[Histonet] RE: Bubble problems

2014-08-08 Thread Wolfe, Christina
Hi Melissa,
We have intermittently had problems with our Leica Glass coverslipper over the 
years.  During troubleshooting with their tech support, we found our issue was 
most likely with the coverslips.  We use the VWR micro glass 24x50 mm cat# 
48404-452 coverslips and the Tissue-Tek Glas mounting medium cat #6419.  There 
are a couple of settings on the instrument that control the amount of mounting 
medium flow onto the glass slide - you can call me if you want our settings 
that work well for us.  Also, the xylene in the 'holding area' on the 
coverslipper should be changed regularly and needs to be pretty full (there is 
a line on ours).  We have found if that xylene is too old and if the volume of 
xylene drops below the fill line - we start getting micro bubbles.  One last 
thing, it is important to also push the button to drop the mounting medium 
dispenser at the end of the day to be submerged in the little container of 
xylene and to make sure and clean the rubber stopper that picks up the glass 
coverslips with the brush before each day of use.  We have found that all of 
these elements affect air bubbles on the slides.  Feel free to give me a call 
if you like.
Christina Wolfe
Bristol Myers Squibb
812-307-2093


From: histonet-boun...@lists.utsouthwestern.edu 
histonet-boun...@lists.utsouthwestern.edu on behalf of Melissa Ng 
melissa...@spdscientific.com
Sent: Thursday, August 07, 2014 7:59 AM
To: Histonet@lists.utsouthwestern.edu
Subject: [Histonet] Bubble problems!

Dear Histonet,



I am having a major problem with bubbles in my mounting media. I am currently 
using the Leica coverslipper and Cellpath mounting media (Phalate free). 
Micro-bubbles are observed on the slides after sometime of coverslipping. In 
fact, when I checked the media, I found lots of bubbles inside the bottle too! 
There were no bubbles in the bottle prior to placing the mounting media with 
the machine (I checked!). Also, we do not see any bubbles in the tubes that 
lead to the probe!



Previously, we were using the Leica mounting media (yellowish liquid) and there 
were no bubble issues. I have checked with other labs using the Cellpath 
mounting media (they use Leica coverslipper too) and no one seems to be having 
the same problem as us!





The Cellpath mounting media is way more viscous compared to that of the Leica 
media.



Has anyone here experienced a similar issue, and how have you resolved it?



Appreciate all the help I can get!



Best regards,

Melissa
Application Specialist
___
**

This message (including any attachments) may contain confidential, proprietary, 
privileged and/or private information.  The information is intended to be for 
the use of the individual or entity designated above.  If you are not the 
intended recipient of this message, please notify the sender immediately, and 
delete the message and any attachments.  Any disclosure, reproduction, 
distribution or other use of this message or any attachments by an individual 
or entity other than the intended recipient is prohibited.

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: Bubble problems!

2014-08-07 Thread Barbara Tibbs
Try adding xylene to the mounting medium to thin it.  About 10 ml of xylene to 
90 ml of mounting medium should do it.
Then, leave the bottle of mounting medium for about a day or two to allow any 
air that may have gotten mixed into to dissipate.

Barbara S. Tibbs
Histology Supervisor
Accurate Diagnostic Labs
South Plainfield, NJ
barbara.ti...@accuratediagnosticlabs.com
732-839-3374
Cell: 610-809-6508



From: histonet-boun...@lists.utsouthwestern.edu 
histonet-boun...@lists.utsouthwestern.edu on behalf of Melissa Ng 
melissa...@spdscientific.com
Sent: Thursday, August 07, 2014 7:59 AM
To: Histonet@lists.utsouthwestern.edu
Subject: [Histonet] Bubble problems!

Dear Histonet,



I am having a major problem with bubbles in my mounting media. I am currently 
using the Leica coverslipper and Cellpath mounting media (Phalate free). 
Micro-bubbles are observed on the slides after sometime of coverslipping. In 
fact, when I checked the media, I found lots of bubbles inside the bottle too! 
There were no bubbles in the bottle prior to placing the mounting media with 
the machine (I checked!). Also, we do not see any bubbles in the tubes that 
lead to the probe!



Previously, we were using the Leica mounting media (yellowish liquid) and there 
were no bubble issues. I have checked with other labs using the Cellpath 
mounting media (they use Leica coverslipper too) and no one seems to be having 
the same problem as us!



The Cellpath mounting media is way more viscous compared to that of the Leica 
media.



Has anyone here experienced a similar issue, and how have you resolved it?



Appreciate all the help I can get!



Best regards,

Melissa
Application Specialist
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: Bubble problems!

2014-08-07 Thread Horn, Hazel V
Melissa we are having the same problem.  We use a different mounting medium.  I 
will try Barbara's advice and try adding a bit of xylene to the media.

Hazel Horn
Supervisor of Histology/Autopsy/Transcription
Anatomic Pathology
Arkansas Children's Hospital
1 Children's Way | Slot 820| Little Rock, AR 72202
501.364.4240 direct | 501.364.1241 fax
hor...@archildrens.org
archildrens.org






-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Barbara Tibbs
Sent: Thursday, August 07, 2014 8:12 AM
To: Melissa Ng; Histonet@lists.utsouthwestern.edu
Subject: [Histonet] RE: Bubble problems!

Try adding xylene to the mounting medium to thin it.  About 10 ml of xylene to 
90 ml of mounting medium should do it.
Then, leave the bottle of mounting medium for about a day or two to allow any 
air that may have gotten mixed into to dissipate.

Barbara S. Tibbs
Histology Supervisor
Accurate Diagnostic Labs
South Plainfield, NJ
barbara.ti...@accuratediagnosticlabs.com
732-839-3374
Cell: 610-809-6508



From: histonet-boun...@lists.utsouthwestern.edu 
histonet-boun...@lists.utsouthwestern.edu on behalf of Melissa Ng 
melissa...@spdscientific.com
Sent: Thursday, August 07, 2014 7:59 AM
To: Histonet@lists.utsouthwestern.edu
Subject: [Histonet] Bubble problems!

Dear Histonet,



I am having a major problem with bubbles in my mounting media. I am currently 
using the Leica coverslipper and Cellpath mounting media (Phalate free). 
Micro-bubbles are observed on the slides after sometime of coverslipping. In 
fact, when I checked the media, I found lots of bubbles inside the bottle too! 
There were no bubbles in the bottle prior to placing the mounting media with 
the machine (I checked!). Also, we do not see any bubbles in the tubes that 
lead to the probe!



Previously, we were using the Leica mounting media (yellowish liquid) and there 
were no bubble issues. I have checked with other labs using the Cellpath 
mounting media (they use Leica coverslipper too) and no one seems to be having 
the same problem as us!



The Cellpath mounting media is way more viscous compared to that of the Leica 
media.



Has anyone here experienced a similar issue, and how have you resolved it?



Appreciate all the help I can get!



Best regards,

Melissa
Application Specialist
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

**
The information contained in this message may be privileged and confidential
and protected from disclosure. If the reader of this message is not the 
intended recipient, or an employee or agent responsible for delivering this 
message to the intended recipient, you are hereby notified that any 
dissemination, distribution or copying of this communication is strictly 
prohibited. If you have received this communication in error, please notify 
us immediately by replying to the message and deleting it from your computer.
Thank you.

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: Bubble problems!

2014-08-07 Thread Melissa Ng
Dear Hazel and Barbara,

Thank you for your feedback. Wouldn't adding too much xylene result in drying 
effects post-coverslipping?

Best thanks and regards,
Melissa

-Original Message-
From: Horn, Hazel V [mailto:hor...@archildrens.org] 
Sent: Thursday, 7 August, 2014 10:10 PM
To: 'Barbara Tibbs'; Melissa Ng; Histonet@lists.utsouthwestern.edu
Subject: RE: Bubble problems!

Melissa we are having the same problem.  We use a different mounting medium.  I 
will try Barbara's advice and try adding a bit of xylene to the media.

Hazel Horn
Supervisor of Histology/Autopsy/Transcription
Anatomic Pathology
Arkansas Children's Hospital
1 Children's Way | Slot 820| Little Rock, AR 72202
501.364.4240 direct | 501.364.1241 fax
hor...@archildrens.org
archildrens.org






-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Barbara Tibbs
Sent: Thursday, August 07, 2014 8:12 AM
To: Melissa Ng; Histonet@lists.utsouthwestern.edu
Subject: [Histonet] RE: Bubble problems!

Try adding xylene to the mounting medium to thin it.  About 10 ml of xylene to 
90 ml of mounting medium should do it.
Then, leave the bottle of mounting medium for about a day or two to allow any 
air that may have gotten mixed into to dissipate.

Barbara S. Tibbs
Histology Supervisor
Accurate Diagnostic Labs
South Plainfield, NJ
barbara.ti...@accuratediagnosticlabs.com
732-839-3374
Cell: 610-809-6508



From: histonet-boun...@lists.utsouthwestern.edu 
histonet-boun...@lists.utsouthwestern.edu on behalf of Melissa Ng 
melissa...@spdscientific.com
Sent: Thursday, August 07, 2014 7:59 AM
To: Histonet@lists.utsouthwestern.edu
Subject: [Histonet] Bubble problems!

Dear Histonet,



I am having a major problem with bubbles in my mounting media. I am currently 
using the Leica coverslipper and Cellpath mounting media (Phalate free). 
Micro-bubbles are observed on the slides after sometime of coverslipping. In 
fact, when I checked the media, I found lots of bubbles inside the bottle too! 
There were no bubbles in the bottle prior to placing the mounting media with 
the machine (I checked!). Also, we do not see any bubbles in the tubes that 
lead to the probe!



Previously, we were using the Leica mounting media (yellowish liquid) and there 
were no bubble issues. I have checked with other labs using the Cellpath 
mounting media (they use Leica coverslipper too) and no one seems to be having 
the same problem as us!




The Cellpath mounting media is way more viscous compared to that of the Leica 
media.



Has anyone here experienced a similar issue, and how have you resolved it?



Appreciate all the help I can get!



Best regards,

Melissa
Application Specialist
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

**
The information contained in this message may be privileged and confidential
and protected from disclosure. If the reader of this message is not the 
intended recipient, or an employee or agent responsible for delivering this 
message to the intended recipient, you are hereby notified that any 
dissemination, distribution or copying of this communication is strictly 
prohibited. If you have received this communication in error, please notify 
us immediately by replying to the message and deleting it from your computer.
Thank you.

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet