Re: [time-nuts] Datum 4065A Cs Tube Response

2014-05-05 Thread J. L. Trantham
Ed,

If I have the math correct, and you are measuring the voltage to ground
through a 10 MegOhm input impedance DMM, you have about 7.5 nA beam current
which seems a bit low compared to what I remember of the HP 5061A.  However,
you still have a definable 'peak' with a 'peak to valley' voltage of about
60 mV or a 'useful signal current' of about 6 nA.  If your unit's circuitry
can properly amplify that and keep it a clean signal, it should work.
However, I would recommend setting the OCXO precisely on frequency with a
GPSDO before trying to close the loop and 'locking' the signal to the CS
tube.  It will dramatically lower the work load of getting everything
adjusted properly, particularly in a setting of low beam current.

Somehow, the value of 40 nA sticks in my mind from the 5061A.  The 5061A
manual says end of life of the HP CS tube is a peak beam current of 8 nA or
less.  However, I have units with less current and they still lock.  The HP
manual also says to measure the voltage at the output of the tube with a 100
MegOhm or higher input impedance DMM.  If yours is less, that may
artificially lower your values.

EOL of the tube is a multifactor issue, including Signal to Noise ratio and
the 'useful signal current' to 'background current' ratio.  The 'background
current' is what you see with no RF signal applied to the tube.  Have you
measured that?  A ratio of 1 is EOL per the HP manual.  If yours is about
4.5 nA, as suggested by the 'off peak' values shown, or less, you still have
a useful signal and, hopefully, a useful tube.

I'd recommend continuing with the repair.

Good luck.

Joe

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Ed Palmer
Sent: Monday, May 05, 2014 12:46 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Datum 4065A Cs Tube Response

I'm playing with my first Cs standard.  It's a Datum 4065A which appears to
have a dead STEL-1173 synthesizer.  Before I put too much effort into
replacing that, I thought I'd check the tube and see if it has any life
left.  I've attached a chart showing the response of the central peak.

My methodology was similar to TVB's as shown here: 
http://www.leapsecond.com/pages/cspeak except that I measured the tube
output directly with a digital voltmeter.  The system is reporting wildly
varying levels for the beam current so I didn't want to use any of it's
circuitry.

Does this look like a usable tube?  Healthy or on it's last legs? What
response levels are typical for a Datum 7504A tube?  I see that these levels
are somewhat lower than those shown on leapsecond for the 5061A tube, but
that could just be the specifics of the measurement.

Thanks,

Ed


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Datum 4065A Cs Tube Response

2014-05-05 Thread paul swed
I will agree with Joe. I have a CS tube thats darn near impossible to read
the beam current and yet it still locks. That truly amazes me. I seem to
recall other comments ages ago about that chip failing. There should be a
way to emulate it these days with all of the DDS chips and such that are
available.
Good luck.
Regards
Paul.
WB8TSL


On Mon, May 5, 2014 at 8:20 AM, J. L. Trantham jlt...@att.net wrote:

 Ed,

 If I have the math correct, and you are measuring the voltage to ground
 through a 10 MegOhm input impedance DMM, you have about 7.5 nA beam current
 which seems a bit low compared to what I remember of the HP 5061A.
  However,
 you still have a definable 'peak' with a 'peak to valley' voltage of about
 60 mV or a 'useful signal current' of about 6 nA.  If your unit's circuitry
 can properly amplify that and keep it a clean signal, it should work.
 However, I would recommend setting the OCXO precisely on frequency with a
 GPSDO before trying to close the loop and 'locking' the signal to the CS
 tube.  It will dramatically lower the work load of getting everything
 adjusted properly, particularly in a setting of low beam current.

 Somehow, the value of 40 nA sticks in my mind from the 5061A.  The 5061A
 manual says end of life of the HP CS tube is a peak beam current of 8 nA or
 less.  However, I have units with less current and they still lock.  The HP
 manual also says to measure the voltage at the output of the tube with a
 100
 MegOhm or higher input impedance DMM.  If yours is less, that may
 artificially lower your values.

 EOL of the tube is a multifactor issue, including Signal to Noise ratio and
 the 'useful signal current' to 'background current' ratio.  The 'background
 current' is what you see with no RF signal applied to the tube.  Have you
 measured that?  A ratio of 1 is EOL per the HP manual.  If yours is about
 4.5 nA, as suggested by the 'off peak' values shown, or less, you still
 have
 a useful signal and, hopefully, a useful tube.

 I'd recommend continuing with the repair.

 Good luck.

 Joe

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Ed Palmer
 Sent: Monday, May 05, 2014 12:46 AM
 To: Discussion of precise time and frequency measurement
 Subject: [time-nuts] Datum 4065A Cs Tube Response

 I'm playing with my first Cs standard.  It's a Datum 4065A which appears to
 have a dead STEL-1173 synthesizer.  Before I put too much effort into
 replacing that, I thought I'd check the tube and see if it has any life
 left.  I've attached a chart showing the response of the central peak.

 My methodology was similar to TVB's as shown here:
 http://www.leapsecond.com/pages/cspeak except that I measured the tube
 output directly with a digital voltmeter.  The system is reporting wildly
 varying levels for the beam current so I didn't want to use any of it's
 circuitry.

 Does this look like a usable tube?  Healthy or on it's last legs? What
 response levels are typical for a Datum 7504A tube?  I see that these
 levels
 are somewhat lower than those shown on leapsecond for the 5061A tube, but
 that could just be the specifics of the measurement.

 Thanks,

 Ed


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Datum 4065A Cs Tube Response

2014-05-05 Thread Magnus Danielson

Hi Ed,

On 05/05/2014 07:45 AM, Ed Palmer wrote:

I'm playing with my first Cs standard.  It's a Datum 4065A which appears
to have a dead STEL-1173 synthesizer.  Before I put too much effort into
replacing that, I thought I'd check the tube and see if it has any life
left.  I've attached a chart showing the response of the central peak.


Looks nice.

I had the same issue with one of my cesiums, so I got lucky and got a 
STEL 1173 in 44 pin PLCC, replaced it and then it worked.


What happens is that the output drivers for the STEL 1173 dies one after 
another and eventually you end up noise.



My methodology was similar to TVB's as shown here:
http://www.leapsecond.com/pages/cspeak except that I measured the tube
output directly with a digital voltmeter.  The system is reporting
wildly varying levels for the beam current so I didn't want to use any
of it's circuitry.


I found the suitable test-point and multiplied with the network 
analyzers output signal as replacing the 12.6 MHz signal, and then the 
output of that into the network analyzer and that way got a nice way of 
checking the tube.



Does this look like a usable tube?  Healthy or on it's last legs? What
response levels are typical for a Datum 7504A tube?  I see that these
levels are somewhat lower than those shown on leapsecond for the 5061A
tube, but that could just be the specifics of the measurement.


Well, looks worth trying replaing the STEL for.

Good luck and hope it works out for you!

Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Datum 4065A Cs Tube Response

2014-05-05 Thread Ed Palmer

Hi Joe,

I didn't realize that the info in the 5061A manual was somewhat 
'generic'.  Thanks for the pointer.


On 5/5/2014 6:20 AM, J. L. Trantham wrote:

Ed,

If I have the math correct, and you are measuring the voltage to ground
through a 10 MegOhm input impedance DMM, you have about 7.5 nA beam current
which seems a bit low compared to what I remember of the HP 5061A.  However,
you still have a definable 'peak' with a 'peak to valley' voltage of about
60 mV or a 'useful signal current' of about 6 nA.  If your unit's circuitry
can properly amplify that and keep it a clean signal, it should work.
However, I would recommend setting the OCXO precisely on frequency with a
GPSDO before trying to close the loop and 'locking' the signal to the CS
tube.  It will dramatically lower the work load of getting everything
adjusted properly, particularly in a setting of low beam current.


Easier said than done.  This oscillator has no manual frequency 
adjustment and if there's an electrical one, I haven't found it yet.  
But the oscillator's response seems very vague as though it's just kind 
of wandering around so I think there's something not right there.  It's 
likely related to the bad synthesizer chip.



Somehow, the value of 40 nA sticks in my mind from the 5061A.  The 5061A
manual says end of life of the HP CS tube is a peak beam current of 8 nA or
less.  However, I have units with less current and they still lock.  The HP
manual also says to measure the voltage at the output of the tube with a 100
MegOhm or higher input impedance DMM.  If yours is less, that may
artificially lower your values.


I reconfigured my test setup to use a 100 Mohm meter and converted the 
results into nanoamps.  The 10 Mohm meter was loading down the tube a 
bit.  The new graph looks somewhat better.



EOL of the tube is a multifactor issue, including Signal to Noise ratio and
the 'useful signal current' to 'background current' ratio.  The 'background
current' is what you see with no RF signal applied to the tube.  Have you
measured that?  A ratio of 1 is EOL per the HP manual.  If yours is about
4.5 nA, as suggested by the 'off peak' values shown, or less, you still have
a useful signal and, hopefully, a useful tube.


The ratio of signal level to background level looks quite good.  The 
background current is only 1.5 na for a ratio of 6:1.  I checked that 
measurement twice because it looked suspiciously good.



I'd recommend continuing with the repair.


I will, but it could be world's slowest repair as I try to find one of 
those synthesizer chips.  I've seen places that claim to have them, but 
only in the 48 pin DIP package rather than the 44 pin PLCC.  I REALLY 
don't want to do that conversion!


In any case, I have many other toys to amuse me! :)

Thanks for all the tips!

Ed


Good luck.

Joe

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Ed Palmer
Sent: Monday, May 05, 2014 12:46 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Datum 4065A Cs Tube Response

I'm playing with my first Cs standard.  It's a Datum 4065A which appears to
have a dead STEL-1173 synthesizer.  Before I put too much effort into
replacing that, I thought I'd check the tube and see if it has any life
left.  I've attached a chart showing the response of the central peak.

My methodology was similar to TVB's as shown here:
http://www.leapsecond.com/pages/cspeak except that I measured the tube
output directly with a digital voltmeter.  The system is reporting wildly
varying levels for the beam current so I didn't want to use any of it's
circuitry.

Does this look like a usable tube?  Healthy or on it's last legs? What
response levels are typical for a Datum 7504A tube?  I see that these levels
are somewhat lower than those shown on leapsecond for the 5061A tube, but
that could just be the specifics of the measurement.

Thanks,

Ed


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Datum 4065A Cs Tube Response

2014-05-05 Thread Ed Palmer
I know very little about today's DDS chips, but I think that emulating 
the STEL-1173 would be a challenge.  It provides 48 bit frequency 
resolution.  I counted 13 different frequencies that are used to monitor 
the signal to make sure that it's on frequency. Based on the manual, 
they were shifting it many times per second. I'm not surprised the chip 
died - they worked it to death!


Ed

On 5/5/2014 6:49 AM, paul swed wrote:

I will agree with Joe. I have a CS tube thats darn near impossible to read
the beam current and yet it still locks. That truly amazes me. I seem to
recall other comments ages ago about that chip failing. There should be a
way to emulate it these days with all of the DDS chips and such that are
available.
Good luck.
Regards
Paul.
WB8TSL


On Mon, May 5, 2014 at 8:20 AM, J. L. Trantham jlt...@att.net wrote:


Ed,

If I have the math correct, and you are measuring the voltage to ground
through a 10 MegOhm input impedance DMM, you have about 7.5 nA beam current
which seems a bit low compared to what I remember of the HP 5061A.
  However,
you still have a definable 'peak' with a 'peak to valley' voltage of about
60 mV or a 'useful signal current' of about 6 nA.  If your unit's circuitry
can properly amplify that and keep it a clean signal, it should work.
However, I would recommend setting the OCXO precisely on frequency with a
GPSDO before trying to close the loop and 'locking' the signal to the CS
tube.  It will dramatically lower the work load of getting everything
adjusted properly, particularly in a setting of low beam current.

Somehow, the value of 40 nA sticks in my mind from the 5061A.  The 5061A
manual says end of life of the HP CS tube is a peak beam current of 8 nA or
less.  However, I have units with less current and they still lock.  The HP
manual also says to measure the voltage at the output of the tube with a
100
MegOhm or higher input impedance DMM.  If yours is less, that may
artificially lower your values.

EOL of the tube is a multifactor issue, including Signal to Noise ratio and
the 'useful signal current' to 'background current' ratio.  The 'background
current' is what you see with no RF signal applied to the tube.  Have you
measured that?  A ratio of 1 is EOL per the HP manual.  If yours is about
4.5 nA, as suggested by the 'off peak' values shown, or less, you still
have
a useful signal and, hopefully, a useful tube.

I'd recommend continuing with the repair.

Good luck.

Joe

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Ed Palmer
Sent: Monday, May 05, 2014 12:46 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Datum 4065A Cs Tube Response

I'm playing with my first Cs standard.  It's a Datum 4065A which appears to
have a dead STEL-1173 synthesizer.  Before I put too much effort into
replacing that, I thought I'd check the tube and see if it has any life
left.  I've attached a chart showing the response of the central peak.

My methodology was similar to TVB's as shown here:
http://www.leapsecond.com/pages/cspeak except that I measured the tube
output directly with a digital voltmeter.  The system is reporting wildly
varying levels for the beam current so I didn't want to use any of it's
circuitry.

Does this look like a usable tube?  Healthy or on it's last legs? What
response levels are typical for a Datum 7504A tube?  I see that these
levels
are somewhat lower than those shown on leapsecond for the 5061A tube, but
that could just be the specifics of the measurement.

Thanks,

Ed

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Datum 4065A Cs Tube Response

2014-05-05 Thread Tom Knox
Hi 
Is this the right chip:
http://www.oemstrade.com/search/STEL-1173%252FCL/
At $65 it is pricy but a bargain if it fixes your clock.
Thomas Knox



 Date: Mon, 5 May 2014 15:14:16 -0600
 From: ed_pal...@sasktel.net
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] Datum 4065A Cs Tube Response
 
 I know very little about today's DDS chips, but I think that emulating 
 the STEL-1173 would be a challenge.  It provides 48 bit frequency 
 resolution.  I counted 13 different frequencies that are used to monitor 
 the signal to make sure that it's on frequency. Based on the manual, 
 they were shifting it many times per second. I'm not surprised the chip 
 died - they worked it to death!
 
 Ed
 
 On 5/5/2014 6:49 AM, paul swed wrote:
  I will agree with Joe. I have a CS tube thats darn near impossible to read
  the beam current and yet it still locks. That truly amazes me. I seem to
  recall other comments ages ago about that chip failing. There should be a
  way to emulate it these days with all of the DDS chips and such that are
  available.
  Good luck.
  Regards
  Paul.
  WB8TSL
 
 
  On Mon, May 5, 2014 at 8:20 AM, J. L. Trantham jlt...@att.net wrote:
 
  Ed,
 
  If I have the math correct, and you are measuring the voltage to ground
  through a 10 MegOhm input impedance DMM, you have about 7.5 nA beam current
  which seems a bit low compared to what I remember of the HP 5061A.
However,
  you still have a definable 'peak' with a 'peak to valley' voltage of about
  60 mV or a 'useful signal current' of about 6 nA.  If your unit's circuitry
  can properly amplify that and keep it a clean signal, it should work.
  However, I would recommend setting the OCXO precisely on frequency with a
  GPSDO before trying to close the loop and 'locking' the signal to the CS
  tube.  It will dramatically lower the work load of getting everything
  adjusted properly, particularly in a setting of low beam current.
 
  Somehow, the value of 40 nA sticks in my mind from the 5061A.  The 5061A
  manual says end of life of the HP CS tube is a peak beam current of 8 nA or
  less.  However, I have units with less current and they still lock.  The HP
  manual also says to measure the voltage at the output of the tube with a
  100
  MegOhm or higher input impedance DMM.  If yours is less, that may
  artificially lower your values.
 
  EOL of the tube is a multifactor issue, including Signal to Noise ratio and
  the 'useful signal current' to 'background current' ratio.  The 'background
  current' is what you see with no RF signal applied to the tube.  Have you
  measured that?  A ratio of 1 is EOL per the HP manual.  If yours is about
  4.5 nA, as suggested by the 'off peak' values shown, or less, you still
  have
  a useful signal and, hopefully, a useful tube.
 
  I'd recommend continuing with the repair.
 
  Good luck.
 
  Joe
 
  -Original Message-
  From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
  Behalf Of Ed Palmer
  Sent: Monday, May 05, 2014 12:46 AM
  To: Discussion of precise time and frequency measurement
  Subject: [time-nuts] Datum 4065A Cs Tube Response
 
  I'm playing with my first Cs standard.  It's a Datum 4065A which appears to
  have a dead STEL-1173 synthesizer.  Before I put too much effort into
  replacing that, I thought I'd check the tube and see if it has any life
  left.  I've attached a chart showing the response of the central peak.
 
  My methodology was similar to TVB's as shown here:
  http://www.leapsecond.com/pages/cspeak except that I measured the tube
  output directly with a digital voltmeter.  The system is reporting wildly
  varying levels for the beam current so I didn't want to use any of it's
  circuitry.
 
  Does this look like a usable tube?  Healthy or on it's last legs? What
  response levels are typical for a Datum 7504A tube?  I see that these
  levels
  are somewhat lower than those shown on leapsecond for the 5061A tube, but
  that could just be the specifics of the measurement.
 
  Thanks,
 
  Ed
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
  
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Datum 4065A Cs Tube Response

2014-05-05 Thread Magnus Danielson

Ed,

On 05/05/2014 10:56 PM, Ed Palmer wrote:

Hi Joe,

I didn't realize that the info in the 5061A manual was somewhat
'generic'.  Thanks for the pointer.

On 5/5/2014 6:20 AM, J. L. Trantham wrote:

Ed,

If I have the math correct, and you are measuring the voltage to ground
through a 10 MegOhm input impedance DMM, you have about 7.5 nA beam
current
which seems a bit low compared to what I remember of the HP 5061A.
However,
you still have a definable 'peak' with a 'peak to valley' voltage of
about
60 mV or a 'useful signal current' of about 6 nA.  If your unit's
circuitry
can properly amplify that and keep it a clean signal, it should work.
However, I would recommend setting the OCXO precisely on frequency with a
GPSDO before trying to close the loop and 'locking' the signal to the CS
tube.  It will dramatically lower the work load of getting everything
adjusted properly, particularly in a setting of low beam current.


Easier said than done.  This oscillator has no manual frequency
adjustment and if there's an electrical one, I haven't found it yet. But
the oscillator's response seems very vague as though it's just kind of
wandering around so I think there's something not right there.  It's
likely related to the bad synthesizer chip.


The 4065 use the STEL DDS to jump around in an interesting patten to 
measure the Zeeman position of the main lobe and of the first side-lobes 
to steer the C-field in a separate control-loop. This means that the 
C-field is completely under CPU control, and no manual trimming. I think 
you can disable this and steer it manual over the serial interface, but 
I don't recall the details.


The 6 side-lobes separate with heavy dependence on the C-field, so 
servo-loop on their position relative the main lobe works really well.
With the C-field servo-controlled like this, the C-field shift of the 
main lobe, which is a much weaker dependence and also one of the major 
aspects in selecting Cesiums in the first place, a more accurate 
Si-realization becomes possible, as well as with much better repeat in 
realization as the arbitrary C-field shift can be almost completely 
canceled.


As the Zeeman steering and centering is done, locking to Rabi feature is 
done in a similar sense, by measuring multiple points. In fact, this is 
where it spends most of it's time and the for some of the TDM slots jump 
around for the Zeeman measuring points every once in a while.


All that is described in a patent, which then describes this being 
driven by the 1802 CPU, but that was later replaced by the 6800-series.
Precision time with whopping 8-bitters. It gives the same warm and fuzzy 
feeling as rubidium and cesiums with 741s in PSUs and control-loops :)


So, this core looks pretty brainless with not as much fun trimmers as 
the 5061A for instance, but it has quite better brains, it's just that 
the faded STEL 1173 drivers makes it blind.


The STEL 1173 DDS functionality isn't very hard to copy good enough, but 
it is still some work do be done. I don't remember where the datasheet 
went for it.



Somehow, the value of 40 nA sticks in my mind from the 5061A.  The 5061A
manual says end of life of the HP CS tube is a peak beam current of 8
nA or
less.  However, I have units with less current and they still lock.
The HP
manual also says to measure the voltage at the output of the tube with
a 100
MegOhm or higher input impedance DMM.  If yours is less, that may
artificially lower your values.


I reconfigured my test setup to use a 100 Mohm meter and converted the
results into nanoamps.  The 10 Mohm meter was loading down the tube a
bit.  The new graph looks somewhat better.


Great.


EOL of the tube is a multifactor issue, including Signal to Noise
ratio and
the 'useful signal current' to 'background current' ratio.  The
'background
current' is what you see with no RF signal applied to the tube.  Have you
measured that?  A ratio of 1 is EOL per the HP manual.  If yours is about
4.5 nA, as suggested by the 'off peak' values shown, or less, you
still have
a useful signal and, hopefully, a useful tube.


The ratio of signal level to background level looks quite good.  The
background current is only 1.5 na for a ratio of 6:1.  I checked that
measurement twice because it looked suspiciously good.


Great.


I'd recommend continuing with the repair.


I will, but it could be world's slowest repair as I try to find one of
those synthesizer chips.  I've seen places that claim to have them, but
only in the 48 pin DIP package rather than the 44 pin PLCC.  I REALLY
don't want to do that conversion!

In any case, I have many other toys to amuse me! :)


While you wait, don't waste that tube.


Thanks for all the tips!


I wish I had a 44 pin PLCC variant of the STEL 1173 lying around to help 
you out.


Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow 

Re: [time-nuts] Datum 4065A Cs Tube Response

2014-05-05 Thread Ed Palmer

Hi Tom,

Yes, that's the right chip, but the 'L' at the end means it's the DIP 
package.  The PLCC package has an 'M' instead of the 'L'.  I also found 
a lot of sellers on a far east site who have the DIP package (probably 
used chips pulled from sockets), but none who have the PLCC package.  
The price is also a lot less than $65, but of course, you're taking a 
risk that you might be buying a dead chip.


Ed

On 5/5/2014 3:18 PM, Tom Knox wrote:

Hi
Is this the right chip:
http://www.oemstrade.com/search/STEL-1173%252FCL/
At $65 it is pricy but a bargain if it fixes your clock.
Thomas Knox




Date: Mon, 5 May 2014 15:14:16 -0600
From: ed_pal...@sasktel.net
To: time-nuts@febo.com
Subject: Re: [time-nuts] Datum 4065A Cs Tube Response

I know very little about today's DDS chips, but I think that emulating
the STEL-1173 would be a challenge.  It provides 48 bit frequency
resolution.  I counted 13 different frequencies that are used to monitor
the signal to make sure that it's on frequency. Based on the manual,
they were shifting it many times per second. I'm not surprised the chip
died - they worked it to death!

Ed

On 5/5/2014 6:49 AM, paul swed wrote:

I will agree with Joe. I have a CS tube thats darn near impossible to read
the beam current and yet it still locks. That truly amazes me. I seem to
recall other comments ages ago about that chip failing. There should be a
way to emulate it these days with all of the DDS chips and such that are
available.
Good luck.
Regards
Paul.
WB8TSL


On Mon, May 5, 2014 at 8:20 AM, J. L. Trantham jlt...@att.net wrote:


Ed,

If I have the math correct, and you are measuring the voltage to ground
through a 10 MegOhm input impedance DMM, you have about 7.5 nA beam current
which seems a bit low compared to what I remember of the HP 5061A.
   However,
you still have a definable 'peak' with a 'peak to valley' voltage of about
60 mV or a 'useful signal current' of about 6 nA.  If your unit's circuitry
can properly amplify that and keep it a clean signal, it should work.
However, I would recommend setting the OCXO precisely on frequency with a
GPSDO before trying to close the loop and 'locking' the signal to the CS
tube.  It will dramatically lower the work load of getting everything
adjusted properly, particularly in a setting of low beam current.

Somehow, the value of 40 nA sticks in my mind from the 5061A.  The 5061A
manual says end of life of the HP CS tube is a peak beam current of 8 nA or
less.  However, I have units with less current and they still lock.  The HP
manual also says to measure the voltage at the output of the tube with a
100
MegOhm or higher input impedance DMM.  If yours is less, that may
artificially lower your values.

EOL of the tube is a multifactor issue, including Signal to Noise ratio and
the 'useful signal current' to 'background current' ratio.  The 'background
current' is what you see with no RF signal applied to the tube.  Have you
measured that?  A ratio of 1 is EOL per the HP manual.  If yours is about
4.5 nA, as suggested by the 'off peak' values shown, or less, you still
have
a useful signal and, hopefully, a useful tube.

I'd recommend continuing with the repair.

Good luck.

Joe

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Ed Palmer
Sent: Monday, May 05, 2014 12:46 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Datum 4065A Cs Tube Response

I'm playing with my first Cs standard.  It's a Datum 4065A which appears to
have a dead STEL-1173 synthesizer.  Before I put too much effort into
replacing that, I thought I'd check the tube and see if it has any life
left.  I've attached a chart showing the response of the central peak.

My methodology was similar to TVB's as shown here:
http://www.leapsecond.com/pages/cspeak except that I measured the tube
output directly with a digital voltmeter.  The system is reporting wildly
varying levels for the beam current so I didn't want to use any of it's
circuitry.

Does this look like a usable tube?  Healthy or on it's last legs? What
response levels are typical for a Datum 7504A tube?  I see that these
levels
are somewhat lower than those shown on leapsecond for the 5061A tube, but
that could just be the specifics of the measurement.

Thanks,

Ed


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Datum 4065A Cs Tube Response

2014-05-05 Thread Brooke Clarke

Hi Ed:

Some links to the related patents are on my 4060  4065 web page:
http://www.prc68.com/I/FTS4060.shtml#Pat

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.end2partygovernment.com/2012Issues.html

Magnus Danielson wrote:

Ed,

On 05/05/2014 10:56 PM, Ed Palmer wrote:

Hi Joe,

I didn't realize that the info in the 5061A manual was somewhat
'generic'.  Thanks for the pointer.

On 5/5/2014 6:20 AM, J. L. Trantham wrote:

Ed,

If I have the math correct, and you are measuring the voltage to ground
through a 10 MegOhm input impedance DMM, you have about 7.5 nA beam
current
which seems a bit low compared to what I remember of the HP 5061A.
However,
you still have a definable 'peak' with a 'peak to valley' voltage of
about
60 mV or a 'useful signal current' of about 6 nA.  If your unit's
circuitry
can properly amplify that and keep it a clean signal, it should work.
However, I would recommend setting the OCXO precisely on frequency with a
GPSDO before trying to close the loop and 'locking' the signal to the CS
tube.  It will dramatically lower the work load of getting everything
adjusted properly, particularly in a setting of low beam current.


Easier said than done.  This oscillator has no manual frequency
adjustment and if there's an electrical one, I haven't found it yet. But
the oscillator's response seems very vague as though it's just kind of
wandering around so I think there's something not right there. It's
likely related to the bad synthesizer chip.


The 4065 use the STEL DDS to jump around in an interesting patten to measure the Zeeman position of the main lobe and 
of the first side-lobes to steer the C-field in a separate control-loop. This means that the C-field is completely 
under CPU control, and no manual trimming. I think you can disable this and steer it manual over the serial interface, 
but I don't recall the details.


The 6 side-lobes separate with heavy dependence on the C-field, so servo-loop on their position relative the main lobe 
works really well.
With the C-field servo-controlled like this, the C-field shift of the main lobe, which is a much weaker dependence and 
also one of the major aspects in selecting Cesiums in the first place, a more accurate Si-realization becomes 
possible, as well as with much better repeat in realization as the arbitrary C-field shift can be almost completely 
canceled.


As the Zeeman steering and centering is done, locking to Rabi feature is done in a similar sense, by measuring 
multiple points. In fact, this is where it spends most of it's time and the for some of the TDM slots jump around for 
the Zeeman measuring points every once in a while.


All that is described in a patent, which then describes this being driven by the 1802 CPU, but that was later replaced 
by the 6800-series.
Precision time with whopping 8-bitters. It gives the same warm and fuzzy feeling as rubidium and cesiums with 741s in 
PSUs and control-loops :)


So, this core looks pretty brainless with not as much fun trimmers as the 5061A for instance, but it has quite better 
brains, it's just that the faded STEL 1173 drivers makes it blind.


The STEL 1173 DDS functionality isn't very hard to copy good enough, but it is still some work do be done. I don't 
remember where the datasheet went for it.



Somehow, the value of 40 nA sticks in my mind from the 5061A.  The 5061A
manual says end of life of the HP CS tube is a peak beam current of 8
nA or
less.  However, I have units with less current and they still lock.
The HP
manual also says to measure the voltage at the output of the tube with
a 100
MegOhm or higher input impedance DMM.  If yours is less, that may
artificially lower your values.


I reconfigured my test setup to use a 100 Mohm meter and converted the
results into nanoamps.  The 10 Mohm meter was loading down the tube a
bit.  The new graph looks somewhat better.


Great.


EOL of the tube is a multifactor issue, including Signal to Noise
ratio and
the 'useful signal current' to 'background current' ratio. The
'background
current' is what you see with no RF signal applied to the tube.  Have you
measured that?  A ratio of 1 is EOL per the HP manual.  If yours is about
4.5 nA, as suggested by the 'off peak' values shown, or less, you
still have
a useful signal and, hopefully, a useful tube.


The ratio of signal level to background level looks quite good. The
background current is only 1.5 na for a ratio of 6:1.  I checked that
measurement twice because it looked suspiciously good.


Great.


I'd recommend continuing with the repair.


I will, but it could be world's slowest repair as I try to find one of
those synthesizer chips.  I've seen places that claim to have them, but
only in the 48 pin DIP package rather than the 44 pin PLCC.  I REALLY
don't want to do that conversion!

In any case, I have many other toys to amuse me! :)


While you wait, don't waste that tube.


Thanks for all the tips!


I wish I had a 44 pin PLCC variant of the STEL 1173 

Re: [time-nuts] Datum 4065A Cs Tube Response

2014-05-05 Thread Ed Palmer

Hi Magnus,

On 5/5/2014 3:30 PM, Magnus Danielson wrote:

Ed,

On 05/05/2014 10:56 PM, Ed Palmer wrote:

Hi Joe,

I didn't realize that the info in the 5061A manual was somewhat
'generic'.  Thanks for the pointer.

On 5/5/2014 6:20 AM, J. L. Trantham wrote:

Ed,

If I have the math correct, and you are measuring the voltage to 
ground through a 10 MegOhm input impedance DMM, you have about 7.5 
nA beam current which seems a bit low compared to what I remember of 
the HP 5061A. However, you still have a definable 'peak' with a 
'peak to valley' voltage of
about 60 mV or a 'useful signal current' of about 6 nA.  If your 
unit's circuitry can properly amplify that and keep it a clean 
signal, it should work. However, I would recommend setting the OCXO 
precisely on frequency with a GPSDO before trying to close the loop 
and 'locking' the signal to the CS tube.  It will dramatically lower 
the work load of getting everything adjusted properly, particularly 
in a setting of low beam current.


Easier said than done.  This oscillator has no manual frequency 
adjustment and if there's an electrical one, I haven't found it yet. 
But the oscillator's response seems very vague as though it's just 
kind of wandering around so I think there's something not right 
there. It's likely related to the bad synthesizer chip.


The 4065 use the STEL DDS to jump around in an interesting patten to 
measure the Zeeman position of the main lobe and of the first 
side-lobes to steer the C-field in a separate control-loop. This means 
that the C-field is completely under CPU control, and no manual 
trimming. I think you can disable this and steer it manual over the 
serial interface, but I don't recall the details.


The 6 side-lobes separate with heavy dependence on the C-field, so 
servo-loop on their position relative the main lobe works really well.
With the C-field servo-controlled like this, the C-field shift of the 
main lobe, which is a much weaker dependence and also one of the major 
aspects in selecting Cesiums in the first place, a more accurate 
Si-realization becomes possible, as well as with much better repeat in 
realization as the arbitrary C-field shift can be almost completely 
canceled.


I was talking about adjusting the EFC for the OCXO, not the C-field, 
although I agree with what you've said.  I hope that there's some kind 
of adjustment to compensate for crystal aging, but maybe not. Obviously, 
this tuning loop is also under CPU control and it appears that the 
failed DDS chip results in no RF to the tube at the right frequency so 
no signal for the main tuning loop to use to discipline the OCXO.  
Hence, the OCXO kind of wanders around.  Even replacing the DDS with a 
fixed synthesizer wasn't enough to coax the CPU to discipline the OCXO.


As the Zeeman steering and centering is done, locking to Rabi feature 
is done in a similar sense, by measuring multiple points. In fact, 
this is where it spends most of it's time and the for some of the TDM 
slots jump around for the Zeeman measuring points every once in a while.


All that is described in a patent, which then describes this being 
driven by the 1802 CPU, but that was later replaced by the 6800-series.
Precision time with whopping 8-bitters. It gives the same warm and 
fuzzy feeling as rubidium and cesiums with 741s in PSUs and 
control-loops :)


So, this core looks pretty brainless with not as much fun trimmers as 
the 5061A for instance, but it has quite better brains, it's just that 
the faded STEL 1173 drivers makes it blind.


The STEL 1173 DDS functionality isn't very hard to copy good enough, 
but it is still some work do be done. I don't remember where the 
datasheet went for it.


It wasn't hard to find the datasheet.  I was surprised to find that 
Intel took over the part.  I don't know how long they carried it, but 
the datasheet was from them.




Somehow, the value of 40 nA sticks in my mind from the 5061A.  The 
5061A manual says end of life of the HP CS tube is a peak beam 
current of 8 nA or less.  However, I have units with less current 
and they still lock. The HP manual also says to measure the voltage 
at the output of the tube with a 100 MegOhm or higher input 
impedance DMM.  If yours is less, that may artificially lower your 
values.


I reconfigured my test setup to use a 100 Mohm meter and converted 
the results into nanoamps.  The 10 Mohm meter was loading down the 
tube a bit.  The new graph looks somewhat better.


Great.

EOL of the tube is a multifactor issue, including Signal to Noise 
ratio and the 'useful signal current' to 'background current' ratio. 
The 'background current' is what you see with no RF signal applied 
to the tube.  Have you measured that?  A ratio of 1 is EOL per the 
HP manual.  If yours is about 4.5 nA, as suggested by the 'off peak' 
values shown, or less, you still have a useful signal and, 
hopefully, a useful tube.


The ratio of signal level to background level looks quite good. The 

Re: [time-nuts] Datum 4065A Cs Tube Response

2014-05-05 Thread EWKehren
The original price of the STEL1173CL was $ 125 in quantity 10 so $ 65 is a  
bargain.  I have no idea as to how it is used in the 4065A but if it is  
limited to 13 frequencies it may be worth wile to take a look at the AD 9913 
it  is capable to produce very precise frequencies exceeding its 32 bits, tvb 
did  run tests on a board we build and I have extra boards available. Has a 
PIC on it  that may help to emulate. Most likely will it not be the only 
STEL  failure.
Bert Kehren
 
 
In a message dated 5/5/2014 5:15:01 P.M. Eastern Daylight Time,  
ed_pal...@sasktel.net writes:

I know very  little about today's DDS chips, but I think that emulating 
the STEL-1173  would be a challenge.  It provides 48 bit frequency 
resolution.   I counted 13 different frequencies that are used to monitor 
the signal to  make sure that it's on frequency. Based on the manual, 
they were shifting  it many times per second. I'm not surprised the chip 
died - they worked it  to death!

Ed

On 5/5/2014 6:49 AM, paul swed wrote:
 I  will agree with Joe. I have a CS tube thats darn near impossible to  
read
 the beam current and yet it still locks. That truly amazes me. I  seem to
 recall other comments ages ago about that chip failing. There  should be a
 way to emulate it these days with all of the DDS chips and  such that are
 available.
 Good luck.
 Regards
  Paul.
 WB8TSL


 On Mon, May 5, 2014 at 8:20 AM,  J. L. Trantham jlt...@att.net wrote:

  Ed,

 If I have the math correct, and you are measuring  the voltage to ground
 through a 10 MegOhm input impedance DMM, you  have about 7.5 nA beam 
current
 which seems a bit low compared to  what I remember of the HP 5061A.
   However,
  you still have a definable 'peak' with a 'peak to valley' voltage of  
about
 60 mV or a 'useful signal current' of about 6 nA.  If  your unit's 
circuitry
 can properly amplify that and keep it a  clean signal, it should work.
 However, I would recommend setting  the OCXO precisely on frequency with 
a
 GPSDO before trying to  close the loop and 'locking' the signal to the CS
 tube.  It  will dramatically lower the work load of getting everything
  adjusted properly, particularly in a setting of low beam  current.

 Somehow, the value of 40 nA sticks in my mind  from the 5061A.  The 5061A
 manual says end of life of the HP  CS tube is a peak beam current of 8 
nA or
 less.  However, I  have units with less current and they still lock.  
The HP
  manual also says to measure the voltage at the output of the tube with  
a
 100
 MegOhm or higher input impedance DMM.  If  yours is less, that may
 artificially lower your  values.

 EOL of the tube is a multifactor issue,  including Signal to Noise ratio 
and
 the 'useful signal current' to  'background current' ratio.  The 
'background
 current' is what  you see with no RF signal applied to the tube.  Have 
you
  measured that?  A ratio of 1 is EOL per the HP manual.  If yours is  
about
 4.5 nA, as suggested by the 'off peak' values shown, or  less, you still
 have
 a useful signal and, hopefully, a  useful tube.

 I'd recommend continuing with the  repair.

 Good luck.

  Joe

 -Original Message-
 From:  time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
  Behalf Of Ed Palmer
 Sent: Monday, May 05, 2014 12:46  AM
 To: Discussion of precise time and frequency  measurement
 Subject: [time-nuts] Datum 4065A Cs Tube  Response

 I'm playing with my first Cs standard.   It's a Datum 4065A which 
appears to
 have a dead STEL-1173  synthesizer.  Before I put too much effort into
 replacing  that, I thought I'd check the tube and see if it has any life
  left.  I've attached a chart showing the response of the central  peak.

 My methodology was similar to TVB's as shown  here:
 http://www.leapsecond.com/pages/cspeak except that I  measured the tube
 output directly with a digital voltmeter.   The system is reporting 
wildly
 varying levels for the beam current  so I didn't want to use any of it's
  circuitry.

 Does this look like a usable tube?   Healthy or on it's last legs? What
 response levels are typical for  a Datum 7504A tube?  I see that these
 levels
 are  somewhat lower than those shown on leapsecond for the 5061A tube,  
but
 that could just be the specifics of the  measurement.

 Thanks,

  Ed
___
time-nuts mailing  list -- time-nuts@febo.com
To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Datum 4065A Cs Tube Response

2014-05-05 Thread Daniel Mendes


The datasheet can be easily found with google. The functionality is very 
simple, it could be emulated with a CPLD+small memory or with an small 
FPGA (maybe the newer CPLDs from Altera.. they are FPGAs inside... and I 
think they have block RAM). But if you can still buy it in another 
package it would be cheaper and easier to make an adapter and be done 
with it.


Daniel

Em 05/05/2014 18:40, Ed Palmer escreveu:

Hi Tom,

Yes, that's the right chip, but the 'L' at the end means it's the DIP 
package.  The PLCC package has an 'M' instead of the 'L'.  I also 
found a lot of sellers on a far east site who have the DIP package 
(probably used chips pulled from sockets), but none who have the PLCC 
package.  The price is also a lot less than $65, but of course, you're 
taking a risk that you might be buying a dead chip.


Ed

On 5/5/2014 3:18 PM, Tom Knox wrote:

Hi
Is this the right chip:
http://www.oemstrade.com/search/STEL-1173%252FCL/
At $65 it is pricy but a bargain if it fixes your clock.
Thomas Knox




Date: Mon, 5 May 2014 15:14:16 -0600
From: ed_pal...@sasktel.net
To: time-nuts@febo.com
Subject: Re: [time-nuts] Datum 4065A Cs Tube Response

I know very little about today's DDS chips, but I think that emulating
the STEL-1173 would be a challenge.  It provides 48 bit frequency
resolution.  I counted 13 different frequencies that are used to 
monitor

the signal to make sure that it's on frequency. Based on the manual,
they were shifting it many times per second. I'm not surprised the chip
died - they worked it to death!

Ed

On 5/5/2014 6:49 AM, paul swed wrote:
I will agree with Joe. I have a CS tube thats darn near impossible 
to read
the beam current and yet it still locks. That truly amazes me. I 
seem to
recall other comments ages ago about that chip failing. There 
should be a
way to emulate it these days with all of the DDS chips and such 
that are

available.
Good luck.
Regards
Paul.
WB8TSL


On Mon, May 5, 2014 at 8:20 AM, J. L. Trantham jlt...@att.net wrote:


Ed,

If I have the math correct, and you are measuring the voltage to 
ground
through a 10 MegOhm input impedance DMM, you have about 7.5 nA 
beam current

which seems a bit low compared to what I remember of the HP 5061A.
   However,
you still have a definable 'peak' with a 'peak to valley' voltage 
of about
60 mV or a 'useful signal current' of about 6 nA.  If your unit's 
circuitry

can properly amplify that and keep it a clean signal, it should work.
However, I would recommend setting the OCXO precisely on frequency 
with a
GPSDO before trying to close the loop and 'locking' the signal to 
the CS

tube.  It will dramatically lower the work load of getting everything
adjusted properly, particularly in a setting of low beam current.

Somehow, the value of 40 nA sticks in my mind from the 5061A.  The 
5061A
manual says end of life of the HP CS tube is a peak beam current 
of 8 nA or
less.  However, I have units with less current and they still 
lock.  The HP
manual also says to measure the voltage at the output of the tube 
with a

100
MegOhm or higher input impedance DMM.  If yours is less, that may
artificially lower your values.

EOL of the tube is a multifactor issue, including Signal to Noise 
ratio and
the 'useful signal current' to 'background current' ratio.  The 
'background
current' is what you see with no RF signal applied to the tube.  
Have you
measured that?  A ratio of 1 is EOL per the HP manual.  If yours 
is about
4.5 nA, as suggested by the 'off peak' values shown, or less, you 
still

have
a useful signal and, hopefully, a useful tube.

I'd recommend continuing with the repair.

Good luck.

Joe

-Original Message-
From: time-nuts-boun...@febo.com 
[mailto:time-nuts-boun...@febo.com] On

Behalf Of Ed Palmer
Sent: Monday, May 05, 2014 12:46 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Datum 4065A Cs Tube Response

I'm playing with my first Cs standard.  It's a Datum 4065A which 
appears to

have a dead STEL-1173 synthesizer.  Before I put too much effort into
replacing that, I thought I'd check the tube and see if it has any 
life
left.  I've attached a chart showing the response of the central 
peak.


My methodology was similar to TVB's as shown here:
http://www.leapsecond.com/pages/cspeak except that I measured the 
tube
output directly with a digital voltmeter.  The system is reporting 
wildly
varying levels for the beam current so I didn't want to use any of 
it's

circuitry.

Does this look like a usable tube?  Healthy or on it's last legs? 
What

response levels are typical for a Datum 7504A tube?  I see that these
levels
are somewhat lower than those shown on leapsecond for the 5061A 
tube, but

that could just be the specifics of the measurement.

Thanks,

Ed


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions

[time-nuts] Datum 4065A Cs Tube Response

2014-05-04 Thread Ed Palmer
I'm playing with my first Cs standard.  It's a Datum 4065A which appears 
to have a dead STEL-1173 synthesizer.  Before I put too much effort into 
replacing that, I thought I'd check the tube and see if it has any life 
left.  I've attached a chart showing the response of the central peak.


My methodology was similar to TVB's as shown here: 
http://www.leapsecond.com/pages/cspeak except that I measured the tube 
output directly with a digital voltmeter.  The system is reporting 
wildly varying levels for the beam current so I didn't want to use any 
of it's circuitry.


Does this look like a usable tube?  Healthy or on it's last legs? What 
response levels are typical for a Datum 7504A tube?  I see that these 
levels are somewhat lower than those shown on leapsecond for the 5061A 
tube, but that could just be the specifics of the measurement.


Thanks,

Ed