Re: [time-nuts] Racal-Dana 1992 switches

2008-03-28 Thread Bruce Griffiths
Charles S. Osborne wrote:
 Now the real question is... is there a clever way to make the Racal 1992
 readout the difference in µHz between two GPS disciplined oscillators? My
 only other counter, an HP5384A, is offscale at 10.000 000.000 MHz . I'm
 referencing the counter with one Lucent RFTG-m-XO and clocking a Lucent
 RFTGm-II-XO. Things are working well enough to be beyond my counter's
 ability to see any jitter. The Racal says nanosecond time interval counter,
 so I bet there's a way to subtract and increase the resolution similar to an
 HP53131?

 tnx,
 Charles
 K4CSO
 Duluth, GA
   
Charles

One way to achieve microHertz resolution at 10MHz is to use a dual mixer 
time difference system or a variant thereof.

The output from each GPDSDO is mixed with the same low noise offset source.
This produces a pair of beat frequencies (one from each mixer) outputs.

The zero crossing from one beat frequency output is used connected start 
a time interval counter whilst the zero crossing of the other beat 
frequency output is used to stop the time interval counter.

The trick is to amplify the low slew rate beat frequency signal in such 
a way that the time interval counter sees 2 high slew rate low jitter 
signals.
Just connecting the low pass filtered beat frequency outputs to the 
counter inputs isnt effective as the measured timing jitter will 
inevitably be much greater than the actual zero crossing jitter in the 
beat frequency signal itself.
A resolution approaching 1E-14/sec is possible with this technique.

Another approach is to use a series of cascaded frequency difference 
multipliers, this is the approach used by Quartzlock in their precision 
frequency comparators.
This approach can also have a resolution approaching 1E-14/sec.

If you are willing to wait a little longer to see microHertz frequency 
differences then an analog phase comparator can be used.
A resolution approaching 1E-11/sec is possible with this technique if a 
high resolution ADC (eg  a high res DVM) is connected to the phase 
detector output.

Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Racal-Dana 1992 switches

2008-03-28 Thread Chris Erickson
Actually, after two days of studying catalogs and datasheets, I think I
found a suitable replacement switch. The trick is to find the ones made for
tape packaging, which only have 2 pins -- the others have 4. The Omron
B3F-6050, available through Digikey as part no. SW798-ND is $0.28 each for
100 ($0.39 each for singles). It has nearly the same dimensions as the
original Toko, except it is about 3mm shorter so it will need to either be
shimmed up or left sitting up on its pins and the plunger is the standard
2.4mm square instead of the cross, so the keycaps will need some
modification. I'll let you all know how they work out once I receive and
install them. 

 

Chris

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Tboltmon Serial Port Selection...??

2008-03-28 Thread Stephen Tompsett
My version of Tboltmon (version 2.60)accepts a command line argument 
which bypasses the selection menu e.g.

TBOLTMON.EXE -c1

This does not appear to be documented.

Michael Baker wrote:
 Hello, All--
 
 Can anyone tell me why, when I click on the Tboltmon
 icon associated with my Trimble Thunderbolt,
 that a little Serial Port Selecton window pops up
 and why does my T-bolt only run after selecting
 COM-4?  Why can't this be selected once and then
 forgotten about?  It seems like an un-necessary step.
 
 Thanks!!
 
 Mike Baker
 WA4HFR
 Micanopy, FL
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 

-- 
Stephen Tompsett


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Tboltmon Serial Port Selection...??

2008-03-28 Thread John Miles
Thanks!  How'd you find that option?

-- john, KE5FX

 -Original Message-
 From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED]
 Behalf Of Stephen Tompsett
 Sent: Friday, March 28, 2008 1:12 AM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Tboltmon Serial Port Selection...??
 
 
 My version of Tboltmon (version 2.60)accepts a command line argument 
 which bypasses the selection menu e.g.
 
   TBOLTMON.EXE -c1
 
 This does not appear to be documented.
 
 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Tboltmon Serial Port Selection...??

2008-03-28 Thread Ken Winterling
That works for me, too!  I also have version 2.60.

Ken, WA2LBI

On Fri, Mar 28, 2008 at 4:11 AM, Stephen Tompsett
[EMAIL PROTECTED] wrote:
 My version of Tboltmon (version 2.60)accepts a command line argument
  which bypasses the selection menu e.g.

 TBOLTMON.EXE -c1

  This does not appear to be documented.



  Michael Baker wrote:
   Hello, All--
  
   Can anyone tell me why, when I click on the Tboltmon
   icon associated with my Trimble Thunderbolt,
   that a little Serial Port Selecton window pops up
   and why does my T-bolt only run after selecting
   COM-4?  Why can't this be selected once and then
   forgotten about?  It seems like an un-necessary step.
  
   Thanks!!
  
   Mike Baker
   WA4HFR
   Micanopy, FL
  
   ___
   time-nuts mailing list -- time-nuts@febo.com
   To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
   and follow the instructions there.
  
  

  --
  Stephen Tompsett




  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] TADD-1 six channel RF distribution amplifier

2008-03-28 Thread John Ackermann N8UR
Steve said the following on 03/27/2008 09:35 PM:
 I'm thinking about using the TAPR TADD-1 six channel RF distribution 
 amplifier to distribute the 10MHz output of my Trimble Thunderbolt to a 
 few pieces of test equipment.
 
 What does the TADD-1 do to the phase noise characteristics of its 
 outputs as compared to the phase noise characteristics of  its input? 
 Improve? Degrade? Little or no change?
 
 Thanks.
 
 Steve K8JQ
 

I did some testing a while ago on a TADD-1 as well as an HP 5087A and
Spectracom 8140T line amplifier.  At 5 MHz, for offsets above 1 Hertz,
the TADD-1 is better than the HP amp (the 5087A is better at very low
offsets, though).  The results are at
http://www.febo.com/pages/amplifier_phase_noise/

I've learned a bit about technique since I did this test, and want to
rerun it one of these days.  But the data shouldn't be more than a few
dB off.

John

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Racal-Dana 1992 switches

2008-03-28 Thread Dan Rae
Back to the switches... I found a Jameco bag for the ones I bought way 
back, which fitted pretty well and work.  The part number was 115086, 
then quoted for KIE22, but not the same as the current ones with that 
number.   Jameco might know who made the 115086, but there are no clues 
from the few remaining in the bag...

Dan


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] TADD-1 six channel RF distribution amplifier

2008-03-28 Thread Didier Juges
The phase plots on tvb's web site
http://www.leapsecond.com/pages/tadd-1/3phase2.gif show  that the device's
temperature was never stabilized. The closest to being stable is at the 29
minute point on that plot.

Particularly when the phase value goes up (not sure if it's when temperature
goes up or down), because the phase was still changing at a pretty high rate
when the compressor changed state. I think Tom should rerun that test with a
much slower cycling, to make sure the phase was stable before the compressor
was turned on (or off). Since the phase excursion was not complete, the
150pS/degree is probably underestimated.

My guess is that the period should be at least doubled.

Didier KO4BB

 -Original Message-
 From: [EMAIL PROTECTED] 
 [mailto:[EMAIL PROTECTED] On Behalf Of Jeff Mock
 Sent: Friday, March 28, 2008 12:29 AM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] TADD-1 six channel RF distribution amplifier
 
 
 Bruce Griffiths wrote:
  Steve wrote:
  I'm thinking about using the TAPR TADD-1 six channel RF 
 distribution 
  amplifier to distribute the 10MHz output of my Trimble 
 Thunderbolt to 
  a few pieces of test equipment.
 
  What does the TADD-1 do to the phase noise characteristics of its 
  outputs as compared to the phase noise characteristics of  
 its input?
  Improve? Degrade? Little or no change?
 
  Thanks.
 
  Steve K8JQ
 
 
 I've been using a TADD-1 for about 6-months now and I'm quite 
 happy with it.  You certainly can't beat the price.  I added 
 the optional 10MHz BPF and tuned it with a tracking generator 
 and spectrum analyzer.  The outputs are very clean.
 
 I've never tried to measure phase noise, but the op amps and 
 the rest of the circuit seem to be good choices to me.  tvb 
 did some analysis of the box you might find interesting:
 http://www.leapsecond.com/pages/tadd-1/
 
 jeff
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, 
 go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 No virus found in this incoming message.
 Checked by AVG. 
 Version: 7.5.519 / Virus Database: 269.21.8/1339 - Release 
 Date: 3/22/2008 4:43 PM
  
 

No virus found in this outgoing message.
Checked by AVG. 
Version: 7.5.519 / Virus Database: 269.21.8/1339 - Release Date: 3/22/2008
4:43 PM
 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] TADD-1 six channel RF distribution amplifier

2008-03-28 Thread Steve
Thanks for the info Bruce.

I'm planning to use the Trimble Thunderbolt as the 10MHz source.

The phase noise characteristics of that device are on the Trimble web 
site. I'm not versed enough in the specifications to call it good, bad 
or mediocre with respect to phase noise. Thoughts?

Thanks.

Steve K8JQ

Bruce Griffiths wrote:
 Steve wrote:
   
 I'm thinking about using the TAPR TADD-1 six channel RF distribution 
 amplifier to distribute the 10MHz output of my Trimble Thunderbolt to a 
 few pieces of test equipment.

 What does the TADD-1 do to the phase noise characteristics of its 
 outputs as compared to the phase noise characteristics of  its input? 
 Improve? Degrade? Little or no change?

 Thanks.

 Steve K8JQ

   
 
 Like any distribution amplifier it always degrades the signal phase noise.
 Whether the amount of degradation is significant or not depends on the 
 phase noise characteristics of your source.

 If your signal source phase noise is 10dB greater than the intrinsic 
 phase noise of the distribution amplifier than the increase in the phase 
 noise at the distribution amplifier is relatively insignificant.

 Thus you need to know the phase noise characteristics of your source and 
 of the distribution amplifier (when set to the gain you need).

 The phase noise characteristics of the distribution amplifier vary with 
 its gain.

 The phase noise of your source and a distribution amplifier vary with 
 the frequency offset at which they are measured.

 What source are you intending to use?
 Do you know its phase noise characteristics?

 Bruce

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

   

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] TADD-1 six channel RF distribution amplifier

2008-03-28 Thread Tom Van Baak
 The phase plots on tvb's web site
 http://www.leapsecond.com/pages/tadd-1/3phase2.gif show  that the device's
 temperature was never stabilized. The closest to being stable is at the 29
 minute point on that plot.
 
 Particularly when the phase value goes up (not sure if it's when temperature
 goes up or down), because the phase was still changing at a pretty high rate
 when the compressor changed state. I think Tom should rerun that test with a
 much slower cycling, to make sure the phase was stable before the compressor

Agreed. Actually, I don't have that A/C anymore so future
tempco tests won't have that rapid cycling.

 was turned on (or off). Since the phase excursion was not complete, the
 150pS/degree is probably underestimated.
 
 My guess is that the period should be at least doubled.

Yeah, I think so.

The main conclusion I had when I made those plots was that
the input filter was a bad idea, at least if the TADD was to be
used for precision work.

 Didier KO4BB

/tvb


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] TADD-1 six channel RF distribution amplifier

2008-03-28 Thread Tom Van Baak
 Thanks for the info Bruce.
 
 I'm planning to use the Trimble Thunderbolt as the 10MHz source.
 
 The phase noise characteristics of that device are on the Trimble web 
 site. I'm not versed enough in the specifications to call it good, bad 
 or mediocre with respect to phase noise. Thoughts?
 
 Thanks.
 
 Steve K8JQ

Here's an old plot of a random Thunderbolt I tested:
Ignore the spurs (test setup problem).

http://www.leapsecond.com/museum/tbolt/tbolt_pn.gif

/tvb


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] TADD-1 six channel RF distribution amplifier

2008-03-28 Thread Bruce Griffiths
Steve wrote:
 Thanks for the info Bruce.

 I'm planning to use the Trimble Thunderbolt as the 10MHz source.

 The phase noise characteristics of that device are on the Trimble web 
 site. I'm not versed enough in the specifications to call it good, bad 
 or mediocre with respect to phase noise. Thoughts?

 Thanks.

 Steve K8JQ

   
Steve

If one compares TvB's plot of the phase noise of a Thunderbolt with John 
Ackermann's plot of the phase noise of a TADD-1 at 10MHz.
Then the phase noise floor of this particular Thunderbolt is lower than 
the phase noise floor of the TADD-1 at the particular gain used in 
John's test.
Thus using the TADD-1 at this gain will degrade the noise floor of that 
particular Thunderbolt by about 7dB.
However at 1Hz and 10Hz the phase noise of the Thunderbolt exceeds that 
of the TADD-1 by 15dB or more so the phase noise of that Thunderbolt 
will not be significantly degraded by the TADD-1 at those offset 
frequencies.

Distribution amplifiers with lower noise floors and equivalent or better 
reverse isolation do exist but tend to be expensive.
A design employing cascaded common base stages can have a significantly 
lower phase noise floor.

Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] TADD-1 six channel RF distribution amplifier

2008-03-28 Thread Steve
Bruce and others,

Thank you for the assistance. Seems that the TADD-1 will meet my needs, 
I will probably acquire one in the near future.

Steve K8JQ

Bruce Griffiths wrote:
 Steve wrote:
   
 Thanks for the info Bruce.

 I'm planning to use the Trimble Thunderbolt as the 10MHz source.

 The phase noise characteristics of that device are on the Trimble web 
 site. I'm not versed enough in the specifications to call it good, bad 
 or mediocre with respect to phase noise. Thoughts?

 Thanks.

 Steve K8JQ

   
 
 Steve

 If one compares TvB's plot of the phase noise of a Thunderbolt with John 
 Ackermann's plot of the phase noise of a TADD-1 at 10MHz.
 Then the phase noise floor of this particular Thunderbolt is lower than 
 the phase noise floor of the TADD-1 at the particular gain used in 
 John's test.
 Thus using the TADD-1 at this gain will degrade the noise floor of that 
 particular Thunderbolt by about 7dB.
 However at 1Hz and 10Hz the phase noise of the Thunderbolt exceeds that 
 of the TADD-1 by 15dB or more so the phase noise of that Thunderbolt 
 will not be significantly degraded by the TADD-1 at those offset 
 frequencies.

 Distribution amplifiers with lower noise floors and equivalent or better 
 reverse isolation do exist but tend to be expensive.
 A design employing cascaded common base stages can have a significantly 
 lower phase noise floor.

 Bruce

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

   

-- 
Read The Patriot   It's Right -- It's Free
http://PatriotPost.US/subscribe/ Veritas vos Liberabit



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Racal-Dana 1992 switches

2008-03-28 Thread Bill Hawkins
Charles S. Osborne said, in part,

Now the real question is... is there a clever way to make the Racal
1992 readout the difference in µHz between two GPS disciplined
oscillators? My only other counter, an HP5384A, is offscale at 10.000
000.000 MHz . I'm referencing the counter with one Lucent RFTG-m-XO and
clocking a Lucent RFTGm-II-XO. Things are working well enough to be
beyond my counter's ability to see any jitter. The Racal says nanosecond
time interval counter, so I bet there's a way to subtract and increase
the resolution similar to an HP53131?

Haven't seen the answer on this list, so perhaps it occurred privately.

The Racal 1992 is able to read the phase error between two 10 MHz
signals (A and B) in degrees. I have done this with the outputs of two
3801s, which are the only pair of frequency sources that I have. This
would be sub-nanosecond accuracy except that the display shows 3-10
degrees of jitter (difference between two successive readings). This,
however, is only 10E-9. Most people on this list are investigating areas
at least two orders of magnitude lower.

I find that the phase method gives me comparative drift errors soon
enough. An hour gets you near 10E-12. Others require measurement
intervals much shorter than that, but the phase angle method is more
than adequate for time errors that humans will notice. A drift of one
second per year is on the order of 10E-8.

It all depends on your reason for pursuing accurate time/frequency
measurement.

Bill Hawkins



 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Odetics 425: Damn... It's really dead now...

2008-03-28 Thread Bruce Lane
Fellow time-tickers,

Well, I suppose it was due to happen. The GPS receiver in my Odetics 
425 finally died. Hard. The thing's been intermittent for nearly the entire two 
years since I became its new owner, so I guess I shouldn't be surprised that 
the thing finally keeled.

I would REALLY like to restore this grand old beastie, not just because 
I like the workmanship but because it has a unique history. Its last home, 
before I got it, was the Amundsen Antarctic Research Station as part of the 
AST/RO program.

So -- Before I start stretching my brain cells, and those of a close 
friend, and try to come up with a way to interface a generic NMEA receiver 
module into this wonderful old beast, I have a couple of questions to put to 
the group, especially those who used to (or still do? Hopefully?) work for 
Odetics/Zyfer.

(1) Does ANYone have a working Magellan OEM 5000 receiver module that 
they would be willing to part with? If so, please contact me off-list.

(2) The schematics I have are poor-quality copies that were reduced to 
letter size from the original ANSI size D pages. Does anyone have schematics of 
this unit's MPU board that are more readable?

(3) Failing finding a Magellan receiver -- Fitting a standard receiver 
to this beast is going to require some reverse-engineering, and perhaps 
modification of the firmware. The section that communicates with the GPS 
receiver uses a 68B09 CPU.

Suggestions on what tools would be most useful to this effort, as well 
as any hints those who have worked with 6809's might have. I do have available 
a Fluke 9100 series CPU troubleshooting system, with a 6809 pod.

(3) Was there EVER a case where a generic (as in NMEA-speaking) 
receiver was used with the 325/425 series? If so, please provide details. Or 
did they get retired before NMEA became the defacto standard?

Keep in mind that if I can come up with a viable way to do this, it 
will likely mean a new lease on equipment life for all owners of 325/425 
systems, because in the event it all works I will make the details of the 
process freely available.

Looking forward to hearing anything that might help.

Thanks much.


-=-=-=-=-=-=-=-=-=-=-=-
Bruce Lane, Owner  Head Hardware Heavy,
Blue Feather Technologies -- http://www.bluefeathertech.com
kyrrin (at) bluefeathertech do/t c=o=m
Quid Malmborg in Plano...


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.