Re: [volt-nuts] HP 3457A

2013-08-11 Thread Charles Steinmetz

Orin wrote:

They send a calibration certificate which claims that their 
standards are traceable to NIST.  I have no reason to doubt that.


Owning standards and instruments with traceable calibrations is 
necessary but not sufficient for making traceable calibrations with 
them.  For a calibration to be traceable, the lab that did the 
calibration must be accredited.  If the lab is accredited, the 
calibration certificate would certainly identify the accrediting body.


Does the certificate you received identify an accrediting body?

Best regards,

Charles



___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] HP 3457A

2013-08-11 Thread J. L. Trantham
Joe,

I looked at the listing you posted. The unit looks very nice.  However, the
units are selling in the $150 -  $250 range, although I note some of the
units you linked to have sold for the price asked.

There has been considerable discussion on the list about the 3457A and some
differences between units, depending on serial number, having to do with the
circuitry about the backup battery for the RAM.  I would recommend you
search the archives about that.  Also, there are various options for the
rear compartment that may or may not be useful so being able to see the back
of the unit to see what is installed and what the serial number is would be
useful.

I prefer to use the original manufacturer's calibration services when it
comes to Agilent or Fluke just to be able to eliminate the 'trust
uncertainty' with 'outside labs', since I have little experience in using
various calibration labs.

Therefore, my choice would be to look for a desirable 'used' unit then clean
it up myself, check the date on the backup battery and possibly change it
(with the need to keep voltage applied to the unit while unsoldering the
battery lest losing the CAL data - careful there - 'Danger, Danger, Will
Robinson') then send it to Agilent for calibration ($204.22 for the Agilent
Calibration).

If you are looking for a currently supported unit, I would consider the
Fluke 8846A.  Can be had for under $1000.  A bit more expensive but
currently available and currently supported.

Good luck.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Joseph Gray
Sent: Saturday, August 10, 2013 9:20 PM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] HP 3457A

If I want to buy a used 3457A, is it better to buy one as-is and send it
for calibration, or perhaps spend a bit more and get one already calibrated?
Just one example:

http://www.ebay.com/itm/HP-3457A-6-5-Digit-Lab-Quality-Front-Terms-Fully-Ref
urbished-NIST-Cal-/300928374155

Joe Gray
W5JG
___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] HP 3457A

2013-08-11 Thread Chuck Harris

I think the statement is awkward English, but what I think he is trying
to say is that:

1) the instrument will be recalibrated
2) after the instrument is verified, by the calibration lab,
   to be working within specs,
3) a NIST traceable calibration certificate will be mailed.

-Chuck Harris

Charles Steinmetz wrote:

Orrin wrote:


Well, lets discuss the 3456A I got from goldenrubi, calibrated.  They send a
calibration certificate which claims that their standards are traceable to 
NIST.  I
have no reason to doubt that.


I happened upon one of the seller's auctions, so I checked that one and several
others.  In the boilerplate of each one I found this:


WE WILL RE-CALIBRATED THE UNIT BEFORE SHIPPING AND A NIST TRACEABLE CALIBRATION
CERTIFICATE, WILL BE PROVIDED AFTER ITEM IS RECEIVED AND IS ACCEPTED AS WORKING
WITH IN SPECS. WE WILL MAIL OUT THE CERTIFICATE.


If I understand this correctly, I buy an instrument, the seller calibrates it
before shipment, but doesn't send the calibration certificate with the item.  He
mails it to me after I accept[] [the instrument] as working within specs.

So, I have to verify the calibration myself, and then this seller will send me a
NIST traceable calibration certificate???  Words cannot express how irregular 
that
sounds to me.  I'd be interested to hear about any other NIST traceable 
calibration
facility that works this way.

Best regards,

Charles




___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] HP 3457A

2013-08-11 Thread Orin Eman
They shipped the certificate with the 3456A I got from them.  The cal
certificate is from AAA Calibration Equipment Specialist Inc. at a SW
Albuquerque address and is issued to AAA Equipment Resources Inc.
 at a NW Albuquerque address.

I suspect some internal accounting reason or a defense against dishonest
buyers for this.

BTW, they don't list any accreditation on the certificate.

I did once ask if they could cal an 8484A and they said no.

As for the local labs here, I'm not happy with the local Tek lab that I
sent my TDS210 to in its original packaging.  They returned it in a small
box with two layers of bubble wrap.

The Fluke HQ and cal lab is twenty miles from here... I think that's about
as good as it gets, so they might be getting my business in the future.

Orin.



On Sun, Aug 11, 2013 at 5:57 AM, Charles Steinmetz csteinm...@yandex.comwrote:

 Orrin wrote:

  Well, lets discuss the 3456A I got from goldenrubi, calibrated.  They
 send a calibration certificate which claims that their standards are
 traceable to NIST.  I have no reason to doubt that.


 I happened upon one of the seller's auctions, so I checked that one and
 several others.  In the boilerplate of each one I found this:

  WE WILL RE-CALIBRATED THE UNIT BEFORE SHIPPING AND A NIST TRACEABLE
 CALIBRATION CERTIFICATE, WILL BE PROVIDED AFTER ITEM IS RECEIVED AND IS
 ACCEPTED AS WORKING WITH IN SPECS. WE WILL MAIL OUT THE CERTIFICATE.


 If I understand this correctly, I buy an instrument, the seller
 calibrates it before shipment, but doesn't send the calibration
 certificate with the item.  He mails it to me after I accept[] [the
 instrument] as working within specs.

 So, I have to verify the calibration myself, and then this seller will
 send me a NIST traceable calibration certificate???  Words cannot express
 how irregular that sounds to me.  I'd be interested to hear about any other
 NIST traceable calibration facility that works this way.


 Best regards,

 Charles




 __**_
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/**
 mailman/listinfo/volt-nutshttps://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] HP 3457A

2013-08-11 Thread Dr. David Kirkby
On 11 August 2013 03:19, Joseph Gray jg...@zianet.com wrote:
 If I want to buy a used 3457A, is it better to buy one as-is and send it
 for calibration, or perhaps spend a bit more and get one already
 calibrated? Just one example:

 http://www.ebay.com/itm/HP-3457A-6-5-Digit-Lab-Quality-Front-Terms-Fully-Refurbished-NIST-Cal-/300928374155

 Joe Gray
 W5JG

If you look at the feedback for the seller, you will find this auction#

http://www.ebay.com/itm/221234066046

He bought one of the meters for $305. At first I see best offer
accepted and no price. But if you print the auction to a PDF file,
then look at the PDF, you will find the actual cost it sold at !!
Funny thing was, after doing that I could see the price anyway, so
perhaps I was mistaken first time I thought the price was a best
offer. Anyway, he seems happy to buy at $305. so it might be worth
sending him a mesage with a lower offer.

Dave
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] HP 3457A

2013-08-11 Thread Charles Steinmetz

Orin wrote:

The cal certificate is from AAA Calibration Equipment Specialist 
Inc. at a SW Albuquerque address and is issued to AAA Equipment 
Resources Inc. at a NW Albuquerque address.


Those are both the seller:

http://www.bizapedia.com/nm/AAA-CALIBRATION-EQUIPMENT-SPECIALIST-INC.html
https://aaaequipmentresources.3dcartstores.com/crm.asp?action=contactus


BTW, they don't list any accreditation on the certificate.


As I suspected.  So it is vanishingly unlikely that they do traceable 
calibrations, contrary to their claim.


The Fluke HQ and cal lab is twenty miles from here... I think that's 
about as good as it gets, so they might be getting my business in the future.


It's hard to do better than that.  The Sandia Primary Standards Lab 
is right up there, too.


Best regards,

Charles



___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Calibration and Certification - Trust and detail

2013-08-11 Thread Dr. David Kirkby
On 11 August 2013 15:39, Robert Atkinson robert8...@yahoo.co.uk wrote:
 On a related issue to the 3457A calibration, I was asked to review an item at 
 work last week (sorry can't say what or why). Looking at compliance 
 certification by two fully qualified, internationally recognised labs, all 
 looked well until I looked at the detail. Things like ..all RELEVANT 
 requirements of.. (my capitals) with no list of what was relevant or any 
 test report., details of only one test when the standard needed a number 
 under different conditions, no record of the part or serial numbers of the 
 unit tested. Others had accepted the certification at face value.

I think it depens why you want the cal certificate.

1) If it to keep the BSI person happy to keep your ISO 9001
acreditation, then I guess as long as it has a cal certificate that is
ok, so use the cheapest dodgy cal lab.

2) If the purpose of the cal certificate is to get the highest price
when selling something, then a cal certificate by some dodgy cal lab
is probably all you need. 99% of buyers are not going to question who
did the cal, and their ability to do it.

3) If you want to know the instrument works correctly, and have it
adjusted for best performance, then it is a very different matter. One
is probably better sending it to the manufacturer in many cases.

I want to know my VNA works properly, so that is going to Agilent this
week. The cal cost on my VNA is about 5% of what I paid for the VNA.
It would be much more difficult to justify sending my 3457A to
Agilent, when the cal cost will probably be more than what I paid for
the instrument.

How useful is this

http://www.ebay.com/itm/281149723636 ?

On the fact of it, the device would give one a reasonly high
confidence something is working readlably well. I wonder if that is
good enough for a 3457A.

 Unfortunatly if the test house customer says just do this bit and the 
 test house is happy to put weasely words on the certifcate, then user beware.

As I wrote earlier, I think there is a very cosy relationship between
cal labs and test equipment dealers. It is in both their interests to
get cal certificates on items even if they are not 100%. A T+M dealer
is not likely to use a cal lab that keeps sending items back marked
Out of specification or similar. A cal lab does not want to lose a
customer.

 Robert G8RPI.

Dave, G8WRB.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Calibration and Certification - Trust and detail

2013-08-11 Thread Charles Steinmetz

David wrote:


How useful is this

http://www.ebay.com/itm/281149723636 ?

On the fact of it, the device would give one a reasonly high
confidence something is working readlably well. I wonder if that is
good enough for a 3457A.


Let's assume that it is still working exactly the same as it was when 
it was calibrated.  The calibration values are recorded to 4-1/2 
digits.  So the uncertainty is greater than the two LSDs of a 
3457A.  Of course, it is almost certainly not working exactly the 
same as it was when it was calibrated, which is going to put at least 
the third LSD of a 3457A in question.  So it is good enough for a 
3457A only if one is content with a 3-1/2 digit verification on a 
6-1/2 digit instrument.


Best regards,

Charles



___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts