> Well I have been looking at the data in the 
> 
> https://www.researchgate.net/publication/325499937_A_portable_cold_87_Rb_
> atomic_clock_with_frequency_instability_at_one_day_in_the_10-15_range
> 
> link and find that maybe they are overstating their performance!
> 
> In the plot of figure 5 my HP 5065A almost perfectly matches the Allan
> deviation 
> out to 1000 Sec.! (Although mine does have VERY good performance)

And you don't think this is impressive? Short term performance of an
atomic clock depends on the SNR of the signal. Which in turn depends
on the number of atoms being probed. A vapor cell standard probes
billions of atoms even for tiny cells, 1e15 to 1e18 for large cells.
A MOT usually traps a few 10k to at most a few  tens of million of atoms.
Ie we are at least a factor of 1000 below a Rb vapor cell, and more like
a factor of billions. And as usual, noise goes down with the square root.
Which in turn means the theoretical SNR in a cold atom Rb standard is a
factor of 10'000 lower than your 5065! Yet it achieves the same short
term performance! (The real differnence in SNR is probably lower than
that, but I doubt it's less than a factor of 100)

And mind you, we are still talking about quite new products where little
optimization has been done. And we aren't talking about big companies
like HP either. µQuans was started by a few students, who made their 
research into a product. SDI is a very small business. Neither
have millions to spend on optimizing the product until it's perfect
before they announce it to the world. They had to get a product out
as soon as possible to make sure they can recapture the cost of development.
And don't worry, both are working on improving the system.

To make things worse, the paper you are citing is actually the first
one SDI published. Meaning that was (one of?) the first complete system
they had built.

 
> Also if you look at figure 7 all the Maser data shown seems to be from
> poorly
> operating Masers so any judgement of better clock performance versus the
> Masers
> is a bit much!

Calling the masers of NIST poorly maintained is a bit much, isn't it?

Besides, they do not measure against the masers, but against UTC(NIST),
though that's not that clear from the paper, as they only say "NIST
measurement system." Other papers state it explicitly.


> Something does not add up in their data!

I think it clear from what I wrote above and what others wrote before me
that you are comparing apples and oranges. While it is true that both
the cold atom clocks and your 5065 use Rubidium as a species, the techniques
involved are so differnt that you have to treat them as a totally different
kind of clock. If you look at them as a active H maser or Cs beam standard
replacement, that would be closer to what they are.


On Sat, 09 Nov 2019 00:07:43 +
"Poul-Henning Kamp"  wrote:

> Given that these two cold-Rb devices are 1st generation of their
> kind and given how little time they have had to collect data on
> them yet, both with respect to performance but also day-to-day
> gremlin-wrangling, I wouldnt be at all surprised if the next
> two generations of that concept delivers almost two orders of
> magnitude improved performance.

While both companies are working on improvements. I don't expect
two orders of magnitude. Both systems work close to what the
theoretical limit of the number of atoms they are probing are.
I wouldn't be surprised if they get a factor of 10 out of the
system, after all improvements have been done. But I would find
it unlikely that the reach a factor of 100 without significantly
increasing the number of atoms they probe or increasing the
interogation time. And both of these things are quite hard to do
in these architectures.

On Sat, 09 Nov 2019 07:05:10 +
"Poul-Henning Kamp"  wrote:

> It is not like people have not been improving the Rb-Gas-cell, but if
> you want to improve it an order of magnitude all the myriad of "small
> issues" to fix add up real fast, and when you are done with all of
> it, you will still have frequency drift.


If you want to know what the limit of Rb vapor cell standards are,
have a look at the research done by the group of Mileti and Affolderbach.
They are at ~1e-13/sqrt(tau) and are trying to push it down to the shot
noise limit (which is at a few parts in 1e-14). That's the limit you
can do. And that's still not accounting for drift and other things which
limit the ADEV to a few parts in 1e-14 somewhere around a tau of 1k to 10k.



On Sat, 9 Nov 2019 02:38:45 +0100
Magnus Danielson  wrote:

> This is well understood techniques now, so that it matures into
> commercial products is not strange. It's worth noting that optically
> probed cesium also exists from Oscilloquartz, altering the beam standard
> techniques

It is also worth noting that it took Oscilloquartz several years to get
it working. And not only that, a lot of people in industry 

Re: [time-nuts] Cold Rubidium over hyped?

2019-11-09 Thread Bob kb8tq
Hi

> On Nov 8, 2019, at 11:18 PM, cdel...@juno.com wrote:
> 
> The pressure sensitivity for a classic design can easily be eliminated.
> 
> I'm pretty sure that the old Varian R20 cell was mostly immune.
> 
> Either using the R20 design or stiffer convex ends in a modern design
> would work.
> 
> Also a sealed optical unit has been used to get rid of the pressure
> problem.

OCXO’s have a pressure sensitivity as well. There, the only real answer 
is a welded enclosure. Anything short of that simply integrates the pressure
via low leak rates and makes things even stranger …..

Bob

> 
> Any of these methods would be trivial in a new design.
> 
> The Rb/glass absorption is mostly in the lamp, when using the proper
> glasses in the cell and using similar lasers to what the cold standard
> uses eliminates that.
> 
> Newer RF architecture would also help, NLDL multiplier, low noise
> synthesis, etc.
> 
> My main gripe was that either they were either intentionally reducing the
> performance of their masers or using a setup that did not do justice to
> them.
> 
> Cheers,
> 
> Corby
> 
> 
> ___
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to 
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.


___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


Re: [time-nuts] Cold Rubidium over hyped?

2019-11-09 Thread Magnus Danielson
Hi,

On 2019-11-09 08:05, Poul-Henning Kamp wrote:
> 
> In message , cdel...@juno.com writes:
>
> [long list of all the work to be done before the classical gas-cell Rb
> improves materially]
>
> You are sort of making my point here Corby.
>
> It is not like people have not been improving the Rb-Gas-cell, but if
> you want to improve it an order of magnitude all the myriad of "small
> issues" to fix add up real fast, and when you are done with all of
> it, you will still have frequency drift.
>
... and these devices address a different aspect as per Michael's point.
It's not for the phase-noise, but for long-term stability, so that would
be instead of cesium beams or fountains. Gas-cell standards cannot
really solve that because the myriad of issues turn out an unknown
device frequency shift, which altought can be stable is not suitable as
standard. Rather, these devices could allow for much more affordable
contribution to the second/TAI rate in international timekeeping.

Cheers,
Magnus


___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


Re: [time-nuts] Cold Rubidium over hyped?

2019-11-09 Thread Poul-Henning Kamp

In message , cdel...@juno.com writes:

[long list of all the work to be done before the classical gas-cell Rb
improves materially]

You are sort of making my point here Corby.

It is not like people have not been improving the Rb-Gas-cell, but if
you want to improve it an order of magnitude all the myriad of "small
issues" to fix add up real fast, and when you are done with all of
it, you will still have frequency drift.

-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
p...@freebsd.org | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


[time-nuts] Cold Rubidium over hyped?

2019-11-08 Thread cdelect
The pressure sensitivity for a classic design can easily be eliminated.
 
I'm pretty sure that the old Varian R20 cell was mostly immune.
 
Either using the R20 design or stiffer convex ends in a modern design
would work.
 
Also a sealed optical unit has been used to get rid of the pressure
problem.
 
Any of these methods would be trivial in a new design.
 
The Rb/glass absorption is mostly in the lamp, when using the proper
glasses in the cell and using similar lasers to what the cold standard
uses eliminates that.

Newer RF architecture would also help, NLDL multiplier, low noise
synthesis, etc.
 
My main gripe was that either they were either intentionally reducing the
performance of their masers or using a setup that did not do justice to
them.
 
Cheers,
 
Corby


___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


Re: [time-nuts] Cold Rubidium over hyped?

2019-11-08 Thread jimlux

On 11/8/19 5:38 PM, Magnus Danielson wrote:



So, while hydrogen masers may work and large cell rubidiums can be
cleaned up, eventually we come to the point where these new techniques
will take over for many good reasons. Maybe one day we will have
time-nuts with cold rubidium clocks and maybe even optical clocks.



Or Mercury or Strontium ion?

___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


Re: [time-nuts] Cold Rubidium over hyped?

2019-11-08 Thread Magnus Danielson
Hi,

On 2019-11-09 01:07, Poul-Henning Kamp wrote:
> 
> In message , cdel...@juno.com writes:
>
>> Just wonder if some  5065A can get so impressive that they don't just
>> make a modern large/cool cell classic Rubidium with modern electronics
>> technology! Certainly would be cheaper and more long lived also.
> But isn't that essentially what they have done ?
Yes.
>
> The Rb-lamp-filter thing has been beaten to death.  Its not like
> people have not researched it in the last 50 years, but nobody
> anywhere have found a way to get rid of the Rb/glass absorption
> related aging or the pressure-sensitivity for that matter, so 5065
> performance pretty much is the best you can ever hope for there
> as long as your customers are terrestial.

The wall shift, the buffer gas shift, the resonator shift and how these
relate to environmentals in addition to the lamp intensity shift. All
well researched. Rubidium as such is not a bad species to measure, on
the contrary, we learned that it excels over cesium in laser cooled
state, as the cross-section is smaller.

>
> Given that these two cold-Rb devices are 1st generation of their
> kind and given how little time they have had to collect data on
> them yet, both with respect to performance but also day-to-day
> gremlin-wrangling, I wouldnt be at all surprised if the next
> two generations of that concept delivers almost two orders of
> magnitude improved performance.

Long-term measurements is being done at NIST and OP/SYRTE, none being
known to be extremely wreckless in this regard, to say the least.

This is well understood techniques now, so that it matures into
commercial products is not strange. It's worth noting that optically
probed cesium also exists from Oscilloquartz, altering the beam standard
techniques.

Common to all three clocks is the fact that they use semiconductor
lasers to pump and interact with the element, and therein lays also the
challenge of having such laser-systems operating long-term.

This is for sure the way forward, and these are just the for-runners of
the continiously operating optical clocks that may one day become
commercial products.

So, while hydrogen masers may work and large cell rubidiums can be
cleaned up, eventually we come to the point where these new techniques
will take over for many good reasons. Maybe one day we will have
time-nuts with cold rubidium clocks and maybe even optical clocks.

Cheers,
Magnus




___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


Re: [time-nuts] Cold Rubidium over hyped?

2019-11-08 Thread Michael Wouters
I think you are missing the key selling point of this device, namely
it's long term stability and accuracy of a few parts in 10^15. From my
point of view as a national timekeeper, this is much more useful than
good short term stability. UTC reporting is at 5 day intervals so what
the clock does at 1000 s is not so important. By the way, there is
another comparison with 12 NIST masers in the current brochure:
https://spectradynamics.com/product-sheets/cRb-Clock-2019.pdf
There are evidently two models of maser here ,which politely, they do
not identify.

Cheers
Michael

On Sat, Nov 9, 2019 at 11:00 AM  wrote:
>
> Well I have been looking at the data in the
>
> https://www.researchgate.net/publication/325499937_A_portable_cold_87_Rb_
> atomic_clock_with_frequency_instability_at_one_day_in_the_10-15_range
>
> link and find that maybe they are overstating their performance!
>
> In the plot of figure 5 my HP 5065A almost perfectly matches the Allan
> deviation
> out to 1000 Sec.! (Although mine does have VERY good performance)
>
> Also if you look at figure 7 all the Maser data shown seems to be from
> poorly
> operating Masers so any judgement of better clock performance versus the
> Masers
> is a bit much!
>
> The figure 5 data looks much better but still is not beating a good
> active Maser.
>
> The EFOS2 a 1982 vintage Maser as well as the MHM 2010 and several other
> modern Masers
> I could find data for show between 2 and 5X10-15th at 1000 Sec.
> The MHM 2010 specs at 5X10-15th at 100Sec and 2.0X10-15th at 1000Sec
> All the Masers shown are worse than that spec!
> Also my old Kvarz passive maser has  2x10-13th at 100 Sec just shy of
> matching the top two Masers at 100Sec, which I don't believe!
>
> Something does not add up in their data!
>
> So I'm not really that impressed and would take an Active or Passive
> Maser anyday.
>
> Just wonder if some  5065A can get so impressive that they don't just
> make a modern large/cool cell classic Rubidium with modern electronics
> technology! Certainly would be cheaper and more long lived also.
>
> Just my thoughts!
>
> Cheers,
>
> Corby
> ___
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to 
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.

___
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.


[time-nuts] Cold Rubidium over hyped?

2019-11-08 Thread cdelect
Well I have been looking at the data in the 

https://www.researchgate.net/publication/325499937_A_portable_cold_87_Rb_
atomic_clock_with_frequency_instability_at_one_day_in_the_10-15_range

link and find that maybe they are overstating their performance!

In the plot of figure 5 my HP 5065A almost perfectly matches the Allan
deviation 
out to 1000 Sec.! (Although mine does have VERY good performance)

Also if you look at figure 7 all the Maser data shown seems to be from
poorly
operating Masers so any judgement of better clock performance versus the
Masers
is a bit much!

The figure 5 data looks much better but still is not beating a good
active Maser.

The EFOS2 a 1982 vintage Maser as well as the MHM 2010 and several other
modern Masers
I could find data for show between 2 and 5X10-15th at 1000 Sec.
The MHM 2010 specs at 5X10-15th at 100Sec and 2.0X10-15th at 1000Sec
All the Masers shown are worse than that spec!
Also my old Kvarz passive maser has  2x10-13th at 100 Sec just shy of 
matching the top two Masers at 100Sec, which I don't believe!

Something does not add up in their data!

So I'm not really that impressed and would take an Active or Passive
Maser anyday.

Just wonder if some  5065A can get so impressive that they don't just
make a modern large/cool cell classic Rubidium with modern electronics
technology! Certainly would be cheaper and more long lived also.