Re: [time-nuts] Need Austron 2110 manual

2009-05-27 Thread tomknox

Hi Skip;
I think I have a 2110A manual if you can't find one let me know.
I purpose I would drop it at a FedEx Kinkos where they could copy it  
and ship the copy to you.
You could pay them directly by credit card. Be aware this could end up  
expensive
I think this manual has fold out pages, and other things that require  
hand placement.

Best Wishes;
Thomas Knox
NIST
4475 Whitney Place
Boulder Colorado 80305
1-303-554-0307
tomk...@nist.gov



Quoting Chuck Harris cfhar...@erols.com:


Try PRC68.com

Skip Withrow wrote:

Hello Time-Nuts,

I am in need of a manual for the Austron 2110 Disciplined Frequency
Standard.  Anyone have an electronic copy?  I have already tried KO4BB's
site and BAMA.

I would be more than happy to pay shipping so that I could copy/scan a
hard-copy if you have one.  Guaranteed to be returned promptly.

Regards,
Skip Withrow

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Need Austron 2110 manual

2009-05-27 Thread Thomas Linbeck
Tom,

You might inquire of Brooke Clark whether or not he has a copy.  He does
provide a CD of a number of Austron models.

Brooke Clarke
http://www.prc68.com

Tom Linbeck

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of tomk...@nist.gov
Sent: Wednesday, May 27, 2009 12:27 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Need Austron 2110 manual

Hi Skip;
I think I have a 2110A manual if you can't find one let me know.
I purpose I would drop it at a FedEx Kinkos where they could copy it  
and ship the copy to you.
You could pay them directly by credit card. Be aware this could end up  
expensive
I think this manual has fold out pages, and other things that require  
hand placement.
Best Wishes;
Thomas Knox
NIST
4475 Whitney Place
Boulder Colorado 80305
1-303-554-0307
tomk...@nist.gov



Quoting Chuck Harris cfhar...@erols.com:

 Try PRC68.com

 Skip Withrow wrote:
 Hello Time-Nuts,

 I am in need of a manual for the Austron 2110 Disciplined Frequency
 Standard.  Anyone have an electronic copy?  I have already tried KO4BB's
 site and BAMA.

 I would be more than happy to pay shipping so that I could copy/scan a
 hard-copy if you have one.  Guaranteed to be returned promptly.

 Regards,
 Skip Withrow

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to  
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Need Austron 2110 manual

2009-05-27 Thread Stanley Reynolds

The copy of Brooke Clarke's CD I have has the 2110B manual.

Stanley



- Original Message 
From: Thomas Linbeck tlinb...@sbcglobal.net
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Wednesday, May 27, 2009 12:54:59 PM
Subject: Re: [time-nuts] Need Austron 2110 manual

Tom,

You might inquire of Brooke Clark whether or not he has a copy.  He does
provide a CD of a number of Austron models.

Brooke Clarke
http://www.prc68.com

Tom Linbeck

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of tomk...@nist.gov
Sent: Wednesday, May 27, 2009 12:27 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Need Austron 2110 manual

Hi Skip;
I think I have a 2110A manual if you can't find one let me know.
I purpose I would drop it at a FedEx Kinkos where they could copy it  
and ship the copy to you.
You could pay them directly by credit card. Be aware this could end up  
expensive
I think this manual has fold out pages, and other things that require  
hand placement.
Best Wishes;
Thomas Knox
NIST
4475 Whitney Place
Boulder Colorado 80305
1-303-554-0307
tomk...@nist.gov



Quoting Chuck Harris cfhar...@erols.com:

 Try PRC68.com

 Skip Withrow wrote:
 Hello Time-Nuts,

 I am in need of a manual for the Austron 2110 Disciplined Frequency
 Standard.  Anyone have an electronic copy?  I have already tried KO4BB's
 site and BAMA.

 I would be more than happy to pay shipping so that I could copy/scan a
 hard-copy if you have one.  Guaranteed to be returned promptly.

 Regards,
 Skip Withrow

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to  
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Need Austron 2110 manual

2009-05-27 Thread Stanley Reynolds

oops that should be 2010B on the CD 

Sorry Stanley



- Original Message 
From: Stanley Reynolds stanley_reyno...@yahoo.com
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Wednesday, May 27, 2009 1:59:26 PM
Subject: Re: [time-nuts] Need Austron 2110 manual


The copy of Brooke Clarke's CD I have has the 2110B manual.

Stanley



- Original Message 
From: Thomas Linbeck tlinb...@sbcglobal.net
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Wednesday, May 27, 2009 12:54:59 PM
Subject: Re: [time-nuts] Need Austron 2110 manual

Tom,

You might inquire of Brooke Clark whether or not he has a copy.  He does
provide a CD of a number of Austron models.

Brooke Clarke
http://www.prc68.com

Tom Linbeck

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of tomk...@nist.gov
Sent: Wednesday, May 27, 2009 12:27 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Need Austron 2110 manual

Hi Skip;
I think I have a 2110A manual if you can't find one let me know.
I purpose I would drop it at a FedEx Kinkos where they could copy it  
and ship the copy to you.
You could pay them directly by credit card. Be aware this could end up  
expensive
I think this manual has fold out pages, and other things that require  
hand placement.
Best Wishes;
Thomas Knox
NIST
4475 Whitney Place
Boulder Colorado 80305
1-303-554-0307
tomk...@nist.gov



Quoting Chuck Harris cfhar...@erols.com:

 Try PRC68.com

 Skip Withrow wrote:
 Hello Time-Nuts,

 I am in need of a manual for the Austron 2110 Disciplined Frequency
 Standard.  Anyone have an electronic copy?  I have already tried KO4BB's
 site and BAMA.

 I would be more than happy to pay shipping so that I could copy/scan a
 hard-copy if you have one.  Guaranteed to be returned promptly.

 Regards,
 Skip Withrow

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to  
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Need Austron 2110 manual

2009-05-27 Thread ewkehren
I think I have one and did lend it out recently. I would be glad to ship it to 
you but I am in Europe and will not be back till the 12th of June. Check recent 
dialog on t-n I thing the person I lend it to did do a digital copy. If every 
thing else fails wait for three weeks. Bert


-Original Message-
From: tomk...@nist.gov
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Wed, 27 May 2009 1:27 pm
Subject: Re: [time-nuts] Need Austron 2110 manual


Hi Skip;?
I think I have a 2110A manual if you can't find one let me know.?
I purpose I would drop it at a FedEx Kinkos where they could copy it and ship 
the copy to you.?
You could pay them directly by credit card. Be aware this could end up 
expensive?
I think this manual has fold out pages, and other things that require hand 
placement.?
Best Wishes;?
Thomas Knox?
NIST?
4475 Whitney Place?
Boulder Colorado 80305?
1-303-554-0307?
tomk...@nist.gov?
?
?
Quoting Chuck Harris cfhar...@erols.com:?
?
 Try PRC68.com?
?
 Skip Withrow wrote:?
 Hello Time-Nuts,?
?
 I am in need of a manual for the Austron 2110 Disciplined Frequency?
 Standard. Anyone have an electronic copy? I have already tried KO4BB's?
 site and BAMA.?
?
 I would be more than happy to pay shipping so that I could copy/scan a?
 hard-copy if you have one. Guaranteed to be returned promptly.?
?
 Regards,?
 Skip Withrow?
?
 ___?
 time-nuts mailing list -- time-n...@febo.com?
 To unsubscribe, go to  
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts?
 and follow the instructions there.?
?
?
 ___?
 time-nuts mailing list -- time-n...@febo.com?
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts?
 and follow the instructions there.?
?
?
?
___?
time-nuts mailing list -- time-n...@febo.com?
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts?
and follow the instructions there.?

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP 3320B manual

2009-05-27 Thread Magnus Danielson

Fellow time-nuts,

Does anyone just happends to have a HP 3320B operations and service 
manual lying around?


I got a unit and wants to become more familiar with it.

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] FE 5680A supply voltage

2009-05-27 Thread Leigh L. Klotz, Jr. WA5ZNU
Is anyone familiar with the hazards of an undervoltage supply on an FE 5680A?

I've seen reports of users running it at a more commonly available 13.2V
instead of 15V.

What module might be most affected?  The lamp?

Thanks,
Leigh.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 3320B manual

2009-05-27 Thread Stanley Reynolds

PDF here : hxxp://www.hparchive.com/Manuals/HP-3320AB-Manual.pdf  (change xx to 
tt).

Stanley



- Original Message 
From: Magnus Danielson mag...@rubidium.dyndns.org
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Wednesday, May 27, 2009 4:54:51 PM
Subject: [time-nuts] HP 3320B manual

Fellow time-nuts,

Does anyone just happends to have a HP 3320B operations and service manual 
lying around?

I got a unit and wants to become more familiar with it.

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] FE 5680A supply voltage

2009-05-27 Thread Mark Sims

The FEI units use two paralleled low dropout regulators to drop the 15V input 
voltage down to 14V.  There is also a series protection diode that drops some 
more voltage.  Combined you need at least 14.7V input.  Anything less and the 
regulators will not regulate and the unit will not work within spec.  Anything 
much more than 15V and the regulators will overheat (you can supposedly go to 
18V,  but I would not do it without lots of cooling).
_
Windows Live™: Keep your life in sync.
http://windowslive.com/explore?ocid=TXT_TAGLM_BR_life_in_synch_052009
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] PCI IRIG receiver card for AIX?

2009-05-27 Thread Joe Gwinn
Does anybody know of any PCI cards that will receive IRIG-B time 
signals and come with an I/O driver for AIX (IBM's flavor of UNIX)? 
Industrial-grade commercial products are preferred.


Thanks,

Joe Gwinn

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] PCI IRIG receiver card for AIX?

2009-05-27 Thread Harlan Stenn
 Does anybody know of any PCI cards that will receive IRIG-B time 
 signals and come with an I/O driver for AIX (IBM's flavor of UNIX)? 
 Industrial-grade commercial products are preferred.

In the past I have, on one occasion, been able to use a regular audio
card for this, and I used the IRIG_AUDIO refclock with ntpd for this.

I don't know if that is what you have in mind but it should be an
option.

H

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] PCI IRIG receiver card for AIX?

2009-05-27 Thread Lux, James P
Symmetricom bc635pci-V2 or -U?  PCI-SG 2U
(some of the old TrueTime products, I'm sure)

It does IRIG... You'll have to call Symmetricom to see about drivers.. It has 
Win, Linux, and Solaris.


On 5/27/09 7:26 PM, Joe Gwinn joegw...@comcast.net wrote:

Does anybody know of any PCI cards that will receive IRIG-B time
signals and come with an I/O driver for AIX (IBM's flavor of UNIX)?
Industrial-grade commercial products are preferred.

Thanks,

Joe Gwinn

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Broken Ovenaire OSC 85-50

2009-05-27 Thread J. L. Trantham
Bruce,

Thanks for your analysis.  Is this all in your head or do you use a computer
program?

If I understand correctly, you are using the term 'buffer' to describe what
I call the 'Output Board'.  If so, then the Output Board seems to be doing
what it is supposed to do and the problem is a low output from the
Oscillator Board.  Any thoughts as to the cause of the low output?
Capacitors changing value, resistors changing value, inductors changing
value (by the way, I need a refresher on how to read the color code of
inductors), transistors loosing gain?  I have not measured the temperature
but the crystal has '79 C' written on it.  The capacitors appear to be
dipped silvered mica and the resistors appear to be carbon.

I can send a picture of both boards if that would help.

Joe

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Bruce Griffiths
Sent: Tuesday, May 26, 2009 10:43 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50

Joe

A quick analysis if the buffer indicates that the gain with a 50 ohm
load is about 1x.
The gain into a 1M load should be about 4x.
Simulating the output stage indicates that the maximum output achievable
is about 1.6V pp into 1M or 0.4Vpp into 50 ohms.
However distortion is a bit high. This requires about 0.4Vpp at the
input to the buffer.

The dc level on the red wire is reasonably consistent with quick
simulation of the level without oscillation at an oven temperature of
90-100C.
When oscillating the current drawn by the oscillator may increase
slightly, bringing the value closer to that observed.

Bruce

J. L. Trantham wrote:
 Well, a little work with a scalpel and a small flat bladed screw driver
and
 I was able to open the package.  Inside is what I call an 'Output Board'
 that contains the output circuit and the 'FINE' frequency adjust along
with
 the 'Oven' which has the 'COARSE' frequency adjust.  Attached to the
'Oven'
 is what I call the 'Oven Controller Board'.  With the removal of 3 screws
 and the 'nut' on the 'COARSE' frequency adjust, the 'Oscillator Board' can
 be removed.

 I have 'derived' the schematic from inspections of the 'Output Board' and
 the 'Oscillator Board'.  I made no attempt at the 'Oven Controller Board'
 since it seems to be operating ok.  More on this below.

 I have attached the schematics of the output board and oscillator boards.
I
 made some measurements of DC and RF voltages on the output board.  I have
 multiple pictures that are too large to attach.  If anyone would like to
see
 these, I can send them.

 At start up, the 85-50 draws about 0.5A then falls to about 0.25A as the
 'oven' heats up suggesting the temperature regulator circuit is
functional.
 The 'COARSE' and 'FINE' frequency adjusts work and the frequency can be
 brought spot on as compared to my GPSDO.

 I have ordered new transistors in the hope that these are the problem.  I
am
 a little puzzled by the drop in voltage at the 'Red' wire to the oven.  It
 serves both the 'Oscillator Board' and the 'Oven Controller Board'.  2.5
VDC
 seems a little low for the oscillator.

 Also, the gain of the output stage of the output board seems a little low.
 0.1 V P-P goes to 0.36 V P-P.

 I have a second, good, 85-50 but would like to avoid opening it if
possible.

 Any help would be appreciated.  I hope I made no mistakes on the
schematics.

 Joe




 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Gordon Batey
 Sent: Sunday, May 24, 2009 6:50 PM
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50

 I had an earlier model 20 mhx ovenaire osc that had very low output.  I
 opened it up with a scalpel and found the TTL chip had failed and was only
 letting a small amount of the osc signal thru.  I was able to use an
 external TTl (74LS20 I think) to restore it.  As I recall I was unable to
 readily get to the chip so I replaced it with an external one.  

 Good luck.  

 Gordon



 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
   
 

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Broken Ovenaire OSC 85-50

2009-05-27 Thread J. L. Trantham
Bruce,

I sent you two pictures directly.  I hope I did not choke your inbox.  They
were a total of about 6 MB.

If so, I can probably reduce the quality and get a smaller file.

Joe

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Bruce Griffiths
Sent: Wednesday, May 27, 2009 11:03 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50

Joe

Low output can be caused by a high impedance path, an open circuit
component, an open solder joint, a low value cap, low gain transistor, etc.
The oscillator limits by either saturating or better cutting off the
oscillator transistor (would need to check to see which but its more
likely to be the latter).
If oscillator transistor cutoff is the limiting mechanism then too low
transistor collector current could be one cause.


Bruce

J. L. Trantham wrote:
 Bruce,

 Thanks for your analysis.  Is this all in your head or do you use a
computer
 program?

   
Both.
I used LTSpice to check.

 If I understand correctly, you are using the term 'buffer' to describe
what
 I call the 'Output Board'.  
Yes.
 If so, then the Output Board seems to be doing
 what it is supposed to do and the problem is a low output from the
 Oscillator Board.  Any thoughts as to the cause of the low output?
 Capacitors changing value, resistors changing value, inductors changing
 value (by the way, I need a refresher on how to read the color code of
 inductors), transistors loosing gain?  I have not measured the temperature
 but the crystal has '79 C' written on it.  The capacitors appear to be
 dipped silvered mica and the resistors appear to be carbon.

 I can send a picture of both boards if that would help.

   
Yes that may help.

 Joe

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Bruce Griffiths
 Sent: Tuesday, May 26, 2009 10:43 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50

 Joe

 A quick analysis if the buffer indicates that the gain with a 50 ohm
 load is about 1x.
 The gain into a 1M load should be about 4x.
 Simulating the output stage indicates that the maximum output achievable
 is about 1.6V pp into 1M or 0.4Vpp into 50 ohms.
 However distortion is a bit high. This requires about 0.4Vpp at the
 input to the buffer.

 The dc level on the red wire is reasonably consistent with quick
 simulation of the level without oscillation at an oven temperature of
 90-100C.
 When oscillating the current drawn by the oscillator may increase
 slightly, bringing the value closer to that observed.

 Bruce

 J. L. Trantham wrote:
   
 Well, a little work with a scalpel and a small flat bladed screw driver
 
 and
   
 I was able to open the package.  Inside is what I call an 'Output Board'
 that contains the output circuit and the 'FINE' frequency adjust along
 
 with
   
 the 'Oven' which has the 'COARSE' frequency adjust.  Attached to the
 
 'Oven'
   
 is what I call the 'Oven Controller Board'.  With the removal of 3 screws
 and the 'nut' on the 'COARSE' frequency adjust, the 'Oscillator Board'
can
 be removed.

 I have 'derived' the schematic from inspections of the 'Output Board' and
 the 'Oscillator Board'.  I made no attempt at the 'Oven Controller Board'
 since it seems to be operating ok.  More on this below.

 I have attached the schematics of the output board and oscillator boards.
 
 I
   
 made some measurements of DC and RF voltages on the output board.  I have
 multiple pictures that are too large to attach.  If anyone would like to
 
 see
   
 these, I can send them.

 At start up, the 85-50 draws about 0.5A then falls to about 0.25A as the
 'oven' heats up suggesting the temperature regulator circuit is
 
 functional.
   
 The 'COARSE' and 'FINE' frequency adjusts work and the frequency can be
 brought spot on as compared to my GPSDO.

 I have ordered new transistors in the hope that these are the problem.  I
 
 am
   
 a little puzzled by the drop in voltage at the 'Red' wire to the oven.
It
 serves both the 'Oscillator Board' and the 'Oven Controller Board'.  2.5
 
 VDC
   
 seems a little low for the oscillator.

 Also, the gain of the output stage of the output board seems a little
low.
 0.1 V P-P goes to 0.36 V P-P.

 I have a second, good, 85-50 but would like to avoid opening it if
 
 possible.
   
 Any help would be appreciated.  I hope I made no mistakes on the
 
 schematics.
   
 Joe




 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Gordon Batey
 Sent: Sunday, May 24, 2009 6:50 PM
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50

 I had an earlier model 20 mhx ovenaire osc that had very low output.  I
 opened it up with a scalpel and found the TTL chip had failed and was
only
 letting a small amount 

Re: [time-nuts] Broken Ovenaire OSC 85-50

2009-05-27 Thread Bruce Griffiths
Joe

In box is fine, 6MB is relatively small and not a problem.
The brown caps look like they might be silver mica.
It may be worth resoldering all the solder joints on these caps as they
can sometimes go high impedance over time.

The ceramic caps marked 103 are 10,000pF = 10nF = 0.01uF.

If you have a current probe and oscilloscope it would be useful to look
at the RF current flowing in the crystal, oscillator transistor etc.

Bruce

J. L. Trantham wrote:
 Bruce,

 I sent you two pictures directly.  I hope I did not choke your inbox.  They
 were a total of about 6 MB.

 If so, I can probably reduce the quality and get a smaller file.

 Joe

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Bruce Griffiths
 Sent: Wednesday, May 27, 2009 11:03 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50

 Joe

 Low output can be caused by a high impedance path, an open circuit
 component, an open solder joint, a low value cap, low gain transistor, etc.
 The oscillator limits by either saturating or better cutting off the
 oscillator transistor (would need to check to see which but its more
 likely to be the latter).
 If oscillator transistor cutoff is the limiting mechanism then too low
 transistor collector current could be one cause.


 Bruce

 J. L. Trantham wrote:
   
 Bruce,

 Thanks for your analysis.  Is this all in your head or do you use a
 
 computer
   
 program?

   
 
 Both.
 I used LTSpice to check.

   
 If I understand correctly, you are using the term 'buffer' to describe
 
 what
   
 I call the 'Output Board'.  
 
 Yes.
   
 If so, then the Output Board seems to be doing
 what it is supposed to do and the problem is a low output from the
 Oscillator Board.  Any thoughts as to the cause of the low output?
 Capacitors changing value, resistors changing value, inductors changing
 value (by the way, I need a refresher on how to read the color code of
 inductors), transistors loosing gain?  I have not measured the temperature
 but the crystal has '79 C' written on it.  The capacitors appear to be
 dipped silvered mica and the resistors appear to be carbon.

 I can send a picture of both boards if that would help.

   
 
 Yes that may help.

   
 Joe

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Bruce Griffiths
 Sent: Tuesday, May 26, 2009 10:43 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Broken Ovenaire OSC 85-50

 Joe

 A quick analysis if the buffer indicates that the gain with a 50 ohm
 load is about 1x.
 The gain into a 1M load should be about 4x.
 Simulating the output stage indicates that the maximum output achievable
 is about 1.6V pp into 1M or 0.4Vpp into 50 ohms.
 However distortion is a bit high. This requires about 0.4Vpp at the
 input to the buffer.

 The dc level on the red wire is reasonably consistent with quick
 simulation of the level without oscillation at an oven temperature of
 90-100C.
 When oscillating the current drawn by the oscillator may increase
 slightly, bringing the value closer to that observed.

 Bruce

 J. L. Trantham wrote:
   
 
 Well, a little work with a scalpel and a small flat bladed screw driver
 
   
 and
   
 
 I was able to open the package.  Inside is what I call an 'Output Board'
 that contains the output circuit and the 'FINE' frequency adjust along
 
   
 with
   
 
 the 'Oven' which has the 'COARSE' frequency adjust.  Attached to the
 
   
 'Oven'
   
 
 is what I call the 'Oven Controller Board'.  With the removal of 3 screws
 and the 'nut' on the 'COARSE' frequency adjust, the 'Oscillator Board'
   
 can
   
 be removed.

 I have 'derived' the schematic from inspections of the 'Output Board' and
 the 'Oscillator Board'.  I made no attempt at the 'Oven Controller Board'
 since it seems to be operating ok.  More on this below.

 I have attached the schematics of the output board and oscillator boards.
 
   
 I
   
 
 made some measurements of DC and RF voltages on the output board.  I have
 multiple pictures that are too large to attach.  If anyone would like to
 
   
 see
   
 
 these, I can send them.

 At start up, the 85-50 draws about 0.5A then falls to about 0.25A as the
 'oven' heats up suggesting the temperature regulator circuit is
 
   
 functional.
   
 
 The 'COARSE' and 'FINE' frequency adjusts work and the frequency can be
 brought spot on as compared to my GPSDO.

 I have ordered new transistors in the hope that these are the problem.  I
 
   
 am
   
 
 a little puzzled by the drop in voltage at the 'Red' wire to the oven.
   
 It
   
 serves both the 'Oscillator Board' and the 'Oven Controller Board'.  2.5
 
   
 VDC
   
 
 seems a little low for the oscillator.

 Also, the gain of