Re: [volt-nuts] Finally got around to modifying my Fluke 845ab with LED's

2014-09-09 Thread Chuck Harris

One thing that might be hosing things up a bit is white led's don't
switch quickly.  They are really UV LED's with a bit of fluorescent
material between the LED and the world.  The fluorescent material will
also luminesce for hours after the LED is turned off.  Perhaps because
the led never shuts off completely, it is biasing the meter up a few
tens of microvolts?

-Chuck Harris

Dallas Smith wrote:

Finally got around to modify my Fluke 845ab with LED 's for the chopper 
circuit. Used
the 17 volt windings for LED's (Mouser 941-C513AMSNCW0Y0511 Warm White Round 
LED)
instead of the 130 volt, move red wire on transformer pin 9 to pin 7.This 
winding is
180 degrees out of phase, so I reversed the steering diodes (CR106  CR107) I 
left in
to help make sure the phase was correct for the LED's when connecting. Change 
R154 to
6K to set the brightness, selected for good operation of the zero control. Then
install jumper  to replace C119. Also changed the filter integration response 
caps
C111 to .022uF and C116 to 47uF, this stabilized the jitter to a manageable 
mode of
operation. Meter now works as well or better when the original neon's worked.

  As the meter originally had this problem, why is the offset reading different 
when
polarity is reversed at the meter input? About 10uV's.

Lamp Blocks.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] Finally got around to modifying my Fluke 845ab with LED's

2014-09-09 Thread Dallas Smith



Hi Chuck,


Its only 84Hz, It had the offset
problem long before I made this mod. The new meter at work many decades ago
also had this problem.


I pick white so the full spectrum would
be covered by the Csd's. Checked waveforms with scope, has sharp rising and
falling edges. This mod may not be as elegant as a Linear Devices copper mod,
but at lest I will not have to mess with the shielded front end.


The
offset may be due to the bias current in the front end reacting with the 
different
polarities.


Dallas


 
 
 Date: Tue, 9 Sep 2014 15:48:01 -0400
 From: cfhar...@erols.com
 To: volt-nuts@febo.com
 Subject: Re: [volt-nuts] Finally got around to modifying my Fluke 845ab with 
 LED's
 
 One thing that might be hosing things up a bit is white led's don't
 switch quickly.  They are really UV LED's with a bit of fluorescent
 material between the LED and the world.  The fluorescent material will
 also luminesce for hours after the LED is turned off.  Perhaps because
 the led never shuts off completely, it is biasing the meter up a few
 tens of microvolts?
 
 -Chuck Harris
 
 Dallas Smith wrote:
  Finally got around to modify my Fluke 845ab with LED 's for the chopper 
  circuit. Used
  the 17 volt windings for LED's (Mouser 941-C513AMSNCW0Y0511 Warm White 
  Round LED)
  instead of the 130 volt, move red wire on transformer pin 9 to pin 7.This 
  winding is
  180 degrees out of phase, so I reversed the steering diodes (CR106  CR107) 
  I left in
  to help make sure the phase was correct for the LED's when connecting. 
  Change R154 to
  6K to set the brightness, selected for good operation of the zero control. 
  Then
  install jumper  to replace C119. Also changed the filter integration 
  response caps
  C111 to .022uF and C116 to 47uF, this stabilized the jitter to a manageable 
  mode of
  operation. Meter now works as well or better when the original neon's 
  worked.
 
As the meter originally had this problem, why is the offset reading 
  different when
  polarity is reversed at the meter input? About 10uV's.
 
  Lamp Blocks.
 
  ___
  volt-nuts mailing list -- volt-nuts@febo.com
  To unsubscribe, go to 
  https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
  
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] Finally got around to modifying my Fluke 845ab with LED's

2014-09-09 Thread Chuck Harris

Hi Dallas,

The thing is, CdS cells peak sharply at about 560nm, which is green/yellow.
They are blind to most of white light... especially to the blues which are
prominent in white LED's.

White LED's are designed purely for bulk illumination.  They have a very
slow response so that they won't flicker when used for domestic lighting
off of a 50/60Hz power source.

There are also green yellow and red LED's that are designed the same way,
but there are many that switch quickly too.

-Chuck Harris

Dallas Smith wrote:




Hi Chuck,


Its only 84Hz, It had the offset
problem long before I made this mod. The new meter at work many decades ago
also had this problem.


I pick white so the full spectrum would
be covered by the Csd's. Checked waveforms with scope, has sharp rising and
falling edges. This mod may not be as elegant as a Linear Devices copper mod,
but at lest I will not have to mess with the shielded front end.


The
offset may be due to the bias current in the front end reacting with the 
different
polarities.


Dallas

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

[volt-nuts] Finally got around to modifying my Fluke 845ab with LED's

2014-09-09 Thread Mark Sims
I would probably use green or yellow LEDs,  but the white ones should not be a 
problem. 
 I built an LED analyzer/integrating sphere and one of the features is a 
circuit that optically measures the LED driver PWM frequency.  It can also 
detect the minute variance in LED intensity from an LED driven by a 950 kHz 
boost converter.  It also had no problems with a white LED driven at 4 MHz from 
a signal generator.  You see the long persistence phosphors mainly in large 
lighting LEDs and not in small indicator LEDs.  
  
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] Finally got around to modifying my Fluke 845ab with LED's

2014-09-09 Thread Dallas Smith



I thought that might be the case, the scope waveform shows
no problem. The white LED has multiple spectra output, thought it might have a
better chance having the right wave length, the extra brightness seems to help
with the deficiencies.


 
 From: hol...@hotmail.com
 To: volt-nuts@febo.com
 Date: Tue, 9 Sep 2014 22:33:45 +
 Subject: [volt-nuts] Finally got around to modifying my Fluke 845ab   with
 LED's
 
 I would probably use green or yellow LEDs,  but the white ones should not be 
 a problem. 
  I built an LED analyzer/integrating sphere and one of the features is a 
 circuit that optically measures the LED driver PWM frequency.  It can also 
 detect the minute variance in LED intensity from an LED driven by a 950 kHz 
 boost converter.  It also had no problems with a white LED driven at 4 MHz 
 from a signal generator.  You see the long persistence phosphors mainly in 
 large lighting LEDs and not in small indicator LEDs.  
 
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
  
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] Finally got around to modifying my Fluke 845ab with LED's

2014-09-09 Thread Chuck Harris

Hi Mark,

I have an itty bitty white led, probably called a T-1 size, and it
glows dimly for a long time after you shut it down.  They all pretty
well have to.

I think what you are seeing with your analyzer is the blue/UV component
that drives the phosphor element.  It will run just as fast as any old
style LED.

I also think that the blue/UV component, which is very bright, is what
is driving Dallas's chopper.

But enough on that.  I wasn't trying to criticize, just make mention of
something I noticed in my own work.
Where I got to thinking of this is when I burned out the strobe tube
in my G-R strobotach.  Replacement tubes were in the $300 range, so I
knew that would never happen, so I was thinking of making a solid state
replacement using a handful of those wonderful little white LED's.
They would certainly be bright enough, but I'm pretty sure the phosphor
hang would make them unsuitable for stopping motion... your eye would
see a blur instead of the razor sharp image you get with the strobe tube.

-Chuck Harris



Mark Sims wrote:

I would probably use green or yellow LEDs,  but the white ones should not be a
problem. I built an LED analyzer/integrating sphere and one of the features is a
circuit that optically measures the LED driver PWM frequency.  It can also 
detect
the minute variance in LED intensity from an LED driven by a 950 kHz boost
converter.  It also had no problems with a white LED driven at 4 MHz from a 
signal
generator.  You see the long persistence phosphors mainly in large lighting LEDs
and not in small indicator LEDs. ___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the
instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] Finally got around to modifying my Fluke 845ab with LED's

2014-09-09 Thread Brooke Clarke

Hi Chuck:

Is there something special about the GR strobe tube that prevents you from 
using a more modern strobe tube?

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.end2partygovernment.com/2012Issues.html
http://www.prc68.com/I/DietNutrition.html

Chuck Harris wrote:

Hi Mark,

I have an itty bitty white led, probably called a T-1 size, and it
glows dimly for a long time after you shut it down.  They all pretty
well have to.

I think what you are seeing with your analyzer is the blue/UV component
that drives the phosphor element.  It will run just as fast as any old
style LED.

I also think that the blue/UV component, which is very bright, is what
is driving Dallas's chopper.

But enough on that.  I wasn't trying to criticize, just make mention of
something I noticed in my own work.
Where I got to thinking of this is when I burned out the strobe tube
in my G-R strobotach.  Replacement tubes were in the $300 range, so I
knew that would never happen, so I was thinking of making a solid state
replacement using a handful of those wonderful little white LED's.
They would certainly be bright enough, but I'm pretty sure the phosphor
hang would make them unsuitable for stopping motion... your eye would
see a blur instead of the razor sharp image you get with the strobe tube.

-Chuck Harris



Mark Sims wrote:

I would probably use green or yellow LEDs,  but the white ones should not be a
problem. I built an LED analyzer/integrating sphere and one of the features is a
circuit that optically measures the LED driver PWM frequency. It can also detect
the minute variance in LED intensity from an LED driven by a 950 kHz boost
converter.  It also had no problems with a white LED driven at 4 MHz from a 
signal
generator.  You see the long persistence phosphors mainly in large lighting LEDs
and not in small indicator LEDs. ___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the
instructions there.


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.




___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] Finally got around to modifying my Fluke 845ab with LED's

2014-09-09 Thread Dallas Smith
Thank you Chuck,

I think you have something there, that is probably driving the choppers.
The meter had a little more jitter than before after the mod, that is why I 
modified the integration filter caps.
I like the slower response anyway. 
 
When I have time I'll share my mods for the Fluke 540b, Much better than 
before, can transfer at lightning speed.
The suggestion of using a voltage regulator (LM1117) for the 1.35v mercury 
battery from the EEV Blogs
Didn’t pan out. Its not a reference, it’s a voltage regulator. You don't need 
accuracy but stability.

Dallas

 
 
 Date: Tue, 9 Sep 2014 20:30:30 -0400
 From: cfhar...@erols.com
 To: volt-nuts@febo.com
 Subject: Re: [volt-nuts] Finally got around to modifying my Fluke 845ab with 
 LED's
 
 Hi Mark,
 
 I have an itty bitty white led, probably called a T-1 size, and it
 glows dimly for a long time after you shut it down.  They all pretty
 well have to.
 
 I think what you are seeing with your analyzer is the blue/UV component
 that drives the phosphor element.  It will run just as fast as any old
 style LED.
 
 I also think that the blue/UV component, which is very bright, is what
 is driving Dallas's chopper.
 
 But enough on that.  I wasn't trying to criticize, just make mention of
 something I noticed in my own work.
 Where I got to thinking of this is when I burned out the strobe tube
 in my G-R strobotach.  Replacement tubes were in the $300 range, so I
 knew that would never happen, so I was thinking of making a solid state
 replacement using a handful of those wonderful little white LED's.
 They would certainly be bright enough, but I'm pretty sure the phosphor
 hang would make them unsuitable for stopping motion... your eye would
 see a blur instead of the razor sharp image you get with the strobe tube.
 
 -Chuck Harris
 
 
 
 Mark Sims wrote:
  I would probably use green or yellow LEDs,  but the white ones should not 
  be a
  problem. I built an LED analyzer/integrating sphere and one of the features 
  is a
  circuit that optically measures the LED driver PWM frequency.  It can also 
  detect
  the minute variance in LED intensity from an LED driven by a 950 kHz boost
  converter.  It also had no problems with a white LED driven at 4 MHz from a 
  signal
  generator.  You see the long persistence phosphors mainly in large lighting 
  LEDs
  and not in small indicator LEDs. 
  ___
  volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the
  instructions there.
 
 ___
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
  
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] Finally got around to modifying my Fluke 845ab with LED's

2014-09-09 Thread NeonJohn


On 09/09/2014 09:00 PM, Brooke Clarke wrote:
 Hi Chuck:
 
 Is there something special about the GR strobe tube that prevents you
 from using a more modern strobe tube?

Extremely special.  Google GR guided arc flash tube.  There are some
old GR tech reports on the tube.

It uses a large bulb so it won't melt at very high flash rates and uses
a series of electrodes between the main ones, biased like in a
photomultiplier tube, to keep the arc nice and tight.  It quenches so
fast that the xenon afterglow is the main thing that limits its flash rate.

I have two of the earlier GR neon tube based Strobotrons.  The neon
actually has less afterglow than xenon but they hadn't discovered the
guided arc principle yet so it isn't as bright.  But it's real fast.  In
one of them I replaced the tube oscillator with a little microprocessor
board to give it crystal-controlled accuracy.  I retained the cool
multi-turn analog speed control to keep it looking authentic.
Linearizing that thing in software was a chore!

John


-- 
John DeArmond
Tellico Plains, Occupied TN
http://www.fluxeon.com  -- THE source for induction heaters
http://www.neon-john.com-- email from here
http://www.johndearmond.com -- Best damned Blog on the net
PGP key: wwwkeys.pgp.net: BCB68D77
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

[volt-nuts] Finally got around to modifying my Fluke 845ab with LED's

2014-09-09 Thread Mark Sims
No web page,  but a little description here.  It has changed quite a bit from 
the original description.  Now uses TAOS color sensors. Suppors Melexis IR 
thermometer chips.  Has 16 bit A/Ds.  Processor is an ATMEGA 1284.  The control 
program is based on Lady Heather.  Besides LEDs it can also 
charge/discharge/analyze batteries,  measure power converter/supply efficiency, 
etc
http://budgetlightforum.com/node/12296
CIrcuit board:
http://img69.imageshack.us/img69/3782/luxor.png
Sphere in action:
http://img515.imageshack.us/img515/4288/spherew.jpg
BTW,  the PWM sensor is a silicon phototransistor with not much blue light 
sensitivity...

-

 Do you have a web page on the LED analyzer/integrating sphere?

  
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] Finally got around to modifying my Fluke 845ab with LED's

2014-09-09 Thread Chuck Harris

Kind of.  It is a 7 pin miniature tube, with about
that many electrodes and triggers.  Very, very
fast.

I did find a good used tube so I am set for now.

-Chuck

Brooke Clarke wrote:

Hi Chuck:

Is there something special about the GR strobe tube that prevents you from 
using a
more modern strobe tube?

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.end2partygovernment.com/2012Issues.html
http://www.prc68.com/I/DietNutrition.html

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Re: [volt-nuts] Finally got around to modifying my Fluke 845ab with LED's

2014-09-09 Thread Chuck Harris

Hi Mark,

What would your PWM sensor circuit do if all of the AC was in the blue, and the
white light appeared as a DC bias on the phototransistor?

Mark Sims wrote:

BTW,  the PWM sensor is a silicon phototransistor with not much blue light 
sensitivity...