Re: [time-nuts] Considerations When Using The SR620

2012-12-17 Thread Volker Esper


I'm curious, which way you went and which accuracy you achieved... :-)
Can you tell us?
Volker


Am 06.12.2012 19:10, schrieb Paul DeStefano:

On Tuesday, 4 December 2012, Tom Van Baak wrote:

We are using the SR620 to measure the interval between 1PPS signals from
two clocks. One is the Septentrio PolaRx4 GPS receiver and the other
is a
Rubidium clock.

Many Thanks,
Paul

1) If you are making frequency measurements, the warm-up of the
internal oscillator is the major factor limiting accuracy. ...
Plotting digits of precision as a function of warm-up time would make
a very educational graph you could tape to the top of your SR620.

2) If you are making time interval measurements and using an external
standard, the warm-up time will also affect the accuracy of your TI
measurements, but to a far lesser degree. Here are informal results
for TI (time interval) mode after a 5 minute power-down (see also
attached plots):

- if you need 1 ns accuracy, you can use the SR620 immediately after
power-up
- if you need 100 ps accuracy, wait 2+ minutes
- if you need 10 ps accuracy, wait 15+ minutes
- if you need 1 ps accuracy, you need a seriously stable lab
environment or a different counter.

Given that you plan to use the SR620 with high-end GPS gear I would
suggest you try this quick experiment for yourself to see what *your*
SR620 does, with *your* inputs, in *your* environment. Your numbers
will come out different than mine; but the methodology is the same.
Your procedures can then be based on measurement and confidence
instead of guesswork and folklore.


Tom  Co.,

Thank you! These plots are excellent and will be very helpful. You are
quite right; we should do the test ourselves. We will definitely do
that. Obviously, there is not need to worry, as we can characterize the
instrument behavior ourselves, which is probably necessary anyway if
we're going to publish these measurements with error values.

Many Thanks,
Paul





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Considerations When Using The SR620

2012-12-06 Thread Paul DeStefano

On Tuesday, 4 December 2012, Tom Van Baak wrote:

We are using the SR620 to measure the interval between 1PPS signals from
two clocks.  One is the Septentrio PolaRx4 GPS receiver and the other is a
Rubidium clock.

Many Thanks,
Paul
1) If you are making frequency measurements, the warm-up of the internal 
oscillator is the major factor limiting accuracy. ... Plotting digits of 
precision as a function of warm-up time would make a very educational 
graph you could tape to the top of your SR620.


2) If you are making time interval measurements and using an external 
standard, the warm-up time will also affect the accuracy of your TI 
measurements, but to a far lesser degree. Here are informal results for 
TI (time interval) mode after a 5 minute power-down (see also attached 
plots):


- if you need 1 ns accuracy, you can use the SR620 immediately after power-up
- if you need 100 ps accuracy, wait 2+ minutes
- if you need 10 ps accuracy, wait 15+ minutes
- if you need 1 ps accuracy, you need a seriously stable lab environment or a 
different counter.

Given that you plan to use the SR620 with high-end GPS gear I would 
suggest you try this quick experiment for yourself to see what *your* 
SR620 does, with *your* inputs, in *your* environment. Your numbers will 
come out different than mine; but the methodology is the same. Your 
procedures can then be based on measurement and confidence instead of 
guesswork and folklore.


Tom  Co.,

	Thank you!  These plots are excellent and will be very helpful. 
You are quite right; we should do the test ourselves.  We will definitely 
do that.  Obviously, there is not need to worry, as we can characterize 
the instrument behavior ourselves, which is probably necessary anyway if 
we're going to publish these measurements with error values.


Many Thanks,
Paul

--
Paul DeStefano


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Considerations When Using The SR620

2012-12-06 Thread Paul DeStefano

Volker,

	That's a great question and I'm afraid I don't have a good answer 
for you.  If pressed, I would estimate less than 100ps.
	The error of this measurement contributes to the error in our 
final measurement which has many components.  I haven't worked out an 
error budget for each contributor.  Our goal is a final error of less than 
10ns.  But a few contributors are already expected to be around 1ns, so 
this error really needs to be less than 1ns, preferably less than 100ps.


Cheers,
Paul

On Tuesday, 4 December 2012, Volker Esper wrote:
I agree. Since Paul want's to use an SR620 I presume he needs precision. 
Otherwise almost any TIC with a fairly stable osc would do, for example 
one with a battery backup. So I further presume that he needs nearly the 
full accuracy / stability. But that's just speculation, surely Paul can 
answere this question?


--
Paul DeStefano


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Considerations When Using The SR620

2012-12-04 Thread Tom Van Baak
 We are using the SR620 to measure the interval between 1PPS signals from 
 two clocks.  One is the Septentrio PolaRx4 GPS receiver and the other is a 
 Rubidium clock.
 
 Many Thanks,
 Paul

Hi Paul,

1) If you are making frequency measurements, the warm-up of the internal 
oscillator is the major factor limiting accuracy. This doesn't mean you have 
wait 12 hours. For example, if all you need is 6 digits of precision, a 
one-minute warm-up may be sufficient. For each further digit of precision you 
wait longer. You can probably get 9 digits with 1/2 hour of warm-up. It depends 
on the oscillator. Plotting digits of precision as a function of warm-up time 
would make a very educational graph you could tape to the top of your SR620.

2) If you are making time interval measurements and using an external standard, 
the warm-up time will also affect the accuracy of your TI measurements, but to 
a far lesser degree. Here are informal results for TI (time interval) mode 
after a 5 minute power-down (see also attached plots):

- if you need 1 ns accuracy, you can use the SR620 immediately after power-up
- if you need 100 ps accuracy, wait 2+ minutes
- if you need 10 ps accuracy, wait 15+ minutes
- if you need 1 ps accuracy, you need a seriously stable lab environment or a 
different counter.

Given that you plan to use the SR620 with high-end GPS gear I would suggest you 
try this quick experiment for yourself to see what *your* SR620 does, with 
*your* inputs, in *your* environment. Your numbers will come out different than 
mine; but the methodology is the same. Your procedures can then be based on 
measurement and confidence instead of guesswork and folklore.

Note also if your data collection is automated, there's really no reason to 
wait after power-up at all. Just collect data as soon as you can and skip the 
predetermined number of samples. I can send you the SR620 GPIB scripts I used 
for my test. This way you have a record of the warm-up settling time itself, 
which further gives you confidence in the data that follows.

/tvbattachment: sr620-warmup-tic-500ps.pngattachment: sr620-warmup-tic-50ps.png___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Considerations When Using The SR620

2012-12-04 Thread Volker Esper


Tom,

I agree. Since Paul want's to use an SR620 I presume he needs precision. 
Otherwise almost any TIC with a fairly stable osc would do, for example 
one with a battery backup. So I further presume that he needs nearly the 
full accuracy / stability. But that's just speculation, surely Paul can 
answere this question?


Volker


Am 04.12.2012 14:11, schrieb Tom Van Baak:

We are using the SR620 to measure the interval between 1PPS signals from
two clocks.  One is the Septentrio PolaRx4 GPS receiver and the other is a
Rubidium clock.

Many Thanks,
Paul


Hi Paul,

1) If you are making frequency measurements, the warm-up of the internal 
oscillator is the major factor limiting accuracy. This doesn't mean you have 
wait 12 hours. For example, if all you need is 6 digits of precision, a 
one-minute warm-up may be sufficient. For each further digit of precision you 
wait longer. You can probably get 9 digits with 1/2 hour of warm-up. It depends 
on the oscillator. Plotting digits of precision as a function of warm-up time 
would make a very educational graph you could tape to the top of your SR620.

2) If you are making time interval measurements and using an external standard, 
the warm-up time will also affect the accuracy of your TI measurements, but to 
a far lesser degree. Here are informal results for TI (time interval) mode 
after a 5 minute power-down (see also attached plots):

- if you need 1 ns accuracy, you can use the SR620 immediately after power-up
- if you need 100 ps accuracy, wait 2+ minutes
- if you need 10 ps accuracy, wait 15+ minutes
- if you need 1 ps accuracy, you need a seriously stable lab environment or a 
different counter.

Given that you plan to use the SR620 with high-end GPS gear I would suggest you 
try this quick experiment for yourself to see what *your* SR620 does, with 
*your* inputs, in *your* environment. Your numbers will come out different than 
mine; but the methodology is the same. Your procedures can then be based on 
measurement and confidence instead of guesswork and folklore.

Note also if your data collection is automated, there's really no reason to 
wait after power-up at all. Just collect data as soon as you can and skip the 
predetermined number of samples. I can send you the SR620 GPIB scripts I used 
for my test. This way you have a record of the warm-up settling time itself, 
which further gives you confidence in the data that follows.

/tvb



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Considerations When Using The SR620

2012-12-04 Thread Eric Garner
I would be interested in those scripts as well. as soon as the replacement
TCXO for my sr620 arrives I can put it back together and this would be fun
to try.

-eric

On Tue, Dec 4, 2012 at 5:11 AM, Tom Van Baak t...@leapsecond.com wrote:

  We are using the SR620 to measure the interval between 1PPS signals from
  two clocks.  One is the Septentrio PolaRx4 GPS receiver and the other is
 a
  Rubidium clock.
 
  Many Thanks,
  Paul

 Hi Paul,

 1) If you are making frequency measurements, the warm-up of the internal
 oscillator is the major factor limiting accuracy. This doesn't mean you
 have wait 12 hours. For example, if all you need is 6 digits of precision,
 a one-minute warm-up may be sufficient. For each further digit of precision
 you wait longer. You can probably get 9 digits with 1/2 hour of warm-up. It
 depends on the oscillator. Plotting digits of precision as a function of
 warm-up time would make a very educational graph you could tape to the top
 of your SR620.

 2) If you are making time interval measurements and using an external
 standard, the warm-up time will also affect the accuracy of your TI
 measurements, but to a far lesser degree. Here are informal results for TI
 (time interval) mode after a 5 minute power-down (see also attached plots):

 - if you need 1 ns accuracy, you can use the SR620 immediately after
 power-up
 - if you need 100 ps accuracy, wait 2+ minutes
 - if you need 10 ps accuracy, wait 15+ minutes
 - if you need 1 ps accuracy, you need a seriously stable lab environment
 or a different counter.

 Given that you plan to use the SR620 with high-end GPS gear I would
 suggest you try this quick experiment for yourself to see what *your* SR620
 does, with *your* inputs, in *your* environment. Your numbers will come out
 different than mine; but the methodology is the same. Your procedures can
 then be based on measurement and confidence instead of guesswork and
 folklore.

 Note also if your data collection is automated, there's really no reason
 to wait after power-up at all. Just collect data as soon as you can and
 skip the predetermined number of samples. I can send you the SR620 GPIB
 scripts I used for my test. This way you have a record of the warm-up
 settling time itself, which further gives you confidence in the data that
 follows.

 /tvb
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




-- 
--Eric
_
Eric Garner
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Considerations When Using The SR620

2012-12-02 Thread Paul DeStefano

All,

	The following comment appeared on this list recently and it scared 
me a little:


Though the SR620 TIC is a great instrument when hunting the pico seconds 
we have to realize, that it's a thermal design desaster (I have to 
apologize to all sr620 friends). I have to run it for at least 12 
hoursif not 24 to be shure, that every single part is at a more or less 
stationary thermal state. Some (NERC) say ...never switch it off.


I assume this instability is due to the instability of the internal 
frequency standard.  There are two options for the SR620: the standard 
oscillator and an ovenized oscillator.  In fact, in our measurements, we 
plan to use a Cesium frequency standard as the timebase to our SR620. 
Does this anecdotal warning apply generally to the instrument or mainly to 
the use of the internal standard oscillator?


We are using the SR620 to measure the interval between 1PPS signals from 
two clocks.  One is the Septentrio PolaRx4 GPS receiver and the other is a 
Rubidium clock.


Many Thanks,
Paul

--
Paul DeStefano


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Considerations When Using The SR620

2012-12-02 Thread Bob Camp
Hi

Some of the problem comes from the OCXO. Some of it also comes from the 
stability of the internal circuitry. They do some interesting things with delay 
lines and the like in the SR620. Best bet is to let it warm up for a while 
before you need it to perform at it's top level of performance. 

Bob

On Dec 2, 2012, at 2:45 PM, Paul DeStefano 
paul.destef...@willamettealumni.com wrote:

 All,
 
   The following comment appeared on this list recently and it scared me a 
 little:
 
 Though the SR620 TIC is a great instrument when hunting the pico seconds we 
 have to realize, that it's a thermal design desaster (I have to apologize to 
 all sr620 friends). I have to run it for at least 12 hoursif not 24 to be 
 shure, that every single part is at a more or less stationary thermal state. 
 Some (NERC) say ...never switch it off.
 
 I assume this instability is due to the instability of the internal frequency 
 standard.  There are two options for the SR620: the standard oscillator and 
 an ovenized oscillator.  In fact, in our measurements, we plan to use a 
 Cesium frequency standard as the timebase to our SR620. Does this anecdotal 
 warning apply generally to the instrument or mainly to the use of the 
 internal standard oscillator?
 
 We are using the SR620 to measure the interval between 1PPS signals from two 
 clocks.  One is the Septentrio PolaRx4 GPS receiver and the other is a 
 Rubidium clock.
 
 Many Thanks,
 Paul
 
 -- 
 Paul DeStefano
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Considerations When Using The SR620

2012-12-02 Thread Charles P. Steinmetz

Paul wrote:

The following comment appeared on this list recently and it scared 
me a little:


Though the SR620 TIC is a great instrument when hunting the pico 
seconds we have to realize, that it's a thermal design desaster (I 
have to apologize to all sr620 friends). I have to run it for at 
least 12 hoursif not 24 to be shure, that every single part is at a 
more or less stationary thermal state. Some (NERC) say ...never 
switch it off.


I assume this instability is due to the instability of the internal 
frequency standard.  *  *  *  In fact, in our measurements, we plan 
to use a Cesium frequency standard as the timebase to our SR620. 
Does this anecdotal warning apply generally to the instrument or 
mainly to the use of the internal standard oscillator?


I concur with the comment above that the thermal design of the 620 
could have been better -- the sensing thermistor is in an exhaust 
stack between the fan (which is blowing out) and the rear enclosure 
wall.  This means that, instead of trying to maintain the internal 
instrument temerature at a constant level, it tries to maintain the 
exhaust stack temperature constant with a viciously fast response 
time that leads to instability at startup.  I have more than once 
considered moving the thermistor to a location near the TCXO, but 
since the fans always run up to full speed rather quickly at room 
temperature anyway, I have never bothered to try to improve the fan circuit.


Additionally. the TCXO remains powered during standby, but not 
exactly on frequency because the DAC that adjusts it during operation 
is not powered.  So, there is some settling from that adding to the 
temperature drift.  Note also that the DAC steps are not very fine, 
so you cannot expect to get the internal oscillator trimmed to better 
than e-9 or so.  SR apparently thought that most users would connect 
620s to external standards, so there was no reason to make them pay 
for a high-precision internal standard they would not use.


IME -- operating with an external reference that is better than the 
specified accuracy of the 620 -- they meet SR's specifications within 
a few minutes at most after switching on from room temperature 
storage.  (The trigger circuitry may drift a bit as it warms up, so 
you may want to check the trigger drift if your application involves 
slowish sine waves.  I have not investigated this.)  Ideally, you 
would let the instrument warm up for at least an hour and then 
perform an internal calibration before starting your measurements.


All that said, the only way you will know for sure how your 
particular instrument and standard will perform is to characterize 
them before you start your mobile measurements.  In doing so, you 
should observe a protocol that resembles the actual travel between 
measurements, at least with respect to time and temperature.  I 
strongly urge you to do this so you can have confidence in your measurements.


Best regards,

Charles







___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Considerations When Using The SR620

2012-12-02 Thread Magnus Danielson

On 12/02/2012 10:25 PM, Charles P. Steinmetz wrote:

Paul wrote:


The following comment appeared on this list recently and it scared me
a little:


Though the SR620 TIC is a great instrument when hunting the pico
seconds we have to realize, that it's a thermal design desaster (I
have to apologize to all sr620 friends). I have to run it for at
least 12 hoursif not 24 to be shure, that every single part is at a
more or less stationary thermal state. Some (NERC) say ...never
switch it off.


I assume this instability is due to the instability of the internal
frequency standard. * * * In fact, in our measurements, we plan to use
a Cesium frequency standard as the timebase to our SR620. Does this
anecdotal warning apply generally to the instrument or mainly to the
use of the internal standard oscillator?


I concur with the comment above that the thermal design of the 620 could
have been better -- the sensing thermistor is in an exhaust stack
between the fan (which is blowing out) and the rear enclosure wall. This
means that, instead of trying to maintain the internal instrument
temerature at a constant level, it tries to maintain the exhaust stack
temperature constant with a viciously fast response time that leads to
instability at startup. I have more than once considered moving the
thermistor to a location near the TCXO, but since the fans always run up
to full speed rather quickly at room temperature anyway, I have never
bothered to try to improve the fan circuit.


Never thought about that part, other than the fact that the fan is 
annoying like hell.



Additionally. the TCXO remains powered during standby, but not exactly
on frequency because the DAC that adjusts it during operation is not
powered. So, there is some settling from that adding to the temperature
drift. Note also that the DAC steps are not very fine, so you cannot
expect to get the internal oscillator trimmed to better than e-9 or so.
SR apparently thought that most users would connect 620s to external
standards, so there was no reason to make them pay for a high-precision
internal standard they would not use.


Which is why a high stability reference is an option, like most.

In contrast it is interesting to note that the HP5370 had a lower 
stability oscillator as option, so removing the 10811 was thus using a 
negative option.



IME -- operating with an external reference that is better than the
specified accuracy of the 620 -- they meet SR's specifications within a
few minutes at most after switching on from room temperature storage.
(The trigger circuitry may drift a bit as it warms up, so you may want
to check the trigger drift if your application involves slowish sine
waves. I have not investigated this.) Ideally, you would let the
instrument warm up for at least an hour and then perform an internal
calibration before starting your measurements.


Another source of temperature dependence is the analogue interpolator.


All that said, the only way you will know for sure how your particular
instrument and standard will perform is to characterize them before you
start your mobile measurements. In doing so, you should observe a
protocol that resembles the actual travel between measurements, at least
with respect to time and temperature. I strongly urge you to do this so
you can have confidence in your measurements.



This is good advice. Consider what you need to do. You might want to 
consider having a rubidium doing your hold-over. for instance. A PRS-10 
would be an interesting option for instance.


Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Considerations When Using The SR620

2012-12-02 Thread Magnus Danielson

Volker,

On 12/02/2012 11:36 PM, Volker Esper wrote:

Hi,

I was the one who wrote those nasty things about a fantastic instrument.

I bought such a counter a few weeks ago. When I first peeked into it
(because the fan was running at top speed after a few minutes) I
noticed, that the arrangement of the components was anything but optimal.

Examples: The osc is only one or two inches away from the linear voltage
regulators. These are cooled by a 1mm iron sheet. The iron sheet has
some cooling slots where the air can flow through, but these slots are
miles away from the regulators. The sheet gets so hot that you won't
touch it voluntarily. If at least the air could flow directly along the
regulators - but they sit in a corner and the air flows diagonally
through the case, avoiding contact with the regulators.


I am also not very impressed with the location of the linear regulators.

This forms a nice interesting relationship between line voltage and room 
temperature. As line voltage shifts, the rectified voltage varies, and 
then the voltage that the linear regulator needs to burn off varies, 
and thus their heat contribution varies with the line voltage. Now, the 
ambient temperature then controls how easy we get rid of this excess 
heat. The way the airflow goes, this cooling makes sure it couples very 
well with the electronics, so fan speed amongst other things depend on 
ambient temperature... and line voltage. All this assist to give us an 
interesting temperature dependence.


In contrast, I was impressed by how the HP10811A was located in a 
turbolence free corner of the HP5070A/B assembly. While itself also not 
particularly well designed in heat context, at least the heat of the 
linear regulators was supposed to mostly affect an external 
self-convection stream. Forces convection (fan) to stabilize the 
temperature of the heat sink would be a good thing there.


Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.