Re: [time-nuts] HP 5065 Meter readings after restoration and adjustments: 2nd Harmonic Problems ?

2017-09-08 Thread tim...@timeok.it

   Hi ,
   these jumps when tapping is normal when the HP5065A is restored to the life 
ater a long storage time, they disappears after a long warmup (weeks).
   Luciano


   Da "time-nuts" time-nuts-boun...@febo.com
   A "Ulf Kylenfall" ulf_...@yahoo.com,"Discussion of precise time and 
frequency measurement" time-nuts@febo.com
   Cc
   Data Fri, 08 Sep 2017 08:49:12 +0000
   Oggetto Re: [time-nuts] HP 5065 Meter readings after restoration and 
adjustments: 2nd Harmonic Problems ?
   
   In message <125529892.249004.1504854237...@mail.yahoo.com>, Ulf Kylenfall 
via time-nuts writes:

   >The meter is not stable and the needle is somewhatnervous. When tapping the 
instrument with a smallscrewdriver, this value jumps up and down.

   If the jumps are always the same place on the scale, it is probably
   a mechanical issue in the meter, if it is all over the scale it
   could be the rotary switch which selects what the meter shows.

   --
   Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
   p...@freebsd.org | TCP/IP since RFC 956
   FreeBSD committer | BSD since 4.3-tahoe
   Never attribute to malice what can adequately be explained by incompetence.
   ___
   time-nuts mailing list -- time-nuts@febo.com
   To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
   and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5065 Meter readings after restoration and adjustments: 2nd Harmonic Problems ? (II)

2017-09-08 Thread Poul-Henning Kamp

In message <881182040.5268197.1504868987...@mail.yahoo.com>, Ulf Kylenfall via 
time-nuts writes:
>

>2nd Harmonic level has gone down from 18 to 14 in a few hours.Needle is 
>nervous in the way that it wiggles about one needle-widthin a random but slow 
>pace.
>All other readings are stable.
>Rotating the knob back and forth restores the previous readings.
>I am considering a fault in the 137 MHz circuitry.

It can also be the synth which bounces up and down in frequency if the 
delay-line has drifted.


-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
p...@freebsd.org | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP 5065 Meter readings after restoration and adjustments: 2nd Harmonic Problems ? (II)

2017-09-08 Thread Ulf Kylenfall via time-nuts

2nd Harmonic level has gone down from 18 to 14 in a few hours.Needle is nervous 
in the way that it wiggles about one needle-widthin a random but slow pace.
All other readings are stable.
Rotating the knob back and forth restores the previous readings.
I am considering a fault in the 137 MHz circuitry.
/Ulf

    
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5065 Meter readings after restoration and adjustments: 2nd Harmonic Problems ?

2017-09-08 Thread John Miles
> My intention is to leave the instrument on for severalweeks in order to see
> if it is stable. Right nowmy concern is wether the 2nd harmonic
> readingindicates a wear out of the photo cell.
> When repairing the RFV Cavity, all components weretaken out and carefully
> examined. There wereno dimmed or darkened glass surfaces
> whatsoever.Apart from repairing the TX heater, the only other thingwas
> replacing the 1K something resistor in the TXassembly.

You've checked the ESR of all the axial electrolytic caps, right?  If not, it's 
quite unlikely that they're all good.  

-- john, KE5FX
Miles Design LLC

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5065 Meter readings after restoration and adjustments: 2nd Harmonic Problems ?

2017-09-08 Thread Poul-Henning Kamp

In message <125529892.249004.1504854237...@mail.yahoo.com>, Ulf Kylenfall via 
time-nuts writes:

>The meter is not stable and the needle is somewhatnervous. When tapping the 
>instrument with a smallscrewdriver, this value jumps up and down.

If the jumps are always the same place on the scale, it is probably
a mechanical issue in the meter, if it is all over the scale it
could be the rotary switch which selects what the meter shows.

-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
p...@freebsd.org | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP 5065 Meter readings after restoration and adjustments: 2nd Harmonic Problems ?

2017-09-08 Thread Ulf Kylenfall via time-nuts
 Having repaired an old 5065 with help from documentson the KO4BB and this 
list, I took it to my workat Onsala Space Observatory where we have two 
H-Masers in order to fine tune it.
Initially, the meter readings were close to those listed insidethe protective 
lid, dated 1987 when the instrument was decomissioned.
Supply = 40Lamp Oven = 24Cell Oven = 28Osc Oven = 48Photo I = 305 MHz = 
38Control = 0 (After adjustments)Error = 02nd harmonic 18 (was 30 
initially)1MHz/100 kHz = 40
Because it is an early instrument, it did not featurethe thermoelectric device 
that is recommended forformatting the PhotoCell I used a hair dryer to format 
itfollowing the advice of Corby Dawson. The 2nd Harmonicvalue was initially at 
30, but now it has dropped to 18.
"ON"-time is now ~ 24 hours.
The meter is not stable and the needle is somewhatnervous. When tapping the 
instrument with a smallscrewdriver, this value jumps up and down.
The instrument is still locked, and after adjusting theSynthesizer & Magnetic 
field, comparing the 5MHzoutput to that of the H-Maser indicates a phase 
shiftchange of approx. 0.1 degree in 30 seconds. 
I did the alignment yesterday, and 12 hours later the phaseshift "velocity" has 
not changed. I also tooka phase noise plot using a Keysight PXA but I do 
notknow if the plot is a "true" one. There seems to bea pedestal within 10 Hz 
of the 5 MHz signal after whichthe PN drops significantly. No power line 
bumpsor other spurii is visible.
My intention is to leave the instrument on for severalweeks in order to see if 
it is stable. Right nowmy concern is wether the 2nd harmonic readingindicates a 
wear out of the photo cell.
When repairing the RFV Cavity, all components weretaken out and carefully 
examined. There wereno dimmed or darkened glass surfaces whatsoever.Apart from 
repairing the TX heater, the only other thingwas replacing the 1K something 
resistor in the TXassembly.
Comments invited.