Re: [volt-nuts] 3458a with 202: Slave Test Convergence (sigh)

2014-03-27 Thread Tom Knox
The dreaded 202 convergence error, If you need help contact me off list.
Best Wishes;

Thomas Knox



 From: pit...@williams-net.com
 To: volt-nuts@febo.com
 Date: Thu, 27 Mar 2014 14:35:42 -0400
 Subject: [volt-nuts] 3458a with 202: Slave Test Convergence (sigh)
 
 I know that HP will (gladly?) take your sick 3458 and bring it up to current 
 specs, followed by a calibration. Since my sick one is at engineering level 2 
 or 3.something, it must be like it was built in 1989 - the dates on some 
 components.
 
 Should I try to do any troubleshooting, or just send it to HP?  
 
 Will they upgrade the current (aged and pretty much not drifting) boards?
 
 Or do they take a good board from their stock and replace the current board 
 with one that may be new and 'drifty'?
 
 I have read here that there are some unlisted upgrades that they do, also.
 
 The RAM batteries will be replaced also, right? 
 
 The biggest positive I see is the ability to go on the service plan and never 
 have to worry about a failure again after the 3458 is brought up to current 
 standards.
 
 Willy
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
  
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


[volt-nuts] 3458a with 202: Slave Test Convergence (sigh)

2014-03-27 Thread Frank Stellmach

I know that HP will (gladly?) take your sick 3458 and bring it up to current specs, 
followed by a calibration. Since my sick one is at engineering level 2 or 3.something, it 
must be like it was built in 1989 - the dates on some components.

Should I try to do any troubleshooting, or just send it to HP?


Hi,

agilent simply replaces boards, but will not desolder and replace components.
Therefore, they will definitely replace the processor board completely (several 
hundred $), but then the latest firmware 9 will be inside. Additionally, a new 
calibration will have to be paid (+ several hundred $).

Alternatively, simply replace the 3 NVRAMs (DALLAS/maxim or digikey, 50$), and 
the 6 EPROMs for firmware 9 (agilent, 180$), if you are able to copy the CAL RAM 
and/or to recalibrate on your own.
And on the old processor board you can upgrade opt. 01 by DIY (10$), instead 
of agilent (+ several 100 $).
Another fellow volt-nuts did that on a 1989 unit, came up fresh like a daisy. 
I.e. no need for further replacement of any other boards.


@Tom Knox: Would you mind sharing your knowledge about error 202 here on that 
list, instead?
THX.

Frank

 


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] 3458a RAM batteries - longer life if unit is continually powered?

2014-03-27 Thread Tom Miller
Only on the newer (AB  AD) Dallas (Maxim) parts. All the old ones (Y) do 
not have that feature.


Tom M.

- Original Message - 
From: Tom Knox act...@hotmail.com

To: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Thursday, March 27, 2014 12:59 PM
Subject: Re: [volt-nuts] 3458a RAM batteries - longer life if unit is 
continually powered?



Date Codes can be misleading because most of this style of chip does not 
connect to the battery until the first power up cycle. So evl


Thomas Knox




Date: Thu, 27 Mar 2014 17:33:17 +0100
From: frank.stellm...@freenet.de
To: volt-nuts@febo.com
Subject: [volt-nuts] 3458a RAM batteries - longer life if unit is 
continually powered?


Hi,

the DALLAS datasheet specifies 10 years data retention. Quite obviously,
that's the unpowered time.

That time includes self discharge and RAM sustaining currents.

self discharge and RAM supply current will be of the same order, i.e. 
1µA, I assume from experience.
And I assume also, that those currents will vary over samples, i.e. over
samples of battery and RAM.

Recently, there was the case, where the CAL RAM was still alive after 24
years (instrument vintage 1989, same DALLAS date code!), but with
unknown powering history.

Additionally, a typical 10 years of lifetime is given also for those
batteries.

Therefore, in practise there will be no pronounced difference between
powered / unpowered state.
On the other hand, DALLAS implemented an electronic seal, so that the
RAM inside the package is initially powered only after first power up.
But there's no parameter in the datasheet, specifying data retention
according to date code or to initial power up.


The waste of energy just to extend the batteries lifetime, does not pay 
off.

On the contrary, the MTBF time of the instrument will be consumed, the
display will get bad, and here in Germany, this energy consumption would
cost a fortune (meanwhile: 25+ Cents / kWh, crazy).


Yes, as long the instrument is powered, the CAL RAM will not lose its
content.. but what about unexpected power outages?


If you save the CAL RAM and repogram a fresh one after failure, there's
no need to send it to agilent.


No, not any RAM error needs recalibration.
Those error messages are not documented.

So I've scanned the text fragments in the firmware file:

RAM TEST 1
RAM TEST 2

probably refer to either to the configuration RAMs (2x32kB), andthe CAL
RAM (8kB).

Eror messages are:
NONVOLATILE RAM HiGH /LOW
CAL RAM FAILURE

and known ones from other users:

RAM TEST 1 HIGH / LOW

So I assume, that a real failure of the cal constants only apply on
message: CAL RAM FAILURE and perhaps RAM TEST 2 , as RAM TEST 1
should refer to the configuration RAMs only, HIGH to the high byte RAM,
LOW to the low byte RAM.

Errors of the cal RAM will manifest in additional ACAL or CAL failures.


Frank
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts

and follow the instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there. 


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] 3458a with 202: Slave Test Convergence (sigh)

2014-03-27 Thread JF PICARD
Hi,
 
What do you mean by  DIY  about the upgrade to opt 01 ?
thanks



From: Frank Stellmach frank.stellm...@freenet.de
To: volt-nuts@febo.com 
Sent: Thursday, March 27, 2014 8:54 PM
Subject: [volt-nuts] 3458a with 202: Slave Test Convergence (sigh)


I know that HP will (gladly?) take your sick 3458 and bring it up to current 
specs, followed by a calibration. Since my sick one is at engineering level 2 
or 3.something, it must be like it was built in 1989 - the dates on some 
components.

Should I try to do any troubleshooting, or just send it to HP?


Hi,

agilent simply replaces boards, but will not desolder and replace components.
Therefore, they will definitely replace the processor board completely (several 
hundred $), but then the latest firmware 9 will be inside. Additionally, a new 
calibration will have to be paid (+ several hundred $).

Alternatively, simply replace the 3 NVRAMs (DALLAS/maxim or digikey, 50$), and 
the 6 EPROMs for firmware 9 (agilent, 180$), if you are able to copy the CAL 
RAM and/or to recalibrate on your own.
And on the old processor board you can upgrade opt. 01 by DIY (10$), instead 
of agilent (+ several 100 $).
Another fellow volt-nuts did that on a 1989 unit, came up fresh like a daisy. 
I.e. no need for further replacement of any other boards.


@Tom Knox: Would you mind sharing your knowledge about error 202 here on that 
list, instead?
THX.

Frank

  

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] 3458a with 202: Slave Test Convergence (sigh)

2014-03-27 Thread J. L. Trantham
Willy,

I would encourage you to call Gary Bierman at Agilent/Keysight, Loveland,
CO, and ask him your questions.

He is very easy to talk to and open to any question or thought about how to
proceed.  IMO, he takes a personal interest in these instruments and wants
to make them the best that HP/Agilent/Keysight can produce.

From my conversations with him, they prefer to do component level repair,
not assembly level repair, and take advantage of all the 'aging' that has
occurred over the years.  Obviously, if there is something that is
destroyed, they will do an assembly replacement, not component level repair.

If you opt for their 'Repair Per Incident', at $2740.46, they will bring it
up to 'spec', provide the 'Agilent Cal' and return it to you, eligible for
their Repair Agreement at $178.68 per year, with a discount available for
muti-year agreements, up to 5 when I did that.  I don't know if that would
include compliance with all Service Notes or not but I'm sure Gary can
answer that question.  They do not routinely replace the Dallas chips but,
if they have failed, they will be replaced.  If you get the Repair Agreement
and the Dallas chips subsequently fail, they will replace them and provide a
fresh Agilent Cal, at no charge (except probably for shipping) except for
the annual rate of $178.68 that you would have already paid.  It is easy to
remove the Dallas chips and install sockets.  You can read the chips,
archive the data, then program a new chip and install it.  Just don't try to
read the chip when it is still warm from the removal process.  Trust me on
this. :^).

I don't know about the firmware upgrade but, again, Gary Bierman can answer
that question.  One thing I was wondering about recently is whether it is
possible to upgrade the 'inguard' firmware to version 2 or if it has to be
done at Agilent.  The 'outguard' firmware is upgradeable by removing and
replacing the EPROM (or six EPROM's, depending on which version of A5 you
have) with a new one (or set) from Agilent.

The only thing you need to achieve in order to qualify for eligibility for
the Repair Agreement is to send the meter in and have it able to be Cal'd.
If it passes their calibration process, then it is 'up to spec' and is
eligible for their Repair Agreement, as I understand it.

However, I defer to Gary Bierman regarding all of this.  You can't go wrong
calling him and seeing what he has to say.

My guess is that he will be able to make a recommendation to you about how
to proceed and be able to give you some insight into what the problem is and
how to resolve it.  Depending on what the issues are, it might be
easier/cheaper to fix it yourself or return it to Agilent.

Others on the list have great experience with these DMM's and I'm sure will
provide their insight as well.

Good luck.

Joe

-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of new
Sent: Thursday, March 27, 2014 1:36 PM
To: volt-nuts@febo.com
Subject: [volt-nuts] 3458a with 202: Slave Test Convergence (sigh)

I know that HP will (gladly?) take your sick 3458 and bring it up to current
specs, followed by a calibration. Since my sick one is at engineering level
2 or 3.something, it must be like it was built in 1989 - the dates on some
components.

Should I try to do any troubleshooting, or just send it to HP?  

Will they upgrade the current (aged and pretty much not drifting) boards?

Or do they take a good board from their stock and replace the current board
with one that may be new and 'drifty'?

I have read here that there are some unlisted upgrades that they do, also.

The RAM batteries will be replaced also, right? 

The biggest positive I see is the ability to go on the service plan and
never have to worry about a failure again after the 3458 is brought up to
current standards.

Willy
___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


[volt-nuts] Fluke 515A Batteries Leaked

2014-03-27 Thread Tony Greene
I have aquired an Fluke 515A calibrator.  The calibrator appears to work OK.

Internal inspection reveals the nicad battery pack has leaked.  It has 
outgassed and there is minor corrosion on some of the internal aluminum 
chassis.  The PC board appears to be OK.

What advise is available for cleaning this and what type of cleaners would one 
use ?

Thanks. Tony.


FREE 3D EARTH SCREENSAVER - Watch the Earth right on your desktop!
Check it out at http://www.inbox.com/earth


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] 3458a with 202: Slave Test Convergence (sigh)

2014-03-27 Thread jeffhook


Hi, 

  

I hate to jump into the middle of this as I haven't been following it but I 
have had about twenty 3458As go thru my ham shack. (a real sickness) 

Seven had the error 202 Slave Convergence and all were bad A3 boards. (A/D 
Converter  I ngua rd Logic ) Not saying this is always the case but swapping 
in a good A3 fixed all seven of my 202 errs. 



Jeff    



- Original Message -


From: Frank Stellmach frank.stellm...@freenet.de 
To: volt-nuts@febo.com 
Sent: Thursday, March 27, 2014 12:54:36 PM 
Subject: [volt-nuts] 3458a with 202: Slave Test Convergence (sigh) 

I know that HP will (gladly?) take your sick 3458 and bring it up to current 
specs, followed by a calibration. Since my sick one is at engineering level 2 
or 3.something, it must be like it was built in 1989 - the dates on some 
components. 

Should I try to do any troubleshooting, or just send it to HP? 


Hi, 

agilent simply replaces boards, but will not desolder and replace components. 
Therefore, they will definitely replace the processor board completely (several 
hundred $), but then the latest firmware 9 will be inside. Additionally, a new 
calibration will have to be paid (+ several hundred $). 

Alternatively, simply replace the 3 NVRAMs (DALLAS/maxim or digikey, 50$), and 
the 6 EPROMs for firmware 9 (agilent, 180$), if you are able to copy the CAL 
RAM and/or to recalibrate on your own. 
And on the old processor board you can upgrade opt. 01 by DIY (10$), instead 
of agilent (+ several 100 $). 
Another fellow volt-nuts did that on a 1989 unit, came up fresh like a daisy. 
I.e. no need for further replacement of any other boards. 


@Tom Knox: Would you mind sharing your knowledge about error 202 here on that 
list, instead? 
THX. 

Frank 

   

___ 
volt-nuts mailing list -- volt-nuts@febo.com 
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts 
and follow the instructions there. 
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] 3458A RAM batteries

2014-03-27 Thread Richard Moore
My 3458A tested with marginal RAM when I sent it to Loveland for cal. 
Gary advised that I buy the replacement RAM board which uses RAM chips 
that have replaceable batteries (? could they be super caps?) that snap 
onto the tops of the chips. They sent the meter back to me with no 
charge other than for shipping. I bought the upgrade board and new cable 
which is needed because the new board is shorter than the old style, and 
the old cable won't reach. Then I sent it back to Loveland for cal. I 
think the RAM board was around $450 -- old memory in failing RAM chips 
in head bad batteries there too.