Re: [volt-nuts] DMM calibration

2013-08-14 Thread J. L. Trantham
Joe,

Sorry to be so late answering this post but I'm 'tied up' right now and
can't get to all my data.

However, I have sent DMM's to Agilent (3458A), Fluke (289 and 8846A), and
Ametek - Solartron (7081).

Agilent was for their 'Agilent Cal' and it included 'As Found' and 'As Left'
data with 'adjustments' made for meters that were 'in cal' 'As Found'.

Fluke was for their Z540 cal.  On the 289, three types of data were offered,
'As Found', 'As Left', and 'As Found - As Left'.  Both my 289's were 'in
cal' and all I got back was 'As Found - As Left' and no adjustments were
made.  I can't recall what the 8846A was but I think it was for their Z540
cal, some measurements were out of cal, and adjustments were made.

Ametek gave 'As Found', 'As Left', and both repairs and adjustments were
made and included returning the replaced parts.

Comparing these meters to each other, using a Fluke 335A, Fluke 5450A, EDC
CR103J, and a Fluke 5100B, they are amazingly close.  Therefore, I feel that
these three companies provide a good service.

I had occasion (as a result of 'killing' the NVRAM by trying to read it
while still 'hot' from the removal process) to calibrate a 3458A using my
recently calibrated 7081.  I did not have a satisfactory AC source at the
time and used a 3326A, Fluke 335A, and some standard resistors.  I then sent
the 3458A to Agilent and the 'As Found' data was all 'in cal' except for two
of the 'mid-range' AC scales.

Now that I have a Fluke 5100B, I think I could do a good job calibrating any
meter by using my 3458A as a 'transfer standard'.

I have also calibrated a 3478A that I managed to lose the NVRAM data on and
it appears to be accurate as well.  I have not sent it to Agilent though I
have been thinking about doing that.

Good luck.

Joe

-----Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Joseph Gray
Sent: Wednesday, August 07, 2013 11:30 AM
To: Discussion of precise voltage measurement
Subject: [volt-nuts] DMM calibration

I just got the following quote for getting my HP 3478A calibrated.

35.00 for ISO 9001 no data
87.50 for Z540 with data and uncertainties
275.00 for 17025 with data, uncertainties and logo calibration.

I still didn't get any explaination as to exactly what the lab does for each
of these price points. When I asked about the "Level 4" listed next to the
meter, I was told that it is the difficulty level. Still not telling me
much. I have read the procedure in the HP service manual and it is very
easy. You set some volt, amp, ohm values on the standard, then push some
buttons on the meter.

I understand about getting data or not, it is the other aspects that I don't
understand. What is Z540, 17025 and "logo calibration"?


Joe Gray
W5JG
___
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] DMM calibration

2013-08-08 Thread John Phillips
I did some work for a resistor manufacture  and was told about resistor
drift. It does not keep going the same direction forever it does hit a
limit and start back the other way. 4 years is just an approximation. The
material and environment will have a lot to do with the period. Agilent
burns in  their  voltage references and the main difference is how long
they "cook" them. They also screen to get a better reference. evenhandedly
the drift rate and amplitude drops. In a few hundred years things will be
stable or broken. Broken is very stable. Like wire resistance will not keep
going down till it hits 0 or up until it opens.
Agilent says do not turn your 3458A off or it will not be as stable as if
it has been running for years.
Agilent move some reference manufacturing off shore, when the references
met the stability numbers they were harvested and put in stock bins. Latter
when they were used they no longer met the spec. There was a memo sent out
that if you had one of these references its stability should be withing
spec if the meter was not turned off for 6 mo. (that is how I remember it).
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] DMM calibration

2013-08-08 Thread Dr. David Kirkby
On 8 August 2013 18:58, John Phillips  wrote:
> You will find that most old "stuff" does not drift as much as the new
> "stuff".
> Most resistors seem to drift in a damped sine wave rate with a period of
> about 4 years.

Is there any theory behind that? It's the first time I have heard of
the 4 year damped sine wave!

Dave
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] DMM calibration

2013-08-08 Thread John Phillips
You will find that most old "stuff" does not drift as much as the new
"stuff".
Most resistors seem to drift in a damped sine wave rate with a period of
about 4 years.


On Thu, Aug 8, 2013 at 10:10 AM, Didier Juges  wrote:

> I had my eBay special $90 HP3478A calibrated by our internal cal lab
> (traceable to NIST). Based on the sticker, the last cal was 5 years old.
> The instrument was well within all specs by a wide margin (most readings
> were within 20% of the stated tolerance, which after 5 years is pretty good.
>
> Didier KO4BB
>
> Roy Phillips  wrote:
> >
> >
> >-Original Message-
> >From: Electronics and Books
> >Sent: Thursday, August 08, 2013 12:54 PM
> >To: Discussion of precise voltage measurement
> >Subject: Re: [volt-nuts] DMM calibration
> >
> >Specs for a 3478A
> >
> >
> >Brand HP Model 3478A Counts 303.099 TrueRMS X Display LCD  Year 1983
> >Range
> >Stability [ppm] Low   High   Resolution  1y
> >DC Volt 30m  300  100n 60 AC Volt 300m 300  1u  2000 Max Frequency 300k
> >
> >Resistance 30  30M  100u 160 DC Current 300m 3  1u  1500 AC Current
> >300m 3
> >1u  7200 Interface GPIB Battery backup X
> >If you do the adjustments with some hobby equipment wich an accuracy of
> >say
> >20 to 30 ppm. Thats all within limits of this multimeter. You never see
> >the
> >difference.
> >In my opion, expensive adjustments and calibration for not so accurate
> >equipment is a waste of money. Especialy when the calibration price
> >exceeds
> >the equipment price.
> >
> >
> >
> >
> >
> >Regards
> >
> >
> >electronicsandbo...@yahoo.com
> >http://ElectronicsAndBooks.com
> >
> >I would like to say - "you are absolutely correct" - I have recently
> >been
> >quoted for a 3458A which requires a new set of ROM's and re-cal. - the
> >cost
> >was greater than I paid for the S/H instrument just 4 years ago. These
> >prices are OK for large Companies who can recover the VAT for a start !
> >
> >. but cannot be justified for hobbyists. Most of us have items
> >of
> >equipment that have been calibrated in recent times, and other
> >precision
> >items like resistors and higher spec. zeners that when you make
> >comparisons
> >can verify your DVM's . I recently purchased a Keithley 2015 for less
> >than
> >some cal. charges, this was calibrated in 2008 and I would say that it
> >is
> >still in very good calibration now.
> >Roy P
> >
> >
> >
> >From: "Marv @ Home" 
> >To: Discussion of precise voltage measurement ;
> >Discussion of precise voltage measurement 
> >Sent: Wednesday, August 7, 2013 6:53 PM
> >Subject: Re: [volt-nuts] DMM calibration
> >
> >
> >I don't know what logo cal is, but between Z540 and 17025 is a
> >question of procedures and how they are done.
> >In metrology the procedures, and how strict they are, reduce the risk
> >of uncertainty in the final measurement.  For example, some
> >procedures in Z540 can be 'interpreted' where in 17025 is explicit,
> >thus there is more room in Z540 for one lab to be worse than another,
> >or vice versa, which defeats the purpose of standardization.  This
> >'intepretation' is fairly reduced when the device is cal'd by an
> >autocalibrator such for close case calibration DMMs, but becomes a
> >big issue with any manual cal.
> >
> >If these do not mean anything to you and you do not need the data for
> >long term drift calculations, ISO 9001 is good to go.
> >
> >IMHO the difference between all 3 is insignificant for 5.5 digit
> >DMMs, except for one's need for data.
> >
> >
> >
> http://us.flukecal.com/literature/articles-and-education/temperature-calibration/papers-articles/comparison-ansincsl-z540-1?geoip=1
> >
> >At 12:30 PM 8/7/2013, Joseph Gray wrote:
> >>I just got the following quote for getting my HP 3478A calibrated.
> >>
> >>35.00 for ISO 9001 no data
> >>87.50 for Z540 with data and uncertainties
> >>275.00 for 17025 with data, uncertainties and logo calibration.
> >>
> >>I still didn't get any explaination as to exactly what the lab does
> >for
> >>each of these price points. When I asked about the "Level 4" listed
> >next to
> >>the meter, I was told that it is the difficulty level. Still not
> >telling me
> >>much. I have read the procedure in t

Re: [volt-nuts] DMM calibration

2013-08-08 Thread Didier Juges
I had my eBay special $90 HP3478A calibrated by our internal cal lab (traceable 
to NIST). Based on the sticker, the last cal was 5 years old. The instrument 
was well within all specs by a wide margin (most readings were within 20% of 
the stated tolerance, which after 5 years is pretty good.

Didier KO4BB

Roy Phillips  wrote:
>
>
>-Original Message- 
>From: Electronics and Books
>Sent: Thursday, August 08, 2013 12:54 PM
>To: Discussion of precise voltage measurement
>Subject: Re: [volt-nuts] DMM calibration
>
>Specs for a 3478A
>
>
>Brand HP Model 3478A Counts 303.099 TrueRMS X Display LCD  Year 1983
>Range 
>Stability [ppm] Low   High   Resolution  1y
>DC Volt 30m  300  100n 60 AC Volt 300m 300  1u  2000 Max Frequency 300k
>
>Resistance 30  30M  100u 160 DC Current 300m 3  1u  1500 AC Current
>300m 3 
>1u  7200 Interface GPIB Battery backup X
>If you do the adjustments with some hobby equipment wich an accuracy of
>say 
>20 to 30 ppm. Thats all within limits of this multimeter. You never see
>the 
>difference.
>In my opion, expensive adjustments and calibration for not so accurate 
>equipment is a waste of money. Especialy when the calibration price
>exceeds 
>the equipment price.
>
>
>
>
>
>Regards
>
>
>electronicsandbo...@yahoo.com
>http://ElectronicsAndBooks.com
>
>I would like to say - "you are absolutely correct" - I have recently
>been 
>quoted for a 3458A which requires a new set of ROM's and re-cal. - the
>cost 
>was greater than I paid for the S/H instrument just 4 years ago. These 
>prices are OK for large Companies who can recover the VAT for a start !
>
>. but cannot be justified for hobbyists. Most of us have items
>of 
>equipment that have been calibrated in recent times, and other
>precision 
>items like resistors and higher spec. zeners that when you make
>comparisons 
>can verify your DVM's . I recently purchased a Keithley 2015 for less
>than 
>some cal. charges, this was calibrated in 2008 and I would say that it
>is 
>still in very good calibration now.
>Roy P
>
>
>____________
>From: "Marv @ Home" 
>To: Discussion of precise voltage measurement ; 
>Discussion of precise voltage measurement 
>Sent: Wednesday, August 7, 2013 6:53 PM
>Subject: Re: [volt-nuts] DMM calibration
>
>
>I don't know what logo cal is, but between Z540 and 17025 is a
>question of procedures and how they are done.
>In metrology the procedures, and how strict they are, reduce the risk
>of uncertainty in the final measurement.  For example, some
>procedures in Z540 can be 'interpreted' where in 17025 is explicit,
>thus there is more room in Z540 for one lab to be worse than another,
>or vice versa, which defeats the purpose of standardization.  This
>'intepretation' is fairly reduced when the device is cal'd by an
>autocalibrator such for close case calibration DMMs, but becomes a
>big issue with any manual cal.
>
>If these do not mean anything to you and you do not need the data for
>long term drift calculations, ISO 9001 is good to go.
>
>IMHO the difference between all 3 is insignificant for 5.5 digit
>DMMs, except for one's need for data.
>
>
>http://us.flukecal.com/literature/articles-and-education/temperature-calibration/papers-articles/comparison-ansincsl-z540-1?geoip=1
>
>At 12:30 PM 8/7/2013, Joseph Gray wrote:
>>I just got the following quote for getting my HP 3478A calibrated.
>>
>>35.00 for ISO 9001 no data
>>87.50 for Z540 with data and uncertainties
>>275.00 for 17025 with data, uncertainties and logo calibration.
>>
>>I still didn't get any explaination as to exactly what the lab does
>for
>>each of these price points. When I asked about the "Level 4" listed
>next to
>>the meter, I was told that it is the difficulty level. Still not
>telling me
>>much. I have read the procedure in the HP service manual and it is
>very
>>easy. You set some volt, amp, ohm values on the standard, then push
>some
>>buttons on the meter.
>>
>>I understand about getting data or not, it is the other aspects that I
>>don't understand. What is Z540, 17025 and "logo calibration"?
>>
>>
>>Joe Gray
>>W5JG
>
>
>
>Sincerely,
>
>
>
>Marv
>Philadelphia, PA
>
>___
>volt-nuts mailing list -- volt-nuts@febo.com
>To unsubscribe, go to 
>https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>and follow the instructions there.
>___
>volt-nuts mailing list -- volt-nuts@febo.c

Re: [volt-nuts] DMM calibration

2013-08-08 Thread Roy Phillips



-Original Message- 
From: Electronics and Books

Sent: Thursday, August 08, 2013 12:54 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] DMM calibration

Specs for a 3478A


Brand HP Model 3478A Counts 303.099 TrueRMS X Display LCD  Year 1983 Range 
Stability [ppm] Low   High   Resolution  1y
DC Volt 30m  300  100n 60 AC Volt 300m 300  1u  2000 Max Frequency 300k 
Resistance 30  30M  100u 160 DC Current 300m 3  1u  1500 AC Current 300m 3 
1u  7200 Interface GPIB Battery backup X
If you do the adjustments with some hobby equipment wich an accuracy of say 
20 to 30 ppm. Thats all within limits of this multimeter. You never see the 
difference.
In my opion, expensive adjustments and calibration for not so accurate 
equipment is a waste of money. Especialy when the calibration price exceeds 
the equipment price.






Regards


electronicsandbo...@yahoo.com
http://ElectronicsAndBooks.com

I would like to say - "you are absolutely correct" - I have recently been 
quoted for a 3458A which requires a new set of ROM's and re-cal. - the cost 
was greater than I paid for the S/H instrument just 4 years ago. These 
prices are OK for large Companies who can recover the VAT for a start ! 
. but cannot be justified for hobbyists. Most of us have items of 
equipment that have been calibrated in recent times, and other precision 
items like resistors and higher spec. zeners that when you make comparisons 
can verify your DVM's . I recently purchased a Keithley 2015 for less than 
some cal. charges, this was calibrated in 2008 and I would say that it is 
still in very good calibration now.

Roy P



From: "Marv @ Home" 
To: Discussion of precise voltage measurement ; 
Discussion of precise voltage measurement 

Sent: Wednesday, August 7, 2013 6:53 PM
Subject: Re: [volt-nuts] DMM calibration


I don't know what logo cal is, but between Z540 and 17025 is a
question of procedures and how they are done.
In metrology the procedures, and how strict they are, reduce the risk
of uncertainty in the final measurement.  For example, some
procedures in Z540 can be 'interpreted' where in 17025 is explicit,
thus there is more room in Z540 for one lab to be worse than another,
or vice versa, which defeats the purpose of standardization.  This
'intepretation' is fairly reduced when the device is cal'd by an
autocalibrator such for close case calibration DMMs, but becomes a
big issue with any manual cal.

If these do not mean anything to you and you do not need the data for
long term drift calculations, ISO 9001 is good to go.

IMHO the difference between all 3 is insignificant for 5.5 digit
DMMs, except for one's need for data.


http://us.flukecal.com/literature/articles-and-education/temperature-calibration/papers-articles/comparison-ansincsl-z540-1?geoip=1

At 12:30 PM 8/7/2013, Joseph Gray wrote:

I just got the following quote for getting my HP 3478A calibrated.

35.00 for ISO 9001 no data
87.50 for Z540 with data and uncertainties
275.00 for 17025 with data, uncertainties and logo calibration.

I still didn't get any explaination as to exactly what the lab does for
each of these price points. When I asked about the "Level 4" listed next to
the meter, I was told that it is the difficulty level. Still not telling me
much. I have read the procedure in the HP service manual and it is very
easy. You set some volt, amp, ohm values on the standard, then push some
buttons on the meter.

I understand about getting data or not, it is the other aspects that I
don't understand. What is Z540, 17025 and "logo calibration"?


Joe Gray
W5JG




Sincerely,



Marv
Philadelphia, PA

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts

and follow the instructions there.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there. 


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] DMM calibration

2013-08-08 Thread Electronics and Books
Specs for a 3478A


Brand HP Model 3478A Counts 303.099 TrueRMS X Display LCD  Year 1983 Range  
   Stability [ppm] Low   High   Resolution  1y
DC Volt 30m  300  100n 60 AC Volt 300m 300  1u  2000 Max Frequency 300k 
Resistance 30  30M  100u 160 DC Current 300m 3  1u  1500 AC Current 300m 3  1u  
7200 Interface GPIB Battery backup X 
If you do the adjustments with some hobby equipment wich an accuracy of say 20 
to 30 ppm. Thats all within limits of this multimeter. You never see the 
difference.
In my opion, expensive adjustments and calibration for not so accurate 
equipment is a waste of money. Especialy when the calibration price exceeds the 
equipment price.
 



 
Regards


electronicsandbo...@yahoo.com
http://ElectronicsAndBooks.com




 From: "Marv @ Home" 
To: Discussion of precise voltage measurement ; Discussion 
of precise voltage measurement  
Sent: Wednesday, August 7, 2013 6:53 PM
Subject: Re: [volt-nuts] DMM calibration
 

I don't know what logo cal is, but between Z540 and 17025 is a 
question of procedures and how they are done.
In metrology the procedures, and how strict they are, reduce the risk 
of uncertainty in the final measurement.  For example, some 
procedures in Z540 can be 'interpreted' where in 17025 is explicit, 
thus there is more room in Z540 for one lab to be worse than another, 
or vice versa, which defeats the purpose of standardization.  This 
'intepretation' is fairly reduced when the device is cal'd by an 
autocalibrator such for close case calibration DMMs, but becomes a 
big issue with any manual cal.

If these do not mean anything to you and you do not need the data for 
long term drift calculations, ISO 9001 is good to go.

IMHO the difference between all 3 is insignificant for 5.5 digit 
DMMs, except for one's need for data.


http://us.flukecal.com/literature/articles-and-education/temperature-calibration/papers-articles/comparison-ansincsl-z540-1?geoip=1

At 12:30 PM 8/7/2013, Joseph Gray wrote:
>I just got the following quote for getting my HP 3478A calibrated.
>
>35.00 for ISO 9001 no data
>87.50 for Z540 with data and uncertainties
>275.00 for 17025 with data, uncertainties and logo calibration.
>
>I still didn't get any explaination as to exactly what the lab does for
>each of these price points. When I asked about the "Level 4" listed next to
>the meter, I was told that it is the difficulty level. Still not telling me
>much. I have read the procedure in the HP service manual and it is very
>easy. You set some volt, amp, ohm values on the standard, then push some
>buttons on the meter.
>
>I understand about getting data or not, it is the other aspects that I
>don't understand. What is Z540, 17025 and "logo calibration"?
>
>
>Joe Gray
>W5JG



Sincerely,



Marv
Philadelphia, PA 

___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


Re: [volt-nuts] DMM calibration

2013-08-07 Thread Marv @ Home
I don't know what logo cal is, but between Z540 and 17025 is a 
question of procedures and how they are done.
In metrology the procedures, and how strict they are, reduce the risk 
of uncertainty in the final measurement.  For example, some 
procedures in Z540 can be 'interpreted' where in 17025 is explicit, 
thus there is more room in Z540 for one lab to be worse than another, 
or vice versa, which defeats the purpose of standardization.  This 
'intepretation' is fairly reduced when the device is cal'd by an 
autocalibrator such for close case calibration DMMs, but becomes a 
big issue with any manual cal.


If these do not mean anything to you and you do not need the data for 
long term drift calculations, ISO 9001 is good to go.


IMHO the difference between all 3 is insignificant for 5.5 digit 
DMMs, except for one's need for data.



http://us.flukecal.com/literature/articles-and-education/temperature-calibration/papers-articles/comparison-ansincsl-z540-1?geoip=1

At 12:30 PM 8/7/2013, Joseph Gray wrote:

I just got the following quote for getting my HP 3478A calibrated.

35.00 for ISO 9001 no data
87.50 for Z540 with data and uncertainties
275.00 for 17025 with data, uncertainties and logo calibration.

I still didn't get any explaination as to exactly what the lab does for
each of these price points. When I asked about the "Level 4" listed next to
the meter, I was told that it is the difficulty level. Still not telling me
much. I have read the procedure in the HP service manual and it is very
easy. You set some volt, amp, ohm values on the standard, then push some
buttons on the meter.

I understand about getting data or not, it is the other aspects that I
don't understand. What is Z540, 17025 and "logo calibration"?


Joe Gray
W5JG




Sincerely,



Marv
Philadelphia, PA 


___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


[volt-nuts] DMM calibration

2013-08-07 Thread Joseph Gray
I just got the following quote for getting my HP 3478A calibrated.

35.00 for ISO 9001 no data
87.50 for Z540 with data and uncertainties
275.00 for 17025 with data, uncertainties and logo calibration.

I still didn't get any explaination as to exactly what the lab does for
each of these price points. When I asked about the "Level 4" listed next to
the meter, I was told that it is the difficulty level. Still not telling me
much. I have read the procedure in the HP service manual and it is very
easy. You set some volt, amp, ohm values on the standard, then push some
buttons on the meter.

I understand about getting data or not, it is the other aspects that I
don't understand. What is Z540, 17025 and "logo calibration"?


Joe Gray
W5JG
___
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.