Re: [time-nuts] 5370B OCXO

2010-03-12 Thread Bob Camp
Hi

I guess the real question is what a better OCXO would have actually cost. 

If the 60111 was a test ten, get ten sort of thing (I'm guessing it was) - was 
a better part simply a test 10 get 9 issue?

The claim was made that short term stability testing could be done directly in 
the aging racks. It's not real clear what the actual cost of an extended test / 
sort would have been. 

Bob


On Mar 11, 2010, at 10:43 PM, John Miles wrote:

 Many if not most 5370-based measurements are based on differential timing
 between the START and STOP channels, and wouldn't benefit from a better 10
 MHz reference.  If a customer did need something better, they probably
 already had a house standard to pipe in the back... and if not, HP would
 have been able to sell them one.  It made more sense to keep the cost down
 by not including a high-end OCXO that would have gone unappreciated by most
 users.
 
 The 5370's jitter+resolution floor doesn't allow it to reach 1E-11 at t=1s
 in any event, so the -60111 wouldn't have been the limiting factor in the
 short term.
 
 One valid question, though, is why they bothered to put the nicer
 10811-60109 OCXOs in the post-2120 series 5065A models, where its short-term
 performance is hosed by tying it to the rubidium reference with a ~1 Hz
 loop.  Those 5065As would have been OK with a -60111, at least in the
 pre-2632 serial #s with the original integrator board.  I'd be curious to
 know if they lowered the loop BW when they respun the integrator PCB.
 
 -- john, KE5FX
 
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com]on
 Behalf Of Bob Camp
 Sent: Thursday, March 11, 2010 7:11 PM
 To: Discussion of precise time and frequency measurement
 Subject: [time-nuts] 5370B OCXO
 
 
 Hi
 
 The OCXO in the 5370B is a 10811-60111. The only added spec on it
 is a 1x10^-11 ADEV spec at 1 second. By modern standards that's
 not a real tight spec. There are other 10811's with tighter specs
 on them at 1 second. My guess is that it was not a real tight
 spec for the 10811 to hit.
 
 The short term would appear to contribute to the total error on
 the counter. Why not put a better oscillator in it?
 
 Bob
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 5370B OCXO

2010-03-12 Thread paul swed
If I understand this thread correctly.
I would speculate it was simply a business choice. 1 less part type to
manage.

On Fri, Mar 12, 2010 at 7:30 AM, Bob Camp li...@rtty.us wrote:

 Hi

 I guess the real question is what a better OCXO would have actually cost.

 If the 60111 was a test ten, get ten sort of thing (I'm guessing it was) -
 was a better part simply a test 10 get 9 issue?

 The claim was made that short term stability testing could be done directly
 in the aging racks. It's not real clear what the actual cost of an extended
 test / sort would have been.

 Bob


 On Mar 11, 2010, at 10:43 PM, John Miles wrote:

  Many if not most 5370-based measurements are based on differential timing
  between the START and STOP channels, and wouldn't benefit from a better
 10
  MHz reference.  If a customer did need something better, they probably
  already had a house standard to pipe in the back... and if not, HP would
  have been able to sell them one.  It made more sense to keep the cost
 down
  by not including a high-end OCXO that would have gone unappreciated by
 most
  users.
 
  The 5370's jitter+resolution floor doesn't allow it to reach 1E-11 at
 t=1s
  in any event, so the -60111 wouldn't have been the limiting factor in the
  short term.
 
  One valid question, though, is why they bothered to put the nicer
  10811-60109 OCXOs in the post-2120 series 5065A models, where its
 short-term
  performance is hosed by tying it to the rubidium reference with a ~1 Hz
  loop.  Those 5065As would have been OK with a -60111, at least in the
  pre-2632 serial #s with the original integrator board.  I'd be curious to
  know if they lowered the loop BW when they respun the integrator PCB.
 
  -- john, KE5FX
 
 
  -Original Message-
  From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com]on
  Behalf Of Bob Camp
  Sent: Thursday, March 11, 2010 7:11 PM
  To: Discussion of precise time and frequency measurement
  Subject: [time-nuts] 5370B OCXO
 
 
  Hi
 
  The OCXO in the 5370B is a 10811-60111. The only added spec on it
  is a 1x10^-11 ADEV spec at 1 second. By modern standards that's
  not a real tight spec. There are other 10811's with tighter specs
  on them at 1 second. My guess is that it was not a real tight
  spec for the 10811 to hit.
 
  The short term would appear to contribute to the total error on
  the counter. Why not put a better oscillator in it?
 
  Bob
 
 
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 
 
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 5370B OCXO

2010-03-12 Thread Bob Camp
Hi

Tough to believe that HP worried a lot about SKU inflation back when they
did the 5370 :)

I'm assuming that the 5370 was a Santa Clara design. That would put the
counter designers down the hall from the oscillator factory. Unlikely that
there was a communications gap about what could or could not be done.

You may well be correct though. Setting up and managing another part is the
most likely reason why not to add a couple more tests.

Bob

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of paul swed
Sent: Friday, March 12, 2010 8:51 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] 5370B OCXO

If I understand this thread correctly.
I would speculate it was simply a business choice. 1 less part type to
manage.

On Fri, Mar 12, 2010 at 7:30 AM, Bob Camp li...@rtty.us wrote:

 Hi

 I guess the real question is what a better OCXO would have actually
cost.

 If the 60111 was a test ten, get ten sort of thing (I'm guessing it was) -
 was a better part simply a test 10 get 9 issue?

 The claim was made that short term stability testing could be done
directly
 in the aging racks. It's not real clear what the actual cost of an
extended
 test / sort would have been.

 Bob


 On Mar 11, 2010, at 10:43 PM, John Miles wrote:

  Many if not most 5370-based measurements are based on differential
timing
  between the START and STOP channels, and wouldn't benefit from a better
 10
  MHz reference.  If a customer did need something better, they probably
  already had a house standard to pipe in the back... and if not, HP would
  have been able to sell them one.  It made more sense to keep the cost
 down
  by not including a high-end OCXO that would have gone unappreciated by
 most
  users.
 
  The 5370's jitter+resolution floor doesn't allow it to reach 1E-11 at
 t=1s
  in any event, so the -60111 wouldn't have been the limiting factor in
the
  short term.
 
  One valid question, though, is why they bothered to put the nicer
  10811-60109 OCXOs in the post-2120 series 5065A models, where its
 short-term
  performance is hosed by tying it to the rubidium reference with a ~1 Hz
  loop.  Those 5065As would have been OK with a -60111, at least in the
  pre-2632 serial #s with the original integrator board.  I'd be curious
to
  know if they lowered the loop BW when they respun the integrator PCB.
 
  -- john, KE5FX
 
 
  -Original Message-
  From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com]on
  Behalf Of Bob Camp
  Sent: Thursday, March 11, 2010 7:11 PM
  To: Discussion of precise time and frequency measurement
  Subject: [time-nuts] 5370B OCXO
 
 
  Hi
 
  The OCXO in the 5370B is a 10811-60111. The only added spec on it
  is a 1x10^-11 ADEV spec at 1 second. By modern standards that's
  not a real tight spec. There are other 10811's with tighter specs
  on them at 1 second. My guess is that it was not a real tight
  spec for the 10811 to hit.
 
  The short term would appear to contribute to the total error on
  the counter. Why not put a better oscillator in it?
 
  Bob
 
 
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 
 
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 5370B OCXO

2010-03-12 Thread paul swed
Well at least today its a very key reality. Maybe not in the 80s.
But I work for a large company that has 10s of thousands of parts and its a
very real drive to remove different parts for more commonality.
I would bet indeed business drove the position

On Fri, Mar 12, 2010 at 12:18 PM, Bob Camp li...@rtty.us wrote:

 Hi

 Tough to believe that HP worried a lot about SKU inflation back when they
 did the 5370 :)

 I'm assuming that the 5370 was a Santa Clara design. That would put the
 counter designers down the hall from the oscillator factory. Unlikely that
 there was a communications gap about what could or could not be done.

 You may well be correct though. Setting up and managing another part is the
 most likely reason why not to add a couple more tests.

 Bob

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of paul swed
 Sent: Friday, March 12, 2010 8:51 AM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] 5370B OCXO

 If I understand this thread correctly.
 I would speculate it was simply a business choice. 1 less part type to
 manage.

 On Fri, Mar 12, 2010 at 7:30 AM, Bob Camp li...@rtty.us wrote:

  Hi
 
  I guess the real question is what a better OCXO would have actually
 cost.
 
  If the 60111 was a test ten, get ten sort of thing (I'm guessing it was)
 -
  was a better part simply a test 10 get 9 issue?
 
  The claim was made that short term stability testing could be done
 directly
  in the aging racks. It's not real clear what the actual cost of an
 extended
  test / sort would have been.
 
  Bob
 
 
  On Mar 11, 2010, at 10:43 PM, John Miles wrote:
 
   Many if not most 5370-based measurements are based on differential
 timing
   between the START and STOP channels, and wouldn't benefit from a better
  10
   MHz reference.  If a customer did need something better, they probably
   already had a house standard to pipe in the back... and if not, HP
 would
   have been able to sell them one.  It made more sense to keep the cost
  down
   by not including a high-end OCXO that would have gone unappreciated by
  most
   users.
  
   The 5370's jitter+resolution floor doesn't allow it to reach 1E-11 at
  t=1s
   in any event, so the -60111 wouldn't have been the limiting factor in
 the
   short term.
  
   One valid question, though, is why they bothered to put the nicer
   10811-60109 OCXOs in the post-2120 series 5065A models, where its
  short-term
   performance is hosed by tying it to the rubidium reference with a ~1 Hz
   loop.  Those 5065As would have been OK with a -60111, at least in the
   pre-2632 serial #s with the original integrator board.  I'd be curious
 to
   know if they lowered the loop BW when they respun the integrator PCB.
  
   -- john, KE5FX
  
  
   -Original Message-
   From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com
 ]On
   Behalf Of Bob Camp
   Sent: Thursday, March 11, 2010 7:11 PM
   To: Discussion of precise time and frequency measurement
   Subject: [time-nuts] 5370B OCXO
  
  
   Hi
  
   The OCXO in the 5370B is a 10811-60111. The only added spec on it
   is a 1x10^-11 ADEV spec at 1 second. By modern standards that's
   not a real tight spec. There are other 10811's with tighter specs
   on them at 1 second. My guess is that it was not a real tight
   spec for the 10811 to hit.
  
   The short term would appear to contribute to the total error on
   the counter. Why not put a better oscillator in it?
  
   Bob
  
  
  
   ___
   time-nuts mailing list -- time-nuts@febo.com
   To unsubscribe, go to
   https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
   and follow the instructions there.
  
  
  
   ___
   time-nuts mailing list -- time-nuts@febo.com
   To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
   and follow the instructions there.
  
 
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 5370B OCXO

2010-03-12 Thread Bob Camp
Hi

I'm looking at things like the plastic shaft pot with switch next to the very 
similar metal shaft pot with switch on the front panel. Not real likely they 
were into modern supply practices 

Bob


On Mar 12, 2010, at 3:44 PM, paul swed wrote:

 Well at least today its a very key reality. Maybe not in the 80s.
 But I work for a large company that has 10s of thousands of parts and its a
 very real drive to remove different parts for more commonality.
 I would bet indeed business drove the position
 
 On Fri, Mar 12, 2010 at 12:18 PM, Bob Camp li...@rtty.us wrote:
 
 Hi
 
 Tough to believe that HP worried a lot about SKU inflation back when they
 did the 5370 :)
 
 I'm assuming that the 5370 was a Santa Clara design. That would put the
 counter designers down the hall from the oscillator factory. Unlikely that
 there was a communications gap about what could or could not be done.
 
 You may well be correct though. Setting up and managing another part is the
 most likely reason why not to add a couple more tests.
 
 Bob
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of paul swed
 Sent: Friday, March 12, 2010 8:51 AM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] 5370B OCXO
 
 If I understand this thread correctly.
 I would speculate it was simply a business choice. 1 less part type to
 manage.
 
 On Fri, Mar 12, 2010 at 7:30 AM, Bob Camp li...@rtty.us wrote:
 
 Hi
 
 I guess the real question is what a better OCXO would have actually
 cost.
 
 If the 60111 was a test ten, get ten sort of thing (I'm guessing it was)
 -
 was a better part simply a test 10 get 9 issue?
 
 The claim was made that short term stability testing could be done
 directly
 in the aging racks. It's not real clear what the actual cost of an
 extended
 test / sort would have been.
 
 Bob
 
 
 On Mar 11, 2010, at 10:43 PM, John Miles wrote:
 
 Many if not most 5370-based measurements are based on differential
 timing
 between the START and STOP channels, and wouldn't benefit from a better
 10
 MHz reference.  If a customer did need something better, they probably
 already had a house standard to pipe in the back... and if not, HP
 would
 have been able to sell them one.  It made more sense to keep the cost
 down
 by not including a high-end OCXO that would have gone unappreciated by
 most
 users.
 
 The 5370's jitter+resolution floor doesn't allow it to reach 1E-11 at
 t=1s
 in any event, so the -60111 wouldn't have been the limiting factor in
 the
 short term.
 
 One valid question, though, is why they bothered to put the nicer
 10811-60109 OCXOs in the post-2120 series 5065A models, where its
 short-term
 performance is hosed by tying it to the rubidium reference with a ~1 Hz
 loop.  Those 5065As would have been OK with a -60111, at least in the
 pre-2632 serial #s with the original integrator board.  I'd be curious
 to
 know if they lowered the loop BW when they respun the integrator PCB.
 
 -- john, KE5FX
 
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com
 ]On
 Behalf Of Bob Camp
 Sent: Thursday, March 11, 2010 7:11 PM
 To: Discussion of precise time and frequency measurement
 Subject: [time-nuts] 5370B OCXO
 
 
 Hi
 
 The OCXO in the 5370B is a 10811-60111. The only added spec on it
 is a 1x10^-11 ADEV spec at 1 second. By modern standards that's
 not a real tight spec. There are other 10811's with tighter specs
 on them at 1 second. My guess is that it was not a real tight
 spec for the 10811 to hit.
 
 The short term would appear to contribute to the total error on
 the counter. Why not put a better oscillator in it?
 
 Bob
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow

Re: [time-nuts] 5370B OCXO

2010-03-12 Thread Richard (Rick) Karlquist



Bob Camp wrote:

Hi

Tough to believe that HP worried a lot about SKU inflation back when they
did the 5370 :)

I'm assuming that the 5370 was a Santa Clara design. That would put the
counter designers down the hall from the oscillator factory. Unlikely that
there was a communications gap about what could or could not be done.

You may well be correct though. Setting up and managing another part is the
most likely reason why not to add a couple more tests.

Bob


I was at Santa Clara Division in the 1980's.  They worried about both
the paper work for extra part numbers and the inventory problem. 
Especially  the latter.  There was always some manager whose stock

options depended on keeping inventory down (this was during the
Japanese just in time fad).

Also, if you wanted a part number, you had to obtain one from the 
official keeper of numbers, and she would usually give you a lecture

about the perils of wasting numbers.

The counter designers were in a different bldg than the oscillator 
factory, which were at opposite ends of the 55 acre site.   More

importantly, they reported to different managers.  So
you shouldn't assume that communication was great.

Rick Karlquist N6RK

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 5370B OCXO

2010-03-12 Thread paul swed
So little has really changed in 30 years even between completely different
companies
Dilbert strikes again

On Fri, Mar 12, 2010 at 9:57 PM, Richard (Rick) Karlquist 
rich...@karlquist.com wrote:



 Bob Camp wrote:

 Hi

 Tough to believe that HP worried a lot about SKU inflation back when they
 did the 5370 :)

 I'm assuming that the 5370 was a Santa Clara design. That would put the
 counter designers down the hall from the oscillator factory. Unlikely that
 there was a communications gap about what could or could not be done.

 You may well be correct though. Setting up and managing another part is
 the
 most likely reason why not to add a couple more tests.

 Bob


 I was at Santa Clara Division in the 1980's.  They worried about both
 the paper work for extra part numbers and the inventory problem. Especially
  the latter.  There was always some manager whose stock
 options depended on keeping inventory down (this was during the
 Japanese just in time fad).

 Also, if you wanted a part number, you had to obtain one from the official
 keeper of numbers, and she would usually give you a lecture
 about the perils of wasting numbers.

 The counter designers were in a different bldg than the oscillator factory,
 which were at opposite ends of the 55 acre site.   More
 importantly, they reported to different managers.  So
 you shouldn't assume that communication was great.

 Rick Karlquist N6RK

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 5370B OCXO

2010-03-11 Thread John Miles
Many if not most 5370-based measurements are based on differential timing
between the START and STOP channels, and wouldn't benefit from a better 10
MHz reference.  If a customer did need something better, they probably
already had a house standard to pipe in the back... and if not, HP would
have been able to sell them one.  It made more sense to keep the cost down
by not including a high-end OCXO that would have gone unappreciated by most
users.

The 5370's jitter+resolution floor doesn't allow it to reach 1E-11 at t=1s
in any event, so the -60111 wouldn't have been the limiting factor in the
short term.

One valid question, though, is why they bothered to put the nicer
10811-60109 OCXOs in the post-2120 series 5065A models, where its short-term
performance is hosed by tying it to the rubidium reference with a ~1 Hz
loop.  Those 5065As would have been OK with a -60111, at least in the
pre-2632 serial #s with the original integrator board.  I'd be curious to
know if they lowered the loop BW when they respun the integrator PCB.

-- john, KE5FX


 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com]on
 Behalf Of Bob Camp
 Sent: Thursday, March 11, 2010 7:11 PM
 To: Discussion of precise time and frequency measurement
 Subject: [time-nuts] 5370B OCXO


 Hi

 The OCXO in the 5370B is a 10811-60111. The only added spec on it
 is a 1x10^-11 ADEV spec at 1 second. By modern standards that's
 not a real tight spec. There are other 10811's with tighter specs
 on them at 1 second. My guess is that it was not a real tight
 spec for the 10811 to hit.

 The short term would appear to contribute to the total error on
 the counter. Why not put a better oscillator in it?

 Bob



 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.