Bob et all

1) We have a 1st class scan of the R-D 4200, manual with complete 
schematics available at a modest price
2) We have a VERY NICE scan  (NOT a retread of the poor quality of 
Agilent scan) 600dpi scans of the schematics full size up to 27x 11 
unpieced, again at a very modest price
3) I am sorry but we dont have manuals for the EI or Valhalla unit at 
the moment but I have added them to my watch for these list


I have been following your saga on the 3456A repair on the Agilent forum 
and don't have much to add except to say this is not a trivial unit to 
troubleshoot because of the overly (IMO) complex nature of the beast . 
To paraphrase the old nursery rhyme, When they are good they are very 
very good but when they are bad they are horrid. I'd probably put that 
one on the back burner till you get more time , patience and experience. 
Then sit down and a read the theory of operation section of the manual 
till you can recite the whole thing by heart..THEN go digging into 
boards. With a unit this complex there are scores of feedback and 
reference check loops that it is often difficult to ascertain when 
troubleshooting  what is a  root cause and what is an effect of some 
other defect elsewhere in the circuit


ON the erratic Racal unit, by comparison this is an easier unit to 
repair with a fairly clean layout of the main board. I would start by 
carefully unplugging and reseating all cable connectors and the Data I/O 
board, Display board and power supply diode board. next once the unit 
begins its erratic behavior after warm up, look at ( with a scope ) the 
AC ripple  on the DC +5V and +20V supply buses . Most likely suspects 
with any unit more than 25 years old ( aren't they all?) are the 
Electrolytic and Tantalum capacitors followed by solder joints that just 
need to be re-flowed. Given the lower complexity of the Dana main board 
the joints could all be re-flowed in  1/2 hour, One other common source 
of noise are the calibration pots...which need to be turned back and 
forth a few times to clean the wipers.. of course after this brilliant 
move you will need to do a complete recalibration of the unit.


Good luck
Dave
ArtekManuals




OnceOn 2/23/2014 1:07 AM, Bob Albert wrote:

Pete,

The EI is in three parts connected by many massive cables.  When I say it's 
pretty close, I mean it agrees with everything else I feel is close.  The 3456A 
has a -12V reference which the EI measures at -12.0026 or so.  It uses an oven 
so it takes a half hour to settle.  I have pics of the panel and would post 
them if I knew how.


The Valhalla is actually a good meter but one range has a small problem.  The 
Racal-Dana is unfixable because I can't find a manual, short of spending more 
than the meter is worth.  Those two units are very similar; the main difference 
is 1) the Valhalla measures AC current and 2) the Dana has a very sensitive DC 
microammeter such that I can measure diode leakage.

I am befuddled by the HP3456A.  I have replaced some parts but so far it's not 
doing its thing.  I have the manual but it sucks, not having a proper schematic 
diagram nor a board trace layout.

The Calib Stds unit is nice, a manually operated slideback DC volts only unit, 
compact and easy to use.  Again, no manual available and, in fact, no mention 
of the company anywhere online.  They were in Los Angeles area.

The main reason I want to fix these is that I hate owning anything that doesn't 
work.  And I like knowing my meters are accurate as well as precise.

I also have some 10V reference ICs that have been around here a few years.  
They all agree within a few mV.  I made a permanent reference out of them with 
a switch to select each, and they are running whenever I am in the lab.

Bob

TRIM



--
Dave
manu...@artekmanuals.com
www.ArtekManuals.com


---
This email is free from viruses and malware because avast! Antivirus protection 
is active.
http://www.avast.com

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] Newbie here...

2014-02-23 Thread Bob Albert
Well all comments are appreciated and well noted.  The 3456A is indeed a 
Pandora's box and is, as suggested, on the back burner for now.


As for the Dana, I have tried all the above.  I replaced one IC that seemed to 
be the root of the trouble, to no avail.  The unit is very clean.  I spray 
cooled it when it did its trouble thing and it came back to operation but I 
haven't been able to find the offending component or solder joint.  Power 
supply seems good throughout.


I paid so little for this, and I have so many other units, that I am loath to 
spend money on a diagram, especially when it is no guarantee of giving me the 
needed information.  Plus, the 300-page HP manual is free for download, so it 
feels like the tail is wagging the dog here.

I will pay attention to the KO4BB site for when the HP manual will appear.

But if I need to measure something, I certainly have the tools to do it.  If 
the voltage is above 1 kV I have my Simpson 260 and Triplett 630-NA for that.  
And I am ashamed to admit, a couple Radio Schlock VOMs.  The whole thing here 
is that I enjoy playing with accurate stuff even though I have no legitimate 
use for 6-digit voltage measurements.

One thing about the HP that appeals is the accurate Ohmmeter.  I have a GR 1658 
bridge that is very accurate but it doesn't measure at DC, so is cumbersome for 
inductors, etc.

And of course the delightful little toy, the Harbor Freight DVM, is a bargain 
that can't be approached by 'normal' units.  I have several of them and, within 
their limitations, they are great units.  There is a calibrate pot inside that 
takes care of any ordinary reading errors.  And if I blow it up, I pick up 
another without that dreaded feeling of having broken something valuable.

Bob




On Sunday, February 23, 2014 3:55 AM, Artek Manuals manu...@artekmanuals.com 
wrote:
 
Bob et all

1) We have a 1st class scan of the R-D 4200, manual with complete 
schematics available at a modest price
2) We have a VERY NICE scan  (NOT a retread of the poor quality of 
Agilent scan) 600dpi scans of the schematics full size up to 27x 11 
unpieced, again at a very modest price
3) I am sorry but we dont have manuals for the EI or Valhalla unit at 
the moment but I have added them to my watch for these list

I have been following your saga on the 3456A repair on the Agilent forum 
and don't have much to add except to say this is not a trivial unit to 
troubleshoot because of the overly (IMO) complex nature of the beast . 
To paraphrase the old nursery rhyme, When they are good they are very 
very good but when they are bad they are horrid. I'd probably put that 
one on the back burner till you get more time , patience and experience. 
Then sit down and a read the theory of operation section of the manual 
till you can recite the whole thing by heart..THEN go digging into 
boards. With a unit this complex there are scores of feedback and 
reference check loops that it is often difficult to ascertain when 
troubleshooting  what is a  root cause and what is an effect of some 
other defect elsewhere in the circuit

ON the erratic Racal unit, by comparison this is an easier unit to 
repair with a fairly clean layout of the main board. I would start by 
carefully unplugging and reseating all cable connectors and the Data I/O 
board, Display board and power supply diode board. next once the unit 
begins its erratic behavior after warm up, look at ( with a scope ) the 
AC ripple  on the DC +5V and +20V supply buses . Most likely suspects 
with any unit more than 25 years old ( aren't they all?) are the 
Electrolytic and Tantalum capacitors followed by solder joints that just 
need to be re-flowed. Given the lower complexity of the Dana main board 
the joints could all be re-flowed in  1/2 hour, One other common source 
of noise are the calibration pots...which need to be turned back and 
forth a few times to clean the wipers.. of course after this brilliant 
move you will need to do a complete recalibration of the unit.

Good luck
Dave
ArtekManuals





OnceOn 2/23/2014 1:07 AM, Bob Albert wrote:
 Pete,

 The EI is in three parts connected by many massive cables.  When I say it's 
 pretty close, I mean it agrees with everything else I feel is close.  The 
 3456A has a -12V reference which the EI measures at -12.0026 or so.  It uses 
 an oven so it takes a half hour to settle.  I have pics of the panel and 
 would post them if I knew how.


 The Valhalla is actually a good meter but one range has a small problem.  The 
 Racal-Dana is unfixable because I can't find a manual, short of spending more 
 than the meter is worth.  Those two units are very similar; the main 
 difference is 1) the Valhalla measures AC current and 2) the Dana has a very 
 sensitive DC microammeter such that I can measure diode leakage.

 I am befuddled by the HP3456A.  I have replaced some parts but so far it's 
 not doing its thing.  I have the manual but it sucks, not having a proper 
 schematic diagram 

Re: [volt-nuts] Newbie here...

2014-02-23 Thread Pete Lancashire
Again just my few cents

 1 KV on the classics

Be very very careful. I was not uncommon for both of these guys to arc
over. If you do work with voltages in that space, consider a divider probe.

If you do keep playing over about 4 digits at some point you going to need
to decided that your absolute measurements are not  important or get
something like the 3456A calibrated, or the calibrator you have. I've had
good luck with one of the
few meteorology labs where I live having things calibrated for very little
as long as I don't want that little sticker.

Once they found out I had a just acquired set of LN 4000s that I wanted to
cal, they even let me do it myself.

For a few $'s there is a small one man company in Washington state that
sells a nicely done set of voltage sources.
http://www.voltagestandard.com/If you add up the parts cost, etc. they
are a bargain and well built (not a
customer but would be if didn't have the stuff I have).

-pete

There is a H-F about 1 mile from where I live. One day there were about 20
of the Red DMM in the dumpster, I could not resist. Fixed all except 2 put
2 in the truck and gave the rest away. 90% assembly issues.










On Sun, Feb 23, 2014 at 10:21 AM, Bob Albert bob91...@yahoo.com wrote:

 Well all comments are appreciated and well noted.  The 3456A is indeed a
 Pandora's box and is, as suggested, on the back burner for now.


 As for the Dana, I have tried all the above.  I replaced one IC that
 seemed to be the root of the trouble, to no avail.  The unit is very
 clean.  I spray cooled it when it did its trouble thing and it came back to
 operation but I haven't been able to find the offending component or solder
 joint.  Power supply seems good throughout.


 I paid so little for this, and I have so many other units, that I am loath
 to spend money on a diagram, especially when it is no guarantee of giving
 me the needed information.  Plus, the 300-page HP manual is free for
 download, so it feels like the tail is wagging the dog here.

 I will pay attention to the KO4BB site for when the HP manual will appear.

 But if I need to measure something, I certainly have the tools to do it.
 If the voltage is above 1 kV I have my Simpson 260 and Triplett 630-NA for
 that.  And I am ashamed to admit, a couple Radio Schlock VOMs.  The whole
 thing here is that I enjoy playing with accurate stuff even though I have
 no legitimate use for 6-digit voltage measurements.

 One thing about the HP that appeals is the accurate Ohmmeter.  I have a GR
 1658 bridge that is very accurate but it doesn't measure at DC, so is
 cumbersome for inductors, etc.

 And of course the delightful little toy, the Harbor Freight DVM, is a
 bargain that can't be approached by 'normal' units.  I have several of them
 and, within their limitations, they are great units.  There is a calibrate
 pot inside that takes care of any ordinary reading errors.  And if I blow
 it up, I pick up another without that dreaded feeling of having broken
 something valuable.

 Bob




 On Sunday, February 23, 2014 3:55 AM, Artek Manuals
 manu...@artekmanuals.com wrote:

 Bob et all

 1) We have a 1st class scan of the R-D 4200, manual with complete
 schematics available at a modest price
 2) We have a VERY NICE scan  (NOT a retread of the poor quality of
 Agilent scan) 600dpi scans of the schematics full size up to 27x 11
 unpieced, again at a very modest price
 3) I am sorry but we dont have manuals for the EI or Valhalla unit at
 the moment but I have added them to my watch for these list

 I have been following your saga on the 3456A repair on the Agilent forum
 and don't have much to add except to say this is not a trivial unit to
 troubleshoot because of the overly (IMO) complex nature of the beast .
 To paraphrase the old nursery rhyme, When they are good they are very
 very good but when they are bad they are horrid. I'd probably put that
 one on the back burner till you get more time , patience and experience.
 Then sit down and a read the theory of operation section of the manual
 till you can recite the whole thing by heart..THEN go digging into
 boards. With a unit this complex there are scores of feedback and
 reference check loops that it is often difficult to ascertain when
 troubleshooting  what is a  root cause and what is an effect of some
 other defect elsewhere in the circuit

 ON the erratic Racal unit, by comparison this is an easier unit to
 repair with a fairly clean layout of the main board. I would start by
 carefully unplugging and reseating all cable connectors and the Data I/O
 board, Display board and power supply diode board. next once the unit
 begins its erratic behavior after warm up, look at ( with a scope ) the
 AC ripple  on the DC +5V and +20V supply buses . Most likely suspects
 with any unit more than 25 years old ( aren't they all?) are the
 Electrolytic and Tantalum capacitors followed by solder joints that just
 need to be re-flowed. Given the lower complexity of the Dana