Re: [time-nuts] Counter OCXO behaviour

2013-01-15 Thread Fabio Eboli

Il 2013-01-14 22:15 Charles P. Steinmetz ha scritto:


Because the oscillators are sealed assemblies, I'm not aware of
anyone who has taken one apart for analysis -- so the reason for this
behavior must be considered unknown until we know what is going on
inside.


By the way, there is a non working unit on ebay
in US just now, 58h to auction end.
If it was in EU I would have picked it up just
for dissection. Unfortunately for me shipping
charges are a little high for an autopsy:
ebay #251211816860



Best regards,

Charles


Fabio.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Counter OCXO behaviour

2013-01-15 Thread David C. Partridge
Yes the 1992 4E adjustment is a bear. Mine behaves pretty much the same, ISTR 
it took me a few days to get it adjusted to my satisfaction and it was still 
drifting a bit.

Interestingly I saw a 1992 with a built in Rb on eBay (I think in France) some 
years back but the price went way beyong my budget.

Dave
-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Fabio Eboli
Sent: 14 January 2013 23:09
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Counter OCXO behaviour

I answer here to all, thank Ed David and Charles for your thoughts.
David, LOL, you posted the pic of the exact counter in question.
Not a similar unit, I mean it's exactly *that* 1992 that is being measured ;)


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Counter OCXO behaviour

2013-01-15 Thread Robert LaJeunesse
re ebay #251211816860 the seller is one I've dealt with a few times before, and 
had excellent service. I would not hesitate to deal with them again. 

FWIW according to the US Post Office web site a 1 pound package to Italy ships 
for $11.60. 

Bob L.


- Original Message 
 From: Fabio Eboli fabi...@quipo.it
 To: Discussion of precise time and frequency measurement time-nuts@febo.com
 Sent: Tue, January 15, 2013 9:21:21 AM
 Subject: Re: [time-nuts] Counter OCXO behaviour
 
 By the  way, there is a non working unit on ebay
 in US just now, 58h to auction  end.
 If it was in EU I would have picked it up just
 for dissection.  Unfortunately for me shipping
 charges are a little high for an  autopsy:
 ebay #251211816860
 
 Fabio.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Counter OCXO behaviour

2013-01-15 Thread Fabio Eboli

Robert,
I asked the seller for the least expensive way to
send that OCXO here (nothing to lose on a marginal
packaging, let's assume the item is non working)
and he quoted 12USD. I think I must refrain
from spending 20USD (assuming I will be the only
bidder) for a device to be dissected, althoug
I've done worse things with my ebay account :)

If it was a working OCXO I would have happily
bought it, but thinking about it I would not
dissect a working OCXO, for me it would be a crime :)

I will see if he has something interesting (and light)
in other items he listed, maybe I could share shipping
cost between items.

Il 2013-01-15 15:34 Robert LaJeunesse ha scritto:
re ebay #251211816860 the seller is one I've dealt with a few times 
before, and

had excellent service. I would not hesitate to deal with them again.

FWIW according to the US Post Office web site a 1 pound package to 
Italy ships

for $11.60.

Bob L.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Counter OCXO behaviour

2013-01-15 Thread Fabio Eboli

Il 2013-01-15 15:31 David C. Partridge ha scritto:

Yes the 1992 4E adjustment is a bear. Mine behaves pretty much the
same, ISTR it took me a few days to get it adjusted to my 
satisfaction

and it was still drifting a bit.


The strange thing is that mine now is changing frequency at
more than 1x10^-10 per day, measuring a 10MHz reference
this would be one digit per day, but before it was steady
on last digit for long time intervals, last digit moved up
and down only with day/night temperature variation.
Is it possible that the OCO was sealed by so much time,
opening it I changed the internal equilibrium? (humidity?).



Interestingly I saw a 1992 with a built in Rb on eBay (I think in
France) some years back but the price went way beyong my budget.


I've about this on this list, option 04R:
http://www.febo.com/pipermail/time-nuts/2011-November/060516.html

Fabio.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Counter OCXO behaviour

2013-01-15 Thread Robert LaJeunesse
Fabio,

I'm glad to see that their shipping charge is fair. Turns out it is only about 
a 
dollar more than the shipping to my place here in the states.

I did notice that the schematic for the oscillator's small PCB, the 5MHz to 
10MHz doubler, is on KO4BB's site (page 22 of the PDF):
http://www.ko4bb.com/Manuals/Racal/Racal-Dana_1991-1992-ServiceManualSch.pdf

Makes me wonder if the oscillator is good, but the doubler has failed. If so, 
the unit might be easily salvageable. As for what's in the sealed can, here's 
hoping you end up with one, somehow, at a cost you are happy with.

Bob


- Original Message 
 From: Fabio Eboli fabi...@quipo.it
 To: Robert LaJeunesse lajeune...@mail.com; Discussion of precise time and 
frequency measurement time-nuts@febo.com
 Sent: Tue, January 15, 2013 10:04:37 AM
 Subject: Re: [time-nuts] Counter OCXO behaviour
 
 Robert,
 I asked the seller for the least expensive way to
 send that OCXO  here (nothing to lose on a marginal
 packaging, let's assume the item is non  working)
 and he quoted 12USD. I think I must refrain
 from spending 20USD  (assuming I will be the only
 bidder) for a device to be dissected,  althoug
 I've done worse things with my ebay account :)
 
 If it was a  working OCXO I would have happily
 bought it, but thinking about it I would  not
 dissect a working OCXO, for me it would be a crime :)
 
 I will see  if he has something interesting (and light)
 in other items he listed, maybe I  could share shipping
 cost between items.
 
 Il 2013-01-15 15:34 Robert  LaJeunesse ha scritto:
  re ebay #251211816860 the seller is one I've  dealt with a few times 
  before, 
and
  had excellent service. I would not  hesitate to deal with them again.
  
  FWIW according to the US Post  Office web site a 1 pound package to Italy 
ships
  for $11.60.
  
  Bob L.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Counter OCXO behaviour

2013-01-14 Thread Fabio Eboli

Hello,
recently I tried to trim the Racal Dana
1992 04E internal reference, using the
GPS pps as a reference. I'd like to ask
a pair of questions...

- First is about the method.
I'm using the counter TI to measure it's
own OCXO. The GPS is starting the count,
the internal reference 10MHz (on the rear
there is a reference output connector)
stops the count and I log the result
as I've done before with other oscillators.
Is this a correct procedure?

- Second, is about the behaviour of the OCXO
after trimming. The OCXO seem not that stable
after the trimming, like if the crystal started
to age faster than before trimming, and now is
slowly stabilizing.
It's like the retrace of the crystals I've read
about, but the instrument was never powered down
in the last month. And it's oven has been on for
the last year. Will a crystal retrace also
after retrimming?
The frequency of the crystal is slowing, in the
first hours after the trimming rapidly, and now
more slowly. After few days the frequency seem
to be slowing somewhere around 1x10^-10 per day.
Unfortunately I havent logged the counter for
enough time before the trimming, but it measured
the same Rb with less than 3 digits of difference
(3x10^-10) in last 6 monthes.

Thanks,
Fabio.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Counter OCXO behaviour

2013-01-14 Thread Ed Breya
Is it manually adjusted right at the oscillator? If so, just opening it 
up and sticking a screwdriver in there gives it a thermal shock, and the 
adjusted element will have mechanical stress that has to settle out too 
- the value can change for a while.


Ed

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Counter OCXO behaviour

2013-01-14 Thread David
The Racal Dana 1991/1992 OCXO has a big pair of pan head screws
exposed on the back.  One is coarse adjust and one is fine adjust.  I
have the TCXO version which also exposes the adjustment on the back so
you do not have to open anything to get to it.

You can see them here:

http://www.eevblog.com/forum/reviews/racal-dana-1992-teardown/?action=dlattach;attach=20937;image;PHPSESSID=a5230a2beafd27b99c07580940fe1574

On Mon, 14 Jan 2013 12:34:04 -0800, Ed Breya e...@telight.com wrote:

Is it manually adjusted right at the oscillator? If so, just opening it 
up and sticking a screwdriver in there gives it a thermal shock, and the 
adjusted element will have mechanical stress that has to settle out too 
- the value can change for a while.

Ed

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Counter OCXO behaviour

2013-01-14 Thread David
On Mon, 14 Jan 2013 21:14:23 +0100, Fabio Eboli fabi...@quipo.it
wrote:

Hello,
recently I tried to trim the Racal Dana
1992 04E internal reference, using the
GPS pps as a reference. I'd like to ask
a pair of questions...

- First is about the method.
I'm using the counter TI to measure it's
own OCXO. The GPS is starting the count,
the internal reference 10MHz (on the rear
there is a reference output connector)
stops the count and I log the result
as I've done before with other oscillators.
Is this a correct procedure?

I just measured the GPS pulse output directly with the counter.  I got
the same calibration results measuring frequency or period.  With time
interval, I used the delay feature to set the minimum measurement
duration.

- Second, is about the behaviour of the OCXO
after trimming. The OCXO seem not that stable
after the trimming, like if the crystal started
to age faster than before trimming, and now is
slowly stabilizing.
It's like the retrace of the crystals I've read
about, but the instrument was never powered down
in the last month. And it's oven has been on for
the last year. Will a crystal retrace also
after retrimming?
The frequency of the crystal is slowing, in the
first hours after the trimming rapidly, and now
more slowly. After few days the frequency seem
to be slowing somewhere around 1x10^-10 per day.
Unfortunately I havent logged the counter for
enough time before the trimming, but it measured
the same Rb with less than 3 digits of difference
(3x10^-10) in last 6 monthes.

My 1992 has the TCXO and I noticed the least significant digit
drifting after trimming.  At least with the 1992 TCXO, I decided the
last digit was not worth worrying about which is pretty much the case
with all of my other counters although my 1992 is about the best of my
bunch.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Counter OCXO behaviour

2013-01-14 Thread Charles P. Steinmetz

Ed wrote:

Is it manually adjusted right at the oscillator? If so, just opening 
it up and sticking a screwdriver in there gives it a thermal shock, 
and the adjusted element will have mechanical stress that has to 
settle out too - the value can change for a while.


The 04E standard used in the military surplus 1992s (which are by far 
the most common 1992s in the US) is typically labeled 9462 454879, 
Rev. A.  There has been a fair bit of discussion on the list about 
these -- a search of the archive will turn up a number of threads.


Yes, they do tend to take a long while to settle after 
adjustment.  So long, in fact -- and with oscillatory gyrations above 
and below the starting frequency -- that I have speculated on-list 
that the adjustment may not be a direct adjustment of the tuned 
circuit (e.g., with the usual capacitor on the crystal), but rather 
an adjustment to the oven controller setting the crystal temperature.


Because the oscillators are sealed assemblies, I'm not aware of 
anyone who has taken one apart for analysis -- so the reason for this 
behavior must be considered unknown until we know what is going on inside.


Best regards,

Charles





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Counter OCXO behaviour

2013-01-14 Thread Fabio Eboli

I answer here to all, thank Ed David and Charles for your thoughts.
David, LOL, you posted the pic of the exact counter in question.
Not a similar unit, I mean it's exactly *that* 1992 that is being
measured ;)

The pictures on eevblog show the early measurements I made with
the counter. One of the Rb wasmeasured to about 10MHz+60mHz.
After a while (I think a pair of monthes) the reading was slightly
over 10MHz+70mHz and remained there for more than 6 monthes, until
I trimmed the counter few days ago.

The trimming is very touchy, and the simple opening of the screw on
the rear is traumatic: the frequency rises for a pair of minutes and
then it takes a while to return to the previous level.

Talking about this OCXO, how much is it good?
I mean for who has eperience with many OCXO is
this a good unit, or it's an average one?
Is this the kind of stability and reaction that can
be expected by an OCXO useful for GPSDO (less the
EFC control of course)?

Measuring this OCXO, just now I'm starting to feel the precision
of the FE5680, or 5682. It's so easy to set the Rb to follow
GPS closely for days...

I have some pics of the measurements I made, they are a little messy,
I hope can be fun and interesting to see:

This was before touching anything, top right the phase between
the GPS and OCXO:
http://www.flickr.com/photos/14336723@N08/8382126024/
it was slow by about 7x10^-9. This was the first trim attempt:
http://www.flickr.com/photos/14336723@N08/8382125734/
and this the last trims, the dips in phase are evident, they
happen when I open the calibration screw on the rear:
http://www.flickr.com/photos/14336723@N08/8381043177/
and at last these are logged after last trim action:
http://www.flickr.com/photos/14336723@N08/8381044033/

I tried to extract more data from last log:
http://www.flickr.com/photos/14336723@N08/8382126108/
Here I made the first and second derivative in time of the phase,
data are heavily averaged, so are not very accurate. The
10MHz period error (vs. GPS) changed from -0.8x10^-10 soon after
the trim, to +1x10^-10 after two days (top right).
The OCXO period was rising from almost 2x10^-15 soon after the
trim to 1.3x10^-15 after two days (bottom left).
Seem the stabilization will take a while...

Fabio.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Counter OCXO behaviour

2013-01-14 Thread Fabio Eboli

Il 2013-01-15 00:08 Fabio Eboli ha scritto:

Trying to be more clear:


10MHz period error (vs. GPS) changed from -0.8x10^-10 soon after
the trim, to +1x10^-10 after two days (top right).


Top right is change of phase in time, i.e. period error
measured using GPS as reference.


The OCXO period was rising from almost 2x10^-15 soon after the
trim to 1.3x10^-15 after two days (bottom left).


Bottom left is change of period error in time,
1.3x10^-15 per second is around 1.1x10^-10 per day,
if I'm not mistaken, more than typical unit's aging
stated by datasheet.


Seem the stabilization will take a while...

Fabio.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.