Re: [time-nuts] Cs stability

2007-07-17 Thread SAIDJACK
 
In a message dated 7/16/2007 22:26:14 Pacific Daylight Time,  
[EMAIL PROTECTED] writes:

 I  have a military OCXO sample (new product) that is supposed to be   
 g-insensitive and even that one has about E-09 frequency shift per  90 
degrees  tilt.

... and you have not put a GPS diciplining  on it yeat??? 



Of course we did :) But it still takes some time for the loop to fix the  
error.

We will address that in a next generation mobile product...
 
bye,
Said



** Get a sneak peek of the all-new AOL at 
http://discover.aol.com/memed/aolcom30tour
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-17 Thread Tom Van Baak
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

 I guess that is why some GPS antenna cables is temperature-stabilized as well
 as the cement-pidestal for the GPS antenna as it stands on solid rock. The 
 same
 place have controlled environment for the cesiums and hydrogens, together with
 UPS and disel-engine that kicks in for longer runs.
 
 Or I could be wrong... :)

Magnus,

This is true (temperature stabilization) for sites that do mm
level survey and ps level time transfer, using all the tricks
in the GPS book.

On the other hand, I think for most of us that play at the
meter and ns levels with cheap OEM receivers and plastic
L1 antennas the coax cable temperature issue is quite
overblown. Or if I'm wrong, show me the data.

/tvb


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-17 Thread Tom Van Baak
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

 Yes it can make sense.
 Place one Cs clock in a chamber where the ambient temperature can be 
 adjusted to various fixed temperatures. Compare the phase of its 
 5/10MHz  and/or PPS outputs with respect to those of another Cs standard 
 held at constant temperature.

Bruce,

I'm not sure I follow this. Yes, you will see a phase shift but
how can you tell how much of said shift is due to a fixed
phase shift (as if it were cable phase tempco) vs. how much
is due to phase shift due to frequency offset (as if it were
oscillator frequency tempco)? No change in temperature is
instantaneous; during the (slow) change both phase and
frequency (equals phase change over time) may change.

/tvb



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-17 Thread Tom Van Baak
 In the article OBSERVATIONS ON STABILITY MEASUREMENTS
 OF COMMERCIAL ATOMIC CLOCKS, Pekka Eskelinen claims to
 have measured a phase temperature coefficient of 100ns/degree
 for commercial Cs clocks in 1999.

Something is wrong with that claim. There's no way a modern
cesium standard exhibits a phase shift of 100 ns for a one
degree change in ambient temperature. I have Cs standards
that often change in temperature by several degrees and still
keep to nanoseconds. I guess I need to decode that paper
and see what's wrong. Does anyone have contact with the
authors (Finland)?

/tvb


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-17 Thread Magnus Danielson
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

From: Tom Van Baak [EMAIL PROTECTED]
Subject: Re: [time-nuts] Cs stability
Date: Tue, 17 Jul 2007 00:21:37 -0700
Message-ID: [EMAIL PROTECTED]

 ); SAEximRunCond expanded to false
 Errors-To: [EMAIL PROTECTED]
 
  I guess that is why some GPS antenna cables is temperature-stabilized as 
  well
  as the cement-pidestal for the GPS antenna as it stands on solid rock. The 
  same
  place have controlled environment for the cesiums and hydrogens, together 
  with
  UPS and disel-engine that kicks in for longer runs.
  
  Or I could be wrong... :)
 
 Magnus,

Tom,

 This is true (temperature stabilization) for sites that do mm
 level survey and ps level time transfer, using all the tricks
 in the GPS book.
 
 On the other hand, I think for most of us that play at the
 meter and ns levels with cheap OEM receivers and plastic
 L1 antennas the coax cable temperature issue is quite
 overblown. Or if I'm wrong, show me the data.

I never claimed otherwise. I mearly pointed out that _some_ antennas have that
arrangement among a number of others. Clearly few of us happy amatuers play in
that league. Some of us have GPS receivers and antennas at that level.

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-17 Thread Magnus Danielson
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

From: Tom Van Baak [EMAIL PROTECTED]
Subject: Re: [time-nuts] Cs stability
Date: Tue, 17 Jul 2007 00:40:49 -0700
Message-ID: [EMAIL PROTECTED]

 ); SAEximRunCond expanded to false
 Errors-To: [EMAIL PROTECTED]
 
  In the article OBSERVATIONS ON STABILITY MEASUREMENTS
  OF COMMERCIAL ATOMIC CLOCKS, Pekka Eskelinen claims to
  have measured a phase temperature coefficient of 100ns/degree
  for commercial Cs clocks in 1999.
 
 Something is wrong with that claim. There's no way a modern
 cesium standard exhibits a phase shift of 100 ns for a one
 degree change in ambient temperature. I have Cs standards
 that often change in temperature by several degrees and still
 keep to nanoseconds. I guess I need to decode that paper
 and see what's wrong.

I think the paper lacks a number of necessary details in order to fully
understand the conditions under which this exercise took place. There is no
clear list of equipment used. It can be implied that it is 5071A being used.
What Time Interval counters where used is for instance not known.

He have compared PPS outputs between GPS receivers (which model are those?)
and Cesium. 

I'd like to see some investigations into how stable the GPSes and measurement
equipment is to temperature.

In the end, there are a few pages of material missing in there for it to give a
fairly comprehensive picture. Many of these issues could probably be cleared up
after some discussions with the author, but we should not have to rely on side-
channels like that.

It's an interesting article, but it leaves me with a few questionsmarks. It is
a bit unsatisfying.

 Does anyone have contact with the authors (Finland)?

Should not be too hard as his email adderess is at the top of the first page.

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-17 Thread Hal Murray
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]


 I'm not sure I follow this. Yes, you will see a phase shift but how
 can you tell how much of said shift is due to a fixed phase shift (as
 if it were cable phase tempco) vs. how much is due to phase shift due
 to frequency offset (as if it were oscillator frequency tempco)? No
 change in temperature is instantaneous; during the (slow) change both
 phase and frequency (equals phase change over time) may change. 

Measure the phase.  Change the temperature.  Wait and watch.

If the new phase is stable/constant, you have a phase shift.  If it grows 
linear with time you have a frequency shift.



-- 
These are my opinions, not necessarily my employer's.  I hate spam.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-17 Thread Didier Juges

Or see a failed TIC test which actually displayed the delay in a piece of
coax showing the temperature in the room as the air conditioning was turning
on and off over night.

http://www.ko4bb.com/Test_Equipment/data/HP5370A coax cable delay.png

I did not calibrate this thermometer though :-)

Didier KO4BB

-Original Message-
From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
Behalf Of Tom Van Baak
Sent: Monday, July 16, 2007 9:47 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Cs stability

 Who was it that said; every clock is a thermometer?

Even a short length of coax is a thermometer if you look
close enough. See the 50 fs / °C tempco at test #6:

http://www.leapsecond.com/pages/tadd-1/

/tvb 



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-17 Thread Björn Gabrielsson
Achieved:  Acceleration sensitivities better than 2E-12/g

http://www.zyfer.com/briefings/telecom/qz%20low-g%20brfg%20olie%204-04.pdf

--

   Björn

On Tue, July 17, 2007 14:40, Rob Kimberley said:
 Another way that has been incorporated successfully to combat g
 sensitivity
 are accelerometers mounted on each of the x, y, Z axes of the crystal.
 Combining this with spring mounts works well. g sensitivities down to e-10
 per g  can be achieved.

 Rob K

 -Original Message-
 From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
 Behalf Of [EMAIL PROTECTED]
 Sent: 17 July 2007 06:14
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] Cs stability

 ); SAEximRunCond expanded to false
 Errors-To:
 [EMAIL PROTECTED]
 om


 In a message dated 7/16/2007 12:10:54 Pacific Daylight Time,
 [EMAIL PROTECTED] writes:

For  example tilting the Cs unit by 90 degrees will typically give an
error

of about 1E-09 or so.

It would take some time for the  control loops to compensate for this
 error.



 Hi guys,

 I did tilt my 4050 when I first installed the unit, that number is from my
 memory. Notice that Magnus is right of course, the Cs will compensate the
 tilt,  given enough time to do so. I did not mean that the 1E-09 error
 will
 stay permanent of course, as I said in the follow-on sentence.

 Also the 4050 is not very well temperature compensated compared to modern
 units, I was not impressed when I saw the FTS tempco specifications. A
 good
 GPSDO can outperform the 4050.

 I have a military OCXO sample (new product) that is supposed to be
 g-insensitive and even that one has about E-09 frequency shift per 90
 degrees  tilt.

 One of the only ways to get around that is to use three crystals in
 series,
 with the three crystals oriented in the the XYZ axis, so their  errors
 compensate out.

 Vibration is also a big enemy of crystal oscillators of course.

 bye,
 Said



 ** Get a sneak peek of the all-new AOL
 at http://discover.aol.com/memed/aolcom30tour
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.



 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-17 Thread Rob Kimberley
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

I guess I'm a bit out of date!

:-)

Rob 

-Original Message-
From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
Behalf Of Björn Gabrielsson
Sent: 17 July 2007 14:13
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Cs stability

); SAEximRunCond expanded to false
Errors-To:
[EMAIL PROTECTED]

Achieved:  Acceleration sensitivities better than 2E-12/g

http://www.zyfer.com/briefings/telecom/qz%20low-g%20brfg%20olie%204-04.pdf

--

   Björn

On Tue, July 17, 2007 14:40, Rob Kimberley said:
 Another way that has been incorporated successfully to combat g 
 sensitivity are accelerometers mounted on each of the x, y, Z axes of 
 the crystal.
 Combining this with spring mounts works well. g sensitivities down to 
 e-10 per g  can be achieved.

 Rob K

 -Original Message-
 From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] 
 On Behalf Of [EMAIL PROTECTED]
 Sent: 17 July 2007 06:14
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] Cs stability

 ); SAEximRunCond expanded to false
 Errors-To:
 time-nuts-bounces+rk=timing-consultants.com+rk=timing-consultants.com@
 time-nuts-bounces+febo.c
 om


 In a message dated 7/16/2007 12:10:54 Pacific Daylight Time, 
 [EMAIL PROTECTED] writes:

For  example tilting the Cs unit by 90 degrees will typically give an 
error

of about 1E-09 or so.

It would take some time for the  control loops to compensate for this
 error.



 Hi guys,

 I did tilt my 4050 when I first installed the unit, that number is 
 from my memory. Notice that Magnus is right of course, the Cs will 
 compensate the tilt,  given enough time to do so. I did not mean that 
 the 1E-09 error will stay permanent of course, as I said in the 
 follow-on sentence.

 Also the 4050 is not very well temperature compensated compared to 
 modern units, I was not impressed when I saw the FTS tempco 
 specifications. A good GPSDO can outperform the 4050.

 I have a military OCXO sample (new product) that is supposed to be 
 g-insensitive and even that one has about E-09 frequency shift per 
 90 degrees  tilt.

 One of the only ways to get around that is to use three crystals in 
 series, with the three crystals oriented in the the XYZ axis, so their  
 errors compensate out.

 Vibration is also a big enemy of crystal oscillators of course.

 bye,
 Said



 ** Get a sneak peek of the all-new 
 AOL at http://discover.aol.com/memed/aolcom30tour
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.



 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-17 Thread Magnus Danielson
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

From: Dr Bruce Griffiths [EMAIL PROTECTED]
Subject: Re: [time-nuts] Cs stability
Date: Tue, 17 Jul 2007 21:30:03 +1200
Message-ID: [EMAIL PROTECTED]

 ); SAEximRunCond expanded to false
 Errors-To: [EMAIL PROTECTED]
 
 Tom Van Baak wrote:
  I guess that is why some GPS antenna cables is temperature-stabilized as 
  well
  as the cement-pidestal for the GPS antenna as it stands on solid rock. The 
  same
  place have controlled environment for the cesiums and hydrogens, together 
  with
  UPS and disel-engine that kicks in for longer runs.
 
  Or I could be wrong... :)
  
 
  Magnus,
 
  This is true (temperature stabilization) for sites that do mm
  level survey and ps level time transfer, using all the tricks
  in the GPS book.
 
  On the other hand, I think for most of us that play at the
  meter and ns levels with cheap OEM receivers and plastic
  L1 antennas the coax cable temperature issue is quite
  overblown. Or if I'm wrong, show me the data.
 
  /tvb
 
 

 Tom
 
 Calculations are easy:
 
 Cable delay tempco is at worst 100ppm/K.
 with 100ns of cable delay tempco will be 10ps?K or less.
 With a 20K temperature change change in delay will be 200ps or less.
 The group delay tempco of the antenna components (bandpass filters 
 amplifiers etc ) is likely to be much greater.

Tom's point is that while we are toying around with L1 C/A code-tracking
receivers the effective noise is so large that more finegrained compensations
in the GPS solution is note taken out so other finegrained offsets will also be
lost. I agree with him. When you toy with dual frequency carrier-tracking
receivers you gain a much better precision in ionspheric correction. Operating
in mm or sub-mm RMS noise on pseudo-ranges noise sources as change of
temperature in cables and mountning stands certainly comes in, as well as the
phase stability of the antenna, quality sites compensates for antenna phase
deviation as they change over the satelite coarse. But that's not very
meaningfull for a L1 C/A code-tracking only receiver, infact such efforts are
lost in the noise and uncompensated biases you have. Especially considering a
cheezy TCXO.

So while it is easy to calculate and compensate, the meaningfullness depends on
the situation. Compensating for cable delay may however be worth it.

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-17 Thread M. Warner Losh
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

In message: [EMAIL PROTECTED]
Tom Van Baak [EMAIL PROTECTED] writes:
: ); SAEximRunCond expanded to false
: Errors-To: [EMAIL PROTECTED]
: 
:  In the article OBSERVATIONS ON STABILITY MEASUREMENTS
:  OF COMMERCIAL ATOMIC CLOCKS, Pekka Eskelinen claims to
:  have measured a phase temperature coefficient of 100ns/degree
:  for commercial Cs clocks in 1999.
: 
: Something is wrong with that claim. There's no way a modern
: cesium standard exhibits a phase shift of 100 ns for a one
: degree change in ambient temperature. I have Cs standards
: that often change in temperature by several degrees and still
: keep to nanoseconds. I guess I need to decode that paper
: and see what's wrong. Does anyone have contact with the
: authors (Finland)?

We have HP5071A's that keep to about 5ns over the course of a typical
heading/cooling cycle measured relative to GPS.  This is over a range
of maybe 15C in the summer...

Warner


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-17 Thread Richard \(Rick\) Karlquist
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

In case anyone still doesn't know who I am, I need to
mention that I designed the RF electronics in the 
5071A in an earlier life, circa 1990.

Now that we have that out of the way, I will agree
with Magnus that the cited paper is severely flawed.
We don't know what models of cesium were used, and
the paper seems to assume that any deviations with
respect to GPS are due solely to the cesium standards.
We also don't know anything about the GPS equipment.
Citation [3] is a 5071A product note.  I guess this
is thrown in to imply that one of the cesiums is a 
5071A; but we can't be sure.  A statement is taken
out of context from this product note to the effect
that some cesiums don't have an independent means
of frequency setting, which might lead the casual
observer to think that this refers to the 5071A.
Actually, the 5071A does have an independent means
of frequency setting, as opposed to older cesium
clocks where you had to vary the C-field to change
the frequency.  For this and other reasons, I don't
put much stock in the paper.

Having said that, let me speak about how we addressed
some of these issues in the 5071A.  First, the matter
of temperature dependent phase shift.  If you have a
really good frequency standard, you have to start to
worry about a temperature ramp resulting in a phase 
ramp.  Of course, a phase ramp is a frequency step,
by definition.  The previous modele, the 5061B, like most
cesium clocks, had an architecture where the crystal oscillator
flywheel was connected to a splitter that drove
a microwave frequency multiplier from 10 MHz to 9.2
GHz and also an isolation amplifier.  Both of these
were absolutely full of narrow band tuned circuits
with substantial temperature coefficients.  The combination
of these rubber bands caused the 10 MHz at the output
connector to be fairly loosely correlated (in terms of 5071A
specifications) with the microwave signals sloshing
around the Cs cavity.   In the 5071A, I used a combination
of a new architecture and better circuit design to 
drastically reduce the temperature coefficient of phase.
You can read about this in my 1992 FCS paper (available at
ieee.org or www.karlquist.com/FCS92.pdf).  We measured the RF
hardware by itself to determine its phase contribution to
the overall 5071A.  Our resident perfectionist, Len Cutler,
insisted that all electronic error contributions be less than
a part in 10^14, including this one.  I can assure you that Len was
a very happy camper with respect to the RF chain, and, if
you knew Len, you know that is saying something.

Shortly after the introduction of the 5071A, we had some
colleagues at JPL perform environmental tests on an early
5071A.  For the better part of a year, they ramped up
and down the temperature, humidity, and pressure to see
if they could measure these tempcos.  By using correlative
techniques, such as ramping the temperature on a daily
basis and looking for a fourier peak at 1 day, they were
able to establish a measurement floor of something like
a part in 10^15.  They were never able to see any correlation
whatsoever of frequency to the environment for any of the variables.

Regarding magnetic field sensitivity.  The mu metal shields
of the 5071A are so effective that it is not necessary to
demagnetize them.  Even Len Cutler didn't think so.  A standing
joke was the 5071A demagnetizer accessory that we were going
to build one of these days.  I think we finally did build
one for some customer who was even more of a perfectionist than
Len.  He didn't see any improvement from using it.  Anyway,
I take any statements about the earth's magnetic field affecting
5071A with a large grain of salt.

AC magnetic fields like 50 Hz, or 60 Hz, are also not a concern.
The frequencies used internally in the 5071A are carefully chosen
to avoid any correlation with those power line frequencies.  
Magnetic fields even in the milliTesla range have absolutely no
effect, let alone nanoTesla.

Regarding vibration, acceleration, etc:  earlier cesium beam tubes
did have a problem with the beam wandering around when the 
clock was installed on a ship.  In the 5071A, Len carefully 
devised a technique to prevent beam wander from having any
effect on accuracy.  Now if you tipped a 5071A upside down,
the 10811 would experience a 2g turnover frequency step.
The control loop would respond to this within its time constant.
So, admittedly, you might have a temporary frequency shift.
However, there is no known application where a clock needs
to work upside down, so even Len couldn't justify worrying about this.

I hope this clears up any confusion about the 5071A, now
made by my good friends at Symmetricom.

Rick Karlquist N6RK
(now with Agilent)

These are my own opinions and don't represent HP, Agilent,
or Symmetricom.


 


   In the article OBSERVATIONS ON STABILITY MEASUREMENTS
   OF COMMERCIAL ATOMIC CLOCKS, Pekka Eskelinen claims to
 I think the paper 

Re: [time-nuts] Cs stability

2007-07-17 Thread Magnus Danielson
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

From: Richard \(Rick\) Karlquist [EMAIL PROTECTED]
Subject: Re: [time-nuts] Cs stability
Date: Tue, 17 Jul 2007 09:52:22 -0700
Message-ID: [EMAIL PROTECTED]

Rick,

 In case anyone still doesn't know who I am, I need to
 mention that I designed the RF electronics in the 
 5071A in an earlier life, circa 1990.
 
 Now that we have that out of the way, I will agree
 with Magnus that the cited paper is severely flawed.

Good, then I am not the only one seeing that. I must admitt that I didn't give
it a comprehensive readthrough asI was tired as hell when I looked at it, but
I did not get the necessary ques as I browsed through it and even as I dipped
into certain parts in detail it left me unsatisfied.

 We don't know what models of cesium were used, and
 the paper seems to assume that any deviations with
 respect to GPS are due solely to the cesium standards.
 We also don't know anything about the GPS equipment.
 Citation [3] is a 5071A product note.  I guess this
 is thrown in to imply that one of the cesiums is a 
 5071A; but we can't be sure.

There is only an indirect reference to 5071A in a.

  A statement is taken
 out of context from this product note to the effect
 that some cesiums don't have an independent means
 of frequency setting, which might lead the casual
 observer to think that this refers to the 5071A.
 Actually, the 5071A does have an independent means
 of frequency setting, as opposed to older cesium
 clocks where you had to vary the C-field to change
 the frequency.  For this and other reasons, I don't
 put much stock in the paper.

Do note that he in the introductionary part shows the factor deviation of
2.7E-13 and that he adjusted it manually twice which can be seen in the second
Fig 1.

Unfortunately, many atomic oscillators lack the possibility of non-invasive
frequency tuning [3]. Using digital correction, the frequency offset of
2.7 E-13 as delivered from factory is shown in Fig.l to be reducable below
1 E-14 with only two frequency corrections at 1800 and 4O00 hours from
start-up, leaving some environmental factors plus a number of unpredictable
timing discrepancies lasting 50-500 horn.

He actually points out that the 5071A lacks that correction, so he agrees with
you. This correction can be done by him as an initial post-processing step.
We just don't know, again. A little sentense or two more on how that was made
would have enlighen us further on that subject.

But what Cesium he actually used is not known.

 I hope this clears up any confusion about the 5071A, now
 made by my good friends at Symmetricom.

Many, many thanks for the write-up. I greatly enjoyed it!

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Cs stability

2007-07-16 Thread Pablo Alvarez Sanchez
Hi, 

I am curious about the total stability of Cs clocks. Normally producers give 
you an initial accuracy after 30 minutes of power on and a table with the Allan 
deviation for different measurement intervals. 

After that they give you the environmental and physical specifications. For the 
hp5071 you have:

General environment
Temperature
Operating 0°C to 55°C
Non-operating -40°C to 70°C
Humidity 0 to 95%RH (45C max)
Magnetic field dc, 55, 60Hz 0 to 2 gauss peak - any orientation Atmospheric 
pressure £1E-13 change in frequency for pressure down to 19kPa (equivalent to 
an altitude of 12.2km) Shock and vibration Mil-T-28800D, Type III, class 5 
Hammer Blow Shock Test, Mil-S-901C, Grade A, Class 1, Type A Mile-STD, 167-1 
(phase noise)
EMI: Conducted and radiated emissions per CISPR 11/EN 55011, Group 1, Class A
EMC: per MIL-STD-461C, Part 7, Class B dc magnetic field up to 7.8 Gauss



My questions are:

Are the Allan deviation specs also valid for all the environmental range, 
including shock and vibration, or only for lab conditions?

In the article OBSERVATIONS ON STABILITY MEASUREMENTS OF COMMERCIAL ATOMIC 
CLOCKS, Pekka Eskelinen claims to have measured a phase temperature 
coefficient of 100ns/degree for commercial Cs clocks in 1999.

http://ieeexplore.ieee.org/iel5/6762/18075/00840739.pdf (If you cannot read it 
I can try to send you a copy by email)

Has any of you ever measured such a coefficient?

Cheers

Pablo 

 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-16 Thread SAIDJACK
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

In a message dated 7/16/2007 10:05:51 Pacific Daylight Time,  
[EMAIL PROTECTED] writes:

My  questions are:

Are the Allan deviation specs also valid for all the  environmental range, 
including shock and vibration, or only for lab  conditions?
Hi Pablo,
 
I can only speak from experience on my FTS4050 Cs:
 
First of all the output signal of almost all Rb and Cs units is generated  by 
a quartz OCXO (FTS1200 in my case) so you have all of the shock/vibration and 
 other sensitivites that quartz oscillators have.
 
For example tilting the Cs unit by 90 degrees will typically give an error  
of about 1E-09 or so.
 
It would take some time for the control loops to compensate for this  error.
 
Another observation is that Allan Deviation plots/measurements are usually  
done on units that have fairly constant temperature during the measurement, do  
not get moved, and also have had ample time to settle down (weeks, or even  
months to stabalize for some manufacturers!).
 
The problem is that any other conditions (adding vibration etc) are very  
hard to specify and re-create, it's much simpler to let the unit sit in a  nice 
and stable lab environment, and the measurement results will show very good  
performance.
 
Hope this helps,
bye,
Said
 



** Get a sneak peak of the all-new AOL at 
http://discover.aol.com/memed/aolcom30tour
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-16 Thread Dr Bruce Griffiths
Pablo Alvarez Sanchez wrote:
 ); SAEximRunCond expanded to false
 Errors-To: [EMAIL PROTECTED]

 Hi, 

 I am curious about the total stability of Cs clocks. Normally producers give 
 you an initial accuracy after 30 minutes of power on and a table with the 
 Allan deviation for different measurement intervals. 

 After that they give you the environmental and physical specifications. For 
 the hp5071 you have:

 General environment
 Temperature
 Operating 0°C to 55°C
 Non-operating -40°C to 70°C
 Humidity 0 to 95%RH (45C max)
 Magnetic field dc, 55, 60Hz 0 to 2 gauss peak - any orientation Atmospheric 
 pressure £1E-13 change in frequency for pressure down to 19kPa (equivalent to 
 an altitude of 12.2km) Shock and vibration Mil-T-28800D, Type III, class 5 
 Hammer Blow Shock Test, Mil-S-901C, Grade A, Class 1, Type A Mile-STD, 167-1 
 (phase noise)
 EMI: Conducted and radiated emissions per CISPR 11/EN 55011, Group 1, Class A
 EMC: per MIL-STD-461C, Part 7, Class B dc magnetic field up to 7.8 Gauss



 My questions are:

 Are the Allan deviation specs also valid for all the environmental range, 
 including shock and vibration, or only for lab conditions?

 In the article OBSERVATIONS ON STABILITY MEASUREMENTS OF COMMERCIAL ATOMIC 
 CLOCKS, Pekka Eskelinen claims to have measured a phase temperature 
 coefficient of 100ns/degree for commercial Cs clocks in 1999.

 http://ieeexplore.ieee.org/iel5/6762/18075/00840739.pdf (If you cannot read 
 it I can try to send you a copy by email)

 Has any of you ever measured such a coefficient?

 Cheers

 Pablo 
   
Pablo

A Cs clock uses a frequency lock loop to control the frequency of the 
local crystal oscillator, the crystal oscillator phase is arbitrary. The 
phase of the output with respect to the crystal depends on the 
propagation delay of any intervening electronics including amplifier and 
filter phase shifts. The phase shift of tuned circuits and other narrow 
bandwidth filters has a relatively high temperature coefficient and this 
is perhaps what has been measured. Modern isolation amplifier designs 
eschew the use of tuned circuits and other narrow bandwidth filters and 
consequently have much lower phase shift temperature coefficients 
(typically a few picosec/C).

Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-16 Thread Tom Van Baak
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

 After that they give you the environmental and physical
 specifications. For the hp5071 you have:
...
 Are the Allan deviation specs also valid for all the environmental
 range, including shock and vibration, or only for lab conditions?

The given specs are conservative (in typical HP style) but
I would guess the best ADEV numbers are only for laboratory
conditions. Someone from Agilent/Symmetricom might want
to comment on this.

 In the article OBSERVATIONS ON STABILITY MEASUREMENTS
 OF COMMERCIAL ATOMIC CLOCKS, Pekka Eskelinen claims to
 have measured a phase temperature coefficient of 100ns/degree
 for commercial Cs clocks in 1999.

I'll comment after I read it. But the 100ns/degree value doesn't
make sense because that's phase instead of frequency units.

Did he mean 100 ns per day per degree? Or per 200 hours,
or 2000 hours, etc. If the latter, that represents a per-degree
frequency shift of 100 ns / 2000 h = 1.3e-14 which sounds
about right to me for a cesium tempco. It also depends on the
model: the tempco of a vintage hp 5060A or hp 5061A is likely
worse than a modern 5071A, for example.

 Has any of you ever measured such a coefficient?

Yes, I'll take a look over my old data files and see what I have
on cesium. This week I just happen to be measuring the tempco
of an old Russian H-maser (about 2e-14/C). Please see:

http://www.leapsecond.com/pages/phm-temp/

/tvb


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-16 Thread Tom Van Baak
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

 For example tilting the Cs unit by 90 degrees will typically give an error  
 of about 1E-09 or so.

Have you tried this with your FTS 4050? Anyone else do
this with a 4060 or hp 5061?

/tvb


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-16 Thread Dr Bruce Griffiths
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

Tom Van Baak wrote:
 The given specs are conservative (in typical HP style) but
 I would guess the best ADEV numbers are only for laboratory
 conditions. Someone from Agilent/Symmetricom might want
 to comment on this.

   
 In the article OBSERVATIONS ON STABILITY MEASUREMENTS
 OF COMMERCIAL ATOMIC CLOCKS, Pekka Eskelinen claims to
 have measured a phase temperature coefficient of 100ns/degree
 for commercial Cs clocks in 1999.
 

 I'll comment after I read it. But the 100ns/degree value doesn't
 make sense because that's phase instead of frequency units.

 Did he mean 100 ns per day per degree? Or per 200 hours,
 or 2000 hours, etc. If the latter, that represents a per-degree
 frequency shift of 100 ns / 2000 h = 1.3e-14 which sounds
 about right to me for a cesium tempco. It also depends on the
 model: the tempco of a vintage hp 5060A or hp 5061A is likely
 worse than a modern 5071A, for example.
   
Tom

Yes it can make sense.
Place one Cs clock in a chamber where the ambient temperature can be 
adjusted to various fixed temperatures. Compare the phase of its 
5/10MHz  and/or PPS outputs with respect to those of another Cs standard 
held at constant temperature. The observed phase shift sequence may then 
be fitted to both frequency shift and a fixed (for a given temperature) 
phase shift components. Repeat for a range of temperatures and plot the 
(temperature dependent) fixed phase shift component as a function of 
temperature.

Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-16 Thread Jack Hudler
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

Who was it that said; every clock is a thermometer?

-Original Message-
From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
Behalf Of Dr Bruce Griffiths
Sent: Monday, July 16, 2007 8:31 PM
To: Tom Van Baak; Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Cs stability

Tom Van Baak wrote:
 The given specs are conservative (in typical HP style) but
 I would guess the best ADEV numbers are only for laboratory
 conditions. Someone from Agilent/Symmetricom might want
 to comment on this.

   
 In the article OBSERVATIONS ON STABILITY MEASUREMENTS
 OF COMMERCIAL ATOMIC CLOCKS, Pekka Eskelinen claims to
 have measured a phase temperature coefficient of 100ns/degree
 for commercial Cs clocks in 1999.
 

 I'll comment after I read it. But the 100ns/degree value doesn't
 make sense because that's phase instead of frequency units.

 Did he mean 100 ns per day per degree? Or per 200 hours,
 or 2000 hours, etc. If the latter, that represents a per-degree
 frequency shift of 100 ns / 2000 h = 1.3e-14 which sounds
 about right to me for a cesium tempco. It also depends on the
 model: the tempco of a vintage hp 5060A or hp 5061A is likely
 worse than a modern 5071A, for example.
   
Tom

Yes it can make sense.
Place one Cs clock in a chamber where the ambient temperature can be 
adjusted to various fixed temperatures. Compare the phase of its 
5/10MHz  and/or PPS outputs with respect to those of another Cs standard 
held at constant temperature. The observed phase shift sequence may then 
be fitted to both frequency shift and a fixed (for a given temperature) 
phase shift components. Repeat for a range of temperatures and plot the 
(temperature dependent) fixed phase shift component as a function of 
temperature.

Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-16 Thread Chuck Harris
Tom Van Baak wrote:
 ); SAEximRunCond expanded to false
 Errors-To: [EMAIL PROTECTED]
 
 For example tilting the Cs unit by 90 degrees will typically give an error  
 of about 1E-09 or so.
 
 Have you tried this with your FTS 4050? Anyone else do
 this with a 4060 or hp 5061?

I'm pretty sure that back when my FTS4050 was still working
it didn't care one little bit about orientation.  Perhaps
yours has lost its magnetic shielding, and is reacting to
the change in the orientation of the Earth's magnetic field?

The C'beam standard package in the FTS4050 is an FTS5000 which
was designed to be used in all sorts of 28V applications, including
portable standards and avionics.

-Chuck Harris

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-16 Thread Tom Van Baak
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

 Who was it that said; every clock is a thermometer?

Even a short length of coax is a thermometer if you look
close enough. See the 50 fs / °C tempco at test #6:

http://www.leapsecond.com/pages/tadd-1/

/tvb 



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-16 Thread Magnus Danielson
From: Chuck Harris [EMAIL PROTECTED]
Subject: Re: [time-nuts] Cs stability
Date: Mon, 16 Jul 2007 21:55:26 -0400
Message-ID: [EMAIL PROTECTED]

 ); SAEximRunCond expanded to false
 Errors-To: [EMAIL PROTECTED]
 
 Tom Van Baak wrote:
  ); SAEximRunCond expanded to false
  Errors-To: [EMAIL PROTECTED]
  
  For example tilting the Cs unit by 90 degrees will typically give an error 
   
  of about 1E-09 or so.
  
  Have you tried this with your FTS 4050? Anyone else do
  this with a 4060 or hp 5061?
 
 I'm pretty sure that back when my FTS4050 was still working
 it didn't care one little bit about orientation.  Perhaps
 yours has lost its magnetic shielding, and is reacting to
 the change in the orientation of the Earth's magnetic field?

It seems a little high a number still.

 The C'beam standard package in the FTS4050 is an FTS5000 which
 was designed to be used in all sorts of 28V applications, including
 portable standards and avionics.

You don't happend to have more detailed FTS1200 specs and maybe schematics?

I assume that it is an SC-cut crystal in there, but it might be AT-cut.

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-16 Thread Magnus Danielson
From: Tom Van Baak [EMAIL PROTECTED]
Subject: Re: [time-nuts] Cs stability
Date: Mon, 16 Jul 2007 19:47:10 -0700
Message-ID: [EMAIL PROTECTED]

 ); SAEximRunCond expanded to false
 Errors-To: [EMAIL PROTECTED]
 
  Who was it that said; every clock is a thermometer?
 
 Even a short length of coax is a thermometer if you look
 close enough. See the 50 fs / °C tempco at test #6:
 
 http://www.leapsecond.com/pages/tadd-1/

I guess that is why some GPS antenna cables is temperature-stabilized as well
as the cement-pidestal for the GPS antenna as it stands on solid rock. The same
place have controlled environment for the cesiums and hydrogens, together with
UPS and disel-engine that kicks in for longer runs.

Or I could be wrong... :)

Cheers,
Magnus - still have to do some of that fancy stuff with my rig

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Cs stability

2007-07-16 Thread Magnus Danielson
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED]

From: [EMAIL PROTECTED]
Subject: Re: [time-nuts] Cs stability
Date: Tue, 17 Jul 2007 01:13:58 EDT
Message-ID: [EMAIL PROTECTED]

 Hi guys,

Hi Said,

 I did tilt my 4050 when I first installed the unit, that number is from my  
 memory. Notice that Magnus is right of course, the Cs will compensate the 
 tilt, 
  given enough time to do so. I did not mean that the 1E-09 error will stay  
 permanent of course, as I said in the follow-on sentence.

Ah. The it was certainly a crystal detuning which took some time to overcome by
a sluggish control-loop.

On the other hand, few Cesium beams should be required to handle quick 90
degrees tilts during operation. They should sit firmly in the rack.

Touring cesiums is a different buissness, but they should rarely have to handle
quick tilts during operation too.

 Also the 4050 is not very well temperature compensated compared to modern  
 units, I was not impressed when I saw the FTS tempco specifications. A good  
 GPSDO can outperform the 4050.

Several developments have occured since the 4050.

 I have a military OCXO sample (new product) that is supposed to be  
 g-insensitive and even that one has about E-09 frequency shift per 90 
 degrees  tilt.

... and you have not put a GPS diciplining on it yeat??? 

 One of the only ways to get around that is to use three crystals in  series, 
 with the three crystals oriented in the the XYZ axis, so their  errors 
 compensate out.

As I recall it, both gravity force and magnetic fields will contribute to
crystal detuning.

 Vibration is also a big enemy of crystal oscillators of course.

As always.

Cheers,