I have installed TimeLab and am getting good results.   But I have
a question on how to setup my HP 5379B.

There are separate trigger level settings for start and stop.   In
my case I want to measure the period of the 1 PPS signal from my
Resolution T.   If the start and stop trigger levels are not EXACTLY
the same then won't the period measurement be wrong.   It seems
that using only the start trigger level for both start and stop
would be the correct way of doing it.

While TimeLab was running I adjusted both the start and stop trigger
levels.   It is clear that they both are setting their respective
trigger points.   Doesn't this effect the period measurement results? 

My setup;
1 PPS signal from GPS connected to start input.
Positive slope trigger both channels.
50 Ohm both channels.
Divide by 1 both channels.
DC coupling both channels.
Start switch set to COM.
Trigger level knobs set to where the trigger is consistent on both channels.


Thanks,
Pete.
 

-Original Message-
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Peter
Reilley
Sent: Tuesday, June 16, 2015 12:46 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] My HP 5370B reads 6 nS out!

Bob;

It seems reasonable to calibrate my 5370B using the 1 PPS signal from my
Resolution T and assume that my rubidium oscillator is 6 X 10^-9 off.
There is no reason to believe that the GPS 1 PPS signal is wrong.

The FEI FE-5680A devices seem to be hard to determine what options they
actually have.   Perhaps it just needs calibration.

Pete.
 

-Original Message-
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Bob Camp
Sent: Tuesday, June 16, 2015 7:02 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] My HP 5370B reads 6 nS out!

Hi
 On Jun 15, 2015, at 9:51 PM, Peter Reilley pe...@reilley.com wrote:
 
 Frank;
 
 Thanks for your long and detailed explanation.
 
 I was able to get the internal OCXO to that precision but it was 
 probably luck
 to get the trimmer that close.   I worked at it for a while.
 
 I am using T.I. mode with the averaging mode.   I assumed that it took 10K
 readings
 and averaged the results.   Is that not correct?   Is something mode
 complicated
 going on?
 
 I will have to set up the GPIB and give that a try.   I did get TimeLab.
 This will be new territory to me.
 
 I did try measuring it's own 10 MHz frequency with 1 sec gate time.   It
 bounces
 around by about +- 4 in the 10th decimal place.
 
 Is my calibration with the rubidium oscillator valid?   Could it be
 that far off?

An old style analog tune Rb could tune a max of 1x10^-8. It's rare to see
them more than 3x10^-9 off. Most are running under 5x10^-10 when received
from the salvage yard. 

The new style digital tuned Rb's have a range limited by their VCXO (if
it's limited at all). They *should* be  3x10^-9 as received from salvage.
They could be much further off.

Since all of these are tunable devices. The only way to be sure of their
accuracy is to calibrate them against something good.

Bob

 
 I will have to ponder this some more.
 
 
 Pete.
 
 
 -Original Message-
 From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Frank 
 Stellmach
 Sent: Monday, June 15, 2015 5:01 PM
 To: time-nuts@febo.com
 Subject: [time-nuts] My HP 5370B reads 6 nS out!
 
 Pete,
 
 you do not specify, whether you use FREQ or T.I. when you use the 
 averaging function.
 
 First of all, its OCXO can be adjusted to a few parts in 1e-9 only, as 
 the trimmer is too unprecise.
 If the OCXO is running for several weeks already (idle state), its 
 drift may be as low as a few parts in 1e-10 or better.
 
 If you put the instrument to FREQ mode, you may measure and the 10MHz 
 of the GPSDO standard to about 2e-11 resolution, if you use 1sec time 
 base on the 5370B.
 That should work also, if you directly measure 1pps, but you have to 
 properly adjust the trigger level.
 Important: Don't use the 10k statistics, set the 5370A also to 1sec 
 time base!
 Due to this low frequency, jitter should be higher, see specifications.
 
 You better do statistics by means of a PC, over GPIB.
 That will show the 30ps jitter of the 5370B, and the jitter of the 
 GPSDO, on the order of 1e-10.
 
 You may also calibrate the OCXO of the 5370B this way, instead of that 
 oscilloscope method.
 
 I strongly recommend Timelab from John Miles to do these measurements 
 properly.
 http://www.ke5fx.com/timelab/readme.htm
 
 
 If you use the internal 10k statistics 10k, pay attention!!
 
 In this instance, the 5370B will do the frequency measurement in a 
 different manner..
 Not 100% sure, it will be a sort of a T.I. measurement, calculated to 
 frequency.
 And that may produce a constant offset, if the internal T.I. 
 calibration is not done properly.
 Look into the specs, its absolute T.I. uncertainty is 1ns only, 
 although it resolves 20ps.
 
 
 You may check that behaviour

Re: [time-nuts] My HP 5370B reads 6 nS out!

2015-06-16 Thread Bob Camp
Hi
 On Jun 15, 2015, at 9:51 PM, Peter Reilley pe...@reilley.com wrote:
 
 Frank;
 
 Thanks for your long and detailed explanation.
 
 I was able to get the internal OCXO to that precision but it was probably
 luck
 to get the trimmer that close.   I worked at it for a while.
 
 I am using T.I. mode with the averaging mode.   I assumed that it took 10K
 readings
 and averaged the results.   Is that not correct?   Is something mode
 complicated
 going on?
 
 I will have to set up the GPIB and give that a try.   I did get TimeLab.
 This will be new territory to me.
 
 I did try measuring it's own 10 MHz frequency with 1 sec gate time.   It
 bounces 
 around by about +- 4 in the 10th decimal place.
 
 Is my calibration with the rubidium oscillator valid?   Could it be
 that far off?

An “old style” analog tune Rb could tune a max of 1x10^-8. It’s rare
to see them more than 3x10^-9 off. Most are running under 5x10^-10 when
received from the salvage yard. 

The “new style” digital tuned Rb’s have a range limited by their VCXO (if
it’s limited at all). They *should* be  3x10^-9 as received from salvage.
They could be much further off.

Since all of these are tunable devices. The only way to be sure of their
accuracy is to calibrate them against something good.

Bob

 
 I will have to ponder this some more.
 
 
 Pete.
 
 
 -Original Message-
 From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Frank
 Stellmach
 Sent: Monday, June 15, 2015 5:01 PM
 To: time-nuts@febo.com
 Subject: [time-nuts] My HP 5370B reads 6 nS out!
 
 Pete,
 
 you do not specify, whether you use FREQ or T.I. when you use the averaging
 function.
 
 First of all, its OCXO can be adjusted to a few parts in 1e-9 only, as the
 trimmer is too unprecise.
 If the OCXO is running for several weeks already (idle state), its drift may
 be as low as a few parts in 1e-10 or better.
 
 If you put the instrument to FREQ mode, you may measure and the 10MHz of the
 GPSDO standard to about 2e-11 resolution, if you use 1sec time base on the
 5370B.
 That should work also, if you directly measure 1pps, but you have to
 properly adjust the trigger level.
 Important: Don't use the 10k statistics, set the 5370A also to 1sec time
 base!
 Due to this low frequency, jitter should be higher, see specifications.
 
 You better do statistics by means of a PC, over GPIB.
 That will show the 30ps jitter of the 5370B, and the jitter of the GPSDO, on
 the order of 1e-10.
 
 You may also calibrate the OCXO of the 5370B this way, instead of that
 oscilloscope method.
 
 I strongly recommend Timelab from John Miles to do these measurements
 properly.
 http://www.ke5fx.com/timelab/readme.htm
 
 
 If you use the internal 10k statistics 10k, pay attention!!
 
 In this instance, the 5370B will do the frequency measurement in a 
 different manner..
 Not 100% sure, it will be a sort of a T.I. measurement, calculated to 
 frequency.
 And that may produce a constant offset, if the internal T.I. calibration 
 is not done properly.
 Look into the specs, its absolute T.I. uncertainty is 1ns only, although 
 it resolves 20ps.
 
 
 You may check that behaviour, if you apply its own 10Mhz OCXO ouput to 
 the FREQ input, and measure this frequency first on FREQ, 1sec.
 That should give nearly exactly 10MHz,  1e-10 jitter or deviation.
 Mine reads 9.999 999 999 85 MHz, for example.
 
 If you now switch to AVERAGE, SAMPLE SIZE 1, 100, 1k, 10K, you will see, 
 that you will get big deviations as big as 0.1%, although it should 
 measure its own OCXO to precisely 10.0MHZ.
 Mine reads 9,989 294 5 MHz, for example.
 
 That's due to the different measurement method, and should explain 
 6..7ns deviation on the 1pps signal also.
 
 This averaging should only be used with T.I.!
 
 Frank
 
 
 
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] My HP 5370B reads 6 nS out!

2015-06-16 Thread Peter Reilley
Bob;

It seems reasonable to calibrate my 5370B using the 1 PPS signal from
my Resolution T and assume that my rubidium oscillator is 6 X 10^-9 off.
There is no reason to believe that the GPS 1 PPS signal is wrong.

The FEI FE-5680A devices seem to be hard to determine what options they
actually have.   Perhaps it just needs calibration.

Pete.
 

-Original Message-
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Bob Camp
Sent: Tuesday, June 16, 2015 7:02 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] My HP 5370B reads 6 nS out!

Hi
 On Jun 15, 2015, at 9:51 PM, Peter Reilley pe...@reilley.com wrote:
 
 Frank;
 
 Thanks for your long and detailed explanation.
 
 I was able to get the internal OCXO to that precision but it was 
 probably luck
 to get the trimmer that close.   I worked at it for a while.
 
 I am using T.I. mode with the averaging mode.   I assumed that it took 10K
 readings
 and averaged the results.   Is that not correct?   Is something mode
 complicated
 going on?
 
 I will have to set up the GPIB and give that a try.   I did get TimeLab.
 This will be new territory to me.
 
 I did try measuring it's own 10 MHz frequency with 1 sec gate time.   It
 bounces
 around by about +- 4 in the 10th decimal place.
 
 Is my calibration with the rubidium oscillator valid?   Could it be
 that far off?

An old style analog tune Rb could tune a max of 1x10^-8. It's rare to see
them more than 3x10^-9 off. Most are running under 5x10^-10 when received
from the salvage yard. 

The new style digital tuned Rb's have a range limited by their VCXO (if
it's limited at all). They *should* be  3x10^-9 as received from salvage.
They could be much further off.

Since all of these are tunable devices. The only way to be sure of their
accuracy is to calibrate them against something good.

Bob

 
 I will have to ponder this some more.
 
 
 Pete.
 
 
 -Original Message-
 From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Frank 
 Stellmach
 Sent: Monday, June 15, 2015 5:01 PM
 To: time-nuts@febo.com
 Subject: [time-nuts] My HP 5370B reads 6 nS out!
 
 Pete,
 
 you do not specify, whether you use FREQ or T.I. when you use the 
 averaging function.
 
 First of all, its OCXO can be adjusted to a few parts in 1e-9 only, as 
 the trimmer is too unprecise.
 If the OCXO is running for several weeks already (idle state), its 
 drift may be as low as a few parts in 1e-10 or better.
 
 If you put the instrument to FREQ mode, you may measure and the 10MHz 
 of the GPSDO standard to about 2e-11 resolution, if you use 1sec time 
 base on the 5370B.
 That should work also, if you directly measure 1pps, but you have to 
 properly adjust the trigger level.
 Important: Don't use the 10k statistics, set the 5370A also to 1sec 
 time base!
 Due to this low frequency, jitter should be higher, see specifications.
 
 You better do statistics by means of a PC, over GPIB.
 That will show the 30ps jitter of the 5370B, and the jitter of the 
 GPSDO, on the order of 1e-10.
 
 You may also calibrate the OCXO of the 5370B this way, instead of that 
 oscilloscope method.
 
 I strongly recommend Timelab from John Miles to do these measurements 
 properly.
 http://www.ke5fx.com/timelab/readme.htm
 
 
 If you use the internal 10k statistics 10k, pay attention!!
 
 In this instance, the 5370B will do the frequency measurement in a 
 different manner..
 Not 100% sure, it will be a sort of a T.I. measurement, calculated to 
 frequency.
 And that may produce a constant offset, if the internal T.I. 
 calibration is not done properly.
 Look into the specs, its absolute T.I. uncertainty is 1ns only, 
 although it resolves 20ps.
 
 
 You may check that behaviour, if you apply its own 10Mhz OCXO ouput to 
 the FREQ input, and measure this frequency first on FREQ, 1sec.
 That should give nearly exactly 10MHz,  1e-10 jitter or deviation.
 Mine reads 9.999 999 999 85 MHz, for example.
 
 If you now switch to AVERAGE, SAMPLE SIZE 1, 100, 1k, 10K, you will 
 see, that you will get big deviations as big as 0.1%, although it 
 should measure its own OCXO to precisely 10.0MHZ.
 Mine reads 9,989 294 5 MHz, for example.
 
 That's due to the different measurement method, and should explain 
 6..7ns deviation on the 1pps signal also.
 
 This averaging should only be used with T.I.!
 
 Frank
 
 
 
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time

Re: [time-nuts] My HP 5370B reads 6 nS out!

2015-06-15 Thread Peter Reilley
Frank;

Thanks for your long and detailed explanation.

I was able to get the internal OCXO to that precision but it was probably
luck
to get the trimmer that close.   I worked at it for a while.

I am using T.I. mode with the averaging mode.   I assumed that it took 10K
readings
and averaged the results.   Is that not correct?   Is something mode
complicated
going on?

I will have to set up the GPIB and give that a try.   I did get TimeLab.
This will be new territory to me.

I did try measuring it's own 10 MHz frequency with 1 sec gate time.   It
bounces 
around by about +- 4 in the 10th decimal place.

Is my calibration with the rubidium oscillator valid?   Could it be
that far off?

I will have to ponder this some more.


Pete.
 

-Original Message-
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Frank
Stellmach
Sent: Monday, June 15, 2015 5:01 PM
To: time-nuts@febo.com
Subject: [time-nuts] My HP 5370B reads 6 nS out!

Pete,

you do not specify, whether you use FREQ or T.I. when you use the averaging
function.

First of all, its OCXO can be adjusted to a few parts in 1e-9 only, as the
trimmer is too unprecise.
If the OCXO is running for several weeks already (idle state), its drift may
be as low as a few parts in 1e-10 or better.

If you put the instrument to FREQ mode, you may measure and the 10MHz of the
GPSDO standard to about 2e-11 resolution, if you use 1sec time base on the
5370B.
That should work also, if you directly measure 1pps, but you have to
properly adjust the trigger level.
Important: Don't use the 10k statistics, set the 5370A also to 1sec time
base!
Due to this low frequency, jitter should be higher, see specifications.

You better do statistics by means of a PC, over GPIB.
That will show the 30ps jitter of the 5370B, and the jitter of the GPSDO, on
the order of 1e-10.

You may also calibrate the OCXO of the 5370B this way, instead of that
oscilloscope method.

I strongly recommend Timelab from John Miles to do these measurements
properly.
http://www.ke5fx.com/timelab/readme.htm


If you use the internal 10k statistics 10k, pay attention!!

In this instance, the 5370B will do the frequency measurement in a 
different manner..
Not 100% sure, it will be a sort of a T.I. measurement, calculated to 
frequency.
And that may produce a constant offset, if the internal T.I. calibration 
is not done properly.
Look into the specs, its absolute T.I. uncertainty is 1ns only, although 
it resolves 20ps.


You may check that behaviour, if you apply its own 10Mhz OCXO ouput to 
the FREQ input, and measure this frequency first on FREQ, 1sec.
That should give nearly exactly 10MHz,  1e-10 jitter or deviation.
Mine reads 9.999 999 999 85 MHz, for example.

If you now switch to AVERAGE, SAMPLE SIZE 1, 100, 1k, 10K, you will see, 
that you will get big deviations as big as 0.1%, although it should 
measure its own OCXO to precisely 10.0MHZ.
Mine reads 9,989 294 5 MHz, for example.

That's due to the different measurement method, and should explain 
6..7ns deviation on the 1pps signal also.

This averaging should only be used with T.I.!

Frank






___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] My HP 5370B reads 6 nS out!

2015-06-15 Thread Frank Stellmach

Pete,

you do not specify, whether you use FREQ or T.I. when you use the 
averaging function.


First of all, its OCXO can be adjusted to a few parts in 1e-9 only, as 
the trimmer is too unprecise.
If the OCXO is running for several weeks already (idle state), its drift 
may be as low as a few parts in 1e-10 or better.


If you put the instrument to FREQ mode, you may measure and the 10MHz of 
the GPSDO standard to about 2e-11 resolution, if you use 1sec time base 
on the 5370B.
That should work also, if you directly measure 1pps, but you have to 
properly adjust the trigger level.
Important: Don't use the 10k statistics, set the 5370A also to 1sec time 
base!

Due to this low frequency, jitter should be higher, see specifications.

You better do statistics by means of a PC, over GPIB.
That will show the 30ps jitter of the 5370B, and the jitter of the 
GPSDO, on the order of 1e-10.


You may also calibrate the OCXO of the 5370B this way, instead of that 
oscilloscope method.


I strongly recommend Timelab from John Miles to do these measurements 
properly.

http://www.ke5fx.com/timelab/readme.htm


If you use the internal 10k statistics 10k, pay attention!!

In this instance, the 5370B will do the frequency measurement in a 
different manner..
Not 100% sure, it will be a sort of a T.I. measurement, calculated to 
frequency.
And that may produce a constant offset, if the internal T.I. calibration 
is not done properly.
Look into the specs, its absolute T.I. uncertainty is 1ns only, although 
it resolves 20ps.



You may check that behaviour, if you apply its own 10Mhz OCXO ouput to 
the FREQ input, and measure this frequency first on FREQ, 1sec.

That should give nearly exactly 10MHz,  1e-10 jitter or deviation.
Mine reads 9.999 999 999 85 MHz, for example.

If you now switch to AVERAGE, SAMPLE SIZE 1, 100, 1k, 10K, you will see, 
that you will get big deviations as big as 0.1%, although it should 
measure its own OCXO to precisely 10.0MHZ.

Mine reads 9,989 294 5 MHz, for example.

That's due to the different measurement method, and should explain 
6..7ns deviation on the 1pps signal also.


This averaging should only be used with T.I.!

Frank






___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] My HP 5370B reads 6 nS out!

2015-06-13 Thread Peter Reilley
I just got a new (to me) HP 5370B and I wanted to try it out on the 1 PPS
output
from a Trimble Resolution T.   Using 10K samples averaging it is always
about
6 or 7 nS to high.   Shouldn't this average toward zero?   If I use no
averaging
it bounces around within the expected range.
 
I have done this test using the internal oscillator and an external rubidium
frequency
standard.   I get the about the same results in both cases.
 
I did calibrate the OCXO in the 5370B using the rubidium standard and got it
so
that there is less than one cycle difference between the two over a period
of a few
hours.
 
I have tried a number of GPS units with the same result.
 
Could my rubidium oscillator be that far out?   It is a FEI model FE5680A.
 
Am I missing something?
 
Pete.
 
 