Re: [time-nuts] Rubidium Performance: DDS noise effect on 10 MHz

2012-02-04 Thread Rex

On 2/4/2012 12:17 AM, John Miles wrote:

In the current 5680A units, the 10 MHz output comes from the 60 MHz VCXO
(divided by 6 in CPLD) and not direct from a DDS. If my architecture
understanding is right, the DDS signal output is mixed with the VCXO

output

only at the 114th harmonic of 60 MHz, and it's the PLL (looking at the

optical

signal from the Rb) that drives the VCXO to keep it lined up.  As I

understand

it, DDS phase noise should be divided by a factor of 6*114 by the time it
appears at the 10 MHz output, and at larger frequency offsets the

amplitude

should also be (significantly) reduced by the PLL loop filter.

I don't have any phase noise measurement tools myself, so this is just an
academic argument, but if there is significantly more noise on the 10 MHz
than expected for a 60 MHz VCXO, I wonder if it's just inadequate

filtering of

an internal power rail.  Is the unit under test being driven by a linear,

or

switching supply?

What made me suspect the AD9832 is that the PN/spur behavior is so close to
what I'd expect from it.  Analog Devices' own plots contain a lot of strong
spurs, and the chip's actual SFDR spec is in the -70 dBc range that I'm
seeing (blue trace attached).  But yes, you're right, it would make a lot
more sense to use the DDS as an offset generator for the control loop,
rather than as an output device.

I tried it with a couple of different supplies, linear and otherwise, and
the noise didn't change meaningfully.  Since it's so far out of line with
FEI's specs, I have to believe that I either got a couple of bad examples --
although they lock very quickly and look essentially unused inside -- or I
don't have the pinout right, and am using the wrong ground or something like
that.  Seems pretty unlikely that this is normal operation.   Let's see what
JohnA comes up with.

-- john



John,

From your comments, it sounds like you may be measuring one of the 
earlier 5680A's that could be tuned over a large range with a DDS output.


These earlier ones had a 50.25 MHz internal osc which was locked after 
multiplication to the rubidium frequency. Then there was a DDS on the 
output that could be programmed over a wide frequency range.


The newer ones have a 60 MHz internal oscillator. This is multiplied up 
to 6840 MHz and subtractively mixed with ~5.3125 MHz from a DDS to get 
to the rubidium frequency. The feedback from the rubidium cell locks the 
60 Hz which is divided by 6 for the 10 MHz output. No DDS on the output 
generation. The DDS in the loop can digitally adjust the rubidium lock 
frequency that tunes the 60 MHz, to fine tune the output only around 10 
MHz.


Do you know which type of 5680A you are measuring?


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Rubidium Performance: DDS noise effect on 10 MHz

2012-02-04 Thread John Miles
  From your comments, it sounds like you may be measuring one of the
 earlier 5680A's that could be tuned over a large range with a DDS output.
 
 These earlier ones had a 50.25 MHz internal osc which was locked after
 multiplication to the rubidium frequency. Then there was a DDS on the
 output that could be programmed over a wide frequency range.

That sounds plausible.   I haven't taken the time to bring myself 100% up to
speed with the many different variations and options for these little boxes,
but I could swear I'm looking at a low-resolution, unfiltered, uncleaned-up
DDS.

 The newer ones have a 60 MHz internal oscillator. This is multiplied up
 to 6840 MHz and subtractively mixed with ~5.3125 MHz from a DDS to get
 to the rubidium frequency. The feedback from the rubidium cell locks the
 60 Hz which is divided by 6 for the 10 MHz output. No DDS on the output
 generation. The DDS in the loop can digitally adjust the rubidium lock
 frequency that tunes the 60 MHz, to fine tune the output only around 10
 MHz.
 
 Do you know which type of 5680A you are measuring?

This one is marked S/N 0339-65969, purchased from
http://www.ebay.com/itm/260930018124 .  It seems to have the pinout
documented at http://vk2xv.djirra.com/tech_rubidium.htm for serial
0127-96634 (pin 1=V+, pin 2=GND, pin 3=+5V, pin 7=RF out.)   However, VK2XV
claims that s/n 96634 was non-programmable.  I haven't tried hooking up a
serial terminal to mine, but I did notice that there was some negative
voltage on one of the other pins (9?) that might correspond to an RS-232
signal level.

Can't complain too much at $40/each, anyway  I believe the seller's claim
that these are new or nearly so, even though they look like they've been
removed from an installation.  The internal foam insulation is usually
discolored from heat, while it looks great in the two units that I bought
from this seller.

-- john



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Rubidium Performance: DDS noise effect on 10 MHz

2012-02-04 Thread Javier Herrero

Hello, John,

Your phase noise plot seems quite similar to that sent by Said some time 
ago (4 Jan), that looks like a lot of spurs quite evenly separated. It 
is likely from the price and p/n that your unit is one of the newer 
ones with 60MHz xtal and the DDS inside the loop - and I would think 
that the pn from the DDS would be removed by the loop. But since the 
division from 60MHz is done by a CPLD, perhaps the signal is corrupted 
by this CPLD, or perhaps it is corrupted by a switching regulator inside 
the unit.


The 60MHz signal is available inside the unit, in a small coaxial 
connector. Perhaps it would be interesting to also analyze it.


Regards,

Javier

El 04/02/2012 10:43, John Miles escribió:

   From your comments, it sounds like you may be measuring one of the
earlier 5680A's that could be tuned over a large range with a DDS output.

These earlier ones had a 50.25 MHz internal osc which was locked after
multiplication to the rubidium frequency. Then there was a DDS on the
output that could be programmed over a wide frequency range.


That sounds plausible.   I haven't taken the time to bring myself 100% up to
speed with the many different variations and options for these little boxes,
but I could swear I'm looking at a low-resolution, unfiltered, uncleaned-up
DDS.


The newer ones have a 60 MHz internal oscillator. This is multiplied up
to 6840 MHz and subtractively mixed with ~5.3125 MHz from a DDS to get
to the rubidium frequency. The feedback from the rubidium cell locks the
60 Hz which is divided by 6 for the 10 MHz output. No DDS on the output
generation. The DDS in the loop can digitally adjust the rubidium lock
frequency that tunes the 60 MHz, to fine tune the output only around 10
MHz.

Do you know which type of 5680A you are measuring?


This one is marked S/N 0339-65969, purchased from
http://www.ebay.com/itm/260930018124 .  It seems to have the pinout
documented at http://vk2xv.djirra.com/tech_rubidium.htm for serial
0127-96634 (pin 1=V+, pin 2=GND, pin 3=+5V, pin 7=RF out.)   However, VK2XV
claims that s/n 96634 was non-programmable.  I haven't tried hooking up a
serial terminal to mine, but I did notice that there was some negative
voltage on one of the other pins (9?) that might correspond to an RS-232
signal level.

Can't complain too much at $40/each, anyway  I believe the seller's claim
that these are new or nearly so, even though they look like they've been
removed from an installation.  The internal foam insulation is usually
discolored from heat, while it looks great in the two units that I bought
from this seller.

-- john



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Rubidium Performance: DDS noise effect on 10 MHz

2012-02-04 Thread EWKehren
Maybe his are new. But I bought one from ggg*fitting that was listed as new 
 turned out very used, and in conversation with them they had to admit they 
do  not know what they have. If it still has its original factory setting, 
measuring  the frequency will tell you if it had extensive running time or 
not.
Bert Kehren
 
 
In a message dated 2/4/2012 4:44:20 A.M. Eastern Standard Time,  
jmi...@pop.net writes:

  From your comments, it sounds like you may be measuring  one of the
 earlier 5680A's that could be tuned over a large range with  a DDS output.
 
 These earlier ones had a 50.25 MHz internal osc  which was locked after
 multiplication to the rubidium frequency. Then  there was a DDS on the
 output that could be programmed over a wide  frequency range.

That sounds plausible.   I haven't taken the  time to bring myself 100% up 
to
speed with the many different variations  and options for these little 
boxes,
but I could swear I'm looking at a  low-resolution, unfiltered, uncleaned-up
DDS.

 The newer ones  have a 60 MHz internal oscillator. This is multiplied up
 to 6840 MHz  and subtractively mixed with ~5.3125 MHz from a DDS to get
 to the  rubidium frequency. The feedback from the rubidium cell locks the
 60  Hz which is divided by 6 for the 10 MHz output. No DDS on the output
  generation. The DDS in the loop can digitally adjust the rubidium lock
  frequency that tunes the 60 MHz, to fine tune the output only around  10
 MHz.
 
 Do you know which type of 5680A you are  measuring?

This one is marked S/N 0339-65969, purchased  from
http://www.ebay.com/itm/260930018124 .  It seems to have the  pinout
documented at http://vk2xv.djirra.com/tech_rubidium.htm for  serial
0127-96634 (pin 1=V+, pin 2=GND, pin 3=+5V, pin 7=RF  out.)   However, VK2XV
claims that s/n 96634 was  non-programmable.  I haven't tried hooking up a
serial terminal to  mine, but I did notice that there was some negative
voltage on one of the  other pins (9?) that might correspond to an RS-232
signal  level.

Can't complain too much at $40/each, anyway  I believe the  seller's claim
that these are new or nearly so, even though they look like  they've been
removed from an installation.  The internal foam  insulation is usually
discolored from heat, while it looks great in the two  units that I bought
from this seller.

--  john



___
time-nuts  mailing list -- time-nuts@febo.com
To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Rubidium Performance: DDS noise effect on 10 MHz

2012-02-03 Thread beale
  From: John Miles jmi...@pop.net
  So far, the FE-5680A is doing well in the ADEV department, but its AD9832
  DDS chip adds some substantial noise and spurs to the 10 MHz output. [...]
  The overall SSB C/N ratio I'm seeing between 1 Hz - 100 kHz is pretty much
  in line with Analog Devices' data sheet, as are the spur levels.  

In the current 5680A units, the 10 MHz output comes from the 60 MHz VCXO 
(divided by 6 in CPLD) and not direct from a DDS. If my architecture 
understanding is right, the DDS signal output is mixed with the VCXO output 
only at the 114th harmonic of 60 MHz, and it's the PLL (looking at the optical 
signal from the Rb) that drives the VCXO to keep it lined up.  As I understand 
it, DDS phase noise should be divided by a factor of 6*114 by the time it 
appears at the 10 MHz output, and at larger frequency offsets the amplitude 
should also be (significantly) reduced by the PLL loop filter.

I don't have any phase noise measurement tools myself, so this is just an 
academic argument, but if there is significantly more noise on the 10 MHz than 
expected for a 60 MHz VCXO, I wonder if it's just inadequate filtering of an 
internal power rail.  Is the unit under test being driven by a linear, or 
switching supply?

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.