Re: [volt-nuts] What's the probability of a random used 3458A passing a Keysight calibration?
I think the only option for repair, out of warranty or out of repair agreement, is 'Per Incident' which the KS website lists as $2851.00. This is a 'one charge fits all' kind of repair price. It comes with a calibration as well. It's what makes the repair agreement attractive as 'insurance' on such an expensive instrument. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Dr. David Kirkby Sent: Tuesday, January 16, 2018 5:35 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] What's the probability of a random used 3458A passing a Keysight calibration? On 16 January 2018 at 23:06, Dr. Frankwrote: > Surely the AC errors if any must be associated with the A2 AC > Convertor >>> board - 03458-66502? >> Dave >>> >> > Why should this be a hardware error??? > > It's much much more probable, that this is a natural drift phenomenon. > This is 'as found' and the 'as left' report is missing, that would > tell the whole story. > > Anyhow, this can very probably be adjusted by the regular calibration > process. > > Frank > But given that Keysight calibration includes adjustments free of charge, the fact it was not adjusted suggests that it can not be adjusted. The AC board 03458-66502 is obsolete, but the replacement (03458-66512) https://www.keysight.com/my/faces/partDetail.jspx?partNumber=03458-66512 is $2243.00with a $439 trade-in. So the AC board price is $1794, which is not inconsistent with a 2100 Euro repair bill. Dave ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] What's the probability of a random used 3458A passing a Keysight calibration?
Forgot to add. All were purchased on theBay. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of J. L. Trantham Sent: Monday, January 15, 2018 9:12 PM To: 'Discussion of precise voltage measurement' Subject: Re: [volt-nuts] What's the probability of a random used 3458A passing a Keysight calibration? Dave, I have three of these meters that have all been sent to Agilent/Keysight for the 'Agilent/Keysight' Calibration and all have passed. Two of the three were still 'in spec' when arrived. The third lost its calibration when I failed to properly remove the 'CalRam' and read it. I think I tried to read it when still 'hot' from the removal process. There is a process that you can use via HPIB to read the contents of the CalRam but I did not know about it at the time. I wanted to remove the chip, install a socket and install a new chip before I sent it back to Agilent/Keysight. I had previously sent a Solartron 7081 to Ametek in England for calibration and I used it to provide a 'transfer standard' to do my own 'in house' recalibration of that 3458A and it was 'in spec' in all ranges when it went back to Agilent, except a couple of the AC voltage ranges. However, it was able to be successfully 'recalibrated' and was 'in spec' on its next trip for recalibration. That was at a time when the calibrations were done in Loveland. Once a 3458A has passed it's 'recalibration', you can purchase the 'Repair Agreement' for $192 per year with up to 5 years at a time (for a slight discount) that will allow you to send it back to Keysight for repair and recalibration for 'no charge' other than the cost of the 'Repair Agreement' if it ever fails during the covered time frame. Hope this helps. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Dr. David Kirkby Sent: Monday, January 15, 2018 3:56 PM To: Discussion of precise voltage measurement Subject: [volt-nuts] What's the probability of a random used 3458A passing a Keysight calibration? There's an "interesting" $5400 buy-it-now on eBay at the minute. It's described as "Agilent 3458A 8.5 digit Multimeter CALIBRATED". http://www.ebay.co.uk/itm/Agilent-3458A-8-5-Digit-Multimeter-CALIBRATED/ 232593692038 When one reads the description, one can see it is an HP (not Agilent) 3458A, that was sent to Keysight for calibration, but *failed* on the 3 V 4 MHz AC voltage. Apparently it passed on everything else. Personally I would sell that as "spares or repair" rather than "used" and "CALIBRATED", but different people have different ideas of what is good or not. Out of interest, I asked the seller what Keysight had quoted to repair this. He said 2100 Euros, (which about $2600 USD or £1900 GBP). So to get it in shape, it is not exactly peanuts. The seller said to me that he has a *lot* of experience of 3458As, and knows that a used one has less than a 50% chance of passing a Keysight calibration. I wonder if that's the experience of others that have bought used 3458As and sent them to Keysight. If true, and I ever buy a 3458A, I think I would only buy one on the condition that the seller sends it to Keysight, I pay the calibration cost, but if it fails, I get a full refund, including the calibration cost. I can understand a meter going out of spec over time, but if it can't be sorted out without thousands of dollars spent on it, then it is not such a good buy. But are more than 50% of 3458As like this? Dave Dr David Kirkby Ph.D C.Eng MIET Kirkby Microwave Ltd Registered office: Stokes Hall Lodge, Burnham Rd, Chelmsford, Essex, CM3 6DT, United Kingdom Registered in England and Wales as company number 08914892 http://www.kirkbymicrowave.co.uk/ Tel 01621'680100 / +44 1621-680100 ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Solartron 7081 randomly re-initialises
David, I have had a similar issue with mine but only on days when my stand-by generator turns on, takes the load for the house, then switches the load back to the street and shuts down. There is clearly a brief 'power hit' when this occurs. I've never tried to chase this down because it is relatively 'rare'. There is a fairly 'complex' shut-down circuitry in the 7081 described in the service manual and I wonder if you have an issue there. In any event, I suspect a 'power issue'. Good luck. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of David C. Partridge Sent: Thursday, April 14, 2016 6:09 AM To: 'Discussion of precise voltage measurement' Subject: Re: [volt-nuts] Solartron 7081 randomly re-initialises That makes sense though +5V looked OK, I'll look a bit harder! -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of acb...@gmx.de Sent: 14 April 2016 11:26 To: volt-nuts@febo.com Subject: Re: [volt-nuts] Solartron 7081 randomly re-initialises guestimate, executes a cold start caused by low voltage. so would check 5V for drops and ripple. keep us posted on your findings > Gesendet: Donnerstag, 14. April 2016 um 11:45 Uhr > Von: "David C. Partridge"> An: "'Discussion of precise voltage measurement'" > , solartron_schlumberger_equipm...@groups.io > Betreff: [volt-nuts] Solartron 7081 randomly re-initialises > > One of my 7081s will occasionally re-start from running in TRACK mode, > ending up with the INITIALIZED message on the display. > > > > Does anyone have any thoughts as to likely faults that would cause this? > > > > Dave > > ___ > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] HP-3458A questions
Frank, Thanks for the picture of your board. I have been interested in the 'plastic cap' on the LTZ1000A and you provided me with information about this in the past. Do you have a picture of the bottom of the board? I would love to see how the bottom half of the 'cap' looks. Thanks. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Frank Stellmach Sent: Tuesday, June 02, 2015 4:50 PM To: volt-nuts@febo.com Subject: [volt-nuts] HP-3458A questions Hi Randy, in advance, this change only improves the stability of the instrument, when the 3458A is powered. The LTZ1000 will not age (nearly), when the instrument is switched off. Therefore, it makes sense mainly, if your instrument is running for an extended period of time. (Reference: Spreadbury: Ultra Zener..) If you change the oven temperature, you will loose the 10V calibration, as the LTZ1000A will change its reference voltage. Although my instrument is strictly switched off, when not doing measurements, I reduced the oven temp. by connecting a 100k precision wirewound (3ppm/K) in parallel to the 15k BMF resistor. That gives 13k in total, yielding about 65°C like in the LT datasheet. An alternative placement is already foreseen on the PCB; in the CLIP, it's component designator is X411. I'll try to link in a photo of my change. lymex on the bbs35.hot forum obviously was more agressive, and paralled 75k of Vitrohm Thin Film, giving 12.5k for about 55°C. I would recommend my solution above the one of lymex, although his one gives theoretically more stability. Also, the lower stability of TF reistors may affect the stability (T, t) of the 15k BMF. With 100k/65°C, the oven runs about 30°C lower. Therefore, the ambient temperature has to be limited to max. 35°C, the fan has ALWAYS to be kept clean, giving a maximum temperature rise of 15°C, and the instrument may not be put in a rack (which would give additional 10..20°C temperature rise). Under these conditions, (35°C ambient + 15°C 3458A inner heating + 10°C self heating of LTZ1000A + 5°C regulation margin), the oven will always be stable. These 75k / 55°C would be possible, only if you limit the ambient temperature further, to = 25°C, and precisely monitor the 3458A inner temperature (TEMP?). Although my instrument is kept at 20..23°C throughout the year, i wouldn't feel comfortable with 55°C. The only possibility would be the exchange of the reference to the non-A version, as the self heating effect may vanish! Frank ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 731B
John, Thanks for the thoughts. I basically agree with the idea of not 'adjusting' the standards but when I got them, they all had 'problems', including leaking NiCd batteries that required clean up. Not knowing how long they had been off and having only calibrated DMM's (3458A and 7081) I chose to get the standards up and running then 'adjust' them to begin a 'new era' of standards, at least in my set-up. I also have two 732A's that have been brought back to life and they are sitting on now for about 12 to 18 months. I have not tried to read them in a while. I have been collecting other gear including Fluke 720A, 721A, 750A, and 752A and a Fluke 845AB. I also have a Fluke 887AB that I need to get working. My problem is that work is about 80 to 100 hours per week and I have little time to spend on this. But, once retired. I have left my 3458A's on for extended periods of time (weeks) and have been very pleased that when I turn them back on after weeks off, they come back within a couple of uV to the last readings of the 735C suggesting everything is fairly stable or at least drifting at the same rate and in the same direction. I'm a bit nervous about leaving the 3458A's on all the time due to the potential for the displays to fade or fail. So far, that's not been a problem. As far as calibration of the Fluke references, I would think Fluke would be the best bet but there is the problem of getting them there and back while still powered. I've been working on a method to do that using SLA batteries or using the Fluke 732A-7003 battery charger/pack. I think Fluke made a case that would house the 732A and the 732A-7003, both filled with batteries and completely charged, to ship the units in for calibration. My plan is to construct a 'case' to house two 12V 18AH SLA's and a charger that can be plugged in once the unit arrives at the calibration facility. I have acquired a couple of dead, industrial strength, UPS's that, once I get some time, I would like to rebuild and power my 'metrology section' with to deal with any power interruptions. I have a stand-by generator system as well. However, it all takes time to get up and running and I don't have much of that right now. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of John Phillips Sent: Friday, March 27, 2015 1:41 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 731B I would never adjust any pots on a working fluke voltage standard. When the pot moves you introduce more drift. Just keep track of differences. By the same token I do not turn off my 3458As as long as the line power is up. There was a time I had a rack of 15 3458As running waiting to be sold. We have a power outage about 1 time per year long enough to trip our julian date and time clock. We use it to track power up time. I do not see anyone posting 3458A in stat mode data: mean, SDEV, High, and Low, Range, and Number of samples over extended time. then reverse polarity and do it again. Also DC cal values before and after Auto cal and Keysight cal would be nice to know. Also using Before and after cal data from Keysight to correct values. Do not adjust the Flukes but keep a spreadsheet of best guess of its value. Something like mean of present guess + before cal data correction + after cal data correction. Example: Say you present guess (guess1) is 9.468 cal the meter using 9.468 and then read the cal var 1 (I think that is the one for the reference board) Your as found data should indicate how far off your reference is. correct and call that guess2 After you get the meter back read cal var 1 and then use successive approximation to when caling to get the same value in var 1. That should be guess3 So new guess is mean (guess1, guess2, guess3). If you have better confidence in one guess or less confidence in an other the values could be weighted. You never really know what the actual value is a given time but do know it is changing. You can guess at it's value like 9.468 is just a guess even if the real value is 9.4681 or even 9.468 with what we have you just cannot know for sure. There can alway be another digit. Anyone want to poke some holes in what I wrote? Does anyone have a better way to come up with a Cal value for a fluke 10 volt reference without sending it out for calibration? Where could send it to get a better calibration and what would it cost? On Thu, Mar 26, 2015 at 8:03 PM, J. L. Trantham jlt...@att.net wrote: A couple of years ago, I sat my 735C, a (at that time) recently Agilent calibrated 3458A, and two 731B's on a bench and have left the references on, with battery back-up for any power failure. I have turned the 3458A off and on a number of times in the interim. When I set all this up, I adjusted the 735C to 10.00 VDC as best I could using the then recently calibrated 3458A and adjusted
Re: [volt-nuts] Fluke 731B
A couple of years ago, I sat my 735C, a (at that time) recently Agilent calibrated 3458A, and two 731B's on a bench and have left the references on, with battery back-up for any power failure. I have turned the 3458A off and on a number of times in the interim. When I set all this up, I adjusted the 735C to 10.00 VDC as best I could using the then recently calibrated 3458A and adjusted the 731B's to match the 735C using the 3458A as a 'null' meter. Tonight, after the 3458A had been on for several weeks, I did the 'AutoCal All' on the 3458A and read each of the references. The 735C measures about 10.16 VDC, and the 731B's measure about 10.05 VDC each, within about 2 uV of each other. In the next couple of months, I plan to send the 3458A back to Agilent/Keysight for recalibration and see what the meter measures 'as received' and 'as left'. We'll see how close it is to being 'in tolerance' 'as received'. Hope this helps. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Orin Eman Sent: Monday, March 23, 2015 12:36 AM To: Discussion of precise voltage measurement Cc: richiem5...@gmail.com Subject: Re: [volt-nuts] Fluke 731B Richard, Some really good information has already been posted. Here is what I have found. I posted a comparison of the 731B against the Geller Labs SVR-T here: http://www.eevblog.com/forum/testgear/geller-labs-svr-and-svr-t/msg282582/#msg282582 There were within 1ppm of each other over 48 hours and 3.8 deg K. I'm currently running a comparison against the reference from an HP 3458A, using an Agilent 34461A in _ratio_ mode. A screen capture follows if the list lets it through. There is only 5 hours here as I just updated the 34461A firmware (it adds another digit to the scaling limits on the trend chart and is definitely worthwhile). One thing about the 731B is that it if the NiCd pack is dead or fully charged(!), there will be much more noise on the output. I designed a 'high-side' switch to switch between AC and battery power and it improves noise when the NiCd is fully charged. (The original design is a diode switch and if you have a NiCd pack which has a high voltage when fully charged, the half-wave rectified AC that is used to charge the NiCd pack leaks through to the voltage reference power supply.) A big WARNING. If a 731B still has its NiCd pack installed, replace it. I had Batteries Plus make me a replacement pack. About $30 as I recall. If the NiCd pack isn't installed, you need to remove the diode that goes from the battery pack to the voltage reference supply, or you'll get unfiltered half-wave rectified AC on your voltage reference power supply and it does get into the output! My 731B seems to be doing quite well IMO, being within 1ppm of the new at the time Geller SVR-T. It has also been reading 9.6 to 9.7 on the 34461A for the last year (as long as I don't turn on the 8568A that is next to the 34461A... the 8568A blows warm air into the 34461A and then all bets are off). Orin. [image: Inline image 1] On Sat, Mar 21, 2015 at 1:02 PM, Richard Moore richiem5...@gmail.com wrote: Hi all — realisitcally, what sort of stability can be expected from a fully functional 731B regarding variation in ambient temp a few degrees C around 23, and for long-term drift? Fluke’s specs are very conservative, and I think the 731 is far better than the specs would lead you to believe, but I have no practical experience with these units. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] plastic caps on 3458A reference board
Frank, Thanks for the info. I've wondered about that. In the days of 3D printers and CAD/CAM, it might be possible to have a 'run' of these 'made to order', so to speak. I wonder if Linear Technology would have any information about them? Thanks again. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Frank Stellmach Sent: Wednesday, January 28, 2015 5:28 AM To: volt-nuts@febo.com Subject: [volt-nuts] plastic caps on 3458A reference board Hello Joe, yes the cap consists of two parts. The upper one serves for thermal isolation of the LTZ1000A TO99 case against the environment and therefore reduces the power dissipation of the oven. The part on the solder side is much more important, as it covers the solder joints and avoids air draught over these pins. That avoids these low frequency voltage variations, which are mentioned in the LTZ data sheet. Anyhow, there are no further shieldings around the PCB, so the solder joints of the OP Amp and the precision resistors are exposed to that air draught, maybe from the fan. That's a further engineering fault they made on this reference. (The other faults are the 95°C oven temperature, the use of the A version instead of the non A, and the use of R417,200k temperature compensation resistor, which is necessary for the non A version only.) The hat is not included in the BOM inside the CLIP, therefore can not be ordered from HP, obviously. It is a smooth, shiny plastic, resembles the one used around the LM399H. Wasn't latter one something like polysulphone? Maybe suitable pieces of polystyrol foam will do the job also. Frank ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] [Bulk] Re: 3458A reference boards on ebay
When these boards are installed in a 3458A, the LTZ1000 is covered by a small, white, 'hat' that mounts with screws that pass through the two holes on opposite sides of the LTZ1000. I've never removed the board to see if there is anything on the bottom of the board that would cover the bottom of the LTZ1000. Does anyone know what this 'hat' is constructed of or the HP/Agilent/Keysight part number of this item is? Is it available from Keysight? In the 3458A the reference board is under a cover that covers about half of the top of the meter but not 'insulated' beyond that, except for the 'hat' referred to above. It would appear desirable to find what that 'hat' is and see if something could be fabricated to 'insulate' the LTZ1000 for a 'do it yourself' standalone reference. Would appreciate hearing anything about how the reference board is 'insulated' in the 3458A. Thanks. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Orin Eman Sent: Tuesday, January 27, 2015 11:45 AM To: Discussion of precise voltage measurement Subject: [Bulk] Re: [volt-nuts] 3458A reference boards on ebay Dave, I'll be putting it in an enclosure to use as a reference. It will probably be a smallish diecast aluminium box. It's what I did with my Geller SVR-T and it compared very well with my Fluke 731B. I'll likely use Pomona 3770 binding posts for output (I used them to repair the 731B). I'll also use the protection circuits from the Fluke 732 on the output - big reverse diode, gas discharge tube and TVS as I recall (I plan to add these protection circuits to the 731B). I'll likely do a PCB for it, though the local equivalent of veroboard would likely be OK; the socket spacing is 0.9 and the pin spacing 0.1. My meters are also 6 1/2 digit (3456A and 34461A) so I'm not too concerned with thermal EMFs with one count being 10uV on the 10V range. However, I will try to avoid temperature gradients and keep the outputs close together. I'll use star wiring for the 18V and ground as in the following: http://www.maxmcarter.com/vref/ It is debatable whether the -15V supply is needed - I'll test with and without. I'd add a few tantalum capacitors to the above design - on the regulator outputs and on the 3320 resistor, along with protection diodes for the LM317. There is also a HUGE thread about the LTZ1000 on EEVBlog - I'll be reviewing that for ideas: http://www.eevblog.com/forum/projects/ultra-precision-reference-ltz1000/ I actually have a couple of LTZ1000As here that I was going to DIY, but never got around to getting the precision resisters . I think there is a source in England mentioned in the EEVBlog thread. Orin. On Tue, Jan 27, 2015 at 2:24 AM, Dr. David Kirkby (Kirkby Microwave Ltd) drkir...@kirkbymicrowave.co.uk wrote: On 27 January 2015 at 09:16, Orin Eman orin.e...@gmail.com wrote: There is a seller letting a slow trickle of them out there... they seem to be going for about $165. The current batch of two is around $100, but will probably be bid up. I just received one that I won. Seems to be working fine after a quick breadboard lashup. I'll be making an enclosure for it next. A little expensive IMO, but given the trouble getting the precision resistors to DIY with the LTZ1000A, probably worth it. Orin Orin, What do you intend doing with it? I was thinking about getting one of those, with a view to putting it in a box with a couple of terminals to have something to compare with my 6.5 digit 3457A. But what put me off is a lack of knowledge in knowing how to convert a 3458A reference board into a boxed unit with a known output voltage at the terminals. How would I avoid / control thermal EMFs? I'd be interested to hear what your plans are for it, and how you intend tackling those issues. Do you know what the difference in the reference is between a standard 3458A (8 ppm) and the high stability option 002 (4 ppm) model? I'm guessing the chips for the option 002 might be the top performing ones. I wonder if there's any way to tell from your board if it came from a standard 3458A or a 3458A with option 002. Dave ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] HP3458A cal ram data dumper
Mark, I am very interested in this. I downloaded the file, saved it, double clicked the .EXE and got 'The program can't start because gpibib.dll is missing from your computer'. What do I need to do with the .CPP file? Do I need to install John Miles GPIBKIT library? I already have MS Visual Studio 2010 Express installed. Thanks for your help and please my 'EXTREME Newbie' status when it comes to working with 'code'. I'm trying to install on a Win7Pro 64bit Core i7 laptop. I have Prologix USB/GPIB, NI USB-GPIB-HS, and HP 82367B adapters that I can use. Any pointers very much appreciated. Thanks and Merry Christmas. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Mark Sims Sent: Wednesday, December 24, 2014 5:25 PM To: volt-nuts@febo.com Subject: [volt-nuts] HP3458A cal ram data dumper I have uploaded a .ZIP file with the source code and Windows .EXE file for the calibration/data memory data dumper program here: http://s000.tinyupload.com/index.php?file_id=0219779600667150 This output of this version of the program also produces a formatted/commented listing of what the various data values in the cal ram are. Program usage info is available by executing the program without any command line arguments. Also see the comments at the start of the .CPP source code file. Again, this program is to be used in conjunction with John Miles GPIBKIT library. Also many thanks to Poul-Henning Kamp for the hard work of documenting the HP3458 MREAD command and for figuring out everything about the cal ram memory map and contents. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] HP3458A cal ram data dumper
The .ZIP or the two individual files, .CPP and .EXE? Thanks. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Mark Sims Sent: Wednesday, December 24, 2014 6:41 PM To: volt-nuts@febo.com Subject: [volt-nuts] HP3458A cal ram data dumper Yes, the program requires GPIBKIT. Copy the code to the GPIBKIT directory. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Old HP3458A - SN: 2823A 03939
I suspect it has the six EPROM version of the A5 board, where the '5' firmware resides. The Ver 8 files are 'out there' and it should be a simple matter of burning the six chips if you have a chip burner. I don't know where the '1' firmware resides. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of cfo Sent: Tuesday, September 23, 2014 1:57 PM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Old HP3458A - SN: 2823A 03939 On Sun, 21 Sep 2014 15:22:14 +, cfo wrote: Gents I'm looking for a 3458A , and have found one : SN: 2823A 03939 w. opt 002 Seller says it's passing selftest. The seller has returned with this info : hello again too, rev : 5,1 i think CAL-Ram's have been changed on 1999 but i m not sure. In can accept a 30 days warranty no problem. So it's fw: 5,1 CFO ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] 732A drift
I think the 'IN CAL' LED has to do with a 'minimum voltage' but I don't think it has to with a minimum battery voltage. When you remove the battery pack, with the unit plugged in, I don't think the 'IN CAL' LED goes out. Otherwise, you would never be able to swap battery packs. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Todd Micallef Sent: Saturday, August 23, 2014 6:21 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] 732A drift The output impedance of both the 1v outputs is ~1k ohm. I wonder if Randy could check the 3458a input impedance with a 10M ohm resistor and the 10v output of the 732a. I was thinking that the 732a needed a minimum battery voltage to allow the in cal led to turn on with the short to the Lo terminal. I guess I could try it on one of mine that is waiting for repair. Todd Sent from my iPad On Aug 23, 2014, at 6:42, J. L. Trantham jlt...@att.net wrote: Randy, The 'IN CAL' LED is turned on as Todd describes. It goes off if power to the unit is lost and, thus, calibration is lost. It is not related to the battery charge LED. As I understand it, the units are designed to powered on 24/7/365 and are 'IN CAL' once they are powered up, stable (weeks, months?), and have been calibrated by your reference lab. Once power is lost, meaning lost AC and batteries depleted, the 'IN CAL' light goes out and outputs are thereafter unreliable. Once you decide to get the 732A calibrated, you will need to find a way to ship it to the reference lab and get it shipped back while continuously powered, connecting an external battery pack to the connector on the back of the battery pack. The internal battery pack is likely to last only a few hours. There are at least two types of connectors for an external battery, two 5 way binding posts and a Hypertronics connector which is a small black connector about 'dime' sized. I can find the part number for the mating connector if you need it. I think the +/- 1 uV drift with the 10 V output and the 3458A are within specs for both the 732A and 3458A. Not sure what to think about the drift on the other readings unless there is some sort of 'dirt' on the 1 V and 1.018 V binding posts and/or ground. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Todd Micallef Sent: Saturday, August 23, 2014 4:13 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] 732A drift Randy, I looked at the 3458a input impedance, and it is difficult to get the meter set to the FixedZ (10M) mode. However, the 100v range is 10M. It should default to HiZ on reset. I forgot to mention to check the outputs and guard resistance to ground. One of my 732a had some foam baffling under the cover that had dry rotted. There was a lot of leakage to ground. I can't remember if the foam was on the older or newer versions of the 732a. The cal light comes on with a short wire stuck in the hole and connected to a LO output terminal. I don't know, but maybe the battery charge led has to be off. Todd Sent from my iPad On Aug 23, 2014, at 1:16, Randy Evans randyevans2...@gmail.com wrote: I hope someone can help with a strange anomaly on either my 3458A or the 732A. The 732A 10V output as measured on my 3458A seems relatively stable over time (it bounces around about +/-1 uV but is it the 732 or the 3458?). However, the 1.000 VDC output drifts downward at a rate around 1-2 uV per second as soon as I plug the 3458A into the 732 output. If I remove the 3458A and connect it back up after a few 10s of seconds, the reading goes back to what it started at and then drifts downward again. The 1.018V output also drifts downward but at a much slower rate and not as much. Now the question is: is it the 732 or the 3458A? I tried to see it on my Agilent 34401A DMM but it really doesn't have the resolution, but I do seem to see it on the 1.000VDC output. If it is the 732A, what would cause it to drift downward like that? Since the 3458A has an input impedance of 10Gohm on the 1 and 10 V ranges, I wouldn't think the 732 would even see the difference of whether the 3458A is connected of not, but it clearly makes a difference as to how long its connected and how long it's been disconnected. Any one have any conjectures? Also, what will turn on the In Cal light? What does it mean if it doesn't come on? Thanks, Randy ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin
Re: [volt-nuts] HP 3458A Mem test 1 hight. How do you read 32K memory chips?
Mark, Would love to try it. I think the main reason to hang onto the calibration data is to be able to send the meter to Keysight and find out if the meter was 'in spec' when it arrived for calibration. If so, gives great confidence that the meter is a 'good one'. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Mark Sims Sent: Thursday, August 21, 2014 1:53 AM To: volt-nuts@febo.com Subject: [volt-nuts] HP 3458A Mem test 1 hight. How do you read 32K memory chips? I'm porting over my HP3458A NVRAM data dumper program to use John Miles GPIBKIT routines. It should make it usable with most GPIB interfaces out there and be able to run under more modern versions of Windoze. My first crack at it is working, but it is quite a bit slower than my original program. Not sure if anything can be done about that... BTW, my program does verify the checksum bytes in the CAL ram so there is a check that the dump went OK. Without Poul-Henning's work none of this would be possible. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] HP 3458A Mem test 1 hight. How do you read 32K memorychips?
Bill, Thanks for the insight. I always thought that I lost my CALRAM data by trying to read the chip in my programmer while the chip was still warm. What you are saying is that you did not try to read your chip, just unsoldered, installed a socket, plugged it back in and the data was corrupted. Very interesting. I have done this to three 3458A's and only lost the data on one of the units. This further supports the need to be able to read the contents before unsoldering and have that data be able to be used to program a new chip. Thanks for the insight and I would love to try some programs to do that. I only have Windows systems but I have several USB to GPIB adapter choices, PCI to HPIB adapter choices, and RS232 to GPIB adapter choices. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Bill Gold Sent: Thursday, August 21, 2014 9:33 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] HP 3458A Mem test 1 hight. How do you read 32K memorychips? When I removed the original NVRAM devices and then put in sockets, I plugged the original devices back in just to see if everything worked. Apparently some cal data was lost just desoldering the devices as I got error messages. It really didn't concern me as I intended to completely recal the meter with my local standards anyhow. It was hours before I plugged the original devices back in so they were completely at room temp. My guess was that since the date codes were 1992 and I did this in 2010 the batteries were almost dead after 18 years and just the shock of desoldering caused the end. Bill - Original Message - From: Mark Sims hol...@hotmail.com To: volt-nuts@febo.com Sent: Wednesday, August 20, 2014 7:20 PM Subject: [volt-nuts] HP 3458A Mem test 1 hight. How do you read 32K memorychips? The problem with removing the chip and copying it is the very real chance of glitching the contents in the process. Much better to make a backup copy first. The 32Kx8 NVRAMS are used for system memory and things like storage of user programs and data. Note that the two 32Kx8 devices are form a 16-bit word. The 2kx8 cal ram is only on the high byte of the data bus. MREAD returns memory contents as a signed integer value (-32768..32767) in ASCII. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] HP 3458A Mem test 1 hight. How do you read 32K memory chips?
John, I don't think I would worry too much about messing up the 'CALRAM'. If you 'mess up' the 'CALRAM', and your meter is, otherwise, OK, you will spend about $500 with Keysight for a 'Keysight' calibration, which I would recommend. If they get your meter and it is not 'suitable' for calibration (meaning that, in some way, it's 'broken'), you will be notified and offered the option of having them 'repair' the meter (for about $2700, which includes the 'Keysight' calibration) or returning the meter to you for your repair at a cost of half the calibration charge, IIRC. The key point about reading these NVRAM's is to, first and foremost, let the chips cool after removal before trying to read them. I learned about that the hard way. Second, make sure your programmer specifically supports the chip you are trying to read and program. If so, reading then archiving the data should be straight forward and programming a new chip should be equally straight forward. I am not familiar with the 'Mem test 1 high' issue but I suspect the Assembly Level Repair Manual or the Component Level Repair Manual should address the issue. I agree with others that this likely means that, in some way, your meter is 'broken'. If you can chase that down to a simple problem and repair it, getting the calibration with Keysight would be the next order of business. Once it passes calibration, you would be able to purchase the 'Repair Agreement' for a couple hundred dollars per year, with a discount for multiple year purchases up to a maximum of 5, IIRC, which gives you a meter with a 'factory warranty' for that time period. Personally, I would look for a 'professional' programmer rather than one of the 'cheaper' ones, especially if you are anticipating getting into vintage equipment and programming EPROM's, etc. BP Micro, Advin, DATA I/O, Elnec, etc., would ones I would look for. I'm sure others on the list would have their own recommendations. The prices on theBay right now are particularly ridiculous but good deals can be had. The majority of these units use parallel port connections and need WinXP as the OS. The later versions, starting around $500 will connect via USB and be able to utilize Win7 or Win8. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of John Phillips Sent: Wednesday, August 20, 2014 8:41 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] HP 3458A Mem test 1 hight. How do you read 32K memory chips? I do have a ni gpib-usb-hs http://sine.ni.com/nips/cds/view/p/lang/en/nid/201586 I was hoping to remove the chip, read the data with a usb programer and wright the data back into a new chip before installing it. Then do a full calibration. If the 16k chip has the cal vars what do the other 32k chips have? Thanks for the input. I should be able to read to read the data from a good meter with the MREAD and wright it with a programer. I have been using excl to read program 3458As I will see if I can get MREAD to work there. On Wed, Aug 20, 2014 at 5:33 PM, Mark Sims hol...@hotmail.com wrote: You REALLY don't want to mess with that 24 pin NVRAM chip... it contains the calibration memory. If you bugger the data in it, you get to spend a couple of grand getting the meter back into working order. I know a couple of people that did just that. The solution is to first make a backup copy of the memory contents using the (undocumented) MREAD command over the GPIB bus. Poul-Henning Kamp figured out how to do it... buy the man a beer... even better a case of beer... Search the archives for details. I just backed up all the memory in my 3 HP3458A's for when the inevitable BIG BAD DAY comes and the backup batteries go bye-bye. If you happen to have a NI GPIB-232CV-A (or possibly a Prologix) RS-232 to GPIB converter and can run a DOS or WIN98 program, I have a program that dumps the memory chips. It has code for the Prologix in it, but I've only used it with my a Prologix compatible unit I built... I don't have a real Prologix to test it with. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. -- *John Phillips* ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] New HP3458A
Randy, Congratulations on your 'new' 3458A. The 'CALRAM' back-up was just recently covered. There are two options: 1. Remove the NVRAM, read its contents with a programmer, program a new NVRAM and install the new NVRAM. I chose that option, installing sockets for the 'CALRAM' and the two other NVRAM's that are used for other, non-calibration, purposes. Be sure to let the NVRAM cool all the way down before trying to read it. Learned by experience with that one. 2. Read the 'CALRAM' via HPIB and the MREAD command (many steps involved but can be done as a program with your laptop/desktop), use that data to program a new NVRAM (I suspect there is a way to 'program' the NVRAM by HPIB as well but I don't recall ever reading anything about it on the list), remove the old NVRAM and install the new NVRAM. When I read the contents of the 'CALRAM' with a programmer and with the MREAD command, they matched exactly. However, when you do an ACAL, the CALRAM contents change. Does your unit have the single EPROM or the 6 EPROM version of the firmware? Also what version of firmware do you have? The latest is REV 9. REV 8 is 'out there' on the net. If you purchase an exchange display, and you have the single EPROM version of the A5 Outguard Controller Assembly, the REV 9 EPROM is $24.10 per the Keysight website. Might be worth adding to the order. The 6 EPROM version is $220. Also, what is the serial number? You can get a good idea of age by the serial number. Also, the various covers inside often will have dates of manufacture on them and the chips, etc., will have date codes, including the NVRAM's. Also, you can add option 001 by plugging some memory chips, specific details I don't recall right now. As far as the reading when shorted, I recall the Calibration Manual calls for a piece of copper wire to short the input terminals. It looks like a flying 'U' when properly fashioned. The Calibration Manual gives the details, IIRC. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Randy Evans Sent: Friday, August 15, 2014 8:40 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] New HP3458A I would like to thank everyone that replied to my query on what to look for in my new HP-3458. I did find that the AUTO ZERO button does work fine, thanks to Bill's comments. I had looked in the manual, just not far enough. The display is still perfectly readable but i would like it to be perfect. I am particularly interested in the exchange display for $272 (better than $700 for a new one, as Todd suggested). Does someone have a contact number at Keysight that I could call (my experience is that one can spend a lot of time calling around until the right person is finally found). I certainly need to figure out how to copy and replace the NVRAM - I lost the calibration on my Datron 1082 by not realizing the memory backup battery died and now I need to calibrate it myself (the HP3458A should make that doable, I hope). Any suggestions would be greatly appreciated. I believe it has been covered before on Volt-Nuts and I need to go through the past discussions to find it. One question I have for the group is what should the display typically show with the input shorted? I see a reading of about -.0025 mV. That seems rather high. I tried several different banana cables (gold plated, tin plated) used to short the input terminals to see if thermocouple effects might be responsible but there was no change in the reading. Still learning. Thanks, Randy On Fri, Aug 15, 2014 at 8:06 AM, Jason Watson watson.ja...@gmail.com wrote: I've also seen excessive Guard to Low leakage when varistor RV501 has gone bad (it's located on the front/rear switch pcb and it's possible to replace it while leaving the circuit board in place if you are careful). HP/Agilent/Keysight Part number is 0837-0196, cross referenced to a Harris Corp. V430MA3A. On Fri, Aug 15, 2014 at 10:46 AM, Stephen Grady grady.st...@gmail.com wrote: Randy, I have come across a few 3458A's that had leakage between Guard and Low when te guard is in external guard position. This was due to a leaky external guard switch and/or leaky front rear switch. This can be quickly determined by measuring resistance between guard and low with guard external. This normally in not an issue except when you are using 3458A to measure voltage with low above earth potential say in a bridge the guard low leakage will be loading other arm of the bridge. Kind Regards, Steve Grady Sydney, Australia -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Randy Evans Sent: Friday, 15 August 2014 1:22 PM To: volt-nuts@febo.com Subject: [volt-nuts] New HP3458A I picked up a used HP3458A today, which I needed for some precision DC measurements i need to
Re: [volt-nuts] HP3458A calibration memory backup
Poul, I would love to know where the CALNUM value is stored. Any idea? Thanks. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Poul-Henning Kamp Sent: Thursday, August 07, 2014 12:36 AM To: Discussion of precise voltage measurement; Mark Sims Subject: Re: [volt-nuts] HP3458A calibration memory backup In message blu170-w4732a073be87463ff74529ce...@phx.gbl, Mark Sims writes: I have three HP3458A's and decided it was time to back up the battery backed static RAM chips. [...] I dumped the third machine from a cold start and one byte was different between each dump. If you mail those files to me, I can try to see if that bit matters. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 p...@freebsd.org | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Buying HP-3458A
Randy, Yes, that is my understanding. The person who would be able to definitively answer this question is Gary Bierman at Agilent at the Loveland Calibration Lab. A very knowledgeable person regarding the 3458A. I had a long talk with him before undertaking the purchase of my 3458A's. I wound up with three 3458A's for about the price of one new one. All three 3458A's turned out to be functional with no failures. All three were of similar 1990's to early 2000's vintage with the same major components. Two were HP and one was Agilent, all made in the US. I had no Malaysian units. I removed the three Dallas chips in each of them, installed sockets, and installed new Dallas chips that I programmed with the data from their respective removed chips. A total of 9 Dallas chips for the three units. I also removed the single EPROM in each of them, installed a socket, and updated the firmware to the latest version I could find. I managed to lose the CAL data from one of the units by, I think, trying to read it while still warm from the unsoldering process. I calibrated that unit using some 'home standards' and a recently calibrated Solartron 7081 as the 'transfer standard'. I then sent the units in for Agilent Calibration, all three of which passed. Then I was able to avail myself of their 'service agreement' for each of the units. They all agree very closely when measuring my now somewhat upgraded 'home standards'. They are truly impressive units. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Randy Evans Sent: Tuesday, July 08, 2014 9:29 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Buying HP-3458A Joe, I imply from your e-mail that if I get a cheap 3458A that is complete but doesn't work, Agilent will repair it and calibrate it for a fixed fee of $2740.46, regardless of what is wrong with it? Of course that means i would need to insure the unit is complete with no missing parts. Randy On Sun, Jul 6, 2014 at 3:15 PM, J. L. Trantham jlt...@att.net wrote: It is worth mentioning (re-mentioning) that once your 3458A passes calibration with HP/Agilent/Keysight, etc., it becomes eligible for their 'repair agreement' at $178.68 per year (with a small discount if you purchase multiple years, up to a maximum of five years). This means that if it 'breaks' within that time period, it is repaired and recalibrated for no additional charge. Per the Agilent US site, the 'repair per incident' charge is $2740.46. So, if you happen to get a 'good' 3458A that Agilent calibrates, you can then get the service agreement (insurance policy?), which applies even if you do not change the NVRAM's. If they die, send it back to Agilent and they will repair and recalibrate. In my shop, the 3458A is the 'house standard' that I use to measure the resistance standards and voltage standards that I have. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of R.Phillips Sent: Sunday, July 06, 2014 2:08 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Buying HP-3458A Randy I'm already forgetting some of the facts that made up the price that I paid for the repair and re-cal. for my 3458A, as I have been involved in a number of other restoration/repairs, but it was remarkably similar in £ sterling, against your US $ price - so I guess it would seem to be a fixed charge ? My instrument failed when the 'classic' RAM/ROM 's ran out of battery support and further, they found another section which had gone bad. There's nothing to give your more confidence than the Agilent Calibration Certificate that comes with the package,as this is probably the prime instrument in the collection of we lucky owners, so it has to be good. As I have stated in a previous exchange, the 'new' processor/ROM/RAM board is now fitted with sockets, and the devices that were formerly fitted, have been replaced with the new type that only connect the battery support power when they are plugged in to their sockets. Roy -Original Message- From: Randy Evans Sent: Sunday, July 06, 2014 6:44 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Buying HP-3458A That e-mail I referenced is several years old. i believe the current repair price is just over $2800. Randy On Sat, Jul 5, 2014 at 10:49 PM, John Phillips john.philli...@gmail.com wrote: They do really offer that service as long as you send them a complete meter... No missing parts. Well they will take a few missing screws and such. My experience with them is they charge the same no mater what need to be fixed. Not sure what the price is, there prices are country dependent. I thought it was a little more than $2200 last I checked. On Sat, Jul 5, 2014 at 4:08 PM
Re: [volt-nuts] AC Voltage Measurement Standards
John, Thanks for the reply. I have both of those. What I'm thinking about is a measurement standard. That is, how do I know that my 'standards' are accurate? That brings to mind the 'thermal converter' issue. The 8506A apparently uses a 'thermal converter' as part of its measurement process. The individual A55 Thermal Converters can be used for their various ranges but might be more expensive if you collect all of them. So, the question is how would you go about 'proving' that your reference is indeed accurate? As best I can tell, this has to do with comparing it to some DC reference by virtue of an RMS (thermal, as best I can tell) comparison. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of John Phillips Sent: Tuesday, July 08, 2014 8:57 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] AC Voltage Measurement Standards I like the 5200A for a good stable unit up to 110 volts but if you go above that you would be better off with a 5100B which will do 1100 volts. On Tue, Jul 8, 2014 at 6:46 PM, J. L. Trantham jlt...@att.net wrote: I've been thinking about adding an AC Voltage Measurement Standard to my shop. It would appear that most of these have to do with thermal converters. Does anyone have any thoughts about this? I've been thinking about a Fluke 540B, 8506A, or a collection of A55 Thermal Converters. I have accurate DC measurement tools and DC standards. I would appreciate any thoughts. Thanks in advance. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. -- John Phillips ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Buying HP-3458A
It is worth mentioning (re-mentioning) that once your 3458A passes calibration with HP/Agilent/Keysight, etc., it becomes eligible for their 'repair agreement' at $178.68 per year (with a small discount if you purchase multiple years, up to a maximum of five years). This means that if it 'breaks' within that time period, it is repaired and recalibrated for no additional charge. Per the Agilent US site, the 'repair per incident' charge is $2740.46. So, if you happen to get a 'good' 3458A that Agilent calibrates, you can then get the service agreement (insurance policy?), which applies even if you do not change the NVRAM's. If they die, send it back to Agilent and they will repair and recalibrate. In my shop, the 3458A is the 'house standard' that I use to measure the resistance standards and voltage standards that I have. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of R.Phillips Sent: Sunday, July 06, 2014 2:08 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Buying HP-3458A Randy I'm already forgetting some of the facts that made up the price that I paid for the repair and re-cal. for my 3458A, as I have been involved in a number of other restoration/repairs, but it was remarkably similar in £ sterling, against your US $ price - so I guess it would seem to be a fixed charge ? My instrument failed when the 'classic' RAM/ROM 's ran out of battery support and further, they found another section which had gone bad. There's nothing to give your more confidence than the Agilent Calibration Certificate that comes with the package,as this is probably the prime instrument in the collection of we lucky owners, so it has to be good. As I have stated in a previous exchange, the 'new' processor/ROM/RAM board is now fitted with sockets, and the devices that were formerly fitted, have been replaced with the new type that only connect the battery support power when they are plugged in to their sockets. Roy -Original Message- From: Randy Evans Sent: Sunday, July 06, 2014 6:44 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Buying HP-3458A That e-mail I referenced is several years old. i believe the current repair price is just over $2800. Randy On Sat, Jul 5, 2014 at 10:49 PM, John Phillips john.philli...@gmail.com wrote: They do really offer that service as long as you send them a complete meter... No missing parts. Well they will take a few missing screws and such. My experience with them is they charge the same no mater what need to be fixed. Not sure what the price is, there prices are country dependent. I thought it was a little more than $2200 last I checked. On Sat, Jul 5, 2014 at 4:08 PM, Randy Evans randyevans2...@gmail.com wrote: I am in the market for a used HP-3458A and, in researching past messages, I came across an interesting question: *Chris Erickson* ericksonc2 at comcast.net volt-nuts% 40febo.com?Subject=Re%3A%20%5Bvolt-nuts%5D%203458A%20questionsIn-Repl y-To=%3C001b01cc7960%24a15d2840%24e41778c0%24%40net%3E *Thu Sep 22 19:48:41 UTC 2011* If Agilent will fix everything, bring it up to specs, and calibrate it no matter what's wrong with it for $2200, then why would I spend $5000-6000 or more for one that is in good shape - even if calibrated? Just grab the first bargain that comes along for $1500-3000, send it straight to Agilent and be money ahead. Seems the better course of action to me if that price is accurate - do they really offer this service? I could not find an answer to this question but it does beg the question. Does nayone have an answer as to why not buy the cheapest unit and send it in for repair? Thanks, Randy Evans ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. -- John Phillips ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] A Fluke 732A: Return it or keep it? - 3458A
Roy, The 3458A is quite the instrument. I'm not familiar with the latest 'plug-in/power-up' devices of which you speak. I suspect you have Rev 9.2. Mine have the single chip EPROM, through-hole, A5 board. I suspect yours has the SMT version. If you have just had Agilent calibrate the meter, it is now eligible for the Repair Agreement for $178.68 per year. You can buy up to 5 year's worth and get a slight discount, 4% or so, IIRC. Might be worth considering. Gary Bierman at Agilent Loveland told me about that. When I called Agilent to purchase that, they were a bit skeptical. However, when I told them that Gary Bierman told me it was possible, they got on the phone to talk to someone and then readily sold me the agreement. Not a bad price to cover a potentially expensive repair, especially if you plan to leave it on 24/7/365 and you don't know it's history. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of R.Phillips Sent: Saturday, June 14, 2014 4:00 AM To: 'Discussion of precise voltage measurement' Subject: Re: [volt-nuts] A Fluke 732A: Return it or keep it? Joe, Bill and Charlie Thank you all for your advice, and for the short term I will leave the 732A as is. Like you Joe, my 732A arrived with a set of 'dead' SLA batteries - these were replaced, and its been powered 24/7 since last October, again it came with no history. I bought mine from a dealer on the West Coast who had a batch of 14, so I guess that they came from a lab. Its stability is impressive, and the Thermistor reading is currently 4.5194 K ohms, with only small variations over time. My 3458A had a new processor/RAM/ROM board fitted by Agilent - so it has the latest 'plug-in/power-up' devices which I understand are a great improvement, and was calibrated just about six weeks ago. I take your point that the 3458A should be power continuously, but of course the display is of the fluorescent type which deteriorate over time. This is now less important, as I have had a new display installed, and the improvement is great. I certainly agree that the 3458A is very sensitive to temperature changes and as you advise, I frequently run 'Auto Cal' . I consider the cost of its refurbishment/calibration to be well worth while, considering the cost of a new instrument. Best regards. Roy ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] A Fluke 732A: Return it or keep it?
another small change you won't have to go through the same problem of trying again and again to engage. I have found that turning the pots may cause a drift which will show up days later due to the pot being dirty or whatever. So what I usually do is to turn the pot back and forth over several revolutions so that I can clean the contact wiper. I usually make an initial adjustment and then turn the pot about 1/32 of a turn in the opposite direction just to relieve any stress that might be there and cause a small change in the output voltage. I usually have very good results with these methods. But I will agree with Joe that sometimes it is better just to leave things alone and just use the known offset from 10 volts when checking or doing cals. Bill - Original Message - From: Todd Micallef tmical...@gmail.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Thursday, June 12, 2014 7:20 AM Subject: Re: [volt-nuts] A Fluke 732A: Return it or keep it? Roy, The 732A uses the 3059Y-1-101 type cermet pot for the 10V adjustment. You can use a simple plastic tweaking tool for adjustment. I believe you will need a fairly long tool. You can shine a light into one of the other holes to see the orientation of the pot. However, many people may prefer to leave the pot as is and just record the readings. It is going to drift anyway and won't stay adjusted for long. The drift may worsen once the pot position has changed too. I have adjusted mine, but only after the broken pots were replaced. Todd On Thu, Jun 12, 2014 at 9:57 AM, R.Phillips phill...@btinternet.com wrote: Hi Joe Sorry if I am retracing steps, but I have recently had my 3458A restored/recal'd by Agilent - its good to have it back. I also have a Fluke 732A which I can now check. Currently it is 10.24 volts, I now feel confident to trim it to the 3458A. I have tried to see the trimmers in each of the three holes - they appear to be some way within, and I am wondering if you could confirm the type of pot. that is used and just how long a trimming device is required. Is this a special tool only available from Fluke ? Regards Roy Phillips. -Original Message- From: J. L. Trantham Sent: Saturday, March 08, 2014 10:09 PM To: 'Discussion of precise voltage measurement' Subject: Re: [volt-nuts] A Fluke 732A: Return it or keep it? Charles, What size batteries did you use for your external pack and where are you located? In other words, how long can your external battery pack 'survive' keeping the 732A 'hot'? Is the external pack recharged at Fluke or does it have to make a round trip on the initial charge? Once I am convinced my unit is stable (which will probably be another six months or so, after I get my 3458A's re-calibrated by Agilent), I hope to be able to send it to Fluke for calibration. Fluke made a transit case (732A-7002) and a Battery Charger and Auxiliary Battery Case (732A-7003). However, I have never seen a picture of either of these. Does anyone have any information on these? How did you make your case? Did you include a charger? Combination of battery pack and shipping container or separate battery pack and shipping container? I was thinking of building a case with built-in charger, fuse, and AC connector and cord such that when the unit arrived at the CAL facility all that would be needed would be to plug it in and allow the external battery pack to recharge while the unit being calibrated was plugged in, recharge the internal batteries, and calibrated. Or does the 732A also charge the external battery pack along with the internal batteries? The connector you need (complete with female contacts) is a Hypertronics P/N D01PB306FSTAH and is in stock at Kensington Electronics at $8.96 each (plus tax and shipping). Only problem is their $50 minimum order. The 'shell' only is P/N D01PB306NT and the female contact is P/N YSK006-010ANH (three needed). http://ecommerce.keiconn.com/hypertronics/D01PB306FSTAH I ordered two of the connectors from Fluke using a P/N that one of the Fluke folks in their eCal facility gave me, noting that someone else had asked the same question a couple of weeks earlier, with Fluke Item# 2181497, described as '100-166, PLUG - MALE, HYPERTRONICS'. They were $12.31 each (plus tax and shipping) and arrived as the shell only, no contacts. If Fluke solves their supply chain problem, I sure would like to get the six female contacts I need to make my connectors 'complete'. When I got on the phone with Fluke, I could never find anyone that had any knowledge about this. Is there anyone there I should ask for that might understand my question? I look forward to your posting the update from Fluke. Thanks. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com
Re: [volt-nuts] 3458a with 202: Slave Test Convergence (sigh)
Willy, I would encourage you to call Gary Bierman at Agilent/Keysight, Loveland, CO, and ask him your questions. He is very easy to talk to and open to any question or thought about how to proceed. IMO, he takes a personal interest in these instruments and wants to make them the best that HP/Agilent/Keysight can produce. From my conversations with him, they prefer to do component level repair, not assembly level repair, and take advantage of all the 'aging' that has occurred over the years. Obviously, if there is something that is destroyed, they will do an assembly replacement, not component level repair. If you opt for their 'Repair Per Incident', at $2740.46, they will bring it up to 'spec', provide the 'Agilent Cal' and return it to you, eligible for their Repair Agreement at $178.68 per year, with a discount available for muti-year agreements, up to 5 when I did that. I don't know if that would include compliance with all Service Notes or not but I'm sure Gary can answer that question. They do not routinely replace the Dallas chips but, if they have failed, they will be replaced. If you get the Repair Agreement and the Dallas chips subsequently fail, they will replace them and provide a fresh Agilent Cal, at no charge (except probably for shipping) except for the annual rate of $178.68 that you would have already paid. It is easy to remove the Dallas chips and install sockets. You can read the chips, archive the data, then program a new chip and install it. Just don't try to read the chip when it is still warm from the removal process. Trust me on this. :^). I don't know about the firmware upgrade but, again, Gary Bierman can answer that question. One thing I was wondering about recently is whether it is possible to upgrade the 'inguard' firmware to version 2 or if it has to be done at Agilent. The 'outguard' firmware is upgradeable by removing and replacing the EPROM (or six EPROM's, depending on which version of A5 you have) with a new one (or set) from Agilent. The only thing you need to achieve in order to qualify for eligibility for the Repair Agreement is to send the meter in and have it able to be Cal'd. If it passes their calibration process, then it is 'up to spec' and is eligible for their Repair Agreement, as I understand it. However, I defer to Gary Bierman regarding all of this. You can't go wrong calling him and seeing what he has to say. My guess is that he will be able to make a recommendation to you about how to proceed and be able to give you some insight into what the problem is and how to resolve it. Depending on what the issues are, it might be easier/cheaper to fix it yourself or return it to Agilent. Others on the list have great experience with these DMM's and I'm sure will provide their insight as well. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of new Sent: Thursday, March 27, 2014 1:36 PM To: volt-nuts@febo.com Subject: [volt-nuts] 3458a with 202: Slave Test Convergence (sigh) I know that HP will (gladly?) take your sick 3458 and bring it up to current specs, followed by a calibration. Since my sick one is at engineering level 2 or 3.something, it must be like it was built in 1989 - the dates on some components. Should I try to do any troubleshooting, or just send it to HP? Will they upgrade the current (aged and pretty much not drifting) boards? Or do they take a good board from their stock and replace the current board with one that may be new and 'drifty'? I have read here that there are some unlisted upgrades that they do, also. The RAM batteries will be replaced also, right? The biggest positive I see is the ability to go on the service plan and never have to worry about a failure again after the 3458 is brought up to current standards. Willy ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] A Fluke 732A: Return it or keep it?
Charles, Thanks for the info on batteries. I agree, some are definitely better than others. However, I'm no 'connoisseur' of batteries yet. The 732A I got was 'unknown working condition' and I wanted the cheapest batteries I could find to test the unit and see if I could get it up and running. So far, it looks good. Do you wind up having to re-solder some of the battery wires in the battery module in order to use the 12 V 5 AH batteries or will they just swing to the other side since it appears that the (+) and (-) terminals will wind up on the same side, instead of opposite sides with the 6 V 4.5 AH batteries? I have not faced the shipping question yet, but, as David pointed out, there is an issue with shipping batteries. I suspect it could be shipped but would be ground and, therefore, need a huge battery in order to get it to the other side of the country 'hot'. I recall reading some thoughts on current drain for the unit when on batteries but I do not remember what it was. Clearly, it should be at a minimum, and therefore the battery should last the longest, when already 'hot' when removed from AC power. Has anyone done that experiment? I live in NW FL and I am not aware of Cal Labs in my neighborhood that could do the work. Atlanta, perhaps? Does anyone have a recommendation of a capable Cal facility 'close by' to NW FL? Short of sending it to Fluke or another facility, I will be limited to using my Agilent Cal'd 3458A to 'adjust' the 732A, then 'tracking' the 732A from year to year as the 3458A comes back from Cal. I paid less than Michael and more than you for my unit but it seems to work. I have seen these sell on theBay for $450 and up. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] A Fluke 732A: Return it or keep it?
Charles, What size batteries did you use for your external pack and where are you located? In other words, how long can your external battery pack 'survive' keeping the 732A 'hot'? Is the external pack recharged at Fluke or does it have to make a round trip on the initial charge? Once I am convinced my unit is stable (which will probably be another six months or so, after I get my 3458A's re-calibrated by Agilent), I hope to be able to send it to Fluke for calibration. Fluke made a transit case (732A-7002) and a Battery Charger and Auxiliary Battery Case (732A-7003). However, I have never seen a picture of either of these. Does anyone have any information on these? How did you make your case? Did you include a charger? Combination of battery pack and shipping container or separate battery pack and shipping container? I was thinking of building a case with built-in charger, fuse, and AC connector and cord such that when the unit arrived at the CAL facility all that would be needed would be to plug it in and allow the external battery pack to recharge while the unit being calibrated was plugged in, recharge the internal batteries, and calibrated. Or does the 732A also charge the external battery pack along with the internal batteries? The connector you need (complete with female contacts) is a Hypertronics P/N D01PB306FSTAH and is in stock at Kensington Electronics at $8.96 each (plus tax and shipping). Only problem is their $50 minimum order. The 'shell' only is P/N D01PB306NT and the female contact is P/N YSK006-010ANH (three needed). http://ecommerce.keiconn.com/hypertronics/D01PB306FSTAH I ordered two of the connectors from Fluke using a P/N that one of the Fluke folks in their eCal facility gave me, noting that someone else had asked the same question a couple of weeks earlier, with Fluke Item# 2181497, described as '100-166, PLUG - MALE, HYPERTRONICS'. They were $12.31 each (plus tax and shipping) and arrived as the shell only, no contacts. If Fluke solves their supply chain problem, I sure would like to get the six female contacts I need to make my connectors 'complete'. When I got on the phone with Fluke, I could never find anyone that had any knowledge about this. Is there anyone there I should ask for that might understand my question? I look forward to your posting the update from Fluke. Thanks. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] A Fluke 732A: Return it or keep it?
Michael, I would recommend you download the manual if you have not already done so. (Sorry, just noticed you had looked at the manual.) I don't think you are going to have enough resolution with the 3457A to characterize the 732A, other than a rough idea of stability. You will need a 3458A, a Solartron 7081 or some other 8 1/2 digit DMM. The unit uses four 6 V SLA batteries that are relatively cheap. I ordered 8 of them since shipping was the same, whether I got 4 or 8. Unless they were recently replaced, you will need to replace them. The 'In Cal' light needs to be 'reset' after the unit is powered up, stabilized, and calibrated. It remains on as long as there is continuous power applied to the unit, either from AC, the battery pack, or an external battery attached to the connector on the back. I have a spare connector, yours for cost plus shipping, if you need one. Two others on the list have sent me the money for one but work has really interfered with my life this week and I have not yet shipped them. It will be tomorrow. (Sorry, guys, my apologies.) The batteries I ordered are these: http://www.batteryclerk.com/store/p/80492-Panasonic-LC-RB064P-Sealed-Lead-Ac id-AGM-VRLA-Battery.html $5.69 each plus shipping. If power is lost, power must be re-attached and the 'In Cal' light 'reset'. The light is 'reset' by connecting the negative terminal to a spot in the 'reset' hole just below the 'In Cal' light. If the battery charge light is not on, and the unit is plugged in, the batteries are either completely charged, not connected, dead, or there is a problem with the charger circuit. My bet is a battery issue since I bet you had to plug in the unit to get it to work. There is a multi-turn pot inside each of the three holes and they are difficult to get to but can be seen if you have a flashlight to shine in essentially parallel to your line of sight. If the 10 V can't be brought to exactly 10.000 V, there are 'jumpers' under the top layer of insulation for the 'heated unit' that can be adjusted to bring it 'in range'. At least I was able to bring mine 'in range' and 'calibrated' against my Agilent Cal'd 3458A. Mine has been on continuously now for about 2 months. I 'calibrated' it after it was on for about a week. It now reads about 9.920 V and seems extremely stable (3458A with NDIG=8 and NPLC=60). I'm thinking of 're-calibrating' it and see how close it will stay to 10.000 VDC. I think it will be more stable the longer it is on. I have no idea of how long it was off before I turned it on. The schematic of the A5 Reference PCB Assembly shows the calibration adjustments and the drawing of that assembly shows the location of the adjustments. They are deep inside the unit. You'll need a small flat bladed tool, a flashlight, and probably a magnifier. I used a length of 12 ga. copper wire with one end flattened to make a screwdriver. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Michael Hong Sent: Friday, March 07, 2014 6:19 PM To: volt-nuts@febo.com Subject: [volt-nuts] A Fluke 732A: Return it or keep it? Hi nuts, This is my first message. Just reading the posted messages is a big learning to me so I didn't post but instead trying to read all. Here is my issue. I received a Fluke 732A yesterday which was purchased from an eBay seller. (1) I plugged in and started measuring the 3 voltages and thermistor resistance value. After about 2 hours later all 3 voltages were stabilized and thermistor value stabilized after 24 hours. All voltages and thermistor values were measured on a HP 3457 which was calibrated and certified by a local cal lab 10 months ago. Here is how I measure: After changing the connection to a different voltage I wait 30 second. I start STAT and wait 1 minute and R MEAN. 10V (10.000360 36.0ppm, 360uV) 1V (1.383 38.3ppm, 38.3uV) 1.018V (1.0180200 20ppm, 20uV) According to the manual adjustable(calibration) ranges are: 10V 5ppm 50uV 1V 5ppm 5uV 1.018 50ppm 50uV So obviously either 732A or 3457A is wrong. I tried to contact the person who calibrated the HP 3457A. No response. (2) No light on BTRY CHG and IN CAL Rear panel Battery Operation switch is on. (3) Front panel calibration hole I put a thin flat blade driver into the holes. I couldn't feel anything until 4 1/4 inside where I felt something but not the potentiometer. I offered to the seller that if I find the 732A is no problem but only requires the new battery, refunding $150 for the battery will close the deal. All other case, I return the item. He responded Just return it. My questions: (1) Which one is wrong the DMM or 732A? (2) What is the cheapest battery replacement cost? (3) Were the calibration mechanism removed or they are there? In manual I couldn't find any detailed description or diagram of it. Last, I paid $650 plus $35 shipping. Do you think it is worth
Re: [volt-nuts] A Fluke 732A: Return it or keep it?
Michael, Definitely need to keep it on, ship it hot, and have an external battery pack attached. IIRC, the internal batteries are good only for about 4 hours. I have not tested that though. I think I read this in the manual or one of the list members mentioned it. That's why the connector is on the back, to allow attaching an external battery pack for longer duration away from AC power. I would love to get mine calibrated with Fluke but I think I want to wait a few more months before undertaking this. Any suggestions from the group on how long to leave it on? What 'drift' rate I would want before concluding it is 'stable'? I have a 3458A Opt 002 that I can use to track it with and another 3458A that I can track the thermistor with. Removing the battery pack is easy, just remove four screws and slide it out. Certainly worth doing if you are going to keep it. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Michael Hong Sent: Friday, March 07, 2014 10:22 PM To: volt-nuts@febo.com Subject: Re: [volt-nuts] A Fluke 732A: Return it or keep it? Thank you all for your concerns. I will reply out of order, sorry. IN CAL right came on as I poked(turned) with a slim screw driver. Yes, I know it supposed to be on after calibration. AC PWR light turned and stayed on ever since I plugged in the main yesterday. When I pulled out the main plug as you said, voltages went down immediately. Probably the battery is completely gone. I have been asking quote on this standard from Fluke down to local no name guy long before I buy it. I have no intention of calibrating this by myself. That defeats the purpose of the purchase. Because it will be my primary standard. I just wanted to turn slightly to both directions to see if the potentiometer is working since the 732A's volts looked out of ranges. I calibrated 8 720As with Nanovolt Null meter and Keithley 155 last week, so I have no fun left to turning the small thing. The battery is big concern to me. When I called Fluke, one of the rep said it should be shipped to them hot Because 732A's battery life is very limited, it should be shipped using Morning Service. I was going to call FedEx if they have 12 hours service. And I belive most cal lab require the 732A to be delivered hot. The manual says so. Thermistor value is now 4470.1132 ohm It changes less than 50mmohm/hour now. Once I purchase one more 732A, I will experiment how close it come back to before turn-off value. Again thank you all. Best Regards, Michael ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Question for Fluke 732A owners
Oh!! THAT's what they mean by 'male' and 'female'. I get it! :^)) My 732A has male pins on the battery module and female sockets on the mating connector/plug. BTW, the mating connector is not easy to find. Several phone calls to Fluke led me to Fluke Item Number 2181497, Description: 100-166, PLUG - MALE, H. I ordered two but all that arrived was the shell, no contacts. I tried calling Fluke but was unable to find anyone that had a clue about the plug or it's contacts. Perhaps I'll try again when I get some spare time to spend another hour on the phone. An effort with the Hypertronics website led to P/N D01PB306FSTAH which is the mating plug for the Fluke 732A, contacts and all. The plug is composed of two components: HYP D01PB306NT, PLUG, 3P BLACK, and HYP YSK006-010ANH, SOCKET, SOLDER CUP 50 UIN, but it is ordered with the P/N listed. I found them at Kensington Electronics in Austin, TX, with a minimum $50 order. I ordered 6 after reviewing the specifics with the sales person who confirmed that I had the correct part number. They work perfectly. I have some spares if anyone is interested. If you wanted male contacts for the mating plug for the 732A, as best I can tell, the P/N would be D01PB306MSTH. The 'TAH' relates to 50 uin gold over nickel plating over female sockets and the 'TH' relates to 50 uin gold over nickel plating over male pins. A little time with the data sheet for the Hypertronics connectors should make it all clear. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Charles Steinmetz Sent: Wednesday, February 26, 2014 4:47 PM To: VoltNuts Subject: [volt-nuts] Question for Fluke 732A owners Fluke used two different connector schemes for the DC input on 732A battery packs. The oldest units have a pair of banana jacks, and later units use a 3-pin connector made by Hypertronics. My question concerns only the latter (732A battery packs with the Hypertronics connectors). The Hypertronics contacts are interchangeable between the panel mount connector body and the plug body. It is customary to build the panel mount connectors with female contacts, and the plugs with male contacts. However, it appears that Fluke may have made at least some 732A battery packs with male contacts in the panel mount connectors. I'd be interested to know what is out there in the field. If you have a 732A battery module with the Hypertronics 3-pin connector, could you please look at the connector and let me know if the contacts are male or female? Just to be painfully clear, I'm referring to the gold-plated metal contacts only, not to the black plastic parts of the connector. The gold-plated metal contacts can be male (i.e., the metal part you can see is a solid pin) or female (i.e., the metal part you can see is a hole that will accept a pin). Thank you, Charles ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] Fluke 732A Questions
I just spoke with Fluke about calibration options for the 732A and how to connect an external battery pack. There are three calibration options: 1.Z540 $700 2. Accredited $1000 3. Primary Standards Lab $1485 These are described on their website: http://us.flukecal.com/support/calibration-services/electrical-rf The Fluke part number for the male plug to connect to the socket on the back of the battery module, 732A-7005, is 2181497 and is available for $12.75. I ordered two. With that, I can construct a transit case with room for external batteries and be able to ship it overnight to Fluke for calibration. The folks in calibration service suggested I send them an email requesting the value of the Thermistor and whether old/original calibration certificates might still be available and they would forward it to the folks that, at one time, were in charge of keeping that information. At one time, that information was kept but, according to them, no one wanted it. Therefore, they no longer keep that information. However, perhaps it might be available for this 'vintage' equipment. We'll see. Thanks to all for their help. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 732A Questions
Just noticed the 'banana plugs' mounted on the chassis to 'align' the battery module. Duh. Makes perfect sense. Also, thanks for the Pomona recommendations. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Bill Gold Sent: Wednesday, January 01, 2014 10:38 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Joe: A correction on the dropout voltage when the In Cal light goes out, I found my notes and it is 21 volts instead of 22 volts. The way I did this was to remove the battery pack, AC voltage and then just feed in a power supply where the battery pack would have delivered it's voltage to the back-plane mother board. Then I just reduced the voltage until the light turned off. So that means that Fluke was following the battery manufacturers specs of 5.25 volts at the end of the discharge curve. The banana jacks on the battery pack have two jobs. One is to provide a very good chassis ground to the battery pack and the other is probably to guide the pack accurately into the mother board. One other thing I have found out the hard way is that the holes in the top plate of the battery pack where the battery terminals come through are very close to the battery terminals and can short out to the chassis. I have made the holes a LOT bigger to avoid this problem. This really depends upon the brand of batteries you are using but after having to repair the regulator board I would rather be safe than sorry. I have 6 Pomona 1756-48 low thermal leads for when I need to be careful with thermals. I also have made my own low thermal(I think) using Pomona 4892 gold plated double banana plugs and Belden 9272 twisted pair twinax, using a Pomona 1825 for the ground lead. If I need single banana plugs to connect to the 732A I use Pomona 4897 gold plated. I know that this isn't the best possible low thermal because the wire is tin plated but I have looked and I can't find a good source of bare copper twisted pair shielded wire. When I am making measurements with the home made cables I just wait a few minutes for the thermals to settle down. Why Belden 9272? It was the best choice from what was available at work. From what I have seen Pomona was acquired by Fluke and makes special low thermal cables for Fluke that are not available through the Pomona catalog. When you find out from Fluke what services they offer and prices please let all of us know. Bill - Original Message - From: J. L. Trantham jlt...@att.net To: 'Discussion of precise voltage measurement' volt-nuts@febo.com Sent: Tuesday, December 31, 2013 5:13 PM Subject: Re: [volt-nuts] Fluke 732A Questions Bill, I have yet to put my 732A though it's paces but I could connect an external power supply, disconnect the AC supply and the battery module and slowly decrease the supplied voltage and see what happens when. I'll try your measurement algorithm for the battery to see when the 'IN CAL' LED goes out and what the current draws are. If the measured voltage at J10 is 27 V, then the actual battery voltage should be about 27.6 V, accounting for the diode in the circuit (unless the current draw during measurement is so low that no voltage drop occurs across the diode). I found the NIST/NBS Tech Note easily. I think if I decide to send it out for calibration, I'll use Fluke, if they still calibrate the unit. I'll call them later this week of next week to see what services/accessories they might still have available. Yes, I wound up with the 735C. It was easy to replace the NiCd battery pack and bring it 'on scale' by adjusting the 'jumpers' and the front panel adjustment. I now only have to deal with the periodic 'NO CAL' LED indication. I need to study the 732A manual then take a close look at the 735C to see how it works on it. This also brings up the question of 'low EMF cables' for accurate measurement. I have a Fluke 720A K/V Divider on the shelf that I have yet to bring out to check to see if it is operational. Another one of those 'projects' for a later time. Where would one look for 'low EMF cables' to be used in precise measurement experiments and what errors might one expect to find using Pomona gold plated Banana Plug cables (which is as good as I have at this time)? Thanks for the info. Happy New Year to all. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 732A Questions
Charles, I would like to find the mating plug for J10 for several reasons, being able to get it to calibration being just one of them. However, I was hoping it would be good enough to make it to calibration on the internal batteries. I seem to recall reading that the internal batteries were good for a few hours but nothing like what the 732B would do, which sounded like the 732B could make the trip on the internal batteries. Where did you find the resistance of the thermistor 'as shipped'? I saw mention of being supplied to the original owner when shipped as well and I have had the top cover off and the battery module out to replace the batteries and found nothing to suggest a thermistor value. I have also lifted the top layer of insulation off the oven to move the cal-jumpers but did not see anything that suggested a thermistor value. My 'metrology level' standards are my HP 3458A's, calibrated by Agilent, and Solartron (Ametek) 7081's calibrated by Ametek. I have been thinking of getting the 732A calibrated by Fluke but I didn't know if they would still calibrate the unit. I wonder if they would have the 'as shipped' thermistor value on file for my specific serial number? I, too, read about the stability of the thermistor being the most important measurement. That was in the 'change sheets' attached to the manual on Didier's site. The 'cal-jumper' was at '40' when I got the unit. Per my 3458A, I could only get it down from about +220 uV to about +100 uV with the front panel adjustment. By moving the jumper to '20', I could only get it up to about -12 uV below 10 V by turning the adjustment all the way in the opposite direction. However, by connecting jumpers for both the '20' and '10', I was able to get it at 10 V with the front panel adjustment at about center position. I have not 'logged' any values yet but every time I go to the shop, it is about +/- 1 or 2 uV from 10 V. I do the 'AUTO CAL' on the 3458A each morning I go into the shop. I measure the resistance with a Fluke 8050A (simply because it is the closest DMM to the standard) and it is 4229 to 4230 ohms each day. Interestingly, when I got the 732A, undisturbed Fluke Calibration stickers were in place over each of the adjustments on the front panel, on a back screw on the top cover, and 'half' of a sticker on the back over the power supply module, as if the other half was lost when someone removed the battery module. Somehow, I find that comforting, given that it now seems so stable, making me think it is 'normal aging' that is responsible for the values I measured rather than a problem with the oven, etc.. IIRC, the spec is 6 ppm per year for 10 V, or 60 uV per year. I'll have to look at the power supply module to see if I can find some date codes but I suspect the unit is 25 to 30 years old. Unfortunately, there were no 'dates' on the stickers to tell when the unit was last calibrated. I was thinking of sending the 3458A's and the 732A out for calibration simultaneously and seeing what I find when they get back. I'll call Fluke and ask about their Cal services, the plug, and the shipping case/battery pack issues. Thanks everyone for very useful information. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Charles Steinmetz Sent: Monday, December 30, 2013 11:23 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Joe wrote: Also, how do you get your 732A calibrated? Do the batteries last long enough to ship overnight to a facility? No -- I assumed that was why you were looking for a battery connector. You install an external battery for shipment overnight. Fluke can supply a complete external battery with cable and plug ($375, last I knew), or you can make your own. I believe they can also supply a shipping container, which has a dedicated space for the Fluke external battery. (If you don't use Fluke's container, build something very sturdy with wood and padding and with handles that make it hard not to carry it and set it down right side up.) I wouldn't have anybody but Fluke calibrate a 732A. There are 3 levels of calibration, all with data, from about $700 to about $1500 (the last is calibration by Fluke's Primary Standards Lab). As far as thermistor values go, the relevant question is not what is the value in absolute terms, but rather what is the value in relation to the as manufactured value, which was provided by Fluke when the unit shipped. This had been recorded by the original owners of each of my units, and they are all within 2 ohms of the original value today. If you do not know the original value, then your original measurement is a going forward value and it cannot tell you anything about where you are in relation to the original manufactured value. The fact that you had to change cal jumpers suggests that either the oven temperature has shifted since manufacture, the reference
Re: [volt-nuts] Fluke 732A Questions
I was afraid of that. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of R.Phillips Sent: Tuesday, December 31, 2013 7:53 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Joe Reference the thermistor value and its recorded value - I asked Fluke this question and the answer was, they did supply the individual 732A's value in written form to the original purchaser, but unfortunately they did not maintain a separate record. Roy -Original Message- From: J. L. Trantham Sent: Tuesday, December 31, 2013 1:22 PM To: 'Discussion of precise voltage measurement' Subject: Re: [volt-nuts] Fluke 732A Questions Charles, I would like to find the mating plug for J10 for several reasons, being able to get it to calibration being just one of them. However, I was hoping it would be good enough to make it to calibration on the internal batteries. I seem to recall reading that the internal batteries were good for a few hours but nothing like what the 732B would do, which sounded like the 732B could make the trip on the internal batteries. Where did you find the resistance of the thermistor 'as shipped'? I saw mention of being supplied to the original owner when shipped as well and I have had the top cover off and the battery module out to replace the batteries and found nothing to suggest a thermistor value. I have also lifted the top layer of insulation off the oven to move the cal-jumpers but did not see anything that suggested a thermistor value. My 'metrology level' standards are my HP 3458A's, calibrated by Agilent, and Solartron (Ametek) 7081's calibrated by Ametek. I have been thinking of getting the 732A calibrated by Fluke but I didn't know if they would still calibrate the unit. I wonder if they would have the 'as shipped' thermistor value on file for my specific serial number? I, too, read about the stability of the thermistor being the most important measurement. That was in the 'change sheets' attached to the manual on Didier's site. The 'cal-jumper' was at '40' when I got the unit. Per my 3458A, I could only get it down from about +220 uV to about +100 uV with the front panel adjustment. By moving the jumper to '20', I could only get it up to about -12 uV below 10 V by turning the adjustment all the way in the opposite direction. However, by connecting jumpers for both the '20' and '10', I was able to get it at 10 V with the front panel adjustment at about center position. I have not 'logged' any values yet but every time I go to the shop, it is about +/- 1 or 2 uV from 10 V. I do the 'AUTO CAL' on the 3458A each morning I go into the shop. I measure the resistance with a Fluke 8050A (simply because it is the closest DMM to the standard) and it is 4229 to 4230 ohms each day. Interestingly, when I got the 732A, undisturbed Fluke Calibration stickers were in place over each of the adjustments on the front panel, on a back screw on the top cover, and 'half' of a sticker on the back over the power supply module, as if the other half was lost when someone removed the battery module. Somehow, I find that comforting, given that it now seems so stable, making me think it is 'normal aging' that is responsible for the values I measured rather than a problem with the oven, etc.. IIRC, the spec is 6 ppm per year for 10 V, or 60 uV per year. I'll have to look at the power supply module to see if I can find some date codes but I suspect the unit is 25 to 30 years old. Unfortunately, there were no 'dates' on the stickers to tell when the unit was last calibrated. I was thinking of sending the 3458A's and the 732A out for calibration simultaneously and seeing what I find when they get back. I'll call Fluke and ask about their Cal services, the plug, and the shipping case/battery pack issues. Thanks everyone for very useful information. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Charles Steinmetz Sent: Monday, December 30, 2013 11:23 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Joe wrote: Also, how do you get your 732A calibrated? Do the batteries last long enough to ship overnight to a facility? No -- I assumed that was why you were looking for a battery connector. You install an external battery for shipment overnight. Fluke can supply a complete external battery with cable and plug ($375, last I knew), or you can make your own. I believe they can also supply a shipping container, which has a dedicated space for the Fluke external battery. (If you don't use Fluke's container, build something very sturdy with wood and padding and with handles that make it hard not to carry it and set it down right side up.) I wouldn't have anybody but Fluke calibrate a 732A. There are 3 levels of calibration, all with data, from about $700 to about $1500 (the last
Re: [volt-nuts] Fluke 732A Questions
Jeff, Too late. I have already moved the jumpers and adjusted all of the front panel adjustments to achieve the desired output voltages. Since I don't have any 'historical' data about the performance of my specific unit, I will 'start from scratch' on the unit I now have. Whether to send it out for calibration or just keep it, adjusted by the 3458A, and watch from now on is the issue. I don't have enough data to decide about long term stability at this point but I am gaining valuable information about how long it will last on batteries and how to go about preparing it for shipment for calibration if I go that way. Thanks for the input. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of jeffh...@comcast.net Sent: Tuesday, December 31, 2013 5:23 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Hi, Are you guys adjusting your 732As? I have a friend that works in a large CAL lab and he told me to never adjust my 732As, just compare them. They send out their 732 Bs to Fluke but are never adjusted, just compared and sent back with current voltage reading ?? Thanks - Jeff - Original Message - From: Charles Steinmetz csteinm...@yandex.com To: Discussion of precise voltage measurement volt-nuts@febo.com Sent: Tuesday, December 31, 2013 9:56:37 AM Subject: Re: [volt-nuts] Fluke 732A Questions Joe wrote: Where did you find the resistance of the thermistor 'as shipped'? I saw mention of being supplied to the original owner when shipped as well It was recorded on the original factory cal certificate. I have the original certificates for two of mine. The others had rear panel stickers applied by their original owner (a cal lab). A properly working, good example should hold 0.1 ppm or better year in, year out. I've seen some that do not meet that, but lots of them do. I'm not aware of any other non-JJ standard that can match it, including most 732Bs. (Some folks think 732As are generally better than 732Bs because they are older and, thus, more stable because better aged. I think there is more to it -- many 732Bs are plenty old enough to have settled down to that level if they are ever going to, but even the older 732Bs don't, in general, seem to exhibit quite the stability of the 732As.) Best regards, Charles ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 732A Questions
Interesting thought. I love it when guys of that caliber compete. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of John Phillips Sent: Tuesday, December 31, 2013 5:51 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Call Agilent about 3458A calibration then tell them that you really want your fluke calibrated and see what they will do for you. On Tue, Dec 31, 2013 at 3:48 PM, John Phillips john.philli...@gmail.comwrote: Any thing you do other than record the current voltage reading will introduce more uncertainty and increase drift unless you have a defect. Replacing a defective part will start the aging process over again for that part. Do not change anything if you can help it. Agilent is a good cal lab to use as well. They will run it for a few days and take readings a few different times and give you a report. Most of the time the battery is not enough to keep the box hot until they plug it in so they count on warming up a few days. Hope that helps. John On Tue, Dec 31, 2013 at 3:31 PM, Chuck Harris cfhar...@erols.com wrote: The obvious reason for that suggestion is the adjustment pot is likely to be noisy and is likely to introduce uncertainty. The big question that comes to my mind is if you care about that level of uncertainty, why not replace the pot with a voltage divider made of some low thermal coefficient fixed value resistors? -Chuck Harris jeffh...@comcast.net wrote: Hi, Are you guys adjusting your 732As? I have a friend that works in a large CAL lab and he told me to never adjust my 732As, just compare them. They send out their 732 Bs to Fluke but are never adjusted, just compared and sent back with current voltage reading ?? Thanks - Jeff ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/ mailman/listinfo/volt-nuts and follow the instructions there. -- John Phillips -- John Phillips ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 732A Questions
of any 732A units and just relied upon the inter-comparison method to hopefully keep a reasonable value of the Volt. I have looked around the South SF Bay Area for a Cal Lab who could Certifiy one of my units and the only place I could find would be Simco in Sunnyvale. I will have to ask them what they charge for a Certitication of a 732A. The local Fluke place used to charge $500 but then that was using 732As that they had sent into Fluke in E verett WA. But then since this is just a hobby anymore I don't need, and didn't need while I was working, the full blown services and accuracy of Fluke in Everett. I guess a 732B would be a way to get all of this but who needs it right now, not me. Of course I could always get my own Josephson Junction but then I would be a real crazy Volt Nut and certifiable at the same time. I have never even heard about the Fluke 735C until I saw one on fleabay, and I assume you were the one who bought this when I saw you asking about schematics. Happy New Year Bill - Original Message - From: J. L. Trantham jlt...@att.net To: 'Discussion of precise voltage measurement' volt-nuts@febo.com Sent: Monday, December 30, 2013 6:28 PM Subject: Re: [volt-nuts] Fluke 732A Questions Bill, Sounds like you are further along in the 'volt-nut' process than I am. I'll have to chase down the NIST paper. I have been very impressed by the stability of the 732A and the 735C. Unfortunately, I don't have any documentation on the 735C and only the early manual for the 732A. I would be interested in the specifics of the battery check procedure in your manual. Unfortunately, my manual does not have a page 2-9. I note that there are two banana jacks on the front of the battery module, both black, one connected to ground and the other not connected to anything. Any idea of what they are for? I always like to keep equipment as 'original' as possible. If I can't find the correct connector, I'll find something similar that will fit in the same hole and use it instead. We'll see what I hear from Fluke and Hypertronics. Also, how do you get your 732A calibrated? Do the batteries last long enough to ship overnight to a facility? Thanks for the info. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Bill Gold Sent: Monday, December 30, 2013 7:03 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Joe: My 732A units vary in thermistor resistance from 3.408K ohms to 4.514K ohms and have been extremely stable to with in +/- 5 ohms for years now. I did measure the oven temperature of the one with the 3.408K ohms just to see what was causing the thermistor resistance to be a little low and came up with about 47 degrees centigrade, as best as my equipment would allow me to measure it. Equipment was a Fluke 80TK thermocouple module, a Fluke 80 PK-1 thermocouple bead and a HP 3456A DVM. I put shrink wrap around the bead and then threaded the bead into the temperature oven through one of the adjustment holes in the front panel and made sure that it was far into the oven, probably around the middle. Then I checked another unit that read 4.5K ohms and it read around 43 degrees C so all of that made sense. I believe I saw somewhere that the spec was 45 degrees C +/- 2 C. I did have to change the jumpers on the Calibration PCB Assembly on one of my units so that it would match other 732A units that were In Cal and had certified values. But that unit has since drifted down at a rate of around 1.2 ppm per year which is with in specs but a little more than I have seen from other units. Recently this unit has suddenly quit drifting down and seems a lot more stable now. That was S/N 459. I have a S/N 343 which has been proven rock solid for years now at around +/- .3 ppm and just seems to have DC noise stability as it just goes up a little and then down a little and never needed any adjustment. I also have a S/N 460 which is extremely stable also. So I guess that age does seem to factor into stability as well as how long they have been powered up. But the only way to insure that you have a good volt is to have at least 4 units and then inter-compare them periodically following NIST (NBS) Technical Note #430. While 430 was written for Saturated Standard Cells, the technique seems to work just fine at 10 volts also. You can also find this measurement technique in the 1st Fluke Calibration - Philosophy in Practice book published around 1974. I gave up on the unobtainium connector on the back of the battery pack very quickly. I drilled a 1/4 holes on either side of the unobtainium connector, after removing it, and used two single miniature banana jacks to allow me to connect to the batteries for the purpose of checking their performance as described in the 732A manual page 2-9
Re: [volt-nuts] Fluke 732A Questions
I called them this AM and they only had a 'copy' of the manual and not an original. If I understand what you are saying, you got a copy as well. If so, were the 'change sheets' part of the 'copy'? When I asked about that, I was told that there were no 'change sheets' in the copy she was looking at when I called and asked about condition and change sheets. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of T. Micallef Sent: Monday, December 30, 2013 1:06 PM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Fluke 732A Questions My manual came from manualsplus. It was in excellent condition compared to the ones on eBay. You do have to ask them for a picture of every manual since they can vary in condition. I have rescanned the whole manual, which I should have done anyway. I have two missing pages and I am hopeful the seller will scan them and forward them to me. Todd ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 732A Questions
Charles, I just assumed that it was obsolete. I'll contact Fluke. Thanks. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Charles Steinmetz Sent: Monday, December 30, 2013 5:52 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Robert wrote: As Roy has already said, this is a hypertronics connector.* * * Sorry, I don't have any, or a cheap source. The plug that mates with the jack on the back of the newer 732A battery is Fluke part number 2181497. IIRC, I paid about $10 each (some years ago) direct from Fluke. Best regards, Charles ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 732A Questions
Roy, Thanks for the info. Looking at the Hypertronics catalog, it would appear that the panel mounted socket (with male, solder cup, pins) is: P/N D01EEB306MSUTH And the mating plug would be: P/N D01PB306FSUTAH However, the dimensions are a bit off and the appearance of the panel mounted socket is not the same as what is on my 732A, J10, on the back of the 732A-7005 battery module. My connector is exactly as shown in the change sheets with a dimension of 0.475 inches side to side and 0.500 inches top to bottom (as viewed from the back of the unit) rather than 0.512 x 0.512 as suggested in the catalog. However, the diameter of the mounting hole and the presence of a 'notch' appears to be as shown in the catalog, thus suggesting that it, indeed, is a series D01 part. The diameter of the threaded part of J10 is 0.410 inches and the catalog calls for a hole diameter of 0.441 inches. One other observation is the presence of a 'symbol' on the bottom edge of the socket (as viewed from below) that looks like a stylized 'FRE', The 'F' perhaps could be an 'A' and the 'E' perhaps could be a 'B'. The symbol starts low and ends low with a 'peak' of the symbol in the middle of the 'R', sort of a 'roof' shape to the top of the symbol and flat on the bottom of the symbol. Not sure what that means but I have a picture of it if anyone thinks it would help. It does not look like the stylized 'H' symbol shown in the Hpertronic catalog. Perhaps the company was something else before they became Hypertronics. I looked through the change sheets for the 732A and could find no mention of the value of the thermistor resistance at temperature, only comments about what its stability should be. Knowing that you have a value in the mid 4K range is reassuring. My S/N starts with 3. Right now, it indicates 10.005 VDC on the 3458A and 4230 ohms on the Fluke 8050A. Thanks for your help. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of R.Phillips Sent: Sunday, December 29, 2013 8:40 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 732A Questions Hi Joe I also obtained a Fluke 732A back in October last, and I have fitted a new set of SLA batteries and it is working just fine - I am still very impressed by its accuracy and stability, and its build quality. First, the external 3-pin power socket is a HYPERTRONICS 'D' type connector (female) type 100-166 - but I have not found one to date ! I also expected the temperature monitoring thermistor to give a 3 to 4 K ohms, as indicated in the manual P/N 645051 dated May 1983 - but I guess that you have a later model as I do, mine being serial # 4845***, some modifications were made,and I understand that this included a change of thermistor, so the standard reading is in keeping with your result. My unit gives a very stable 4.520 K ohms, being between 4.52080 and 4.52090 K ohms. Judging by your findings, you are also using the 1983 copy of the User Manual - we could both use a more up to date issue. I am awaiting the repair of my 3458A - so I am having to rely upon a recently calibrated Keithley 2015 and my good old 3456A - but they both give very impressive results - The Keithley giving 1.07 volts for the 1 volt output, and 10.0 volts for the 10 volt output (how I miss that final digit). Looking forward to your further results and any other owners comments. Best regards Roy -Original Message- From: J. L. Trantham Sent: Saturday, December 28, 2013 10:51 PM To: volt-nuts@febo.com Subject: [volt-nuts] Fluke 732A Questions I have reviewed the prior postings on Volt-nuts (a very rewarding review, BTW)) regarding the 732A and have two questions: 1.Has anyone determined the part number, or a source, for the mating plug to J10, the external power connector for the 732A-7005 battery pack? I noted some postings recently about this but did not find a definitive response regarding the identity of this plug. The alternative would appear to be a complete replacement of J10 and its plug, as long is it all fits in the opening in the panel. I hate to 'bore holes' in vintage equipment. 2. The oven thermistor in my 'new to me' unit measures 4229 ohms (+/- an ohm or two) after the unit has warmed up for a week or so. I note the manual refers to 3K to 4K for the value of this thermistor when the unit is 'stable' as well as other's posting values in the mid 3500's ohms for their units. Should I be concerned? The unit seems to be stable to within about 2 uV over about a week (as measured by my 3458A - see below). Should I open the unit and try to measure the oven temperature or just be satisfied that the unit seems to be working? I had to replace the four 6V 4AH SLA batteries and they charged up appropriately as judged by the front panel LED's. I had to remove the 'jumper' for the '40' option on the A7 board and connect the jumper to the '20
Re: [volt-nuts] Fluke 732A Questions
Todd, Very good information. Sounds like we are getting close to the correct P/N for the mating plug for the socket. My 732A has only a single transistor mounted to the frame from the A4 Regulator Assembly, in the position of Q4, and the resistance is 4230 ohms after the unit is 'warmed up' and stable. I recently bought a manual as well but it did not have any 'update sheets'. I would love to have a copy of your 'update sheets'. I did a 'clean up' of the update sheets on Didier's site and can send them to anyone that wants one. I printed the .PDF from the Fluke site and added the 'cleaned up' version of the update sheets and it makes a very nice manual. I would love to add the other sheets to this. I also have a 735C that is very similar to the 732A but lot's of mechanical differences. It has two transistors on the Regulator Assembly, presumably A4 and Q4 and Q14. I do not have a manual for the 735C. It's thermistor value is in the mid 3K range. It, too, is very stable and has many similarities to the 732A but has recently developed a 'new problem'. The 'NO CAL' LED comes on after a few days for no apparent reason. I need to open it up again and take a look at that circuit to see what the problem is. On the 732A, the 'IN CAL' LED is continuously lighted when there has been no power interruption. On the 735C, the 'NO CAL' LED illuminates when power is restored after a power failure. I replaced the NiCd battery pack and all seems to be well but the 'NO CAL' LED comes on after having been reset (by inserting a copper wire in the 'adjust' opening). I wonder if my 'reset' process is the problem. On the 732A, you connect to the negative terminal then touch the 'pad' inside the 'RESET' opening. There is no such 'RESET' opening on the 735C. Just placing the wire (that I use to adjust the 10 V adjustment) inside the adjustment opening and aiming to the side would reset the LED. Happy to help with the scanning of the manual and update sheets if needed. Thanks again for the info. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of T. Micallef Sent: Sunday, December 29, 2013 6:05 PM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Fluke 732A Questions Joe, I have the external battery pack (732A-7003) that I picked up a while ago. It is indeed the D01 connector. It is printed clearly on the outside edge along with the three letters that look like ARB. The company's website states they are now owned by Smiths Connectors and the connector is part of Hypertac. No list of distributors was found on the website. I found a seller Astrex Electronics, but they do not have a price listed for the female version, just the male. I searched for D01P306FSTAH. This is the female plug with solder cups. To date, I have reconditioned three 732A's and they have been operating fine with little issues, except when I lost power for an extended time and two batteries in one of the packs died. All four were replaced and they have been working fine. The two older 732A's have thermistor values in the 3k-4k range. The newer unit (based on the newer heater rev) is around 4.25K. All three had their potentiometers changed on the regulator and pre-regulator boards. The datecodes on all appeared to be from 1985. The parts list must have been updated as some of the pots did not have the same value as the ones installed. I cannot remember which ones were different. I did not see a mention in the change/errata. Also, I had to remove the reference ovens to repair broken 10V potentiometers, as someone had broken one (free spinning) and another unit appeared to have a noisy wiper so it was replaced as well. The 1V and 1.018V pots were not changed. The third 10V pot was replaced because I had success with the other two and it was much easier the third time around. The jumpers on the all the A7 boards were adjusted. I highly recommend checking the inline fuse to the heater if you suspect it is not working. You can find it tucked inside behind the oven. It should be covered in heat shrink. Lastly, there was a rev done that does not appear in the downloadable users manual on Didier's website. It appears the heaters were separated into two groups of two instead of one group of four. Some of the parts were mounted under the oven assembly on the A8 Preheater PCA. If you look at the A4 board from the top, you will see either one or two transistors mounted to the frame (Q4,Q14). If you see only one (Q4), chances are you have the newer rev. I bought the manual recently and just noticed these revs were in my book. I will try and scan the extra info and upload them. Todd ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list --
[volt-nuts] Fluke 732A Questions
I have reviewed the prior postings on Volt-nuts (a very rewarding review, BTW)) regarding the 732A and have two questions: 1.Has anyone determined the part number, or a source, for the mating plug to J10, the external power connector for the 732A-7005 battery pack? I noted some postings recently about this but did not find a definitive response regarding the identity of this plug. The alternative would appear to be a complete replacement of J10 and its plug, as long is it all fits in the opening in the panel. I hate to 'bore holes' in vintage equipment. 2. The oven thermistor in my 'new to me' unit measures 4229 ohms (+/- an ohm or two) after the unit has warmed up for a week or so. I note the manual refers to 3K to 4K for the value of this thermistor when the unit is 'stable' as well as other's posting values in the mid 3500's ohms for their units. Should I be concerned? The unit seems to be stable to within about 2 uV over about a week (as measured by my 3458A - see below). Should I open the unit and try to measure the oven temperature or just be satisfied that the unit seems to be working? I had to replace the four 6V 4AH SLA batteries and they charged up appropriately as judged by the front panel LED's. I had to remove the 'jumper' for the '40' option on the A7 board and connect the jumper to the '20' and '10' options (total of '30') in order to get the unit to adjust to 10.000 VDC on my 'Agilent In Cal'd' 3458A. The need to change the jumpers, perhaps, could be just an ageing issue or, on the other hand, a 'temperature' issue with the oven. Should I open the unit and directly measure the temperature (supposedly about 48 degrees C) or just be satisfied with what I have? My recently added 735C also needed moving some 'jumpers' in order to get it 'on scale', as determined by my 3458A, although it's thermistor measures about 3330 ohms (after being on for several weeks). Thanks for everyone's help. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] Fluke 735C - Follow Up
Thanks to everyone for their thoughts about the 735C 10.0 VDC standard. I cleaned up the 'corrosion', rebuilt the NiCd battery packs, and plugged it in. It measured about 10.00010 VDC after being on for 24 hours or so. I reviewed the manual for the 732A and found quite a similarity, including the resistor 'decade' of 2K, 1K, 500, 250, and 125 ohm resistors to select from to achieve calibration. On the 732A, these are brought out to a separate (sort of external) PCB to make selection easier. However, on the 735C, all are mounted on the 'oven unit PCB' and I had to open the oven several times in order to come up with the correct combination of resistors in parallel to achieve calibration. I used my recently Agilent Cal'd 3458A and was able to adjust the 'CAL' pot on the front panel (R20 per the 732A schematic) to bring it 'on scale'. Looking at the output, it seems to be +/- 2 uV when watching it over about a week. Amazingly stable, compared to my 731B's. I wound up using the 125, 250, and 1K ohm resistors in parallel to get the 'CAL' pot to bring on scale as measured by the 3458A. The 'TH' resistance is about 3.3307 K ohms and very stable. The 732A has a front panel LED that illuminates when the unit is 'IN CAL' while the 735C has an LED that illuminates when there is 'NO CAL'. To get the 'NO CAL' LED to go out, I just used a copper 14 ga wire as the 'twiddler' and when I touched the internal 'Oven' case (likely shorting it to the front panel case) the LED went out (after everything had warmed up for a day and adjusted to 10.000 VDC). The 'NO CAL' LED has been out now for about a week and looking at the 3458A it is about +/- 2 uV from 10.000 VDC. I have some pictures of the 'innards' of the oven if anyone is interested. The NiCd's (2800 mAH compared to 2500 mAH original batteries) seem to power the unit for only about 3 hours. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] VS330
Worked fine with Win7Pro and Adobe Reader X. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Didier Juges Sent: Saturday, September 21, 2013 5:13 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] VS330 I do not use Acrobat, but the manual opens fine in the Fox it Reader Didier KO4BB RF PRECHTL K7DFW k7...@clatskanie.com wrote: Downloaded this pdf from KO4BB and my Adobe reader 6.0.6 - 01/08/2007 claims it is damaged and not capable of opening. Is this a case where Adobe is not backwards compatible and the KO4BB Adobe being too new for my 2007 edition? Rolynn ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. -- Sent from my Motorola Droid Razr 4G LTE wireless tracker while I do other things. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] HP 3458A repair.
I agree with your assessment of an 'old' meter being more desirable. I would recommend a conversation with Gary Bierman if you have not already done that. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of John Phillips Sent: Wednesday, September 18, 2013 10:45 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] HP 3458A repair. The meter will be here in a few days. I did buy the ram. As far as the meter age goes, unless it is a very old hardware revision, I would rather have an old meter because it is more stable. That is until it brakes. I do have before and after readings and the 4 readings are the only ones that failed. The after readings are better than the before readings. Because the 8 and 10 MHz failed we did not get a certificate but we do have a good limited calibration up to 2 MHz. Just no official cal. The cal we wanted was $1600. We will still have to pay that after the repair. From what I can tell they did not run SCAL or the before/after readings would not have matched. Like they really did not run after readings or there would have been some mismatch. On Tue, Sep 17, 2013 at 6:50 PM, J. L. Trantham jlt...@att.net wrote: John, Can you give us more information? Serial number, Rev. number, CALNUM? How much to invest will be determined by age and other condition. It doesn't sound like a simple CALRAM issue but changing the CALRAM is relatively easy. I removed all three DALLAS chips in mine and installed sockets. The CALRAM can be read with a chip programmer and the data written to a new DALLAS chip. I would also call Gary Bierman at the Loveland Cal Lab and have a long talk with him. He has a lot of insight into these meters and generally prefers to do a component level repair rather than an assembly level repair. The charge sounds like their standard repair charge, no matter what the problem is, and includes a 'fresh calibration' along with a warranty, a year I think, but Gary will be able to answer that question. Also, once you get the meter calibrated by Agilent (and thus prove it is functioning normally) it will be eligible for their 'repair agreement' which is $178.68 per year. I would consider buying a 5 year agreement after the repair. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of John Phillips Sent: Tuesday, September 17, 2013 3:36 PM To: Discussion of precise voltage measurement Subject: [volt-nuts] HP 3458A repair. Hi, I have a 3458A that we sent to Agilent for calibration which it failed. Before we sent it we calibrated it and it looked good to us. The infor. we revived led us to believe that the cal memory may have caused the failure. We ask that it be sent buck to us and paid half the cal charges (about $800) insted of the $2660.64 they wanted to repair it. We were just going to repalce the ram in try again. When we got the meter back it came with befor and afer data Like before 10 volts read 9.957 and after it read 10.9 so they did something or the meter drifted that much. The problem is 0.1 volt and 1.0 volts failed at 8 and 10 MHz but passed at 4 MHz. 4MHZ 0.1 volt reads 0.097251 Lower Limit is 0.095930 PASSED 8MHZ 0.1 volt reads 0.085712 Lower Limit is 0.0959 2 0 FAILED 10MHZ 0.1 volt reads 0.75569 Lower Limit is 0.084900 FAILED 4MHZ 1 volt reads 0.97272 Lower Limit is 0.95930 PASSED 8MHZ 1 volt reads 0.86389 Lower Limit is 0.95920 FAILED 10MHZ 1 volt reads 0.73514 Lower Limit is 0.84900 FAILED The AC after readings are the same. I do not see how AC after could be that identical even if they did not try to calibrate it. Did they just copy the before data and call it after data? My best guess is that if the 4 MHz is in and the higher frequencies are not the meter requires some kind of mechanical adjustment to get the frequency response withing spec or the AC board needs to be repaid. Are they charging a standard repair charge to do a calibration? I do not see changing the memory to fix this. Where would you go from here if this was your meter? -- John Phillips ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. -- John Phillips ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts
Re: [volt-nuts] HP 3458A repair.
You have to remember he is employed by Agilent and he has considerable energy invested in this meter. I'm not convinced he thinks moving manufacturing to Malaysia was a smart move given the extreme performance specs for this meter and the extreme performance specs required for the parts that went into the meter. It might be that he was giving you some hints about what to change to again see if it would meet spec. What 'standards' do you have that might address the scales in question? Do you have another 'calibrated' meter that you might use as a 'transfer standard'? IIRC, they call for an HP 3325A as the source for the AC scales. I don't have any 'thermal standards' but I did use an HP 3326A as the source and an Ametek calibrated Solartron 7081 as the 'transfer standard' on the 3458A that I calibrated but the two middle AC scales (can't recall right now as I am OOT) did not meet spec in the 'as found' category but were able to be calibrated by HP. All I got was the 'Agilent Cal' for about $500 or so. I have since returned it to Agilent for 're-cal' just over a year later and it was still 'in spec' 'as found'. I also got the 'Repair Agreement' after it was calibrated. That way, if it fails in the next 5 years, it will be repaired and calibrated 'no additional charge'. So, if you have any insight from the conversation with Gary, you might try changing the appropriate parts and see what you get. I now have a Fluke 5100B and 5200A along with my 'in cal' 3458A and happy to help in any way I can. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of John Phillips Sent: Wednesday, September 18, 2013 3:51 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] HP 3458A repair. I talked with Gary... it sounded more like a sales pitch like they replace relays before they failed. The date code on the cal ram is 6 years past replacement. The only problem seems to be the amplifier/attenuator flatness. He said this could be caused by caps aging and changing value. They tried calibrating the AC but it did not come withing spec so they did not update the as left data... I think if they are going to send the as left data it should be real as left or blank. This meter looks like a good meter for what I will use it for so we will not use the AC above 2MHz as if we ever did. On Wed, Sep 18, 2013 at 9:17 AM, J. L. Trantham jlt...@att.net wrote: I agree with your assessment of an 'old' meter being more desirable. I would recommend a conversation with Gary Bierman if you have not already done that. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of John Phillips Sent: Wednesday, September 18, 2013 10:45 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] HP 3458A repair. The meter will be here in a few days. I did buy the ram. As far as the meter age goes, unless it is a very old hardware revision, I would rather have an old meter because it is more stable. That is until it brakes. I do have before and after readings and the 4 readings are the only ones that failed. The after readings are better than the before readings. Because the 8 and 10 MHz failed we did not get a certificate but we do have a good limited calibration up to 2 MHz. Just no official cal. The cal we wanted was $1600. We will still have to pay that after the repair. From what I can tell they did not run SCAL or the before/after readings would not have matched. Like they really did not run after readings or there would have been some mismatch. On Tue, Sep 17, 2013 at 6:50 PM, J. L. Trantham jlt...@att.net wrote: John, Can you give us more information? Serial number, Rev. number, CALNUM? How much to invest will be determined by age and other condition. It doesn't sound like a simple CALRAM issue but changing the CALRAM is relatively easy. I removed all three DALLAS chips in mine and installed sockets. The CALRAM can be read with a chip programmer and the data written to a new DALLAS chip. I would also call Gary Bierman at the Loveland Cal Lab and have a long talk with him. He has a lot of insight into these meters and generally prefers to do a component level repair rather than an assembly level repair. The charge sounds like their standard repair charge, no matter what the problem is, and includes a 'fresh calibration' along with a warranty, a year I think, but Gary will be able to answer that question. Also, once you get the meter calibrated by Agilent (and thus prove it is functioning normally) it will be eligible for their 'repair agreement' which is $178.68 per year. I would consider buying a 5 year agreement after the repair. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of John Phillips
Re: [volt-nuts] HP 3458A repair.
Forgot to ask. Do you have the CLIP? Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of John Phillips Sent: Wednesday, September 18, 2013 3:51 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] HP 3458A repair. I talked with Gary... it sounded more like a sales pitch like they replace relays before they failed. The date code on the cal ram is 6 years past replacement. The only problem seems to be the amplifier/attenuator flatness. He said this could be caused by caps aging and changing value. They tried calibrating the AC but it did not come withing spec so they did not update the as left data... I think if they are going to send the as left data it should be real as left or blank. This meter looks like a good meter for what I will use it for so we will not use the AC above 2MHz as if we ever did. On Wed, Sep 18, 2013 at 9:17 AM, J. L. Trantham jlt...@att.net wrote: I agree with your assessment of an 'old' meter being more desirable. I would recommend a conversation with Gary Bierman if you have not already done that. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of John Phillips Sent: Wednesday, September 18, 2013 10:45 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] HP 3458A repair. The meter will be here in a few days. I did buy the ram. As far as the meter age goes, unless it is a very old hardware revision, I would rather have an old meter because it is more stable. That is until it brakes. I do have before and after readings and the 4 readings are the only ones that failed. The after readings are better than the before readings. Because the 8 and 10 MHz failed we did not get a certificate but we do have a good limited calibration up to 2 MHz. Just no official cal. The cal we wanted was $1600. We will still have to pay that after the repair. From what I can tell they did not run SCAL or the before/after readings would not have matched. Like they really did not run after readings or there would have been some mismatch. On Tue, Sep 17, 2013 at 6:50 PM, J. L. Trantham jlt...@att.net wrote: John, Can you give us more information? Serial number, Rev. number, CALNUM? How much to invest will be determined by age and other condition. It doesn't sound like a simple CALRAM issue but changing the CALRAM is relatively easy. I removed all three DALLAS chips in mine and installed sockets. The CALRAM can be read with a chip programmer and the data written to a new DALLAS chip. I would also call Gary Bierman at the Loveland Cal Lab and have a long talk with him. He has a lot of insight into these meters and generally prefers to do a component level repair rather than an assembly level repair. The charge sounds like their standard repair charge, no matter what the problem is, and includes a 'fresh calibration' along with a warranty, a year I think, but Gary will be able to answer that question. Also, once you get the meter calibrated by Agilent (and thus prove it is functioning normally) it will be eligible for their 'repair agreement' which is $178.68 per year. I would consider buying a 5 year agreement after the repair. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of John Phillips Sent: Tuesday, September 17, 2013 3:36 PM To: Discussion of precise voltage measurement Subject: [volt-nuts] HP 3458A repair. Hi, I have a 3458A that we sent to Agilent for calibration which it failed. Before we sent it we calibrated it and it looked good to us. The infor. we revived led us to believe that the cal memory may have caused the failure. We ask that it be sent buck to us and paid half the cal charges (about $800) insted of the $2660.64 they wanted to repair it. We were just going to repalce the ram in try again. When we got the meter back it came with befor and afer data Like before 10 volts read 9.957 and after it read 10.9 so they did something or the meter drifted that much. The problem is 0.1 volt and 1.0 volts failed at 8 and 10 MHz but passed at 4 MHz. 4MHZ 0.1 volt reads 0.097251 Lower Limit is 0.095930 PASSED 8MHZ 0.1 volt reads 0.085712 Lower Limit is 0.0959 2 0 FAILED 10MHZ 0.1 volt reads 0.75569 Lower Limit is 0.084900 FAILED 4MHZ 1 volt reads 0.97272 Lower Limit is 0.95930 PASSED 8MHZ 1 volt reads 0.86389 Lower Limit is 0.95920 FAILED 10MHZ 1 volt reads 0.73514 Lower Limit is 0.84900 FAILED The AC after readings are the same. I do not see how AC after could be that identical even if they did not try to calibrate it. Did they just copy the before data and call it after data? My best guess is that if the 4 MHz is in and the higher frequencies are not the meter requires some
Re: [volt-nuts] HP 3458A repair.
John, Can you give us more information? Serial number, Rev. number, CALNUM? How much to invest will be determined by age and other condition. It doesn't sound like a simple CALRAM issue but changing the CALRAM is relatively easy. I removed all three DALLAS chips in mine and installed sockets. The CALRAM can be read with a chip programmer and the data written to a new DALLAS chip. I would also call Gary Bierman at the Loveland Cal Lab and have a long talk with him. He has a lot of insight into these meters and generally prefers to do a component level repair rather than an assembly level repair. The charge sounds like their standard repair charge, no matter what the problem is, and includes a 'fresh calibration' along with a warranty, a year I think, but Gary will be able to answer that question. Also, once you get the meter calibrated by Agilent (and thus prove it is functioning normally) it will be eligible for their 'repair agreement' which is $178.68 per year. I would consider buying a 5 year agreement after the repair. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of John Phillips Sent: Tuesday, September 17, 2013 3:36 PM To: Discussion of precise voltage measurement Subject: [volt-nuts] HP 3458A repair. Hi, I have a 3458A that we sent to Agilent for calibration which it failed. Before we sent it we calibrated it and it looked good to us. The infor. we revived led us to believe that the cal memory may have caused the failure. We ask that it be sent buck to us and paid half the cal charges (about $800) insted of the $2660.64 they wanted to repair it. We were just going to repalce the ram in try again. When we got the meter back it came with befor and afer data Like before 10 volts read 9.957 and after it read 10.9 so they did something or the meter drifted that much. The problem is 0.1 volt and 1.0 volts failed at 8 and 10 MHz but passed at 4 MHz. 4MHZ 0.1 volt reads 0.097251 Lower Limit is 0.095930 PASSED 8MHZ 0.1 volt reads 0.085712 Lower Limit is 0.0959 2 0 FAILED 10MHZ 0.1 volt reads 0.75569 Lower Limit is 0.084900 FAILED 4MHZ 1 volt reads 0.97272 Lower Limit is 0.95930 PASSED 8MHZ 1 volt reads 0.86389 Lower Limit is 0.95920 FAILED 10MHZ 1 volt reads 0.73514 Lower Limit is 0.84900 FAILED The AC after readings are the same. I do not see how AC after could be that identical even if they did not try to calibrate it. Did they just copy the before data and call it after data? My best guess is that if the 4 MHz is in and the higher frequencies are not the meter requires some kind of mechanical adjustment to get the frequency response withing spec or the AC board needs to be repaid. Are they charging a standard repair charge to do a calibration? I do not see changing the memory to fix this. Where would you go from here if this was your meter? -- John Phillips ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 735C - Information Needed
Same color scheme on the batteries in the 5315A. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Chuck Harris Sent: Tuesday, September 10, 2013 10:30 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 735C - Information Needed Nope, I was talking about GE and their early Nicads.. And, the cylindrical sealed LEAD acid cells were also originally commercialized by GE. They later spun off that division to became GATES. I had an old SOLA UPS that was made in the mid 1960's that was full of GE SLA's... The color scheme of the cells was exactly the same as the later GATES cells. (Black bottom, white top with red trim...) and I had an old Sulzer 2.5A double oven oscillator that was full of the GE Nicads from about 1960. GE was big in battery research in the late 50's, early 60's. I don't recall the exact order, but there is a big connection between GE, Gould, and GATES batteries. ... both Nicads and SLA's. -Chuck Harris J. L. Trantham wrote: Chuck, Do you mean 'commercialize the cylindrical sealed LEAD cells'? The cylindrical cells in the 6 V battery that powers the HP 5315A is a Gates IIRC. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Chuck Harris Sent: Tuesday, September 10, 2013 9:53 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 735C - Information Needed As I recall, GE was first to commercialize the cylindrical sealed Nicad cells. They spun off their battery division to Gould, then to GATES... or some such memory muddled permutation like that... -Chuck Harris ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 735C - Information Needed
Will, Thanks for the info. I'll download the 732A manual and see how it compares. It is quite 'the beast'. I plugged it in, connected it to my (recently Fluke calibrated) Fluke 8846A and it read 10.0 VDC. This AM it read 10.7 VDC The 'TH' resistance read about 8000 ohms (IIRC) and slowly came down as the unit 'warmed up'. This AM it is about 3330 ohms. I'd love to know what the 'normal value' of this is at operating temperature. I need to replace the NiCd's and 'clean up' some corrosion on the battery charger board but it looks like it will be an interesting addition. It has a 'sticker' relating calibration in 1982 by 'Primary Standards Lab' with a 're-cal' date 3 months later. So, that would fit as a 'transfer standard'. I'll have to get it 'back in shape', allow it to stabilize for a couple of weeks then see what it is on my 3458A. Do you have any 'specifications' or other documents about it? Thanks again. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Will Sent: Saturday, September 07, 2013 7:57 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 735C - Information Needed 735C = prototype version of the 732A. 10V output only, no 1V divider. As far as I know it was used for DC transfer experiments from the Fluke primary lab to their other calibration sites. 2013/9/7, J. L. Trantham jlt...@att.net: Does anyone have any information about the Fluke 735C 10.0 DC Voltage Standard? Thanks. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] Fluke 735C - Information Needed
Does anyone have any information about the Fluke 735C 10.0 DC Voltage Standard? Thanks. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Old Weston cells
Joe, Do you have any data about prior measurements of these cells or their history? I read somewhere there is a predicted decline rate based on time and temperature IIRC. Once you get the meter calibrated, you can 'track' their measurements. Might also be time for a 'differential voltmeter' to add to the 'stable'. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Joseph Gray Sent: Monday, August 26, 2013 9:59 PM To: Discussion of precise voltage measurement Subject: [volt-nuts] Old Weston cells I got my hands on a few very old, unsaturated cells. I figured that they would be no good after all this time. Just out of curiosity, I measured them on my 3457A, which hasn't been sent off for calibration yet. DMM was set to 3 V scale to keep the input impedance high. Room temperature = 26 C. Cell #1 = 1.018288 Cell #2 = 1.018236 Cell #3 = 1.018224 I read that when they get below 1.0183 V, they are no good. Is there any (easy) way to rejuvenate these cells? Joe Gray W5JG ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] HP 3457A calibration?
I would agree, the point of keeping the cal data intact is the ability to tell if the unit is 'in tolerance' on arrival at Agilent, useful information about the 'quality/stability' of the meter, I think. Also, the 'cal number' is stored in the 'cal RAM'. If you send the 3457A in with no cal data stored and it comes back with a 'cal number' of '1', I guess you also get the answer to the question of how the 'cal number' increments after a visit to Agilent. Is there a 'cal number' stored in other 34xxA DMM's? 3478A for instance? Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Dr. David Kirkby Sent: Tuesday, August 20, 2013 12:39 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] HP 3457A calibration? On 20 August 2013 02:59, J. L. Trantham jlt...@att.net wrote: Interesting video. A bit of a silly video IMHO. I can't take anyone seriously who measures the noise on a 6.5 digit multimeter by looking at the digits which change as it is across his standard bench power supply. Personally I would have thought one measured the stability of the bench supply by measuring with the DVM rather than the other way around On the 3458A that I was able to 'kill' the 'cal RAM', the 'cal counter' was at '1' when I received it. When I finished doing my 'home cal' on a 'blank' 'cal RAM', the 'cal counter' was about 20 or so. It appears that every measurement that is calibrated increments the 'cal counter'. I was thinking about that cal ram, and the issues changing the battery. If one intends sending the meter away for cal, is there any point in worrying about if the ram is kept alive? I guess it might be nice to know the performance before it was calibrated as Agilent will give you, but there is a bit of a risk of damaging chaning the battery live. I intend changing the battery in mine then sending it for cal. Dave ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] HP 3457A calibration?
Joe, You might start with just measuring the battery voltage. There are two different batteries used in the 3457A, one that is obsolete and the other still obtainable. Per the manual, there are two resistors that need to be changed if switching from the 'obsolete' battery to the 'new battery'. However, I wonder if the resistors even need to be changed. See message #48954 in the archives and the several messages surrounding that. If the battery is 'OK', I would favor just sending it to Agilent for calibration, unless you can 'SAFELY' change the battery while preserving the cal data. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Joseph Gray Sent: Monday, August 19, 2013 10:16 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] HP 3457A calibration? Someone pointed out in the comments that his DMM was probably calibrated once or twice at the most. From looking at what is on this guys shelf, he has way more money than I do to spend on test equipment. I'm jealous :-) Speaking of too much test equipment, that guy Dave who does the EEVBlog videos also has a lab to drool over. I don't know what he does for a living, but he seems to be more knowledgeable than the guy from South Africa (who is also long winded and boring). As for the cheaper solution, I just bought a 3457A. I expect to see it this week. After checking it out for a bit, I'll be sending it to Agilent. Although it may seem ridiculous to spend $200 getting it calibrated, I figure that I'd be better off having Agilent do it this time, so I know where I stand. I'll have to read the cal counter before and after sending it to Agilent. I'll let you know the results. If the battery looks original, I'll have to change that out before sending it for calibration. I know that I have to keep the SRAM powered, so I don't loose the data already there. Joe Gray W5JG On Mon, Aug 19, 2013 at 7:59 PM, J. L. Trantham jlt...@att.net wrote: Interesting video. His 'cal certificate' shows no data other than 'in tolerance' and the 'standards' used to established that. He also makes a point of the 'calibration counter'. On the 3458A that I was able to 'kill' the 'cal RAM', the 'cal counter' was at '1' when I received it. When I finished doing my 'home cal' on a 'blank' 'cal RAM', the 'cal counter' was about 20 or so. It appears that every measurement that is calibrated increments the 'cal counter'. I don't recall how many ranges the 3457A has but add them all up and that is what I would expect the 'cal counter' to increment by once you finish a 'home cal'. Since Agilent does all this via HPIB (at least that's what I think), it increments the 'cal counter' only by '1' step. I would opt for a 'cheaper' 'seems to be working' 3457A then send it to Agilent for their calibration rather than spend extra for someone else's calibration. Joe WB4BPP -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Joseph Gray Sent: Monday, August 19, 2013 8:20 PM To: Discussion of precise voltage measurement Subject: [volt-nuts] HP 3457A calibration? Recently, we were discussing where to get a DMM calibrated. Someone mentioned an ebay seller in Albuquerque who had their meters calibrated across town by a company with a similar name. I am currently watching a YouTube video about the HP 3457A. At around 7 minutes, take a look at the calibration certificate when he holds it up. http://www.youtube.com/watch?v=HfxpJCdgVwc Joe Gray W5JG ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] DMM calibration
Joe, Sorry to be so late answering this post but I'm 'tied up' right now and can't get to all my data. However, I have sent DMM's to Agilent (3458A), Fluke (289 and 8846A), and Ametek - Solartron (7081). Agilent was for their 'Agilent Cal' and it included 'As Found' and 'As Left' data with 'adjustments' made for meters that were 'in cal' 'As Found'. Fluke was for their Z540 cal. On the 289, three types of data were offered, 'As Found', 'As Left', and 'As Found - As Left'. Both my 289's were 'in cal' and all I got back was 'As Found - As Left' and no adjustments were made. I can't recall what the 8846A was but I think it was for their Z540 cal, some measurements were out of cal, and adjustments were made. Ametek gave 'As Found', 'As Left', and both repairs and adjustments were made and included returning the replaced parts. Comparing these meters to each other, using a Fluke 335A, Fluke 5450A, EDC CR103J, and a Fluke 5100B, they are amazingly close. Therefore, I feel that these three companies provide a good service. I had occasion (as a result of 'killing' the NVRAM by trying to read it while still 'hot' from the removal process) to calibrate a 3458A using my recently calibrated 7081. I did not have a satisfactory AC source at the time and used a 3326A, Fluke 335A, and some standard resistors. I then sent the 3458A to Agilent and the 'As Found' data was all 'in cal' except for two of the 'mid-range' AC scales. Now that I have a Fluke 5100B, I think I could do a good job calibrating any meter by using my 3458A as a 'transfer standard'. I have also calibrated a 3478A that I managed to lose the NVRAM data on and it appears to be accurate as well. I have not sent it to Agilent though I have been thinking about doing that. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Joseph Gray Sent: Wednesday, August 07, 2013 11:30 AM To: Discussion of precise voltage measurement Subject: [volt-nuts] DMM calibration I just got the following quote for getting my HP 3478A calibrated. 35.00 for ISO 9001 no data 87.50 for Z540 with data and uncertainties 275.00 for 17025 with data, uncertainties and logo calibration. I still didn't get any explaination as to exactly what the lab does for each of these price points. When I asked about the Level 4 listed next to the meter, I was told that it is the difficulty level. Still not telling me much. I have read the procedure in the HP service manual and it is very easy. You set some volt, amp, ohm values on the standard, then push some buttons on the meter. I understand about getting data or not, it is the other aspects that I don't understand. What is Z540, 17025 and logo calibration? Joe Gray W5JG ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] HP 3457A
Joe, I looked at the listing you posted. The unit looks very nice. However, the units are selling in the $150 - $250 range, although I note some of the units you linked to have sold for the price asked. There has been considerable discussion on the list about the 3457A and some differences between units, depending on serial number, having to do with the circuitry about the backup battery for the RAM. I would recommend you search the archives about that. Also, there are various options for the rear compartment that may or may not be useful so being able to see the back of the unit to see what is installed and what the serial number is would be useful. I prefer to use the original manufacturer's calibration services when it comes to Agilent or Fluke just to be able to eliminate the 'trust uncertainty' with 'outside labs', since I have little experience in using various calibration labs. Therefore, my choice would be to look for a desirable 'used' unit then clean it up myself, check the date on the backup battery and possibly change it (with the need to keep voltage applied to the unit while unsoldering the battery lest losing the CAL data - careful there - 'Danger, Danger, Will Robinson') then send it to Agilent for calibration ($204.22 for the Agilent Calibration). If you are looking for a currently supported unit, I would consider the Fluke 8846A. Can be had for under $1000. A bit more expensive but currently available and currently supported. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Joseph Gray Sent: Saturday, August 10, 2013 9:20 PM To: Discussion of precise voltage measurement Subject: [volt-nuts] HP 3457A If I want to buy a used 3457A, is it better to buy one as-is and send it for calibration, or perhaps spend a bit more and get one already calibrated? Just one example: http://www.ebay.com/itm/HP-3457A-6-5-Digit-Lab-Quality-Front-Terms-Fully-Ref urbished-NIST-Cal-/300928374155 Joe Gray W5JG ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] EDC VS330 Voltage Standard
Sorry to hear that. I suspect someone has one somewhere. Hopefully you'll get lucky as well. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Joseph Gray Sent: Monday, August 05, 2013 1:17 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] EDC VS330 Voltage Standard Just a follow up. You were lucky getting a manual. I just got a reply from Krohn-Hite, saying that no manual was avaialble. Joe Gray W5JG On Sun, Jul 28, 2013 at 4:11 PM, J. L. Trantham jlt...@att.net wrote: Do you have a manual? I have an EDC CR103 voltage and current standard and was able to find a 'partial' manual on the web. It noted that a complete manual was available from Krohn-Hite and to contact them at (508) 508-1660 or sa...@krohn-hite.com to purchase a complete manual. When I emailed them about obtaining one, I received a complete .PDF manual by return email from one of their service engineers. Others have had a similar experience. When asked, Krohn-Hite asked that the manuals not be 'shared'. So, if you can't find one, you might try contacting them. Good luck. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] EDC VS330 Voltage Standard
Same issue as with the time-nuts and the man with one watch versus more. No way to tell except to 'vote with the majority'. So, the answer is 'probably'. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Joseph Gray Sent: Sunday, July 28, 2013 6:37 PM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] EDC VS330 Voltage Standard When I get the VS330, I think I'll first try setting it for 5 V and comparing the 3478A reading with what I got with the MAX6350. If I get about the same reading of 5.0026 V, is it safe to assume that the 3478A is in error? Joe Gray W5JG On Sun, Jul 28, 2013 at 5:02 PM, Joseph Gray jg...@zianet.com wrote: Thanks for the subject. Actually, I had intended to put that exact subject on it, but forgot and clicked send. I couldn't find even a partial manual for the VS330, but the 1030A partial manual looks like it would be very similar. Would you please send me the email contact at K-H that sent you the manual? Joe Gray W5JG On Sun, Jul 28, 2013 at 4:11 PM, J. L. Trantham jlt...@att.net wrote: Joe, Pardon my adding a 'subject' to the thread. If all you want is a 'house standard', you're probably 'in the ball park'. If you want to set up a home 'metrology lab' (careful, this 'metrology' stuff can be as addicting as the 'time' stuff), you should get something calibrated by an outside lab. I would send the highest resolution meter out for cal. I don't know if calibration services are available for the VS330. If the 3478A is it, I would vote to send the 3478A to Agilent for their 'Agilent' cal ($204.22 per their website) and see where you are. The meters I have sent to Agilent for their 'Agilent' cal, come back with 'as found' and 'as left' values and is very informative about the capabilities of the meters. The meters I have returned for a repeat 'Agilent' cal have all returned with their 'as found' values still in cal. Once you get the 3478A back, you will have a number of 'calibrated' measurement capabilities. Do you have a manual? I have an EDC CR103 voltage and current standard and was able to find a 'partial' manual on the web. It noted that a complete manual was available from Krohn-Hite and to contact them at (508) 508-1660 or sa...@krohn-hite.com to purchase a complete manual. When I emailed them about obtaining one, I received a complete .PDF manual by return email from one of their service engineers. Others have had a similar experience. When asked, Krohn-Hite asked that the manuals not be 'shared'. So, if you can't find one, you might try contacting them. Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Joseph Gray Sent: Sunday, July 28, 2013 4:06 PM To: volt-nuts@febo.com Subject: [volt-nuts] (no subject) I just had a friend buy this for me (I don't have an ebay account). I hope I didn't make a mistake. http://www.ebay.com/itm/Electronic-Development-Corporation-EDC-VS330- VOLTAGE -STANDARD-/161069656213?nma=truesi=PiGtcBMSdSOxgbm9DhiZ6iLo1GQ%253D orig_cv ip=truert=nc_trksid=p2047675.l2557http://www.ebay.com/itm/Electron ic-Development-Corporation-EDC-VS330-VOLTAGE-STANDARD-/161069656213?n ma=truesi=PiGtcBMSdSOxgbm9DhiZ6iLo1GQ%253Dorig_cvip=truert=nc_trk sid=p2047675.l2557 Of course, once I have it, it's the chicken and egg problem. The best I have to check it against is an HP 3478A that was last calibrated an unknown number of years ago. It has a little round sticker over the CAL switch on the front, but no dated calibration label that I can see. It came from a repair facility that closed a few years ago. Actually, the reason I was tempted to buy the VS330 was because I think the 3478A is off a bit. A while back, I built a simple 5V reference with a MAX6350. The 3478A read high. I took the reference to work and the best Fluke handheld we have was dead on (I don't remember the model right now). I made the measurement again last night and the 3478A reads 5.0026 V, with the last digit fluctuating. Room temperature was 26 C. I guess obsessing about these things has brought me here to the Volt Nuts. I have been on Time Nuts for several years. Joe Gray W5JG ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions
Re: [volt-nuts] Solartron Files
I probably have it. Can you give me the specifics of what you are looking for? Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Sam Reaves Sent: Friday, May 10, 2013 7:46 PM To: volt-nuts@febo.com Subject: [volt-nuts] Solartron Files Hello, Does anyone have the two Solartrin 7081 files posted by Mickle? The source I found has corrupt files. Thank you, Sam ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Solartron 7071 beeping about twice a second
Dave, My knowledge of 'FOD' comes from the Navy, in particular, aircraft carriers. It is an acronym standing for 'Foreign Object Damage' or 'Foreign Object Debris'. On an aircraft carrier, there are periodic 'FOD patrols' whereby the crew walk the flight deck, shoulder to shoulder, stern to bow and bow to stern, to identify and remove any object that might get 'sucked up' into the intake of a jet engine as aircraft are being launched. A particularly bad time for an engine failure and usually costs an aircraft and not uncommonly a crew. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of David C. Partridge Sent: Thursday, January 31, 2013 3:24 AM To: 'Discussion of precise voltage measurement' Subject: Re: [volt-nuts] Solartron 7071 beeping about twice a second I presume that when you push the 'DIG-FILT' key, pin 5 and 6 are 'shorted' when measured on the key board connector. Correct. 'FOD' OK, I'll bite, what does that one mean? try to open the keyboard I'm pretty sure the keyboard isn't amenable to being opened without permanent damage. Dave ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Solartron 7071 beeping about twice a second
One other thing. I presume that when you push the 'DIG-FILT' key, pin 5 and 6 are 'shorted' when measured on the key board connector. I think this is what you are saying but just want to make sure. If so, then I would conclude that there is a piece of conductive 'FOD' loose in the keyboard or the key was stuck. You might want to connect your 'beeping' DMM to indicate a short to pins 5 and 6 from the keyboard then shake the keyboard to see if it will 'beep' when you get it in the proper position. If an occasional beep occurs, then I would try to open the keyboard and clean the connectors. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of J. L. Trantham Sent: Wednesday, January 30, 2013 1:50 PM To: 'Discussion of precise voltage measurement' Subject: Re: [volt-nuts] Solartron 7071 beeping about twice a second I wonder if you have a small piece of 'FOD' floating around in the front panel contacts? I have never taken a front panel for one of these apart but would suspect there would be a way to separate the buttons from their contacts and perhaps clean them. If you could do the disassembly in a clean plastic tub then carefully inspect for any small 'pieces' that might fall out, it might explain what the problem was. Glad you are making progress. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of David C. Partridge Sent: Wednesday, January 30, 2013 12:53 PM To: 'Discussion of precise voltage measurement' Subject: Re: [volt-nuts] Solartron 7071 beeping about twice a second And just to make a monkeys uncle out me, it just started working after reconnecting the keyboard for the nth time. Confused ... You bet, especially as I cannot now provoke the fault (except by holding down the DIG-FILT key or shorting pins 5 and 6). Dave -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of David C. Partridge Sent: 30 January 2013 17:49 To: 'Discussion of precise voltage measurement' Subject: Re: [volt-nuts] Solartron 7071 beeping about twice a second Further information: Shorting pins 5 and 6 of PL2 causes the same symptoms, as does pressing and holding the DIG FILT key on another meter. The odd thing is though that an ohms test between 5 and 6 of keyboard is OC - which does not make sense to me? Dave ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] 7081 AC buffer *again*
Is it possible that it is, somehow, 'picking up interference' from your 'house standard'? How close to 10.000 000 000 MHz is it? 100 mV P-P is fairly substantial. Otherwise, is it oscillating? Good luck. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of David C. Partridge Sent: Tuesday, January 29, 2013 9:54 AM To: 'Discussion of precise voltage measurement' Subject: [volt-nuts] 7081 AC buffer *again* I've returned to the 7081 that was giving me troubles in the AC buffer circuit. Probing the output test point (TP 705) I'm seeing about 100mV pp (approx) at about 10MHz. This would likely explain the strange results I was getting last year. However I'm puzzled as to where it can be coming from!!! Any suggestions gratefully received. Regards, David Partridge ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] solartron 7081 rom reading q
MK, I am not aware of a way to read the EPROM's without removing them. Dave Partridge has done a lot of work on these along with Mickle, both on the list. Dave Partridge's EPROM images are on Didier's site. I have read the EPROM's from two of these units, of two different vintages IIRC, and can send them to you if you need them. I used the images from one to upgrade the other to the latest version that I/we know of. I know that there is an undocumented command for the HP 3458A that allows you to read the memory but I don't know if the same sort of 'global' access exists on the 7081. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of m k Sent: Saturday, January 12, 2013 2:40 PM To: volt-nuts@febo.com Subject: [volt-nuts] solartron 7081 rom reading q Hi, Finally able to talk to the meter through RS232. Is there any way to dump the rom contents using the RS232? or do the chips have to be put in a reader to get the values out? MK ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 335A versus HP 740B
I am not familiar with the HP 740B. It looks like it needs a special connector for the front panel to get an output or input. If true, I would score that as a negative. I have a 335A that had some problems when I got it. Turned out it was just some dead electrolytic caps that were identifiable by sight, bulging, etc. I replaced them and all was well. The manual is readily available. It uses binding posts for connection. It recently developed a new problem, the 'Overload' light is continuously illuminated though the output is still there and accurate. I suspect a transistor issue but have not had a chance to explore yet. You will need a 5 1/2 digit DMM, minimum, to be able to see the 1 uV change. Measuring that is a bit of a challenge in that it is not stable, voltage source, meter, temp changes, etc., all playing a role. However, I seem to recall you have a Solartron 7081 so you can watch the 1 uV all the way up to 10 V. I have been looking for an affordable 335D but, so far, have not found one. However, I found a Fluke 5440B/AF. It just arrived recently, seems to work on initial, superficial, check out, but I have not had a chance to incorporate it into my rack yet or investigate it's stability. However, at the current price, why not get all three, 335A, 335D, and 740B? Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of David C. Partridge Sent: Friday, December 14, 2012 8:16 PM To: 'Discussion of precise voltage measurement' Subject: [volt-nuts] Fluke 335A versus HP 740B Both DC Voltage standards up to 1000V or so, and also null differential meters. Also both being sold by the same vendor as the other stuff Do the assembled mavens have a view as to which is preferable? Thanks Dave ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] (New to me) 3458A Trivia
I just added a 'new' (manufacture date late 1998) 3458A to the collection that had Option 001 and 002 noted on the rear panel. When it arrived, all seemed well, passed self test, seemed accurate but with OPT? and ENTER, it returned OPT 1,0 which implies OPT 001 is installed. However, I thought that it did not have OPT 002 installed. I called Agilent, spoke with Gary Biermann, and learned that there never was a way to tell from the front panel if OPT 002 is installed or not. The only way to tell is to open it up and look at the part number on the DC Reference Board. I did this and discovered that it, indeed, had part number 03458-66519 installed, the 4 ppm reference. 03458-66509 is the 8 ppm reference and 03458-66529 is the 'ultra high stability' 2 ppm reference. Thought I would share this 'pearl' but I suspect most know this already. Was news to me though. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] Traveling Standard
Anyone heard from Bob Smither lately? I emailed him direct but have had no response. Thanks. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] HP3457 - Cal Key for Solartron 7061
Can't recall if I posted this here or not but the Cal Key for my 7081 also fits the 7061. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Fabio Eboli Sent: Wednesday, August 22, 2012 2:58 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] HP3457 As for the key, I can duplicate mine and send it to you, if you want to try another key. Fabio. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] HP3457
Fred, 700 euro seems a bit high, at least by US standards, for a 3457A. It is a good meter and I have one. Though no longer supported, Agilent will still calibrate one for $200.22, according to their website. You might want to consider a Fluke 8846A. I have seen them sell for $900 new. I don't have one but it was going to be my next addition until I got a 3458A. What is the issue with the 7061? Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Fred Schneider Sent: Tuesday, August 21, 2012 2:00 PM To: volt-nuts@febo.com Subject: [volt-nuts] HP3457 I have a solatron 7061 but i am not very impressed with it. Must I go looking for an alternative or is this 7061 when i gave it a good checkup / overhaul a good meter ? I am offered a 3457 for 700 euro and the owner, will calibrate it (he has several calibrtors, standards and a couple of 3458 meters. My intuition tells me to trusts him ( and that had never let me down before)) But he also told it is only 7,5 digit over hpib, and has 6,5 digit display. My Keithly 2000 is 6,5 digits and over RS232 i get two digits extra. I once tried that, but that took me a lot of time to get it more ore less working, I am not so into computers and afraid I do not get it to work. Is it really better as my ( one year old 2000 ?) and better as my Solartron can be ? I am afraid a 3458 is outside my budget. Fred Fred PA4TIM ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] Suggestions for Calibrators or Calibration Standards
I am considering adding some 'standards' to my shop. I am interested in the ability to calibrate DMM's, using my Agilent Calibrated 3458A as a 'transfer standard'. I have a Fluke 731B 10 VDC standard, a Fluke 335A DC Voltage Standard that goes to 1000 VDC, and some selected standard resistors from 0.1 ohm to 10 KOhm in decade steps. I also have an EDC CR103 DC Voltage and Current Standard that goes to 11.0 VDC and 111.1110 mA. In order to complete this 'lab', I would need an AC Voltage Standard and an AC Current Standard. What suggestions would there be to address this? I have thought about a Fluke 5100B as well as a 5450A Resistance Standard, and 5400A AC Voltage Standard. If I use my 3458A as the 'transfer standard', I can use it to provide the additional 'digits' beyond the 5100B. I also have some LCR meters but I don't have any capacitance or inductance standards. Once I have this sorted out, I could 'off load' my earlier, redundant, equipment. I would appreciate any and all thoughts about what to add to my 'lab'. Thanks in advance. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] 7061 On Board Clock and Battery
Fabio, I uploaded the EPROM contents from my meter to Didier's site as well. The Floating board I have has 'AA' version firmware while yours is 'CA'. Mine uses TMS2516 chips that are 24 pin and I think yours are 2764 28 pin chips. I'll investigate the statistics function after I make sure it is 'in range' as far as the calibration goes. I don't understand 'SP1 and SP2 on floating board are both open circuit'. To what does this refer? Thanks. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Fabio Eboli Sent: Monday, June 25, 2012 10:54 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] 7061 On Board Clock and Battery Interesting, these roms are mine's: http://www.ko4bb.com/Manuals/04)_ROM_Images_and_Drivers/Solartron_7061_ROMs. zip I have also a problem with statistics, seem that the standard deviation calculation is wrong: http://www.febo.com/pipermail/volt-nuts/2012-April/001793.html If you have time to test yours for this I would be grateful. Another thing is that on mine SP1 and SP2 on floating board are both open circuit. Fabio. J. L. Trantham jlt...@att.net ha scritto: Fabio, Thanks for the link to the pictures. I never seem to remember eevblog when it comes to looking for stuff. Your pictures show that your Earthy board is completely different from mine in the area of 'B101'. You have no place for B101. Also, your Floating board is different in that you have the 'CA' firmware and 28 pin EPROM's. I wonder if anyone has modified my 'version' of the Floating board to use 28 pin EPROM's and if there would be any utility to doing this. I did not get a message directly from you (that I recognized at least). I'll try to send you one directly to make sure we can 'establish communications'. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Fabio Eboli Sent: Monday, June 25, 2012 5:14 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] 7061 On Board Clock and Battery Here are some pics of the mine, only one battery: http://www.eevblog.com/forum/product-reviews-photos-and-discussion/solartron -7061-help-and-come-photos/msg93650/#msg93650 As for the manual I did try to contact you directly, did you read the message? Fabio. J. L. Trantham jlt...@att.net ha scritto: My 7061 has a single battery installed, B501, a 3.6 V Ni-MH Rechargeable 3 pin battery. My 7081 has two batteries installed, the same 3.6 V Ni-MH Rechargeable battery and a ½ AA Lithium Battery that serves as the onboard clock backup battery. It is connected with its negative pin to a 47K resistor to ground and its positive pin to a diode that feeds a connection that is also fed by another diode from the main power supply. The common connection of these diodes feed the clock chip. On my 7061 there is a location, B101 that appears to be for the same type ½ AA Lithium Battery. The + pin connects to a diode, in a manner similar to that described above, and the pin connects to a space for a resistor, R117, with the other end of the resistor going to ground. Both the battery, B101, and the resistor, R117, are missing. My questions are: 1. Does anyone have a 7061 with this battery and resistor installed? 2. If so, what is the value of the resistor, R117, and what type is the battery? It seems that this would be an easy addition to get a live clock without the need to restart the clock each time the meter is turned on. Or, perhaps there is a reason why this in not a good idea. Thanks in advance. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. This message was sent using IMP, the Internet Messaging Program. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. This message was sent using IMP, the Internet Messaging Program. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow
Re: [volt-nuts] Solartron 7061 Service Manual - Does anyone haveone?
Dave, That is always the first place I look but all I found was the ROM images and the Operating Manual. I did not see a Service Manual. I would love to find one, assuming one is 'out there' somewhere. I archived the ROM's in my 7061. The Earthy Board is the same ROM's as on Didier's site. However, my Floating Board uses TI TMS2516's and has version 51AA, 52AA, 53AA, and 54AA firmware. I can send it to anyone that wants it and can upload it to Didier's site if useful. My Floating Board has 24 pin sockets for the TMS2516's but room for a 28 pin socket and 'pads' to solder to tell it if it is using the '16's' that mine uses or the '64's' that are used for the 'CA' firmware. I don't know if there is any advantage to changing this to use the 'later' firmware. I would appreciate any info on this that anyone has. Thanks. Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Dave M Sent: Sunday, June 24, 2012 8:08 AM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Solartron 7061 Service Manual - Does anyone haveone? Date: Sat, 23 Jun 2012 11:12:59 -0500 From: J. L. Trantham jlt...@att.net I am looking for a service manual for the Solartron 7061, paper or .PDF. Does anyone know where I might find one? Thanks in advance. Joe * Joe, Didier has PDF copies of the User manual, Service manual and ROM images on his manual web site at http://www.ko4bb.com/manuals/index.php. Just search for 7061. Dave M A woman has the last word in any argument. Anything a man says after that is the beginning of a new argument. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] 7061 On Board Clock and Battery
My 7061 has a single battery installed, B501, a 3.6 V Ni-MH Rechargeable 3 pin battery. My 7081 has two batteries installed, the same 3.6 V Ni-MH Rechargeable battery and a ½ AA Lithium Battery that serves as the onboard clock backup battery. It is connected with its negative pin to a 47K resistor to ground and its positive pin to a diode that feeds a connection that is also fed by another diode from the main power supply. The common connection of these diodes feed the clock chip. On my 7061 there is a location, B101 that appears to be for the same type ½ AA Lithium Battery. The + pin connects to a diode, in a manner similar to that described above, and the pin connects to a space for a resistor, R117, with the other end of the resistor going to ground. Both the battery, B101, and the resistor, R117, are missing. My questions are: 1. Does anyone have a 7061 with this battery and resistor installed? 2. If so, what is the value of the resistor, R117, and what type is the battery? It seems that this would be an easy addition to get a live clock without the need to restart the clock each time the meter is turned on. Or, perhaps there is a reason why this in not a good idea. Thanks in advance. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] Solartron 7061 Cal Key
The Cal Key I received for my 7081 also works for the 7061. If anyone needs one, let me know. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Calibration Sources for Solartron 7081
Jean-Louis, My DC standards are a Fluke 731B and Fluke 335A DC Voltage Standards and an EDC CR103 DC Current and Voltage Standard. My resistance standards consist of a collection of Leeds and Northrup type standard resistors as well as a collection of General Radio Decade Resistance boxes that will get me up into the MegOhm range. However, AC promises to be a problem. I guess I will try using a signal generator and an amplifier to see what output level and stability I can achieve. But still, what frequency? 60 Hz? 1 kHz? Thanks, Joe -Original Message- From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On Behalf Of Jean-Louis Noel Sent: Friday, June 22, 2012 3:58 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Calibration Sources for Solartron 7081 Hi J. L. From: J. L. Trantham jlt...@att.net What source could/should I use for the ACV and what frequency for the ACV? I can come up with a power transformer and a variac for 60 Hz. The stability is not high enough to be useful! Have a look to http://www.on4jln.be/infos/getfile?file=515A-OM.pdf, it is the Fluke 515A, to learn how they dealt with that problem. If your 10V is good you will get better results too. And it needs to be very good to calibrate the 7081! Stability is more important than everything else and that is the problem. Have a good day. Bye, Jean-Louis ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] Calibration Sources for Solartron 7081
I recently added another Solartron 7081 to my collection that seems to be working correctly, is stable and I have upgraded it to the latest firmware. However, it is clearly not accurate, at least on the 10 VDC scale, measuring my 10.000 000 standard at about 10.000 840 V while my other meters are +/- about 10 uV of 10.000 000 VDC. If I understand the calibration procedure correctly, I will need an input short, an input open, and 'standards' for each of the VDC, VAC, and Ohms ranges to perform a complete calibration procedure. This would appear to be a source for 0.1, 1, 10, 100, and 1000 VDC and VAC along with 0.1, 1, 10, 100, and 1000 kOhm and 10 and 1000 MOhm resistor standards. I can come up with the VDC standards and the resistance standards except the 1000 MOhm. I don't have 'standards' but I have references that are stable and calibrated meters (another 7081 and two 3458A's) that I can use to measure my references and then input their measurements into the DUT 7081 to perform the calibrations. My questions are: What source could/should I use for the ACV and what frequency for the ACV? I can come up with a power transformer and a variac for 60 Hz. I also have a selection of signal generators and synthesizers for amplitudes up to about +10 dBm at pretty much any frequency you prefer. What would be the lowest ACV that I could/should use for the 1000 VAC scale? What suggestions for the 1000 MOhm standard? What is the lowest value that I could/should use for the 1000 MOhm scale? My plan is to test each scale against my other meters and calibrate only the scales that are 'out of bounds'. Good idea or bad? Thanks in advance. Joe ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.