Re: [time-nuts] Why would Keysight UK uncertainty measuring 1 MHz be as high as 7.6 Hz?

2015-09-30 Thread Dr. David Kirkby (Kirkby Microwave Ltd)
On 28 August 2015 at 21:48, Dr. David Kirkby (Kirkby Microwave Ltd) <
drkir...@kirkbymicrowave.co.uk> wrote:

> My LCR meter came back from Keysight  UK last week, where it was
> calibrated. This instrument works at various frequencies from 20 Hz to 1
> MHz, so obviously has some sort of oscillator in it. But I don't think the
> absolute accuracy on frequency is important on this, as it does not even
> have the ability to set to an arbitrary frequency. There are only 8000 or
> so steps, and at the high end, some of those steps are more than 100 kHz
> apart!!!  So clearly frequency accuracy on this instrument is not that
> important.
>
> Anyway, the cal certificate, a copy of which I put here
>
>
> http://www.kirkbymicrowave.co.uk/cal_certificates/Keysight-standard-calibration-with-uncertainties-for-4284A-precison-LCR-meter-18-08-2015.pdf
>
> shows on page 5 that it was checked at 1, 8, 20, 80, 400 kHz, and 1 MHz.
> But the uncertainty reported (7.6 Hz) seems extremely high, given they used
> a 53132A counter as a working standard, and a 5071A primary frequency
> standard. Why should the uncertainty be so high? Am I missing something?
>

I don't know if this thread is closed, but I did get a telephone call from
someone at Keysight today, in response to my query on this matter.
Essentially

1) At the particular calibration centre in the UK, the frequency  counter
used would have been locked to an external reference, as are all the
counters they use at that particular calibration centre in the UK. The
reference is an old HP  5071 cesium, which is regularly compared with GPS.
Their best uncertainty is 5 x 10^-13.

2) The LCR meter tolerance on the 1 MHz is +/- 100 Hz. This means it does
not need a particularly good counter.

3) To enable Keysight to use one calibration procedure world-wide, that
calibration procedure is written to allow the uncertainty to be as high as
the worst case with a counter running on its own internal oven.

So in essence, the meter when calibrated in the UK would have been measured
to far better than the +/- 7.6 Hz indicated on the calibration certificate.
but by specifying +/- 7.6 Hz, it allows the same procedure to be used
anywhere in the world.

Dave
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Why would Keysight UK uncertainty measuring 1 MHz be as high as 7.6 Hz?

2015-08-30 Thread Scott McGrath
Look at the service manual,  the calibration setup listed is usually pretty 
close to what Keysight is using in their local calibration facility.   For an 
LCR meter I doubt they would be using a time nuts grade counter

Content by Scott
Typos by Siri

 On Aug 29, 2015, at 10:02 PM, Bob Camp kb...@n1k.org wrote:
 
 Hi
 
 Well, since we all have made our totally uninformed guesses, the only 
 thing to do now is to give Keysight a call and see what the real answer 
 is. 
 
 Bob
 
 
 On Aug 29, 2015, at 1:46 PM, Dr. David Kirkby (Kirkby Microwave Ltd) 
 drkir...@kirkbymicrowave.co.uk wrote:
 
 On 29 August 2015 at 12:59, Javier Herrero jherr...@hvsistemas.es wrote:
 
 Hello,
 
 The calibration certificate does not indicate that the measurements were
 done with the frequency counters referenced to the 5071A at the time of
 calbiration (if so, it would be listed under the Calibration Equipment Used
 table). It says that the 53132A were calibrated against the 5071A.
 
 If for your calibration they have used 53132A witout the oven oscillator
 option it is very probable that its uncertainity is 7.6ppm as indicated in
 the certificate. Since the maximum error tolerable for the LCR meter is
 100ppm (+/-100Hz @ 1MHz), it makes sense to perform the measurement with an
 instrument with an uncertainity of 7.6ppm, and not to use the better
 counter in the lab for that purpose.
 
 Regards,
 
 Javier
 
 
 I would find it a bit hard to believe they would use a counter without an
 oven in their lab, as it would seriously restrict what they can do with it,
 making it more difficult to replace one counter with another. I would have
 thought that within reason it best to have the lab have reasonably high
 spec kit, so more than one instrument could be done on the same line. They
 did for example use a pair of 3458As, despite I'm sure the voltage accuracy
 requirements could be met with a multimeter with far greater uncertainty
 than an expensive 3458A. It makes more sense (within reason) to have 3458As
 in the cal lab, as it allows a wider range of instruments to be calibrated.
 
 Also, if you consider the spec on the 53132A without an oven, it is 3 x
 10^-7 per month. So after 12 months that could be 12 * 3 10^-7 or 3.6
 10^-6, so if it did drift the maximum amount each month for a year, the
 uncertainty would higher than it actually is.
 
 I intended to contact Keysight about the calibration for a couple of other
 reasons
 
 1) I would like to know if it was adjusted or not. That is not clear from
 the cal certificate, since the
 
 * As received condition  - Not applicable, as this calibration certificate
 applies to the initial calibration of a new, refurbished or upgraded
 equipment.
 * Action taken  - The equipment was upgraded.
 
 I doubt it has seen a cal lab in ages.
 
 The upgrade was just a software one, to enable cable lengths of 2 m and 4 m
 (option 006) to be used to connect the DUT, which they kindly provided free
 of charge, on the condition I paid for the calibration.
 
 2) They never put any stickers over the screws that prevent the covers
 being removed, which struck me as a bit odd.
 
 Since I was going to ask about those two issues, I will ask about the
 uncertainty on frequency too. It will be interesting what response I get.
 I'm just interested -  I realize that this instrument does not demand much
 of the counter used to calibrate it. The demanding calibration devices
 would be the resistance and capacitance standards.
 
 Dave
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Why would Keysight UK uncertainty measuring 1 MHz be as high as 7.6 Hz?

2015-08-29 Thread Dr. David Kirkby (Kirkby Microwave Ltd)
On 28 Aug 2015 23:05, Oz-in-DFW li...@ozindfw.net wrote:

 The uncertainly listed seems to be 7.6 mHz (milliHertz, or .0076 Hz.  A
 bit better that you mention..

No, please look again.

The first line does show an uncertainty of 7.6 mHz, but that is when the
LCR meter was set to 1 kHz. The last line shows an uncertainty of 7.6 Hz
(1000x higher) when measuring 1 MHz, which is obviously 1000x higher in
frequency than 1 kHz. So the original uncertainty I quoted was correct.
The uncertainty rises proportionately with frequency.

Dave.

Dave.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Why would Keysight UK uncertainty measuring 1 MHz be as high as 7.6 Hz?

2015-08-29 Thread Magnus Danielson

Hi,

On 08/29/2015 10:01 AM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote:

On 28 Aug 2015 23:05, Oz-in-DFW li...@ozindfw.net wrote:


The uncertainly listed seems to be 7.6 mHz (milliHertz, or .0076 Hz.  A
bit better that you mention..


No, please look again.

The first line does show an uncertainty of 7.6 mHz, but that is when the
LCR meter was set to 1 kHz. The last line shows an uncertainty of 7.6 Hz
(1000x higher) when measuring 1 MHz, which is obviously 1000x higher in
frequency than 1 kHz. So the original uncertainty I quoted was correct.
The uncertainty rises proportionately with frequency.


I agree. That's how I read it too, and it is very obvious as you look at 
the table.


The calibration is in line with Chapter 10 of the manual:
http://wiki.epfl.ch/carplat/documents/hp4284a_lcr_manual.pdf

It might be that the 7.6 ppm number was a practical uncertainty measure 
found in deeper analysis along the lines of GUM and was added to the 
calibration chart afterwards. For the peak-error of 100 ppm, this 
provide a good margin anyway.


So, it's the instrument and the way it behaves, nothing in their 
test-setup which is way better than needed for this particular instrument.


In short, don't worry about it, it's OK.

Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Why would Keysight UK uncertainty measuring 1 MHz be as high as 7.6 Hz?

2015-08-29 Thread Javier Herrero

Hello,

The calibration certificate does not indicate that the measurements were 
done with the frequency counters referenced to the 5071A at the time of 
calbiration (if so, it would be listed under the Calibration Equipment 
Used table). It says that the 53132A were calibrated against the 5071A.


If for your calibration they have used 53132A witout the oven oscillator 
option it is very probable that its uncertainity is 7.6ppm as indicated 
in the certificate. Since the maximum error tolerable for the LCR meter 
is 100ppm (+/-100Hz @ 1MHz), it makes sense to perform the measurement 
with an instrument with an uncertainity of 7.6ppm, and not to use the 
better counter in the lab for that purpose.


Regards,

Javier

On 28/08/2015 22:48, Dr. David Kirkby (Kirkby Microwave Ltd) wrote:

My LCR meter came back from Keysight  UK last week, where it was
calibrated. This instrument works at various frequencies from 20 Hz to 1
MHz, so obviously has some sort of oscillator in it. But I don't think the
absolute accuracy on frequency is important on this, as it does not even
have the ability to set to an arbitrary frequency. There are only 8000 or
so steps, and at the high end, some of those steps are more than 100 kHz
apart!!!  So clearly frequency accuracy on this instrument is not that
important.

Anyway, the cal certificate, a copy of which I put here

http://www.kirkbymicrowave.co.uk/cal_certificates/Keysight-standard-calibration-with-uncertainties-for-4284A-precison-LCR-meter-18-08-2015.pdf

shows on page 5 that it was checked at 1, 8, 20, 80, 400 kHz, and 1 MHz.
But the uncertainty reported (7.6 Hz) seems extremely high, given they used
a 53132A counter as a working standard, and a 5071A primary frequency
standard. Why should the uncertainty be so high? Am I missing something?

When they done my VNA last year

http://www.kirkbymicrowave.co.uk/cal_certificates/Keysight-standard-calibration-with-uncertainties-for-8720D-vector-network-analyzer-16-09-2014.pdf

the uncertainty on frequency was about 5 orders of magnitude better than
that. The 10 MHz timebase was measured with an uncertainty of 0.0010 Hz.

I checked the Keysight UK accreditation (by UKAS) for frequency

http://www.keysight.com/upload/cmc_upload/All/UKAS_S_2015-08-14_Eng.pdf

and see over the range 0.1 Hz to 500 MHz, which covers the LCR meter, their
accreditation is 6.0 in 10^11 + 0.020 nHz.

I can't believe they are unable to measure better than 7.6 ppm on
frequency, so are wondering why the uncertainty is so high, even though I
am sure such an uncertainly is very acceptable for this application.

It is either an error on the cal certificate, or I am missing something. I
expect it is the latter, and hoping someone here can fill me in.

Dave
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Why would Keysight UK uncertainty measuring 1 MHz be as high as 7.6 Hz?

2015-08-29 Thread Bob Camp
Hi


 On Aug 29, 2015, at 7:09 AM, Magnus Danielson mag...@rubidium.dyndns.org 
 wrote:
 
 Hi,
 
 On 08/29/2015 10:01 AM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote:
 On 28 Aug 2015 23:05, Oz-in-DFW li...@ozindfw.net wrote:
 
 The uncertainly listed seems to be 7.6 mHz (milliHertz, or .0076 Hz.  A
 bit better that you mention..
 
 No, please look again.
 
 The first line does show an uncertainty of 7.6 mHz, but that is when the
 LCR meter was set to 1 kHz. The last line shows an uncertainty of 7.6 Hz
 (1000x higher) when measuring 1 MHz, which is obviously 1000x higher in
 frequency than 1 kHz. So the original uncertainty I quoted was correct.
 The uncertainty rises proportionately with frequency.
 
 I agree. That's how I read it too, and it is very obvious as you look at the 
 table.
 
 The calibration is in line with Chapter 10 of the manual:
 http://wiki.epfl.ch/carplat/documents/hp4284a_lcr_manual.pdf
 
 It might be that the 7.6 ppm number was a practical uncertainty measure found 
 in deeper analysis along the lines of GUM and was added to the calibration 
 chart afterwards. For the peak-error of 100 ppm, this provide a good margin 
 anyway.

If it’s a +/- 100 ppm error and that’s all due to temperature, you get a linear 
“best guess” of 2 ppm / C. 

That would equate to needing to know the temperature to about 3.5 C. That’s 
doing pretty well in an open
room. If you are guessing a temperature inside a piece of gear, that’s doing 
pretty well running inside a chamber. 

Bob

 
 So, it's the instrument and the way it behaves, nothing in their test-setup 
 which is way better than needed for this particular instrument.
 
 In short, don't worry about it, it's OK.
 
 Cheers,
 Magnus
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Why would Keysight UK uncertainty measuring 1 MHz be as high as 7.6 Hz?

2015-08-29 Thread Dr. David Kirkby (Kirkby Microwave Ltd)
On 29 August 2015 at 12:59, Javier Herrero jherr...@hvsistemas.es wrote:

 Hello,

 The calibration certificate does not indicate that the measurements were
 done with the frequency counters referenced to the 5071A at the time of
 calbiration (if so, it would be listed under the Calibration Equipment Used
 table). It says that the 53132A were calibrated against the 5071A.

 If for your calibration they have used 53132A witout the oven oscillator
 option it is very probable that its uncertainity is 7.6ppm as indicated in
 the certificate. Since the maximum error tolerable for the LCR meter is
 100ppm (+/-100Hz @ 1MHz), it makes sense to perform the measurement with an
 instrument with an uncertainity of 7.6ppm, and not to use the better
 counter in the lab for that purpose.

 Regards,

 Javier


I would find it a bit hard to believe they would use a counter without an
oven in their lab, as it would seriously restrict what they can do with it,
making it more difficult to replace one counter with another. I would have
thought that within reason it best to have the lab have reasonably high
spec kit, so more than one instrument could be done on the same line. They
did for example use a pair of 3458As, despite I'm sure the voltage accuracy
requirements could be met with a multimeter with far greater uncertainty
than an expensive 3458A. It makes more sense (within reason) to have 3458As
in the cal lab, as it allows a wider range of instruments to be calibrated.

Also, if you consider the spec on the 53132A without an oven, it is 3 x
10^-7 per month. So after 12 months that could be 12 * 3 10^-7 or 3.6
10^-6, so if it did drift the maximum amount each month for a year, the
uncertainty would higher than it actually is.

I intended to contact Keysight about the calibration for a couple of other
reasons

1) I would like to know if it was adjusted or not. That is not clear from
the cal certificate, since the

* As received condition  - Not applicable, as this calibration certificate
applies to the initial calibration of a new, refurbished or upgraded
equipment.
* Action taken  - The equipment was upgraded.

I doubt it has seen a cal lab in ages.

The upgrade was just a software one, to enable cable lengths of 2 m and 4 m
(option 006) to be used to connect the DUT, which they kindly provided free
of charge, on the condition I paid for the calibration.

2) They never put any stickers over the screws that prevent the covers
being removed, which struck me as a bit odd.

Since I was going to ask about those two issues, I will ask about the
uncertainty on frequency too. It will be interesting what response I get.
I'm just interested -  I realize that this instrument does not demand much
of the counter used to calibrate it. The demanding calibration devices
would be the resistance and capacitance standards.

Dave
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Why would Keysight UK uncertainty measuring 1 MHz be as high as 7.6 Hz?

2015-08-29 Thread Bob Camp
Hi

Well, since we all have made our totally uninformed guesses, the only 
thing to do now is to give Keysight a call and see what the real answer 
is. 

Bob


 On Aug 29, 2015, at 1:46 PM, Dr. David Kirkby (Kirkby Microwave Ltd) 
 drkir...@kirkbymicrowave.co.uk wrote:
 
 On 29 August 2015 at 12:59, Javier Herrero jherr...@hvsistemas.es wrote:
 
 Hello,
 
 The calibration certificate does not indicate that the measurements were
 done with the frequency counters referenced to the 5071A at the time of
 calbiration (if so, it would be listed under the Calibration Equipment Used
 table). It says that the 53132A were calibrated against the 5071A.
 
 If for your calibration they have used 53132A witout the oven oscillator
 option it is very probable that its uncertainity is 7.6ppm as indicated in
 the certificate. Since the maximum error tolerable for the LCR meter is
 100ppm (+/-100Hz @ 1MHz), it makes sense to perform the measurement with an
 instrument with an uncertainity of 7.6ppm, and not to use the better
 counter in the lab for that purpose.
 
 Regards,
 
 Javier
 
 
 I would find it a bit hard to believe they would use a counter without an
 oven in their lab, as it would seriously restrict what they can do with it,
 making it more difficult to replace one counter with another. I would have
 thought that within reason it best to have the lab have reasonably high
 spec kit, so more than one instrument could be done on the same line. They
 did for example use a pair of 3458As, despite I'm sure the voltage accuracy
 requirements could be met with a multimeter with far greater uncertainty
 than an expensive 3458A. It makes more sense (within reason) to have 3458As
 in the cal lab, as it allows a wider range of instruments to be calibrated.
 
 Also, if you consider the spec on the 53132A without an oven, it is 3 x
 10^-7 per month. So after 12 months that could be 12 * 3 10^-7 or 3.6
 10^-6, so if it did drift the maximum amount each month for a year, the
 uncertainty would higher than it actually is.
 
 I intended to contact Keysight about the calibration for a couple of other
 reasons
 
 1) I would like to know if it was adjusted or not. That is not clear from
 the cal certificate, since the
 
 * As received condition  - Not applicable, as this calibration certificate
 applies to the initial calibration of a new, refurbished or upgraded
 equipment.
 * Action taken  - The equipment was upgraded.
 
 I doubt it has seen a cal lab in ages.
 
 The upgrade was just a software one, to enable cable lengths of 2 m and 4 m
 (option 006) to be used to connect the DUT, which they kindly provided free
 of charge, on the condition I paid for the calibration.
 
 2) They never put any stickers over the screws that prevent the covers
 being removed, which struck me as a bit odd.
 
 Since I was going to ask about those two issues, I will ask about the
 uncertainty on frequency too. It will be interesting what response I get.
 I'm just interested -  I realize that this instrument does not demand much
 of the counter used to calibrate it. The demanding calibration devices
 would be the resistance and capacitance standards.
 
 Dave
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Why would Keysight UK uncertainty measuring 1 MHz be as high as 7.6 Hz?

2015-08-28 Thread Oz-in-DFW
The uncertainly listed seems to be 7.6 mHz (milliHertz, or .0076 Hz.  A
bit better that you mention..

On 8/28/2015 3:48 PM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote:
 My LCR meter came back from Keysight  UK last week, where it was
 calibrated. This instrument works at various frequencies from 20 Hz to 1
 MHz, so obviously has some sort of oscillator in it. But I don't think the
 absolute accuracy on frequency is important on this, as it does not even
 have the ability to set to an arbitrary frequency. There are only 8000 or
 so steps, and at the high end, some of those steps are more than 100 kHz
 apart!!!  So clearly frequency accuracy on this instrument is not that
 important.

 Anyway, the cal certificate, a copy of which I put here

 http://www.kirkbymicrowave.co.uk/cal_certificates/Keysight-standard-calibration-with-uncertainties-for-4284A-precison-LCR-meter-18-08-2015.pdf

 shows on page 5 that it was checked at 1, 8, 20, 80, 400 kHz, and 1 MHz.
 But the uncertainty reported (7.6 Hz) seems extremely high, given they used
 a 53132A counter as a working standard, and a 5071A primary frequency
 standard. Why should the uncertainty be so high? Am I missing something?

 When they done my VNA last year

 http://www.kirkbymicrowave.co.uk/cal_certificates/Keysight-standard-calibration-with-uncertainties-for-8720D-vector-network-analyzer-16-09-2014.pdf

 the uncertainty on frequency was about 5 orders of magnitude better than
 that. The 10 MHz timebase was measured with an uncertainty of 0.0010 Hz.

 I checked the Keysight UK accreditation (by UKAS) for frequency

 http://www.keysight.com/upload/cmc_upload/All/UKAS_S_2015-08-14_Eng.pdf

 and see over the range 0.1 Hz to 500 MHz, which covers the LCR meter, their
 accreditation is 6.0 in 10^11 + 0.020 nHz.

 I can't believe they are unable to measure better than 7.6 ppm on
 frequency, so are wondering why the uncertainty is so high, even though I
 am sure such an uncertainly is very acceptable for this application.

 It is either an error on the cal certificate, or I am missing something. I
 expect it is the latter, and hoping someone here can fill me in.

 Dave
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


-- 
mailto:o...@ozindfw.net
Oz
POB 93167 
Southlake, TX 76092 (Near DFW Airport) 



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Why would Keysight UK uncertainty measuring 1 MHz be as high as 7.6 Hz?

2015-08-28 Thread Bob Camp
Hi

If it is an un-compensated crystal oscillator (I’d bet it is) there will be a 
maximum temperature slope to the part.
In order to have absolute calibration, you would need to know both the 
frequency read *and* the temperature 
of the inside of the device. I’d bet somebody did the math, looked at the 
“rated” temperature control in the cal 
lab area and came up with a proper number for the ambiguity of the calibration. 
If the slope is 2 ppm / C then the 
number you have implies a 3.5C temperature control in the area. That seems 
pretty tight. A 1 ppm / C number gets
you to 7.6 C. That seems about right for the room plus the ambiguity in the 
internal temperature rise inside the box. 

Bob

 On Aug 28, 2015, at 4:48 PM, Dr. David Kirkby (Kirkby Microwave Ltd) 
 drkir...@kirkbymicrowave.co.uk wrote:
 
 My LCR meter came back from Keysight  UK last week, where it was
 calibrated. This instrument works at various frequencies from 20 Hz to 1
 MHz, so obviously has some sort of oscillator in it. But I don't think the
 absolute accuracy on frequency is important on this, as it does not even
 have the ability to set to an arbitrary frequency. There are only 8000 or
 so steps, and at the high end, some of those steps are more than 100 kHz
 apart!!!  So clearly frequency accuracy on this instrument is not that
 important.
 
 Anyway, the cal certificate, a copy of which I put here
 
 http://www.kirkbymicrowave.co.uk/cal_certificates/Keysight-standard-calibration-with-uncertainties-for-4284A-precison-LCR-meter-18-08-2015.pdf
 
 shows on page 5 that it was checked at 1, 8, 20, 80, 400 kHz, and 1 MHz.
 But the uncertainty reported (7.6 Hz) seems extremely high, given they used
 a 53132A counter as a working standard, and a 5071A primary frequency
 standard. Why should the uncertainty be so high? Am I missing something?
 
 When they done my VNA last year
 
 http://www.kirkbymicrowave.co.uk/cal_certificates/Keysight-standard-calibration-with-uncertainties-for-8720D-vector-network-analyzer-16-09-2014.pdf
 
 the uncertainty on frequency was about 5 orders of magnitude better than
 that. The 10 MHz timebase was measured with an uncertainty of 0.0010 Hz.
 
 I checked the Keysight UK accreditation (by UKAS) for frequency
 
 http://www.keysight.com/upload/cmc_upload/All/UKAS_S_2015-08-14_Eng.pdf
 
 and see over the range 0.1 Hz to 500 MHz, which covers the LCR meter, their
 accreditation is 6.0 in 10^11 + 0.020 nHz.
 
 I can't believe they are unable to measure better than 7.6 ppm on
 frequency, so are wondering why the uncertainty is so high, even though I
 am sure such an uncertainly is very acceptable for this application.
 
 It is either an error on the cal certificate, or I am missing something. I
 expect it is the latter, and hoping someone here can fill me in.
 
 Dave
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Why would Keysight UK uncertainty measuring 1 MHz be as high as 7.6 Hz?

2015-08-28 Thread Dr. David Kirkby (Kirkby Microwave Ltd)
My LCR meter came back from Keysight  UK last week, where it was
calibrated. This instrument works at various frequencies from 20 Hz to 1
MHz, so obviously has some sort of oscillator in it. But I don't think the
absolute accuracy on frequency is important on this, as it does not even
have the ability to set to an arbitrary frequency. There are only 8000 or
so steps, and at the high end, some of those steps are more than 100 kHz
apart!!!  So clearly frequency accuracy on this instrument is not that
important.

Anyway, the cal certificate, a copy of which I put here

http://www.kirkbymicrowave.co.uk/cal_certificates/Keysight-standard-calibration-with-uncertainties-for-4284A-precison-LCR-meter-18-08-2015.pdf

shows on page 5 that it was checked at 1, 8, 20, 80, 400 kHz, and 1 MHz.
But the uncertainty reported (7.6 Hz) seems extremely high, given they used
a 53132A counter as a working standard, and a 5071A primary frequency
standard. Why should the uncertainty be so high? Am I missing something?

When they done my VNA last year

http://www.kirkbymicrowave.co.uk/cal_certificates/Keysight-standard-calibration-with-uncertainties-for-8720D-vector-network-analyzer-16-09-2014.pdf

the uncertainty on frequency was about 5 orders of magnitude better than
that. The 10 MHz timebase was measured with an uncertainty of 0.0010 Hz.

I checked the Keysight UK accreditation (by UKAS) for frequency

http://www.keysight.com/upload/cmc_upload/All/UKAS_S_2015-08-14_Eng.pdf

and see over the range 0.1 Hz to 500 MHz, which covers the LCR meter, their
accreditation is 6.0 in 10^11 + 0.020 nHz.

I can't believe they are unable to measure better than 7.6 ppm on
frequency, so are wondering why the uncertainty is so high, even though I
am sure such an uncertainly is very acceptable for this application.

It is either an error on the cal certificate, or I am missing something. I
expect it is the latter, and hoping someone here can fill me in.

Dave
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.