Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Magnus Danielson

Volker,

On 02/02/14 01:51, Volker Esper wrote:

Magnus,

I took an EBMPabst 624. Pabst has a good reputation.


Papst is known good brand. I have seen another brand fail miserably so 
this is why I am asking.



The original fan is a Delta Electronics DFB0624 H, Dimensions are 60mm x
60mm x 25mm, DC24V 0.11A


Yes, and the EBM-Papst 624 matches it well. The Papst seems to have much 
lower noise from the datasheet.


BTW. Both my SR620 checks out well. One of them was supposedly diagnosed 
with flicker.


I still need to do the sensitivity part of the measurements, but lacking 
a suitable RF generator I can't do it smoothly.


The fan in the new TIC locks itself into one position, so I had to help 
it. I don't know if it is worth the trouble to open it up, clean and 
lubricate.


Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Magnus Danielson

On 02/02/14 03:43, Charles Steinmetz wrote:

Jarl wrote:


In my SR620 the fan is a Delta DBF0624H.  It is a 60x60x25 mm fan, 24V
/0.11A.


Mine, too.  Does anyone know the airflow rating of the Delta fan?  It
does not appear to move as much air as it should (at least not if the
idea is to hold the interior at a fixed temperature) -- every SR620 I've
used has the fan running at full speed by the time it's been on for 10
minutes, unless the ambient temperature is less than 17-18C.

On the one hand, it would be nice to move more air so the internal
temperature is more tightly regulated; but on the other hand, it would
also be nice if the fan were quieter.  It is unlikely we can have both.
The fact that the original fan seems marginal counsels against replacing
it with one that moves less air.


Well, considering the conditions it can't do much else.


The thermal design of the 620 is far from optimal.  The thermistor is
located in a tunnel between the interior of the instrument and the
exterior, with the fan blowing directly across it through the tunnel.
So, the fan startup is an ugly process of fits and starts as the
instrument warms up.  Also, the air inlets seem to be too small and the
internal airflow was not properly designed to circulate cooling air
where it needs to go.


If the heat sources where well coupled to the air-flow, which they are 
not, and the flow-path as low air-flow resistance, which it also doesn't 
have, requires the fan to work at high rate to get any air move, and to 
get the thermistor happy.



I have toyed with the idea of cutting a slot maybe 4 long and 1/8 tall
in the right wall of the chassis at the rear, above the four TO-220
devices mounted there -- and perhaps another slot about the same size in
the top cover above those devices.  Also, maybe attaching some internal
baffles to the top cover to channel airflow where it is needed.


You can mount a fan to create a steady flow on the right side, and that 
way cool off much of the heat there.



I installed a terminal strip on the GPIB connector mounting screw and
relocated the thermistor there in my SR620s (in the general vicinity of
the oscillators).  (However, note that stabilizing the internal
oscillator temperature is not really very important for most of us,
because time nuts generally use an external time base.  In that case, it
is probably more important that the temperature of the triggers and
interpolators is held constant.)


Well, we *do* care about a stable internal temperature, since it will 
also shift calibration factors, so the stabler the internal temperature 
is and hence various shifts which is being compensate, the more accurate 
it becomes.



With the thermistors relocated as I have described, the fans start up as
they should (monotonically, speeding up smoothly from stopped to full
speed without any fits and starts).  They still reach full speed in 10
minutes or so, so at the end of the day I don't think I've really
changed anything except the aesthetics of the fan startup.  (IMO, the
change is worth it just for that, but there does not seem to be any
operational improvement.)


If you try to achieve the goal I think it will require a different approach.


Perhaps SRS did not intend to regulate the interior temperature of the
SR620 -- maybe they just wanted it to warm up faster (if you did away
with the thermistor and had the fan run full speed whenever the counter
was on, it would presumably take longer to warm up).


Maybe, would make kind of sense, on the other hand, they could have 
achieve both quick heat-up and stable but lower temperature and quieter 
if they wanted.



Anyway -- does anybody have an old Delta catalog or datasheet that
specifies the airflow rating of the original Delta fan?


It would indeed be interesting. The Papst 624 seems quite capable little 
critter and there seem to be some magic to the 624 number matching, 
which doesn't seem to be accidental.


Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Charles Steinmetz

Magnus wrote:

If the heat sources where well coupled to the air-flow, which they 
are not, and the flow-path as low air-flow resistance, which it also 
doesn't have, requires the fan to work at high rate to get any air 
move, and to get the thermistor happy.


My point was, the thermistor is never happy.  It always wants more 
cooling.  So it spins the fan up to full speed and is still too hot 
to reach equlibrium.


You can mount a fan to create a steady flow on the right side, and 
that way cool off much of the heat there.


You mean an internal fan, I take it.  That still doesn't solve the 
problem of too little air moving through the box to lower the 
internal temperature to the target value in normal room ambient temperatures.


Well, we *do* care about a stable internal temperature, since it 
will also shift calibration factors, so the stabler the internal 
temperature is and hence various shifts which is being compensate, 
the more accurate it becomes.


Right.  But unless the thermal design has been optimized (and it has 
not, on the SR620), you have to choose which part of the interior you 
want to be regulated to a constant temperature.  The very worst 
possible choice is to put the thermistor in the exhaust stream of the 
fan (where SRS put it).  It needs to be somewhere inside the box, and 
where you put it determines what part of the interior is 
regulated.  Of course, all of this assumes that you use a fan that 
moves enough air to actually reach equilibrium before it gets to full 
speed.  The stock fan doesn't, so NO place inside the box is 
regulated to a constant temperature, except in very cold ambient temperatures.



Perhaps SRS did not intend to regulate the interior temperature of the
SR620 -- maybe they just wanted it to warm up faster (if you did away
with the thermistor and had the fan run full speed whenever the counter
was on, it would presumably take longer to warm up).


Maybe, would make kind of sense, on the other hand, they could have 
achieve both quick heat-up and stable but lower temperature and 
quieter if they wanted.


Yes, they could have.  So, the question is, do we just replace fans 
when they go bad and live with the poor thermal design, or do we try 
to improve the thermal design?  If we want to improve the thermal 
design, the methods available to us are: (i) use a fan that moves 
more air, so the location of the thermistor is actually regulated to 
a constant temperature in normal ambient temperatures (not just in 
very cold ambient temperatures); (ii) relocate the thermistor to the 
most temperature-critical area inside the instrument; (iii) make 
additional air inlet holes, strategically placed to evenly cool the 
various zones of the interior; and (iv) add air baffles inside the 
box to evenly cool the various zones of the interior.


It would indeed be interesting. The Papst 624 seems quite capable 
little critter and there seem to be some magic to the 624 number 
matching, which doesn't seem to be accidental.


No magic.  624 just means that it is a 60x60mm fan that runs on 
24v.  Most fan manufacturers make a dozen or more fans with all 
different current draws, air movement, noise, number of fan blades, 
bearings, etc., etc., all with 624 in the part number (or whatever 
code that manufacturer uses for 60x60@24v -- 0624, 2406, 
etc.).  Papst makes 624HH, 624N, 624/2H3P, 624H, 624M, 624/2HH, 624L, 
624J, 624F, etc.


Best regards,

Charles



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Magnus Danielson

On 02/02/14 18:47, Charles Steinmetz wrote:

Magnus wrote:


If the heat sources where well coupled to the air-flow, which they are
not, and the flow-path as low air-flow resistance, which it also
doesn't have, requires the fan to work at high rate to get any air
move, and to get the thermistor happy.


My point was, the thermistor is never happy.  It always wants more
cooling.  So it spins the fan up to full speed and is still too hot to
reach equlibrium.


Which only means that the thermistor setup is shifted and needs to be 
adjusted to achieve the goal.



You can mount a fan to create a steady flow on the right side, and
that way cool off much of the heat there.


You mean an internal fan, I take it.  That still doesn't solve the
problem of too little air moving through the box to lower the internal
temperature to the target value in normal room ambient temperatures.


No, I mean an external fan.

The box have far to little air intake, as you have already pointed out, 
and I agree fully.



Well, we *do* care about a stable internal temperature, since it will
also shift calibration factors, so the stabler the internal
temperature is and hence various shifts which is being compensate, the
more accurate it becomes.


Right.  But unless the thermal design has been optimized (and it has
not, on the SR620), you have to choose which part of the interior you
want to be regulated to a constant temperature.  The very worst possible
choice is to put the thermistor in the exhaust stream of the fan (where
SRS put it).  It needs to be somewhere inside the box, and where you put
it determines what part of the interior is regulated.  Of course, all of
this assumes that you use a fan that moves enough air to actually reach
equilibrium before it gets to full speed.  The stock fan doesn't, so NO
place inside the box is regulated to a constant temperature, except in
very cold ambient temperatures.


I agree. There is many things which needs to be fixed to get there, 
there is no single silver-bullet to solve it all.



Perhaps SRS did not intend to regulate the interior temperature of the
SR620 -- maybe they just wanted it to warm up faster (if you did away
with the thermistor and had the fan run full speed whenever the counter
was on, it would presumably take longer to warm up).


Maybe, would make kind of sense, on the other hand, they could have
achieve both quick heat-up and stable but lower temperature and
quieter if they wanted.


Yes, they could have.  So, the question is, do we just replace fans when
they go bad and live with the poor thermal design, or do we try to
improve the thermal design?  If we want to improve the thermal design,
the methods available to us are: (i) use a fan that moves more air, so
the location of the thermistor is actually regulated to a constant
temperature in normal ambient temperatures (not just in very cold
ambient temperatures); (ii) relocate the thermistor to the most
temperature-critical area inside the instrument; (iii) make additional
air inlet holes, strategically placed to evenly cool the various zones
of the interior; and (iv) add air baffles inside the box to evenly cool
the various zones of the interior.


I think a sub-set of these are needed, but you also need to include (v) 
change the balance-point for the thermistor stabilization.


I think the single biggest problem is too little air inlet, which forces 
the fan to run in stupidly high speed without actually do much, as the 
air input has too low cross-section and with several small holes you 
need to create a large pressure difference in order to achieve the air 
flow wanted. Also, those holes help to create noise as they are not 
shaped to avoid turbulence. Best way is just achieve large enough 
cross-section.



It would indeed be interesting. The Papst 624 seems quite capable
little critter and there seem to be some magic to the 624 number
matching, which doesn't seem to be accidental.


No magic.  624 just means that it is a 60x60mm fan that runs on 24v.
Most fan manufacturers make a dozen or more fans with all different
current draws, air movement, noise, number of fan blades, bearings,
etc., etc., all with 624 in the part number (or whatever code that
manufacturer uses for 60x60@24v -- 0624, 2406, etc.).  Papst makes
624HH, 624N, 624/2H3P, 624H, 624M, 624/2HH, 624L, 624J, 624F, etc.


Ah, makes sense. It was a long time I looked at fan details myself.

Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Charles Steinmetz

Magnus wrote:


My point was, the thermistor is never happy.  It always wants more
cooling.  So it spins the fan up to full speed and is still too hot to
reach equlibrium.


Which only means that the thermistor setup is shifted and needs to 
be adjusted to achieve the goal.

 *   *   *

So, the question is, do we just replace fans when
they go bad and live with the poor thermal design, or do we try to
improve the thermal design?  If we want to improve the thermal design,
the methods available to us are:


I think a sub-set of these are needed, but you also need to include 
(v) change the balance-point for the thermistor stabilization.


Well, sure, you can raise the setpoint until the fan slows down and 
regulates the temperature, but then the regulated temperature would 
be too high.  In other words, if the fan cannot keep things cool 
enough by running at full speed, raising the set-point so the fan 
runs slower is not the right answer even if it does result in some 
approximation to isothermy.  [Note that adjusting the setpoint *down* 
will have no effect -- the fan already cannot cool the instrument 
enough to reach equilibrium.]


I think the single biggest problem is too little air inlet, which 
forces the fan to run in stupidly high speed without actually do 
much, as the air input has too low cross-section and with several 
small holes you need to create a large pressure difference in order 
to achieve the air flow wanted. Also, those holes help to create 
noise as they are not shaped to avoid turbulence. Best way is just 
achieve large enough cross-section.


I have experimented with cardboard top covers that have various holes 
in them.  My conclusion is that even with more air inlet area, you 
still want more cfm of air flow (i.e., a stronger fan).  You also 
need to adjust the airflow to the various sections of the interior to 
keep everything sufficiently cool.  As it is, there would be way too 
much temperature variation from one part of the interior to another 
even if you did reach a regulated equilibrium for the thermistor 
location (wherever you put the thermistor).


Best regards,

Charles



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Tom Knox

A little off topic but It seems many instruments (the SR620 and 53132A 
included) would work best with an internal fan. (A closed system, not 
exchanging outside air). Possibly with some sort of internal/external heat sink 
if needed. Or in high power situations outside air would flow through  a hollow 
heatsink (again not exchanging outside and internal air). If nothing else this 
would keep dust out. Since fan filters can really restrict air-flow and without 
a filter dust on circuit boards can act as on thermal insulator leading to 
overheating while conducting static.
Thomas Knox



 Date: Sun, 2 Feb 2014 19:17:03 +0100
 From: mag...@rubidium.dyndns.org
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] Replacement fan in SR620
 
 On 02/02/14 18:47, Charles Steinmetz wrote:
  Magnus wrote:
 
  If the heat sources where well coupled to the air-flow, which they are
  not, and the flow-path as low air-flow resistance, which it also
  doesn't have, requires the fan to work at high rate to get any air
  move, and to get the thermistor happy.
 
  My point was, the thermistor is never happy.  It always wants more
  cooling.  So it spins the fan up to full speed and is still too hot to
  reach equlibrium.
 
 Which only means that the thermistor setup is shifted and needs to be 
 adjusted to achieve the goal.
 
  You can mount a fan to create a steady flow on the right side, and
  that way cool off much of the heat there.
 
  You mean an internal fan, I take it.  That still doesn't solve the
  problem of too little air moving through the box to lower the internal
  temperature to the target value in normal room ambient temperatures.
 
 No, I mean an external fan.
 
 The box have far to little air intake, as you have already pointed out, 
 and I agree fully.
 
  Well, we *do* care about a stable internal temperature, since it will
  also shift calibration factors, so the stabler the internal
  temperature is and hence various shifts which is being compensate, the
  more accurate it becomes.
 
  Right.  But unless the thermal design has been optimized (and it has
  not, on the SR620), you have to choose which part of the interior you
  want to be regulated to a constant temperature.  The very worst possible
  choice is to put the thermistor in the exhaust stream of the fan (where
  SRS put it).  It needs to be somewhere inside the box, and where you put
  it determines what part of the interior is regulated.  Of course, all of
  this assumes that you use a fan that moves enough air to actually reach
  equilibrium before it gets to full speed.  The stock fan doesn't, so NO
  place inside the box is regulated to a constant temperature, except in
  very cold ambient temperatures.
 
 I agree. There is many things which needs to be fixed to get there, 
 there is no single silver-bullet to solve it all.
 
  Perhaps SRS did not intend to regulate the interior temperature of the
  SR620 -- maybe they just wanted it to warm up faster (if you did away
  with the thermistor and had the fan run full speed whenever the counter
  was on, it would presumably take longer to warm up).
 
  Maybe, would make kind of sense, on the other hand, they could have
  achieve both quick heat-up and stable but lower temperature and
  quieter if they wanted.
 
  Yes, they could have.  So, the question is, do we just replace fans when
  they go bad and live with the poor thermal design, or do we try to
  improve the thermal design?  If we want to improve the thermal design,
  the methods available to us are: (i) use a fan that moves more air, so
  the location of the thermistor is actually regulated to a constant
  temperature in normal ambient temperatures (not just in very cold
  ambient temperatures); (ii) relocate the thermistor to the most
  temperature-critical area inside the instrument; (iii) make additional
  air inlet holes, strategically placed to evenly cool the various zones
  of the interior; and (iv) add air baffles inside the box to evenly cool
  the various zones of the interior.
 
 I think a sub-set of these are needed, but you also need to include (v) 
 change the balance-point for the thermistor stabilization.
 
 I think the single biggest problem is too little air inlet, which forces 
 the fan to run in stupidly high speed without actually do much, as the 
 air input has too low cross-section and with several small holes you 
 need to create a large pressure difference in order to achieve the air 
 flow wanted. Also, those holes help to create noise as they are not 
 shaped to avoid turbulence. Best way is just achieve large enough 
 cross-section.
 
  It would indeed be interesting. The Papst 624 seems quite capable
  little critter and there seem to be some magic to the 624 number
  matching, which doesn't seem to be accidental.
 
  No magic.  624 just means that it is a 60x60mm fan that runs on 24v.
  Most fan manufacturers make a dozen or more fans with all different
  current draws, air movement, noise, number of fan

Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Volker Esper
Magnus,

 Yes, and the EBM-Papst 624 matches it well. The Papst seems to have
 much lower noise from the datasheet.

There's yet another problem: the mechanical construction conducts the
fan noise to the case, so the noise is kind of amplified by the case. I
felt somewhat disappointed after replacing the fan, don't expect too much.

I just read the postings of Charles and I, too, was unhappy with the
startup behaviour. I changed the values of the driver circuit, but kept
the sensor in the tunnel. That helped making the motor speed rising
monotonically.

The air inlets are too small - that's true, if you need so much air
flow. We agree that the fan speed is at it's maximum, not at it's
equilibrium. So I tried to separate the heat sources. At the right (and
left) side of the case various voltage regulators heat the side wall. I
placed a heat sink with vertikal fins from the outside on that wall, the
screws of the regulators screwed into the aluminium. I'll  further place
a wall inside the counter right before those heating parts. The
experiment isn't completed yet, but I'm keen enough to claim that it'll
help to stabilize internal temperature particularly at the sensitive
components.

Volker

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Magnus Danielson

On 02/02/14 20:16, Charles Steinmetz wrote:

Magnus wrote:


My point was, the thermistor is never happy.  It always wants more
cooling.  So it spins the fan up to full speed and is still too hot to
reach equlibrium.


Which only means that the thermistor setup is shifted and needs to be
adjusted to achieve the goal.
 *   *   *

So, the question is, do we just replace fans when
they go bad and live with the poor thermal design, or do we try to
improve the thermal design?  If we want to improve the thermal design,
the methods available to us are:


I think a sub-set of these are needed, but you also need to include
(v) change the balance-point for the thermistor stabilization.


Well, sure, you can raise the setpoint until the fan slows down and
regulates the temperature, but then the regulated temperature would be
too high.  In other words, if the fan cannot keep things cool enough by
running at full speed, raising the set-point so the fan runs slower is
not the right answer even if it does result in some approximation to
isothermy.  [Note that adjusting the setpoint *down* will have no effect
-- the fan already cannot cool the instrument enough to reach equilibrium.]


I'm not discussing it as a single measure, but rather a thing you can 
look at only after you got air-flow working much better.



I think the single biggest problem is too little air inlet, which
forces the fan to run in stupidly high speed without actually do much,
as the air input has too low cross-section and with several small
holes you need to create a large pressure difference in order to
achieve the air flow wanted. Also, those holes help to create noise as
they are not shaped to avoid turbulence. Best way is just achieve
large enough cross-section.


I have experimented with cardboard top covers that have various holes in
them.  My conclusion is that even with more air inlet area, you still
want more cfm of air flow (i.e., a stronger fan).  You also need to
adjust the airflow to the various sections of the interior to keep
everything sufficiently cool.  As it is, there would be way too much
temperature variation from one part of the interior to another even if
you did reach a regulated equilibrium for the thermistor location
(wherever you put the thermistor).


I agree. You need both, but just tossing in a stronger fan isn't going 
to cut it either.


Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Magnus Danielson

On 02/02/14 20:28, Tom Knox wrote:


A little off topic but It seems many instruments (the SR620 and 53132A 
included) would work best with an internal fan. (A closed system, not 
exchanging outside air). Possibly with some sort of internal/external heat sink 
if needed. Or in high power situations outside air would flow through  a hollow 
heatsink (again not exchanging outside and internal air). If nothing else this 
would keep dust out. Since fan filters can really restrict air-flow and without 
a filter dust on circuit boards can act as on thermal insulator leading to 
overheating while conducting static.


Good point. A fan in a closed environment will survive longer, or 
rather, won't degrade as fast as it possibly would do.


These days I would assume that heat-pipes would be used to move heat to 
a large external heat-sink. It's fairly cheap these days.


Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Magnus Danielson

On 02/02/14 20:13, Volker Esper wrote:

Magnus,


Yes, and the EBM-Papst 624 matches it well. The Papst seems to have
much lower noise from the datasheet.


There's yet another problem: the mechanical construction conducts the
fan noise to the case, so the noise is kind of amplified by the case. I
felt somewhat disappointed after replacing the fan, don't expect too much.


You mean the fact that the lid acts like a speaker membrane? So true.
The best cases I have seen have suspended mounting for fans and other 
vibration-makers (hard-disks).



I just read the postings of Charles and I, too, was unhappy with the
startup behaviour. I changed the values of the driver circuit, but kept
the sensor in the tunnel. That helped making the motor speed rising
monotonically.

The air inlets are too small - that's true, if you need so much air
flow. We agree that the fan speed is at it's maximum, not at it's
equilibrium. So I tried to separate the heat sources. At the right (and
left) side of the case various voltage regulators heat the side wall. I
placed a heat sink with vertikal fins from the outside on that wall, the
screws of the regulators screwed into the aluminium. I'll  further place
a wall inside the counter right before those heating parts. The
experiment isn't completed yet, but I'm keen enough to claim that it'll
help to stabilize internal temperature particularly at the sensitive
components.


It's indeed clear that the voltage regulators contribute a lot of heat.
That's why I proposed to put a fan on the outside.

As built, it is easy to cook it when it is put too crowded.

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Charles Steinmetz

Magnus wrote:

I agree. You need both, but just tossing in a stronger fan isn't 
going to cut it either.


Very true.  You need to attack all of the problems to fix it.

Best regards,

Charles



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Charles Steinmetz

Magnus wrote:

These days I would assume that heat-pipes would be used to move heat 
to a large external heat-sink. It's fairly cheap these days.


I have looked into heat pipes for several projects, and in the end 
have never used them.  The main problem is that almost every solution 
is customized.  No trouble if you make PC motherboards or 
consumer-grade stereo amplifiers 50,000 at a time.  But most 
instruments are made in quantities of 5,000 or fewer, and it takes 
many years to make (sell) even that number.  With production numbers 
that low, the bean counters want the design and tooling costs 
amortized over the first year's production -- maybe only 100-500 
units -- which can raise the retail price by a factor of 2-10.  At 
least that's what I've found.


Also, most heat-pipe systems are gravity-fed, so if someone turns 
your product on its side they can burn it down.  (I almost always 
operate my portable spectrum analyzer (HP8591E) standing on its 
rear feet with the screen pointing up.  I can foresee someone with 
limited bench space setting an SR620 on its side.)


Best regards,

Charles



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Magnus Danielson

Charles,

On 02/02/14 21:06, Charles Steinmetz wrote:

Magnus wrote:


I agree. You need both, but just tossing in a stronger fan isn't going
to cut it either.


Very true.  You need to attack all of the problems to fix it.


Well, maybe not all of them, but there is several key areas that 
together form the basic problem.


We also have ECL dispersed throughout the right side of the board which 
also needs cooling. Most of that can be handled through general cooling 
if only air flows there.


I just checked the lab, and the SR620 standing on the somewhat crowded 
bench and with a SR620 manual laying on top of it was cooler than the 
SR620 sitting up among the other instruments.


As I re-checked the slots for air inlet I was amazed how little there where.

Letting more air in on the back would work if there is large enough air 
at the back, then setting a central baffle so that it at least sweeps 
over the electronics up to the front before being sucked over the CPU 
into the fan.


Also, the fan-noise did not change a lot by drying to damp the top lid, 
so it seems the fan emits it mostly as a direct mode.


The older unit has a high hiss to its fan while the newer has a more 
rattling sound to it.


Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Charles Steinmetz

Magnus wrote:

Also, the fan-noise did not change a lot by drying to damp the top 
lid, so it seems the fan emits it mostly as a direct mode.


The older unit has a high hiss to its fan


That is what I have observed, as well.  The noise is mostly fan blade 
noise and some bearing noise coming directly out the back of the 
unit.  I did not find that the inlet vents were a significant source 
of noise, or that increasing the inlet area caused the fan itself to 
run quieter.


You probably wouldn't notice a substantial noise reduction until you 
got down to a fan with a 15-18 dBA noise rating, and that would move 
so little air that you'd probably fry the counter.  The Papst fan has 
about the same noise rating as the original Delta, so I wouldn't 
expect changing the fan to have a big effect on the amount of noise 
(but, as you say, the noise spectrum may be different).


Best regards,

Charles



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Magnus Danielson

On 02/02/14 21:34, Charles Steinmetz wrote:

Magnus wrote:


These days I would assume that heat-pipes would be used to move heat
to a large external heat-sink. It's fairly cheap these days.


I have looked into heat pipes for several projects, and in the end have
never used them.  The main problem is that almost every solution is
customized.  No trouble if you make PC motherboards or consumer-grade
stereo amplifiers 50,000 at a time.  But most instruments are made in
quantities of 5,000 or fewer, and it takes many years to make (sell)
even that number.  With production numbers that low, the bean counters
want the design and tooling costs amortized over the first year's
production -- maybe only 100-500 units -- which can raise the retail
price by a factor of 2-10.  At least that's what I've found.

Also, most heat-pipe systems are gravity-fed, so if someone turns your
product on its side they can burn it down.  (I almost always operate my
portable spectrum analyzer (HP8591E) standing on its rear feet with
the screen pointing up.  I can foresee someone with limited bench space
setting an SR620 on its side.)


Good point.
The SR620 doesn't stand on the back very well, maybe on it's side, but 
that helps to close the little air intake there is...


Decided to finally toss the original box for my SIA-3000. It's just big 
and messy. Not tossing the SIA-3000 itself, it's a keeper.


Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Magnus Danielson

Charles,

On 02/02/14 21:58, Charles Steinmetz wrote:

Magnus wrote:


Also, the fan-noise did not change a lot by drying to damp the top
lid, so it seems the fan emits it mostly as a direct mode.

The older unit has a high hiss to its fan


That is what I have observed, as well.  The noise is mostly fan blade
noise and some bearing noise coming directly out the back of the unit.
I did not find that the inlet vents were a significant source of noise,
or that increasing the inlet area caused the fan itself to run quieter.

You probably wouldn't notice a substantial noise reduction until you got
down to a fan with a 15-18 dBA noise rating, and that would move so
little air that you'd probably fry the counter.  The Papst fan has about
the same noise rating as the original Delta, so I wouldn't expect
changing the fan to have a big effect on the amount of noise (but, as
you say, the noise spectrum may be different).


Well, with a ball-bearing noise it's much worse than fan-blade hiss.

Fan-blade hiss is result of running at stupid speeds without achieving much.

Anyway, I will probably get a new fan so I might as well get two.

Finding out why the UPS won't start and parts of my rig with it is 
however higher up on the to-do list now that I figured that much out... 
and more clean-up.


Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Tom Knox
With an internal fan I think the covers actually can make a great heat 
exchanger as well. Often an instrument that is overheating will have portions 
of the case still cool.  Perhaps the worst example of fan cooling is the 
Symetricom 512XA Phase Noise Test set. It is kind of out of place on such an 
otherwise exceptional produce. No offense meant to the Symmetricom engineers. 
It is one of the few product I have purchased new.

Thomas Knox



 Date: Sun, 2 Feb 2014 20:42:54 +0100
 From: mag...@rubidium.dyndns.org
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] Replacement fan in SR620
 
 On 02/02/14 20:28, Tom Knox wrote:
 
  A little off topic but It seems many instruments (the SR620 and 53132A 
  included) would work best with an internal fan. (A closed system, not 
  exchanging outside air). Possibly with some sort of internal/external heat 
  sink if needed. Or in high power situations outside air would flow through  
  a hollow heatsink (again not exchanging outside and internal air). If 
  nothing else this would keep dust out. Since fan filters can really 
  restrict air-flow and without a filter dust on circuit boards can act as on 
  thermal insulator leading to overheating while conducting static.
 
 Good point. A fan in a closed environment will survive longer, or 
 rather, won't degrade as fast as it possibly would do.
 
 These days I would assume that heat-pipes would be used to move heat to 
 a large external heat-sink. It's fairly cheap these days.
 
 Cheers,
 Magnus
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
  
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-02 Thread Brooke Clarke

Hi:

There are some radios where the internal power supply heat degrades performance.  The fix is to remove the power supply 
to an external box.


Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.end2partygovernment.com/2012Issues.html


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Replacement fan in SR620

2014-02-01 Thread Magnus Danielson

Fellow time-nuts,

I heard the fan in one of my SR620s and it didn't really was a nice 
sound. Has someone found a good replacement fan? Quieter would be nice.


Considering performance checks and calibration this evening.

Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-01 Thread Tom Knox

Magnus I have a big collector of fans, mostly for Agilent equipment not that it 
matters, let me know the dimensions I will see what I have that is quiet. 
Cheers; 
Thomas Knox



 Date: Sat, 1 Feb 2014 18:56:48 +0100
 From: mag...@rubidium.dyndns.org
 To: time-nuts@febo.com
 Subject: [time-nuts] Replacement fan in SR620
 
 Fellow time-nuts,
 
 I heard the fan in one of my SR620s and it didn't really was a nice 
 sound. Has someone found a good replacement fan? Quieter would be nice.
 
 Considering performance checks and calibration this evening.
 
 Cheers,
 Magnus
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
  
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-01 Thread Volker Esper
Magnus,

I took an EBMPabst 624. Pabst has a good reputation.
The original fan is a Delta Electronics DFB0624 H, Dimensions are 60mm x
60mm x 25mm, DC24V 0.11A

Volker


Am 01.02.2014 18:56, schrieb Magnus Danielson:
 Fellow time-nuts,

 I heard the fan in one of my SR620s and it didn't really was a nice
 sound. Has someone found a good replacement fan? Quieter would be nice.

 Considering performance checks and calibration this evening.

 Cheers,
 Magnus
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Replacement fan in SR620

2014-02-01 Thread Charles Steinmetz

Jarl wrote:


In my SR620 the fan is a Delta DBF0624H.  It is a 60x60x25 mm fan, 24V /0.11A.


Mine, too.  Does anyone know the airflow rating of the Delta fan?  It 
does not appear to move as much air as it should (at least not if the 
idea is to hold the interior at a fixed temperature) -- every SR620 
I've used has the fan running at full speed by the time it's been on 
for 10 minutes, unless the ambient temperature is less than 17-18C.


On the one hand, it would be nice to move more air so the internal 
temperature is more tightly regulated; but on the other hand, it 
would also be nice if the fan were quieter.  It is unlikely we can 
have both.  The fact that the original fan seems marginal counsels 
against replacing it with one that moves less air.


The thermal design of the 620 is far from optimal.  The thermistor is 
located in a tunnel between the interior of the instrument and the 
exterior, with the fan blowing directly across it through the 
tunnel.  So, the fan startup is an ugly process of fits and starts as 
the instrument warms up.  Also, the air inlets seem to be too small 
and the internal airflow was not properly designed to circulate 
cooling air where it needs to go.


I have toyed with the idea of cutting a slot maybe 4 long and 1/8 
tall in the right wall of the chassis at the rear, above the four 
TO-220 devices mounted there -- and perhaps another slot about the 
same size in the top cover above those devices.  Also, maybe 
attaching some internal baffles to the top cover to channel airflow 
where it is needed.


I installed a terminal strip on the GPIB connector mounting screw and 
relocated the thermistor there in my SR620s (in the general vicinity 
of the oscillators).  (However, note that stabilizing the internal 
oscillator temperature is not really very important for most of us, 
because time nuts generally use an external time base.  In that case, 
it is probably more important that the temperature of the triggers 
and interpolators is held constant.)


With the thermistors relocated as I have described, the fans start up 
as they should (monotonically, speeding up smoothly from stopped to 
full speed without any fits and starts).  They still reach full speed 
in 10 minutes or so, so at the end of the day I don't think I've 
really changed anything except the aesthetics of the fan 
startup.  (IMO, the change is worth it just for that, but there does 
not seem to be any operational improvement.)


Perhaps SRS did not intend to regulate the interior temperature of 
the SR620 -- maybe they just wanted it to warm up faster (if you did 
away with the thermistor and had the fan run full speed whenever the 
counter was on, it would presumably take longer to warm up).


Anyway -- does anybody have an old Delta catalog or datasheet that 
specifies the airflow rating of the original Delta fan?


Best regards,

Charles



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.