Re: [time-nuts] Morion MV89A Repair

2012-02-27 Thread Roberto Barrios
Hi Bob,

I wonder what the other oscillators that others are receiving look like in
the SA. Replacing all the capacitors would take a serious effort, not sure
if it would be justified.

Can I ask how can you know all that by looking at the spectrum plot?

Thank you,
Roberto EB4EQA

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Bob Camp
Sent: sábado, 25 de febrero de 2012 23:55
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Morion MV89A Repair

Hi

The spectrum you show in the middle is the correct one. It shows the 10 MHz
output and the sub-harmonics related to the 5 MHz crystal oscillator. The
other two are a bit broken. One appears to have a stage oscillating. The
other is more interesting. It has a stage oscillating that is injection
locked to the 5 MHz crystal oscillator. I suspect more than one capacitor
went bad in these OCXO's. Somebody may have sold Morion a bad reel of bypass
caps.

Bob



On Feb 25, 2012, at 4:01 PM, Roberto Barrios rbarri...@msn.com wrote:

 Hi all,
 
 
 
 I spent this evening repairing a MV89A that had a weak output, output 
 cap was bad:
 
 
 
 http://www.rbarrios.com/projects/MV89A
 
 
 
 I also found while comparing with other two MV89A's that they show 
 different output spectra, as can be seen.  Test conditions were 
 identical but one appears much cleaner than the others. What are the
reasons for this?
 
 
 
 Regards,
 
 Roberto EB4EQA
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-27 Thread EWKehren
Roberto
It is called tens of years of experience
Bert Kehren
 
 
In a message dated 2/27/2012 5:16:28 A.M. Eastern Standard Time,  
rbarri...@msn.com writes:

Hi  Bob,

I wonder what the other oscillators that others are receiving look  like in
the SA. Replacing all the capacitors would take a serious effort,  not sure
if it would be justified.

Can I ask how can you know all  that by looking at the spectrum plot?

Thank you,
Roberto  EB4EQA

-Original Message-
From: time-nuts-boun...@febo.com  [mailto:time-nuts-boun...@febo.com] On
Behalf Of Bob Camp
Sent: sábado,  25 de febrero de 2012 23:55
To: Discussion of precise time and frequency  measurement
Subject: Re: [time-nuts] Morion MV89A  Repair

Hi

The spectrum you show in the middle is the correct  one. It shows the 10 MHz
output and the sub-harmonics related to the 5 MHz  crystal oscillator. The
other two are a bit broken. One appears to have a  stage oscillating. The
other is more interesting. It has a stage  oscillating that is injection
locked to the 5 MHz crystal oscillator. I  suspect more than one capacitor
went bad in these OCXO's. Somebody may have  sold Morion a bad reel of 
bypass
caps.

Bob



On Feb 25,  2012, at 4:01 PM, Roberto Barrios rbarri...@msn.com wrote:

  Hi all,
 
 
 
 I spent this evening repairing a  MV89A that had a weak output, output 
 cap was bad:
 
  
 
 http://www.rbarrios.com/projects/MV89A
 
  
 
 I also found while comparing with other two MV89A's that  they show 
 different output spectra, as can be seen.  Test  conditions were 
 identical but one appears much cleaner than the  others. What are the
reasons for this?
 
 
 
  Regards,
 
 Roberto EB4EQA
 
  ___
 time-nuts mailing list  -- time-nuts@febo.com To unsubscribe, go to 
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the  instructions  there.

___
time-nuts  mailing list -- time-nuts@febo.com To unsubscribe, go  to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow  the instructions  there.


___
time-nuts  mailing list -- time-nuts@febo.com
To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-27 Thread Roberto Barrios
Hi Bert,

I've googled for that and didn't find a related paper :-)

Regards,
Roberto EB4EQA

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of ewkeh...@aol.com
Sent: lunes, 27 de febrero de 2012 11:44
To: time-nuts@febo.com
Subject: Re: [time-nuts] Morion MV89A Repair

Roberto
It is called tens of years of experience Bert Kehren
 
 
In a message dated 2/27/2012 5:16:28 A.M. Eastern Standard Time,
rbarri...@msn.com writes:

Hi  Bob,

I wonder what the other oscillators that others are receiving look  like in
the SA. Replacing all the capacitors would take a serious effort,  not sure
if it would be justified.

Can I ask how can you know all  that by looking at the spectrum plot?

Thank you,
Roberto  EB4EQA

-Original Message-
From: time-nuts-boun...@febo.com  [mailto:time-nuts-boun...@febo.com] On
Behalf Of Bob Camp
Sent: sábado,  25 de febrero de 2012 23:55
To: Discussion of precise time and frequency  measurement
Subject: Re: [time-nuts] Morion MV89A  Repair

Hi

The spectrum you show in the middle is the correct  one. It shows the 10 MHz
output and the sub-harmonics related to the 5 MHz  crystal oscillator. The
other two are a bit broken. One appears to have a  stage oscillating. The
other is more interesting. It has a stage  oscillating that is injection
locked to the 5 MHz crystal oscillator. I  suspect more than one capacitor
went bad in these OCXO's. Somebody may have  sold Morion a bad reel of
bypass caps.

Bob



On Feb 25,  2012, at 4:01 PM, Roberto Barrios rbarri...@msn.com wrote:

  Hi all,
 
 
 
 I spent this evening repairing a  MV89A that had a weak output, output 
 cap was bad:
 
  
 
 http://www.rbarrios.com/projects/MV89A
 
  
 
 I also found while comparing with other two MV89A's that  they show 
 different output spectra, as can be seen.  Test  conditions were 
 identical but one appears much cleaner than the  others. What are the
reasons for this?
 
 
 
  Regards,
 
 Roberto EB4EQA
 
  ___
 time-nuts mailing list  -- time-nuts@febo.com To unsubscribe, go to  
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the  instructions  there.

___
time-nuts  mailing list -- time-nuts@febo.com To unsubscribe, go  to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow  the instructions  there.


___
time-nuts  mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.
___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-27 Thread Bob Camp
Hi

...or my lack of experience. I think I first saw problems like these in 1975. 
Somehow they keep showing up :)

Bob



On Feb 27, 2012, at 5:44 AM, ewkeh...@aol.com wrote:

 Roberto
 It is called tens of years of experience
 Bert Kehren
 
 
 In a message dated 2/27/2012 5:16:28 A.M. Eastern Standard Time,  
 rbarri...@msn.com writes:
 
 Hi  Bob,
 
 I wonder what the other oscillators that others are receiving look  like in
 the SA. Replacing all the capacitors would take a serious effort,  not sure
 if it would be justified.
 
 Can I ask how can you know all  that by looking at the spectrum plot?
 
 Thank you,
 Roberto  EB4EQA
 
 -Original Message-
 From: time-nuts-boun...@febo.com  [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Bob Camp
 Sent: sábado,  25 de febrero de 2012 23:55
 To: Discussion of precise time and frequency  measurement
 Subject: Re: [time-nuts] Morion MV89A  Repair
 
 Hi
 
 The spectrum you show in the middle is the correct  one. It shows the 10 MHz
 output and the sub-harmonics related to the 5 MHz  crystal oscillator. The
 other two are a bit broken. One appears to have a  stage oscillating. The
 other is more interesting. It has a stage  oscillating that is injection
 locked to the 5 MHz crystal oscillator. I  suspect more than one capacitor
 went bad in these OCXO's. Somebody may have  sold Morion a bad reel of 
 bypass
 caps.
 
 Bob
 
 
 
 On Feb 25,  2012, at 4:01 PM, Roberto Barrios rbarri...@msn.com wrote:
 
 Hi all,
 
 
 
 I spent this evening repairing a  MV89A that had a weak output, output 
 cap was bad:
 
 
 
 http://www.rbarrios.com/projects/MV89A
 
 
 
 I also found while comparing with other two MV89A's that  they show 
 different output spectra, as can be seen.  Test  conditions were 
 identical but one appears much cleaner than the  others. What are the
 reasons for this?
 
 
 
 Regards,
 
 Roberto EB4EQA
 
 ___
 time-nuts mailing list  -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the  instructions  there.
 
 ___
 time-nuts  mailing list -- time-nuts@febo.com To unsubscribe, go  to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow  the instructions  there.
 
 
 ___
 time-nuts  mailing list -- time-nuts@febo.com
 To unsubscribe, go to  
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the  instructions there.
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Morion MV89A Repair

2012-02-26 Thread Bert, VE2ZAZ
Hi Neville,

RoHS 5 likely means RoHS-5/6, which means that all RoHS-banned substances, 
except lead, are removed. This means that regular leaded solder is used.  
RoHS-6/6 means that lead-free solder is used.

Cheers,

Bert, VE2ZAZ




Message: 6
Date: Sun, 26 Feb 2012 11:17:55 +1100
From: Neville Michie namic...@gmail.com
To: Discussion of precise time and frequency measurement
    time-nuts@febo.com
Subject: Re: [time-nuts] Morion MV89A Repair
Message-ID: 2568f2a2-4de2-4dbc-a711-f93c56c21...@gmail.com
Content-Type: text/plain; charset=US-ASCII; delsp=yes; format=flowed

Hi,
I have two of Morion MV89A, one is prominently marked (RoHS 5).
I assume that means lead-less solder. Pure tin solder is much more  
prone to
crystallised joints which break later.
Are these the dodgy units ?


cheers, Neville Michie




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-26 Thread Bob Camp
Hi

The paper mentions HC-40 crystals. They probably have the crystal drive current 
cranked down to optimize short term stability at 1 sec and longer. The phase 
noise floor takes a back seat.

The other interesting thing in the paper is that the short term numbers shown 
aren't into parts in the 13th.

Bob



On Feb 25, 2012, at 9:11 PM, John Miles jmi...@pop.net wrote:

 Are the specs at 5 MHz or at 10? They should be 6 db worse at 10 than at
 5.
 
 Now that I'm looking back at the data sheet posted with the auction, they
 were indeed specified at 5 MHz.  But even after the 6 dB penalty, their spec
 limits versus their real-world numbers are pretty marginal.
 
 Morion actually announced this particular oscillator at IFCS 2002, so
 there's a paper of sorts:
 http://www.ke5fx.com/Morion_MV89A_IEEE_IFCS_2002.pdf .  Figure 4 was taken
 from a 10 MHz part, but all three of the ones I've tested look like the
 attached.  Same as figure 4 until about 200 Hz... then they never go any
 lower.
 
 -- john
 
 mv89a_pn.png
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-26 Thread Magnus Danielson

On 02/26/2012 05:20 PM, Bob Camp wrote:

Hi

The paper mentions HC-40 crystals. They probably have the crystal drive current 
cranked down to optimize short term stability at 1 sec and longer. The phase 
noise floor takes a back seat.

The other interesting thing in the paper is that the short term numbers shown 
aren't into parts in the 13th.


Come to think of it, I never saw a good explanation as to why the 
crystal drive current really has this effect. I can only seem to recall 
hand-waving and experience show stuff. Could be I read it but forgot 
about it as I don't design crystal oscillators, only tend to use the 
once that others have designed and built for me.


Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-26 Thread Tom Knox

Is this the paper you are speaking of?
Google:  Optimization of Drive-Level in High Stability Low-Noise OCXOs

Thomas Knox



 From: li...@rtty.us
 Date: Sun, 26 Feb 2012 11:20:16 -0500
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] Morion MV89A Repair
 
 Hi
 
 The paper mentions HC-40 crystals. They probably have the crystal drive 
 current cranked down to optimize short term stability at 1 sec and longer. 
 The phase noise floor takes a back seat.
 
 The other interesting thing in the paper is that the short term numbers shown 
 aren't into parts in the 13th.
 
 Bob
 
 
 
 On Feb 25, 2012, at 9:11 PM, John Miles jmi...@pop.net wrote:
 
  Are the specs at 5 MHz or at 10? They should be 6 db worse at 10 than at
  5.
  
  Now that I'm looking back at the data sheet posted with the auction, they
  were indeed specified at 5 MHz.  But even after the 6 dB penalty, their spec
  limits versus their real-world numbers are pretty marginal.
  
  Morion actually announced this particular oscillator at IFCS 2002, so
  there's a paper of sorts:
  http://www.ke5fx.com/Morion_MV89A_IEEE_IFCS_2002.pdf .  Figure 4 was taken
  from a 10 MHz part, but all three of the ones I've tested look like the
  attached.  Same as figure 4 until about 200 Hz... then they never go any
  lower.
  
  -- john
  
  mv89a_pn.png
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to 
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
  
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-26 Thread John Miles
 The other interesting thing in the paper is that the short term numbers
 shown aren't into parts in the 13th.

The ADEV plot in figure 3 doesn't exactly show the oscillator in its best
light.   The right side of the 'bathtub' is well beyond 1000 seconds with
these parts, if you actually let them warm up before you measure them.

So the person who generated the plot either didn't allow sufficient warmup
time, or they used a low-quality reference.  The Ch A versus Ch B amplitudes
are almost 3 dB apart, so they probably were not using another MV89A.

-- john



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-26 Thread John Miles


 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-
 boun...@febo.com] On Behalf Of Tom Knox
 Sent: Sunday, February 26, 2012 9:41 AM
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] Morion MV89A Repair
 
 
 Is this the paper you are speaking of?
 Google:  Optimization of Drive-Level in High Stability Low-Noise OCXOs
 
 Thomas Knox

That's not the one Bob (and I) was talking about, but there are some
interesting contradictions between the paper you mentioned:

 
http://www.magicxtal.com/articles/Optimization%20of%20Drive-Level%20in%20Hig
h%20Stability%20Low-Noise%20OCXOs.PDF 

and Rakon's article:

http://www.rakon.com/Products/Public%20Documents/Whitepapers/PHASE%20NOISE%2
0IN%20CRYSTAL%20OSCILLATORS.pdf

The Rakon author states that the flicker and white PM regions, including the
broadband floor, are influenced by factors other than the crystal (Figure
9).   

Meanwhile, the Russian authors of the first paper state repeatedly that the
broadband noise floor is a function of the crystal current, and that
close-in noise isn't affected by crystal current.

It's obvious enough that high levels of crystal current are bad for
long-term stability and aging, but it's not as easy to see why high crystal
current in itself should rule out a good broadband noise floor.  There are a
lot of assertions in the literature, especially in the magicxtal.com paper,
that are backed up by measurements but not by solid analysis. 

-- john


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-26 Thread Tom Knox

I wonder how Wenzel looks at drive level, they seem to have their products 
dialed in.

Thomas Knox



 From: jmi...@pop.net
 To: time-nuts@febo.com
 Date: Sun, 26 Feb 2012 12:48:26 -0800
 Subject: Re: [time-nuts] Morion MV89A Repair
 
 
 
  -Original Message-
  From: time-nuts-boun...@febo.com [mailto:time-nuts-
  boun...@febo.com] On Behalf Of Tom Knox
  Sent: Sunday, February 26, 2012 9:41 AM
  To: time-nuts@febo.com
  Subject: Re: [time-nuts] Morion MV89A Repair
  
  
  Is this the paper you are speaking of?
  Google:  Optimization of Drive-Level in High Stability Low-Noise OCXOs
  
  Thomas Knox
 
 That's not the one Bob (and I) was talking about, but there are some
 interesting contradictions between the paper you mentioned:
 
  
 http://www.magicxtal.com/articles/Optimization%20of%20Drive-Level%20in%20Hig
 h%20Stability%20Low-Noise%20OCXOs.PDF 
 
 and Rakon's article:
 
 http://www.rakon.com/Products/Public%20Documents/Whitepapers/PHASE%20NOISE%2
 0IN%20CRYSTAL%20OSCILLATORS.pdf
 
 The Rakon author states that the flicker and white PM regions, including the
 broadband floor, are influenced by factors other than the crystal (Figure
 9).   
 
 Meanwhile, the Russian authors of the first paper state repeatedly that the
 broadband noise floor is a function of the crystal current, and that
 close-in noise isn't affected by crystal current.
 
 It's obvious enough that high levels of crystal current are bad for
 long-term stability and aging, but it's not as easy to see why high crystal
 current in itself should rule out a good broadband noise floor.  There are a
 lot of assertions in the literature, especially in the magicxtal.com paper,
 that are backed up by measurements but not by solid analysis. 
 
 -- john
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
  
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-26 Thread Bob Camp
Hi

Most people put the best of the batch in the paper. It is indeed odd that the 
very experienced authors of the paper did not do so. With low drive a 5 MHz 
HC-40 should get into parts in the thirteenth over much of the bathtub region.

Bob



On Feb 26, 2012, at 3:22 PM, John Miles jmi...@pop.net wrote:

 The other interesting thing in the paper is that the short term numbers
 shown aren't into parts in the 13th.
 
 The ADEV plot in figure 3 doesn't exactly show the oscillator in its best
 light.   The right side of the 'bathtub' is well beyond 1000 seconds with
 these parts, if you actually let them warm up before you measure them.
 
 So the person who generated the plot either didn't allow sufficient warmup
 time, or they used a low-quality reference.  The Ch A versus Ch B amplitudes
 are almost 3 dB apart, so they probably were not using another MV89A.
 
 -- john
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Morion MV89A Repair

2012-02-25 Thread Roberto Barrios
Hi all,

 

I spent this evening repairing a MV89A that had a weak output, output cap
was bad:

 

http://www.rbarrios.com/projects/MV89A

 

I also found while comparing with other two MV89A's that they show different
output spectra, as can be seen.  Test conditions were identical but one
appears much cleaner than the others. What are the reasons for this?

 

Regards,

Roberto EB4EQA

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-25 Thread Bob Camp
Hi

The spectrum you show in the middle is the correct one. It shows the 10 MHz 
output and the sub-harmonics related to the 5 MHz crystal oscillator. The other 
two are a bit broken. One appears to have a stage oscillating. The other is 
more interesting. It has a stage oscillating that is injection locked to the 5 
MHz crystal oscillator. I suspect more than one capacitor went bad in these 
OCXO's. Somebody may have sold Morion a bad reel of bypass caps.

Bob



On Feb 25, 2012, at 4:01 PM, Roberto Barrios rbarri...@msn.com wrote:

 Hi all,
 
 
 
 I spent this evening repairing a MV89A that had a weak output, output cap
 was bad:
 
 
 
 http://www.rbarrios.com/projects/MV89A
 
 
 
 I also found while comparing with other two MV89A's that they show different
 output spectra, as can be seen.  Test conditions were identical but one
 appears much cleaner than the others. What are the reasons for this?
 
 
 
 Regards,
 
 Roberto EB4EQA
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-25 Thread John Miles

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-
 boun...@febo.com] On Behalf Of Bob Camp
 Sent: Saturday, February 25, 2012 2:55 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Morion MV89A Repair
 
 Hi
 
 The spectrum you show in the middle is the correct one. It shows the 10
MHz
 output and the sub-harmonics related to the 5 MHz crystal oscillator. The
 other two are a bit broken. One appears to have a stage oscillating. The
 other is more interesting. It has a stage oscillating that is injection
locked to
 the 5 MHz crystal oscillator. I suspect more than one capacitor went bad
in
 these OCXO's. Somebody may have sold Morion a bad reel of bypass caps.
 
 Bob

Interesting.  The MV89As are very stable, but they have a lot of AM noise
and they don't meet their PN specs (about -150 dBc/Hz at 10 kHz rather than
-155).  I've tested a few and they all look the same in that respect.
Sounds like replacing all of the SMT caps may be a good move.

-- john


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-25 Thread Neville Michie

Hi,
I have two of Morion MV89A, one is prominently marked (RoHS 5).
I assume that means lead-less solder. Pure tin solder is much more  
prone to

crystallised joints which break later.
Are these the dodgy units ?


cheers, Neville Michie





On 26/02/2012, at 11:02 AM, John Miles wrote:




-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-
boun...@febo.com] On Behalf Of Bob Camp
Sent: Saturday, February 25, 2012 2:55 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Morion MV89A Repair

Hi

The spectrum you show in the middle is the correct one. It shows  
the 10

MHz
output and the sub-harmonics related to the 5 MHz crystal  
oscillator. The
other two are a bit broken. One appears to have a stage  
oscillating. The
other is more interesting. It has a stage oscillating that is  
injection

locked to
the 5 MHz crystal oscillator. I suspect more than one capacitor  
went bad

in
these OCXO's. Somebody may have sold Morion a bad reel of bypass  
caps.


Bob


Interesting.  The MV89As are very stable, but they have a lot of AM  
noise
and they don't meet their PN specs (about -150 dBc/Hz at 10 kHz  
rather than

-155).  I've tested a few and they all look the same in that respect.
Sounds like replacing all of the SMT caps may be a good move.

-- john


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/ 
time-nuts

and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-25 Thread Bob Camp
Hi

Are the specs at 5 MHz or at 10? They should be 6 db worse at 10 than at 5.

Bob



On Feb 25, 2012, at 7:02 PM, John Miles jmi...@pop.net wrote:

 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-
 boun...@febo.com] On Behalf Of Bob Camp
 Sent: Saturday, February 25, 2012 2:55 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Morion MV89A Repair
 
 Hi
 
 The spectrum you show in the middle is the correct one. It shows the 10
 MHz
 output and the sub-harmonics related to the 5 MHz crystal oscillator. The
 other two are a bit broken. One appears to have a stage oscillating. The
 other is more interesting. It has a stage oscillating that is injection
 locked to
 the 5 MHz crystal oscillator. I suspect more than one capacitor went bad
 in
 these OCXO's. Somebody may have sold Morion a bad reel of bypass caps.
 
 Bob
 
 Interesting.  The MV89As are very stable, but they have a lot of AM noise
 and they don't meet their PN specs (about -150 dBc/Hz at 10 kHz rather than
 -155).  I've tested a few and they all look the same in that respect.
 Sounds like replacing all of the SMT caps may be a good move.
 
 -- john
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-25 Thread Bob Camp
Hi

Hit send to soon...

The AM noise could easily be another bypass issue.

Bob



On Feb 25, 2012, at 7:02 PM, John Miles jmi...@pop.net wrote:

 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-
 boun...@febo.com] On Behalf Of Bob Camp
 Sent: Saturday, February 25, 2012 2:55 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Morion MV89A Repair
 
 Hi
 
 The spectrum you show in the middle is the correct one. It shows the 10
 MHz
 output and the sub-harmonics related to the 5 MHz crystal oscillator. The
 other two are a bit broken. One appears to have a stage oscillating. The
 other is more interesting. It has a stage oscillating that is injection
 locked to
 the 5 MHz crystal oscillator. I suspect more than one capacitor went bad
 in
 these OCXO's. Somebody may have sold Morion a bad reel of bypass caps.
 
 Bob
 
 Interesting.  The MV89As are very stable, but they have a lot of AM noise
 and they don't meet their PN specs (about -150 dBc/Hz at 10 kHz rather than
 -155).  I've tested a few and they all look the same in that respect.
 Sounds like replacing all of the SMT caps may be a good move.
 
 -- john
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Morion MV89A Repair

2012-02-25 Thread John Miles
 Are the specs at 5 MHz or at 10? They should be 6 db worse at 10 than at
5.

Now that I'm looking back at the data sheet posted with the auction, they
were indeed specified at 5 MHz.  But even after the 6 dB penalty, their spec
limits versus their real-world numbers are pretty marginal.

Morion actually announced this particular oscillator at IFCS 2002, so
there's a paper of sorts:
http://www.ke5fx.com/Morion_MV89A_IEEE_IFCS_2002.pdf .  Figure 4 was taken
from a 10 MHz part, but all three of the ones I've tested look like the
attached.  Same as figure 4 until about 200 Hz... then they never go any
lower.

-- john

attachment: mv89a_pn.png___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts