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Microtome knives

2012-11-09 Thread Rene J Buesa
Because of your donated knives  you will have to buy a knives sharpener which 
are costly and not very easy to find.
Your best option is to buy a high profile disposable blades holder (that will 
be cheaper), buy disposable blades and avoid all the frustrations and waste of 
time sharpening knives.
René J.




From: Jon Krupp jkr...@deltacollege.edu
To: Histonet@lists.utsouthwestern.edu 
Sent: Friday, November 9, 2012 1:49 PM
Subject: [Histonet] Microtome knives

Greetings

I need some advice regarding microtome knives. I am not  histo tech, I did all 
my sectioning in a plant research lab, but now I find myself needing to learn 
more about histo type methods.

We have microtomes, AO 820's, and we have a bunch of donated knives. I need 
advice about whether it would be better to find a knife sharpener and use the 
microtome knives we have, or check into getting a disposable knife holder. 

When I was sectioning, we just used a simple razor blade holder. Now I see 
references to high profile and low profile blades and holders, and I don't know 
the difference. 

Anyone willing to help me out?

Thanks

Jon

Jonathan Krupp
Delta College
5151 Pacific Ave.
Box 212
Stockton, CA  95207
209-954-5284
jkr...@deltacollege.edu

Find us on Facebook @
Electron Microscopy at SJ Delta College







___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Microtome knives

2012-11-09 Thread Connolly, Brett M
Hi Jon,

Depends on your cash flow. You could get a used sharpener somewhere off the web 
somewhere such as 
http://www.labx.com/v2/adsearch/resultsnew.cfm?sw=sharpenermr=25te=cat , or 
http://www.medwow.com/used-microtome-knife-sharpener-equipment/63.med but 
sharpening knives is a pain IMO and steel knives present more of a safety 
hazard. I would recommend a sharpener that uses the glass honing plates. You 
would also need the coarse and fine abrasives.

Personally, I would opt for a low profile disposable blade holder that fits 
your 820. Low and high profile refer to the size (height) of the blade. We use 
low profile for paraffin block sectioning and high profile for cryostat 
sectioning.

Brett

Brett M. Connolly, Ph.D.
Principal Scientist, Imaging Dept.
Merck  Co., Inc.
PO Box 4, WP-44K
West Point, PA 19486
brett_conno...@merck.com
T- 215-652-2501
F- 215-993-6803

 

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Jon Krupp
Sent: Friday, November 09, 2012 1:50 PM
To: Histonet@lists.utsouthwestern.edu
Subject: [Histonet] Microtome knives

Greetings

I need some advice regarding microtome knives. I am not  histo tech, I did all 
my sectioning in a plant research lab, but now I find myself needing to learn 
more about histo type methods.

We have microtomes, AO 820's, and we have a bunch of donated knives. I need 
advice about whether it would be better to find a knife sharpener and use the 
microtome knives we have, or check into getting a disposable knife holder. 

When I was sectioning, we just used a simple razor blade holder. Now I see 
references to high profile and low profile blades and holders, and I don't know 
the difference. 

Anyone willing to help me out?

Thanks

Jon

Jonathan Krupp
Delta College
5151 Pacific Ave.
Box 212
Stockton, CA  95207
209-954-5284
jkr...@deltacollege.edu

Find us on Facebook @
Electron Microscopy at SJ Delta College







___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
Notice:  This e-mail message, together with any attachments, contains
information of Merck  Co., Inc. (One Merck Drive, Whitehouse Station,
New Jersey, USA 08889), and/or its affiliates Direct contact information
for affiliates is available at 
http://www.merck.com/contact/contacts.html) that may be confidential,
proprietary copyrighted and/or legally privileged. It is intended solely
for the use of the individual or entity named on this message. If you are
not the intended recipient, and have received this message in error,
please notify us immediately by reply e-mail and then delete it from 
your system.


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Dermatology KOH testing

2012-11-09 Thread Trini Nguyen
Does anyone perform KOH tests? If so, could you please email me and give me
information on your process.

Thanks,
Trini
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] test

2012-11-09 Thread AFirst Name Gee
Just testing-new subscriber
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Diane Tokugawa/CA/KAIPERM is out of the office.

2012-11-09 Thread Diane . Tokugawa


I will be out of the office starting  11/09/2012 and will not return until
11/15/2012.

Note:   For Cytology issues, please call Robin (day) at 8-421-5040, Wanda
(day)  8-421-5426, or Eric (swing) 8-421-5405   For Histology  issues,
please call Mario (day) 8-421-4961, Barbara (swing) 8-421-4959,  Bob (IHC)
8-421-4706,   Kiran at 8-421-5404,  or general histology client service at
8-421-5408 or Lotus Notes.___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Mold Release Problems

2012-11-09 Thread AFirst Name Gee
Does anyone know of a reference for the problems from using mold release?  When 
too much is used on the mold it eats into the block and causes the ribbons to 
explode on the water bath and creates an artifact in the tissue.I cannot find a 
reference for it or any published material about the problem but have been to 
seminars where it was discussed.When it is not used the problem never occurs.
___
Histonet mailing list